So last nite was my first time with my new Airsense 10 auto.... the settings were identical with my 5 year old Fisher Paykel 200 auto...
on several occasions i would hear myself snore as i feel asleep - which kept me awake. this NEVER happens with my Fisher-P.. In fact i switched to my Fisher in the middle of the nite and the problem stopped....
Any thoughts/suggestions ( I will increase my baseline setting)
thanks
matt
Airsense 10 Auto -- Snored with First Use!
Re: Airsense 10 Auto -- Snored with First Use!
The 2 brands go about their jobs a little differently so the same pressure setting on one may not work exactly the same as on the other especially if using auto adjusting more. The algorithms that dictate machine response are just different.
Increasing the minimum pressure (that baseline) pressure is what I would do in this situation.
Increasing the minimum pressure (that baseline) pressure is what I would do in this situation.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Airsense 10 Auto -- Snored with First Use!
Snoring is a indicator of A apnea maybe going to occur.
Auto ramp pressure could be set to low or for two long is what I'm thinking. I use a 5 minute ramp starting at 6cm/o2 if that might help.
Auto ramp pressure could be set to low or for two long is what I'm thinking. I use a 5 minute ramp starting at 6cm/o2 if that might help.
Re: Airsense 10 Auto -- Snored with First Use!
Hi. I'm the OP....
I turned up my baseline from 6 to 7.
Plus i reduced the EPR (that which allows the pressure to reduce when exhaling) from 3 to 1.
I think I was reducing the exhale pressure too much at the 3 setting -- at 1 it more closely matched what my Fisher Paykel was like.
Problem seems to have gone away.
Thanks for the suggestions!
I turned up my baseline from 6 to 7.
Plus i reduced the EPR (that which allows the pressure to reduce when exhaling) from 3 to 1.
I think I was reducing the exhale pressure too much at the 3 setting -- at 1 it more closely matched what my Fisher Paykel was like.
Problem seems to have gone away.
Thanks for the suggestions!