I have been a nearly 100% cpap/apap user for over 14 years. I have gone through several machines for home, travel, hiking/canoeing, etc. as well as many mask models. For a few years, I was a regular reader and occasional poster on this site. Cpap just became part of life and faded from importance for me so I haven't kept up with the folks here. My current home machine is an aging Respironics System One Apap which has served me well enough but is getting louder. Technology since that purchase time period has obviously changed a lot. I see that both the ResMed and Respironics leading machines have been deeply discounted recently. I have two questions and one comment.
First, from the current discounts, does anyone think the industry or those businesses are about to take a technological leap and offer a better machine to replace the AirSense and Dreamstation Apaps?
Second, if I were to buy one of those two machines, which is the better in terms of programming and recognition, consistency, and durability/longevity(it won't be moved much)?
My comment is that it took just a brief look at the site and found, like many years ago, people are still helping people get through this lifestyle transition. If you are new to Cpap, you are in good hands here.
Thanks for any guidance or advice you may have.
First, from the current discounts, does anyone think the industry or those businesses are about to take a technological leap and offer a better machine to replace the AirSense and Dreamstation Apaps?
Second, if I were to buy one of those two machines, which is the better in terms of programming and recognition, consistency, and durability/longevity(it won't be moved much)?
My comment is that it took just a brief look at the site and found, like many years ago, people are still helping people get through this lifestyle transition. If you are new to Cpap, you are in good hands here.
Thanks for any guidance or advice you may have.
_________________
| Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Old Everest CPAP for at the cabin. Z2 for travel and backpacking |
"Breathe deep. Seek peace."
(James Gurney, Dinotopia)
(James Gurney, Dinotopia)
- Dog Slobber
- Posts: 4331
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: AirSense Vs Dreamstation
The favourite around here appears to be the ResMed over Respironics, though they're both good machines.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: AirSense Vs Dreamstation
it's always difficult to predict when a new machine will come out. prolly right after i buy a new machine!
peruse the forum a bit. look at the signatures and see for yourself how most people feel.
for me, it's resmed all the way and i have used both.
my main attraction to my current machine is that i can raise my pressure by increments of .2. resperonics uses increments of .5 which, believe it or not, was too much of a leap for me when i was trying to get good therapy while staving off aerophagia.
peruse the forum a bit. look at the signatures and see for yourself how most people feel.
for me, it's resmed all the way and i have used both.
my main attraction to my current machine is that i can raise my pressure by increments of .2. resperonics uses increments of .5 which, believe it or not, was too much of a leap for me when i was trying to get good therapy while staving off aerophagia.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- Okie bipap
- Posts: 3567
- Joined: Thu Oct 15, 2015 4:14 pm
- Location: Central Oklahoma
Re: AirSense Vs Dreamstation
The most highly recommended machine is the ResMed Auto For Her. That being said, you may be happier with the Dream Station auto machine since you are used to how Respironics machines operate. The ResMed machines handle events slightly different from the Respironics machines.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Evora Full Face Mask - Fitpack |
| Additional Comments: IPAP 20-25, ps 4, OSCAR software |
Growing old is mandatory, but growing up is optional.
Re: AirSense Vs Dreamstation
Or, like most (but not all (inserting standard disclaimer)) people that have tried both, they'll never look back at the Respironics.Okie bipap wrote: ↑Wed Apr 29, 2020 6:13 pmThe most highly recommended machine is the ResMed Auto For Her. That being said, you may be happier with the Dream Station auto machine since you are used to how Respironics machines operate. The ResMed machines handle events slightly different from the Respironics machines.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Snoozing Gonzo
- Posts: 199
- Joined: Wed Mar 08, 2006 2:11 pm
- Location: Western Montana
Re: AirSense Vs Dreamstation
You guys are great! Thinks for your time. As far as familiarity, I can go between my old Everest Cpap at the lake place, my older Transcend Apap with no humidifier,, and the System One without noticing too much - except the Transcend sounds like a jet engine. It's fine tuning, relaxation, and habit.
Thanks again. This helps.
Thanks again. This helps.
_________________
| Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Old Everest CPAP for at the cabin. Z2 for travel and backpacking |
"Breathe deep. Seek peace."
