RemStar Plus

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
CowboyDave
Posts: 14
Joined: Mon Nov 20, 2006 11:16 am
Location: Tennessee

RemStar Plus

Post by CowboyDave » Tue Feb 20, 2007 2:02 pm

Can someone please tell me how to increase the pressure on a RemStar Plus C flex CPAP. Thanks


User avatar
Wulfman
Posts: 12316
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Post by Wulfman » Tue Feb 20, 2007 3:14 pm

Dave,

Give this a try. You might want to copy to a file on your computer.

Den


REMstar Plus CPAP.

1. With the machine plugged into a wall outlet, unplug the cord from the back of the CPAP. Then, press and hold down the two user buttons below the display screen while plugging the cord back into the CPAP, and continue to hold the buttons down until the CPAP beeps. This will indicate that the CPAP is in the Therapy Setup Menu.

2. The display will first show the number of hours the machine has been used. To erase this total and begin again at zero, press and hold the ramp or humidifier button. An “X” will appear on the display. Hold the button down until the time change to “0” and the “X” disappears.

To skip this setting or continue, press the right user button.

3. The elevation setting will appear next. To change this setting, press the ramp or humidifier button until the correct setting appears. Here are the elevation settings: 1 = less than 2,500 ft.; 2 = 2,500 to 5,000 ft.; 3 = 5,001 to 7,500 ft.

To skip this setting or continue, press the right user button.

4. The therapy mode will appear, either CPAP or CFLEX. To change this setting, press the ramp or humidifier button until the correct setting appears.

To skip this setting or continue, press the right user button.

5. The CPAP pressure will appear. To change this setting, press the ramp or humidifier button until the correct setting appears.

To skip this setting or continue, press the right user button.

6. The CPAP fine adjustment setting will appear. This setting allows you to calibrate your CPAP pressure to a manometer. We suggest you do not try to adjust this setting.

To skip this setting or continue, press the right user button.

7. The CFLEX setting will appear if you selected the CFLEX mode in number 4 above. A setting of 1 offers the least drop in pressure, 3 offers the most. To change this setting, press the ramp or humidifier button until the correct setting appears.

To skip this setting or continue, press the right user button.

8. The ramp time will appear next. It may be set for 0 to 45 minutes in 5 minute increments. To change this setting, press the ramp or humidifier button until the correct setting appears.

To skip this setting or continue, press the right user button.

9. The ramp starting pressure will appear. To change this setting, press the ramp or humidifier button until the correct setting appears.

To skip this setting or continue, press the right user button.

10. The patient disconnect setting will appear. This sounds an alarm when there is a mask leak and turns the airflow off. 1 = On; 2 = Off. To change this setting, press the ramp or humidifier button until the correct setting appears.

There are no other settings available. Press the pressure start/stop button to exit the menu or press the right user button to scroll through the settings again.

(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

CowboyDave
Posts: 14
Joined: Mon Nov 20, 2006 11:16 am
Location: Tennessee

Post by CowboyDave » Tue Feb 20, 2007 3:52 pm

Thank you very much!!

David

Ron d.

Re: RemStar Plus

Post by Ron d. » Thu May 12, 2011 4:32 pm

thank you I was having the same problem it works for me