My doctor recommended I use a setting of 11 - 16 on my pressure.
Since I just got a new Aircurve10 Vauto machine, I set the mode to VAuto and wanted to know if there any disadvantage to setting the higher end of the pressure to 20 instead of 16?
It just makes sense that if this machine automatically tracts your breathing that you should make the settings wider.
Can someone please explain this to me.
Thanks!
Lennyten
Question on VAuto mode on Resmed Aircurve10 VAuto
- Dog Slobber
- Posts: 4331
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Question on VAuto mode on Resmed Aircurve10 VAuto
Your VAuto max pressure can be set as high as 25.
A fairly common belief here is to set the max pressure to the device maximum, unless there is a compelling reason to limit it. The thought being, let the machine go to where it *needs* to go.
Some reasons to limit the upper pressure are:
I like the idea of a narrow pressure zone, but not between minimum and maximum, instead narrow between minimum and where the device typically goes.
A fairly common belief here is to set the max pressure to the device maximum, unless there is a compelling reason to limit it. The thought being, let the machine go to where it *needs* to go.
Some reasons to limit the upper pressure are:
- aerophagia
- intolerance to higher pressures
- I'm sure thee are more
I like the idea of a narrow pressure zone, but not between minimum and maximum, instead narrow between minimum and where the device typically goes.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: Question on VAuto mode on Resmed Aircurve10 VAuto
That makes sense.
Thanks for the reply.
Lennyten
Thanks for the reply.
Lennyten
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ F10 Full Face Mask with Headgear |