(James Gurney, Dinotopia)
(James Gurney, Dinotopia)
- DreamDiver
- Posts: 3082
- Joined: Thu Oct 04, 2007 11:19 am
Re: AirSense Vs Dreamstation
I started with Respironics with the M-Series Pro.Snoozing Gonzo wrote: ↑Wed Apr 29, 2020 11:54 amSecond, if I were to buy one of those two machines, which is the better in terms of programming and recognition, consistency, and durability/longevity(it won't be moved much)?
I then got a ResMed S9 and now have a ResMed AS10.
I'm a ResMed fanboy. Sorry, but ResMed's algorithm has it all over Respironics, at least for my instance. Respironics' algorithm is coarser in all respects.
That said:
- Yes, I agree that Respironics and Resmed are probably about to drop their latest machines, hence the desire to reduce the current inventory. Sale! It's likely because neither company has introduced anything in five/six years. < conspiracy theory > I wonder if they have intentionally coordinated drop cycles. < / conspiracy theory > The current models will probably be discontinued.
- Theorization - wild flights of fancy - speculation...
- It's possible the next version of both or either will be without any SD card. This would mean either Bluetooth (or preferably WiFi) compatibility would allow us to connect directly to OSCAR from our machines, or unfortunately, they'll try to remove all OSCAR compatibility from the equation, putting the power back into the hands of the DME's. We'll be stuck with apps like myAir that don't give a full story or allow for facilitated self-titration. Ideally, they would be working with OSCAR devs to streamline and insure OSCAR compatibility. Of course OSCAR devs aren't going to tell us if this is the case because ResMed and Respironics both want control over their respective drop dates.
- There may be trade-offs. If the new machines use Artificial Intelligence (AI), it may not be as easy to change settings because the machine is learning your specific needs and remembering and adapting them night after night. Neural networks are unfortunately a little like black box coding, in that even those who programmed them cannot always say exactly why or how the code works. It just does. Neural networks that allow safe external control through a clinician's menu may be more difficult to program safely than simply allowing the AI to take full control of your therapy. AI might require a fast internet connection if the computer at the heart of the machine isn't robust enough to do it alone. (For instance when you use google speech to text, it always works better when you're connected to the internet.) If AI is not part of the new crop of machines, then not allowing SD cards and destroying compatibility with OSCAR would just be a stab at getting patients to go once more through the DME tunnel of suffering.
- An AI machine could also allow for better on-the-fly diagnosis and resultant ideal proper therapy. Unfortunately, they could also set up some whackadoo business model where they make you pay for the special 2-month AI period, then lock your machine into a specific algorithm from the internet, rather than having AI continually monitor and control your settings for the same actual cost. I would not be in favor of this business model.
- I'm hoping if AI is part of the new therapy that they'll offer a machine similar to the AS10 Autoset that allows all the old settings like straight CPAP, Auto, and the new "Auto AI", which can be controlled from the clinician menu, or maybe the patient menu. If they intend to shut patients out of their own therapy, it may backfire, causing a boom in the Open CPAP open source community. It seems to have stalled for now because the current crop of machines are pretty incredible as they are. But intentionally taking control of therapy out of the hands of the patient may cause the open source hardware community to do with CPAP machines what OSCAR has done for the open source CPAP software community.
- ResMed works for me. It may not work for you. Respironics may be better for you.
- Under ideal circumstances, the DME's would rent you a loaner machine, switching out between ResMed and Respironics every week or two for a couple months. From there, you could decide which machine works best for you and buy the one that fits your specific requirements. With most DME's it really is more like buying shoes or pants where they throw you a pair that you have to wear regardless of fit. Kind of senseless.
- While I'm not placing bets, it's entirely possible Respironics could eclipse ResMed's capabilities with the next iteration of cpap. We'll have to wait and see.
Others may have very some good reasons why they prefer Respironics.
_________________
| Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions |
| Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF |

Most members of this forum are wonderful.
However, if you are the target of bullying on this forum, please consider these excellent alternative forums:
Apnea Board
Sleep Apnea Talk Forum
Free CPAP Advice
Be well,
Chris
Re: AirSense Vs Dreamstation
Diagnosed and prescribed a CPAP in 2005. Still using my original Reperonics Remstar 1018547 and humidifier 1005792.
