is AHI under 15 a reasonable goal?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
lorenadam
Posts: 8
Joined: Mon Oct 02, 2017 10:14 am

is AHI under 15 a reasonable goal?

Post by lorenadam » Sat Oct 14, 2017 7:34 pm

my AHI is very rarely below 5 which I understand to be the goal, I notice when I check my results every morning that when I get an AHI greater than 5 such as 14.2 last night, that the res med formula gives me an AHI score of 2 out of 5; this leads me to conclude the goal o a 5 AHI might be the gold standard, but that an AHI under 15 or ok?

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: is AHI under 15 a reasonable goal?

Post by TedVPAP » Sat Oct 14, 2017 7:41 pm

lorenadam wrote:my AHI is very rarely below 5 which I understand to be the goal, I notice when I check my results every morning that when I get an AHI greater than 5 such as 14.2 last night, that the res med formula gives me an AHI score of 2 out of 5; this leads me to conclude the goal o a 5 AHI might be the gold standard, but that an AHI under 15 or ok?
Welcome to the forum. Below 5 is considered by the medical community to be adequate. I feel best when my AHI is below 2. My average is ~0.5. How do you feel?
Add your equipment description to your profile. If you machine is data capable, the three links below will show you how to examine your raw data and share it so knowledgeable people can help guide you to better treatment.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: is AHI under 15 a reasonable goal?

Post by Julie » Sat Oct 14, 2017 8:02 pm

No, it's not ok. You may be having large leak or other problems, but without more to go on, it's hard to say. What is your equipment (specific if possible) and your pressure settings?

User avatar
kteague
Posts: 7784
Joined: Tue May 16, 2006 8:30 pm
Location: West and Midwest

Re: is AHI under 15 a reasonable goal?

Post by kteague » Sat Oct 14, 2017 8:02 pm

No, an AHI of under 15 is not a reasonable goal. Please do as has been suggested regarding getting the software to check your data so you can get some direction on how to improve your treatment. Any number of things cound be going on. If I mask up then end up in a conversation, it will show up as clusters of events. If you're awake a lot it could affect your data. Leaks can too. For some people the treatment triggers centrals. It would be in your best interest to use the data to hone it on how to imrove your treatment. My first inclination was to say increase your pressure, till I thought about centrals need to be ruled out first. Until you can access the data, is there any more info on your machine's screen? Good luck with sorting things out and getting on to successful treatment. It may turn out a small tweak in your settings is all you need.

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Bleep/DreamPort for full nights, Tap Pap for shorter sessions

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: is AHI under 15 a reasonable goal?

Post by Goofproof » Sat Oct 14, 2017 8:11 pm

Only if you don't like a better life, by this time you should no about sleep apnea, under 5 AHI is the government goal under 2 AHI is what we strive for. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
Bill44133
Posts: 1087
Joined: Tue Jul 09, 2013 8:34 pm
Location: North Royalton, OH

Re: is AHI under 15 a reasonable goal?

Post by Bill44133 » Sat Oct 14, 2017 8:13 pm

NO that is NOT reasonable. remember that means you stopped breathing for 10 seconds or longer. Not good.

Take 15 minutes and review this youtube video:

https://www.youtube.com/watch?v=-gie2dhqP2c

Then you will understand why you don't want 15 AHI.

I wish you weil

_________________
Machine: DreamStation BiPAP® Auto Machine
Humidifier: DreamStation Heated Humidifier
Additional Comments: Settings are IPap 23 EPap 19

User avatar
zoocrewphoto
Posts: 3732
Joined: Mon Apr 30, 2012 10:34 pm
Location: Seatac, WA

Re: is AHI under 15 a reasonable goal?

Post by zoocrewphoto » Sat Oct 14, 2017 8:39 pm

An ahi of 5 means that you average a breathing problem every 12 minutes. With 8 hours of sleep, that's 40 disturbances a night! Doesn't sound too good, huh? And that is what the doctors want us to aim for. Most people feel better at 2 or under, and that is quite reasonable to expect with optimized settings. I rarely go over 1.5.

Thinking you are okay by getting below 15 is dooming yourself to still crappy sleep. An ahi of 14 is still being disturbed every 4 minutes. All night long.

That may be much better than untreated sleep apnea, but it is still crap, and it will still cause damage and make you feel like crap.

Let us help you optimize your settings. Tell us your machine (brand AND model name). Also, what are your current pressure settings? Have you heard about sleeyhead? It is a free software that you can use to see tour machine's data. If you post some graphs of a typical night, we can help you dial in on better settings.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 autoset pressure range 11-17
Who would have thought it would be this challenging to sleep and breathe at the same time?

User avatar
49er
Posts: 5624
Joined: Mon Jan 16, 2012 8:18 am

Re: is AHI under 15 a reasonable goal?

Post by 49er » Sun Oct 15, 2017 3:54 am

xxyzx wrote:
Bill44133 wrote:NO that is NOT reasonable. remember that means you stopped breathing for 10 seconds or longer. Not good.

