Hey All,
I recently had a really bad night where I had way too many OAs. I had 50 in the period of 1.5 hours. I actually went to the ER as I had a really bad lump in the throat feeling, this is why I only had 1.5 hours of sleep that night. My numbers had been okay up until then, but now I want to look at tweaking my settings a bit.
After this night in the ER I took my machine to the supply house to get it tested. I told them the EPAP didn't seem to be working, they said my BIPAP was acting more like a CPAP and my numbers looked fine except for that one night. I see they did make a change to the PS setting they raised it from 0 - 8, to 2-8. This pressure change has seemed to give me more CAs. I wonder if it is that whole blowing off the CO2 idea.
Attached is a good night before the pressure change:
A recent night after the pressure change:
I have been reading up on the forums and I will try to do more research. Any advice is greatly appreciated.
Thanks
BIPAP Recent PS increase leading to higher AHI numbers
-
timeforagorism
- Posts: 14
- Joined: Mon Oct 08, 2018 6:51 am
BIPAP Recent PS increase leading to higher AHI numbers
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ F10 Full Face Mask with Headgear |
| Additional Comments: 5 Years of BIPAP Therapy |
Aircurve 10 VAUTO
ResMed AirFit F20
5 Years of BIPAP. Switched from Dreamstation after recall.
ResMed AirFit F20
5 Years of BIPAP. Switched from Dreamstation after recall.
Re: BIPAP Recent PS increase leading to higher AHI numbers
Welcome to the forum.
Before the change in PS the max PS was a lot lower than it is now with the new PS minimum.
For some people more PS does cause the blowing off of too much carbon dioxide so the levels don't get high enough for the brain to send the signal to breathe.
The change in minimum PS from 0 to 2 is minimal but sometimes there is a fine line between what PS causes too much carbon dioxide wash out and what doesn't.
My suggestion would be to go back to the old settings. They worked well apparently except for maybe one night. I never suggest changing something based on a one off weird night.
And yes...when PS minimum is 0 the machine will function more like cpap/apap with a single pressure instead of 2 distinct pressures but that's not necessarily a bad thing.
If you want to keep the bilevel distinct pressure thing...limit the max PS to 4 and see if the centrals reduce.
Before the change in PS the max PS was a lot lower than it is now with the new PS minimum.
For some people more PS does cause the blowing off of too much carbon dioxide so the levels don't get high enough for the brain to send the signal to breathe.
The change in minimum PS from 0 to 2 is minimal but sometimes there is a fine line between what PS causes too much carbon dioxide wash out and what doesn't.
My suggestion would be to go back to the old settings. They worked well apparently except for maybe one night. I never suggest changing something based on a one off weird night.
And yes...when PS minimum is 0 the machine will function more like cpap/apap with a single pressure instead of 2 distinct pressures but that's not necessarily a bad thing.
If you want to keep the bilevel distinct pressure thing...limit the max PS to 4 and see if the centrals reduce.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
-
timeforagorism
- Posts: 14
- Joined: Mon Oct 08, 2018 6:51 am
Re: BIPAP Recent PS increase leading to higher AHI numbers
I will try setting my pressures back. I called the DME therapist yesterday and they said my report is fine. They must not even look at the real data because I have never had flow rate waves like this, and all these CAs. AHI was good this morning, "technically treated", but I felt worse.
I am having these really wide flow waves:
Then these really long CAs:
The Night overall:
This is kind of frustrating to feel worse and see the data all jacked up, and they keep saying you are doing fine, don't change anything.
I am having these really wide flow waves:
Then these really long CAs:
The Night overall:
This is kind of frustrating to feel worse and see the data all jacked up, and they keep saying you are doing fine, don't change anything.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ F10 Full Face Mask with Headgear |
| Additional Comments: 5 Years of BIPAP Therapy |
Aircurve 10 VAUTO
ResMed AirFit F20
5 Years of BIPAP. Switched from Dreamstation after recall.
ResMed AirFit F20
5 Years of BIPAP. Switched from Dreamstation after recall.