_________________
| Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Started 2005 with a Remstar machine. Then a 2020 Dreamstation 1. Now a Resmed Aircurve ST |
- DreamDiver
- Posts: 3082
- Joined: Thu Oct 04, 2007 11:19 am
Re: AirSense Vs Dreamstation
Wow! Straight CPAP since 2005... That machine must be amazingly robust to still be going strong! Newer machines are designed to be replaced about every five years.
_________________
| Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions |
| Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF |

Most members of this forum are wonderful.
However, if you are the target of bullying on this forum, please consider these excellent alternative forums:
Apnea Board
Sleep Apnea Talk Forum
Free CPAP Advice
Be well,
Chris
- Dog Slobber
- Posts: 4331
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: AirSense Vs Dreamstation
Drawing the conclusion that CPAPs are *designed* to be replaced every five years, because insurance will pay for a new one every five years; would be like stating cars are *designed* to be replaced every three years because that's the standard time frame for most leases.DreamDiver wrote: ↑Thu Apr 30, 2020 7:53 am
Wow! Straight CPAP since 2005... That machine must be amazingly robust to still be going strong! Newer machines are designed to be replaced about every five years.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
- DreamDiver
- Posts: 3082
- Joined: Thu Oct 04, 2007 11:19 am
Re: AirSense Vs Dreamstation
I know, right? After twenty years, my truck is still going strong, but my wife's newer car is now saying "Motor Life Exceeded" after five years. So we have to get her a new car because it was clearly designed that way.Dog Slobber wrote: ↑Thu Apr 30, 2020 7:59 amDrawing the conclusion that CPAPs are *designed* to be replaced every five years, because insurance will pay for a new one every five years; would be like stating cars are *designed* to be replaced every three years because that's the standard time frame for most leases.DreamDiver wrote: ↑Thu Apr 30, 2020 7:53 am
Wow! Straight CPAP since 2005... That machine must be amazingly robust to still be going strong! Newer machines are designed to be replaced about every five years.
_________________
| Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions |
| Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF |
Last edited by DreamDiver on Thu Apr 30, 2020 8:04 am, edited 1 time in total.

Most members of this forum are wonderful.
However, if you are the target of bullying on this forum, please consider these excellent alternative forums:
Apnea Board
Sleep Apnea Talk Forum
Free CPAP Advice
Be well,
Chris
Re: AirSense Vs Dreamstation
No they're not, no more than filters are *designed to be replaced" after 2 weeks, or nasal cushions are "designed to be replaced" every 2 weeks, or water reservoirs are "designed to be replaced" every 6 months.DreamDiver wrote: ↑Thu Apr 30, 2020 7:53 amWow! Straight CPAP since 2005... That machine must be amazingly robust to still be going strong! Newer machines are designed to be replaced about every five years.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- DreamDiver
- Posts: 3082
- Joined: Thu Oct 04, 2007 11:19 am
Re: AirSense Vs Dreamstation
Thanks, paleriderpalerider wrote: ↑Thu Apr 30, 2020 8:04 amNo they're not, no more than filters are *designed to be replaced" after 2 weeks, or nasal cushions are "designed to be replaced" every 2 weeks, or water reservoirs are "designed to be replaced" every 6 months.DreamDiver wrote: ↑Thu Apr 30, 2020 7:53 amWow! Straight CPAP since 2005... That machine must be amazingly robust to still be going strong! Newer machines are designed to be replaced about every five years.![]()
_________________
| Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions |
| Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF |

Most members of this forum are wonderful.
However, if you are the target of bullying on this forum, please consider these excellent alternative forums:
Apnea Board
Sleep Apnea Talk Forum
Free CPAP Advice
Be well,
Chris
Re: AirSense Vs Dreamstation
tried to update my profile, Machine selections are incomplete. Respiromics Remstar.
_________________
| Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Started 2005 with a Remstar machine. Then a 2020 Dreamstation 1. Now a Resmed Aircurve ST |
Re: AirSense Vs Dreamstation
If you don't see your model machine in the equipment menu just manually add it to the comments section of your equipment profile.
The choices available in the menu are tied to current inventory at cpap.com and if the machine is an older discontinued model then you won't see it. No big deal. Just add it to the comments section.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.