Take 15 minutes and review this youtube video:

https://www.youtube.com/watch?v=-gie2dhqP2c

Then you will understand why you don't want 15 AHI.

I wish you weil
========

not true

it includes hypops
that are just reduced airflow for ten seconds

desats and RDI mean a lot more than AHI
Any reduced airflow which includes hypopneas is not good.

And as I have mentioned to you before, a person who has sleep apnea may not have O2 desats. So telling them that is more significant is simply incorrect information.

I do agree that the RDI is a more helpful statistic than AHI since it also measures RERAS which restrict airflow. I am not up to date on which machines measure that so I will defer to someone more knowledgeable than me on this situation.

But meanwhile, the OP should follow zoocrew's advice regarding optimizing the settings. An AHI under 15 is totally unacceptable.

lorenadam
Posts: 8
Joined: Mon Oct 02, 2017 10:14 am

Add your equipment description to your profile?

Post by lorenadam » Sun Oct 15, 2017 9:30 am

how do you do this?

User avatar
Pugsy
Posts: 65255
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: is AHI under 15 a reasonable goal?

Post by Pugsy » Sun Oct 15, 2017 9:45 am

lorenadam wrote:how do you do this?
Do what?
Reduce the AHI?

First you start by using the available software and figure out exactly which event category is elevated and then devise a plan to reduce those events...usually by more pressure but since some events aren't fixable with more pressure you first have to figure out what the problem is and then come up with a plan of attack.

Do you know about the available software?
https://sleep.tnet.com/equipment

If you want member input about the software detailed results you are seeing then we have to see what you are seeing.
How to post images
https://sleep.tnet.com/reference/tips/imgur

How to organize the graphs (we don't need all of them)
https://sleep.tnet.com/resources/sleepyhead/shorganize

Examples seen here
viewtopic/t103468/Need-help-with-screen-shots.html

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Arlene1963
Posts: 548
Joined: Thu Nov 05, 2015 5:43 am

Re: is AHI under 15 a reasonable goal?

Post by Arlene1963 » Sun Oct 15, 2017 9:53 am

To add equipment to your profile click on User Control Panel (under the Search Box at the top of the forum page) and then choose Profile. After that it is pretty straight forward. Many prefer to use the text display rather than a visual for the equipment.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: is AHI under 15 a reasonable goal?

Post by Julie » Sun Oct 15, 2017 10:12 am

If the results of all that come down to how to change your pressure settings, we can easily help you do that quickly once we know what equipment you have now.

User avatar
49er
Posts: 5624
Joined: Mon Jan 16, 2012 8:18 am

Re: is AHI under 15 a reasonable goal?

Post by 49er » Sun Oct 15, 2017 11:20 am

xxyxz - ""if there are no desats then apnea did not happen
and if there are no desats then hypops are not a problem unless they cause arousals

so forget AHi fix the spo2 first
then stop arousals

but fix the real problem not the symptoms
and dont worry about AHI which is a statistic created for medicrap bureaucrats to decide objectively if you should get a cpap""

49er - https://singularsleep.com/blogs/news/ah ... apnea-test

""Apneas are a type of abnormal respiratory event that occur during sleep. The word originates from the Greek roots meaning "without" and "to breathe." The American Academy of Sleep Medicine (AASM) provides a scoring manual that delineates how all aspects of a sleep study should be interpreted, including defining what constitutes abnormal respiratory events such as apneas. To score an apnea on a sleep study there must be essentially no airflow (measured through a nasal pressure sensor or thermistor) for at least ten seconds. A blood oxygen drop, known as an oxygen desaturation, is not part of the criteria for an apnea. There are three different types of apneas described in this manual:""

49er

PS - Sorry folks for the confusing format. I am having trouble with the quote function again on my computer.

User avatar
TASmart
Posts: 1071
Joined: Wed Jan 04, 2017 7:23 pm
Location: Eugene, OR

Re: is AHI under 15 a reasonable goal?

Post by TASmart » Sun Oct 15, 2017 1:45 pm

You most certainly can have apneas without desaturations. Sleep deprivation can be fatal just as much as desaturations just a different mechanism of action.
All posts reflect my own opinion based on my experience and reading.
Your mileage may vary
Past performance is no guarantee of future results
Consult with your own physician as people very

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: is AHI under 15 a reasonable goal?

Post by palerider » Sun Oct 15, 2017 2:06 pm

lorenadam wrote:my AHI is very rarely below 5 which I understand to be the goal, I notice when I check my results every morning that when I get an AHI greater than 5 such as 14.2 last night, that the res med formula gives me an AHI score of 2 out of 5; this leads me to conclude the goal o a 5 AHI might be the gold standard, but that an AHI under 15 or ok?
A goal of 1.5 or under is what I'd call 'good enough'.

An AHI of 5 is like being poked with a pointy stick an average of every 12 minutes all night long.

That's nothing more than a series of naps.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Fri Jan 17, 2020 10:52 am, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.