Advise Pressure Setting and Mode

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
wolf911
Posts: 35
Joined: Fri Apr 10, 2015 8:07 am

Advise Pressure Setting and Mode

Post by wolf911 » Mon May 18, 2015 10:17 am

I will be getting my RESMED S10 Autoset today and masks (ordered a Swift and a Nuance pillows) tomorrow and wanted to just get some advise on how i should set the pressure settings. I have my tritration sleep study summary below. Im thinking set it for 7 min and 9 max on auto or should i just go straight 8 CPAP. what are your thoughts on these starting positions. i do sleep on my back almost all the time but occationally on my side.
Also any RESMED s10 users whats a good starting position for the humidity and heat or should i just start with the auto set. im sure ill figure these settings out but just looking for somewhere to start.

INTERPRETATION
this is a cpap titration overnight performed in a sleep laboratory under the direct supervision of a sleep technician. Total recording time 418 minutes, total sleep time is 342 minutes, and sleep efficiency is 82%. Latency to sleep onset is 19 minutes, latency to REM is 81 minutes, and wake after sleep is 56 minutes all normal.

SLEEP ARCHITECTURE
The hypnogram demenstrates 13 arousals, 3 REM episodes, and 7 awakenings. The patient spent in stage N1 6.4% of the time, stage N2 63.8% of the time, stage N3 15.5% of the time, and stage REM 14.3% of the time.

POSITIONAL SUMMARY
The patient spent 100% tme supine.

AROUSALS
There were 13 arousals in total: 1 due to apnea/hypopneas, 5 due to snoring, and 6 spontaneous.

MOTOR EVENTS
There were no motor events.

HEART RATE ANALYSIS
The average heart rate 61 and the nadir 53. No arrthymias were seen.

CPAP TRITRATION
CPAP tritration was started at 4 and tritrated all the way up to 7. At 7 acually the patient had 2 hypopneas with an apnea/hyponea index of 3. CPAP between 7 and 8 is the recomended. Would be better 8 if the patient can tolerate with a Nuance Pillow mask, no spplemental oxygen will be needed.

IMPRESSION
This is a patient with severe obstructive sleep apnea with an apnea/hypopnea index of 30.8. This is a CPAP tritration polysomnography. It is technically adequate, the patient had enough REM, supine, and would treat, although with CPAP 7 the index is acceptable at 3, my preference would be 8 if the patient can tolerate it with Nuance pillow mask. No spplemental oxygen needed.

User avatar
LSAT
Posts: 13406
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Advise Pressure Setting and Mode

Post by LSAT » Mon May 18, 2015 10:20 am

Wherever you are getting your equipment from should set the machine to the pressure indicated on your prescription.

User avatar
wolf911
Posts: 35
Joined: Fri Apr 10, 2015 8:07 am

Re: Advise Pressure Setting and Mode

Post by wolf911 » Mon May 18, 2015 10:26 am

LSAT wrote:Wherever you are getting your equipment from should set the machine to the pressure indicated on your prescription.
im on my own on this

User avatar
Pugsy
Posts: 65360
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Advise Pressure Setting and Mode

Post by Pugsy » Mon May 18, 2015 10:50 am

Base on what the titration report says....I think I would start with a modest apap range of 7 minimum and 9 maximum and see what happens....or split the difference if you want to use cpap mode and do 7.5 cm and see what happens.

EPR use ...your choice or preference...and see what happens on the reports. May or may not need to adjust the pressure a little to compensate for EPR reduction (if you elect to use it).

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
wolf911
Posts: 35
Joined: Fri Apr 10, 2015 8:07 am

Re: Advise Pressure Setting and Mode

Post by wolf911 » Mon May 18, 2015 3:29 pm

Pugsy wrote:Base on what the titration report says....I think I would start with a modest apap range of 7 minimum and 9 maximum and see what happens....or split the difference if you want to use cpap mode and do 7.5 cm and see what happens.

EPR use ...your choice or preference...and see what happens on the reports. May or may not need to adjust the pressure a little to compensate for EPR reduction (if you elect to use it).
if ive never used EPR or any CPAP (except the one night in study) should i start with or without EPR. which would be better to get a base to start from. in the sleep study do you think they use EPR. i didnt feel the pressure much falling asleep but in the morning i could sense the pressure was much higher than when i had started

machine came today, looks new, just as stated. masks arrive tomorrow. good thing this forum exists.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Advise Pressure Setting and Mode

Post by Julie » Mon May 18, 2015 3:37 pm

EPR is generally used for people who are conscious of difficulty exhaling (consistently), and you'd only set it - to whatever level suits you - if you have a problem... some do, some don't.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Advise Pressure Setting and Mode

Post by Wulfman... » Mon May 18, 2015 3:38 pm

wolf911 wrote:I will be getting my RESMED S10 Autoset today and masks (ordered a Swift and a Nuance pillows) tomorrow and wanted to just get some advise on how i should set the pressure settings. I have my tritration sleep study summary below. Im thinking set it for 7 min and 9 max on auto or should i just go straight 8 CPAP. what are your thoughts on these starting positions. i do sleep on my back almost all the time but occationally on my side.
Also any RESMED s10 users whats a good starting position for the humidity and heat or should i just start with the auto set. im sure ill figure these settings out but just looking for somewhere to start.

INTERPRETATION
this is a cpap titration overnight performed in a sleep laboratory under the direct supervision of a sleep technician. Total recording time 418 minutes, total sleep time is 342 minutes, and sleep efficiency is 82%. Latency to sleep onset is 19 minutes, latency to REM is 81 minutes, and wake after sleep is 56 minutes all normal.

SLEEP ARCHITECTURE
The hypnogram demenstrates 13 arousals, 3 REM episodes, and 7 awakenings. The patient spent in stage N1 6.4% of the time, stage N2 63.8% of the time, stage N3 15.5% of the time, and stage REM 14.3% of the time.

POSITIONAL SUMMARY
The patient spent 100% tme supine.

AROUSALS
There were 13 arousals in total: 1 due to apnea/hypopneas, 5 due to snoring, and 6 spontaneous.

MOTOR EVENTS
There were no motor events.

HEART RATE ANALYSIS
The average heart rate 61 and the nadir 53. No arrthymias were seen.

CPAP TRITRATION
CPAP tritration was started at 4 and tritrated all the way up to 7. At 7 acually the patient had 2 hypopneas with an apnea/hyponea index of 3. CPAP between 7 and 8 is the recomended. Would be better 8 if the patient can tolerate with a Nuance Pillow mask, no spplemental oxygen will be needed.

IMPRESSION
This is a patient with severe obstructive sleep apnea with an apnea/hypopnea index of 30.8. This is a CPAP tritration polysomnography. It is technically adequate, the patient had enough REM, supine, and would treat, although with CPAP 7 the index is acceptable at 3, my preference would be 8 if the patient can tolerate it with Nuance pillow mask. No spplemental oxygen needed.
That's where I would start.........7 or 8 centimeters.
I would probably not advise using exhale relief (EPR) at those low pressures. They wouldn't have used it in your titration and the pressure drop on exhale can actually affect the results in actual use.

Use some software and monitor your data.......and adjust if necessary.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Pugsy
Posts: 65360
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Advise Pressure Setting and Mode

Post by Pugsy » Mon May 18, 2015 3:47 pm

EPR is your personal choice or preference. Some people think that 7 cm feels like a tornado and they can't handle exhaling against it and others think "what pressure" so you just need to play with it and see what feels best for you.
I doubt that any sort of exhale relief would have been used at the titration study but we don't know for sure...anyway you are home now and you can use it or not depending on what feels the best to you.
Exhale relief and to use it or how much is up to the person doing the using. I don't have to use it but it sure is nice so my preference is to use whatever exhale relief is available. It feels more like natural breathing rhythm and for me that means more "comfort" and more comfort means better sleep for me. Others don't like the way it "feels" and prefer to not use any.

You will find that pretty much all this cpap stuff comes with a really big YMMV sticker and you have to sort it out yourself...like humidity needs. What my nose likes (Palerider thinks I am a swamp monster because I can't get too much humidity) your nose might hate and again it's personal preference.
When in doubt...pick a middle of the road setting and start from there.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
wolf911
Posts: 35
Joined: Fri Apr 10, 2015 8:07 am

Re: Advise Pressure Setting and Mode

Post by wolf911 » Mon May 18, 2015 3:49 pm

.[/quote]
That's where I would start.........7 or 8 centimeters.
I would probably not advise using exhale relief (EPR) at those low pressures. They wouldn't have used it in your titration and the pressure drop on exhale can actually affect the results in actual use.

Use some software and monitor your data.......and adjust if necessary.


Den

.[/quote]
7 or 8 in CPAP mode?

i will get the software and start monitoring

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Advise Pressure Setting and Mode

Post by Wulfman... » Mon May 18, 2015 4:06 pm

wolf911 wrote:7 or 8 in CPAP mode?
i will get the software and start monitoring
Yes.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

patrick112
Posts: 1
Joined: Tue May 19, 2015 8:30 pm

Re: Advise Pressure Setting and Mode

Post by patrick112 » Tue May 19, 2015 8:37 pm

What software do you recommend Den?

User avatar
Bill44133
Posts: 1087
Joined: Tue Jul 09, 2013 8:34 pm
Location: North Royalton, OH

Re: Advise Pressure Setting and Mode

Post by Bill44133 » Tue May 19, 2015 8:44 pm

patrick112 wrote:What software do you recommend Den?
Most of the people use SleepyHead.. It is free and tutorial links are on the top of the start page here.

_________________
Machine: DreamStation BiPAP® Auto Machine
Humidifier: DreamStation Heated Humidifier
Additional Comments: Settings are IPap 23 EPap 19

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Advise Pressure Setting and Mode

Post by Wulfman... » Tue May 19, 2015 9:03 pm

Bill44133 wrote:
patrick112 wrote:What software do you recommend Den?
Most of the people use SleepyHead.. It is free and tutorial links are on the top of the start page here.
Yeah, what Bill said.
You don't have your machine in your profile, Patrick, but if it's a newer (data-capable) ResMed or P/R machine, then the Sleepyhead software would probably be the way to go. Some still use the ResScan software for ResMed machines and some still use the various Encore software versions for the P/R machines, but sometimes it's a little more problematic to get hold of those. Sleepyhead is readily available.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
wolf911
Posts: 35
Joined: Fri Apr 10, 2015 8:07 am

Re: Advise Pressure Setting and Mode

Post by wolf911 » Wed May 20, 2015 9:07 am

i took a screenshot from sleepyhead and followed the tutorial ON HOW TO upload to imgur(was succesful) but when i click large thumbnail and Linked BBcode(as the tutorial says) and then copy and paste the link to the forum it shows up full size not as a thumbnail. any one know what im missing.

User avatar
Pugsy
Posts: 65360
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Advise Pressure Setting and Mode

Post by Pugsy » Wed May 20, 2015 9:59 am

wolf911 wrote:i took a screenshot from sleepyhead and followed the tutorial ON HOW TO upload to imgur(was succesful) but when i click large thumbnail and Linked BBcode(as the tutorial says) and then copy and paste the link to the forum it shows up full size not as a thumbnail. any one know what im missing.
Large thumbnail is a pretty big thumbnail and often is big enough that we don't need to click on it to see further details.

Put an image here using the large thumbnail code (last line in the box of addresses) and let's see what happens.
It's really kinda nice to have a large thumbnail image that doesn't always require clicking on it to see the details.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.