Can higher than needed pressure cause Hypopnea Apena

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
fdw
Posts: 260
Joined: Wed Aug 27, 2014 7:08 am
Location: D/FW TEXAS

Can higher than needed pressure cause Hypopnea Apena

Post by fdw » Sun Feb 15, 2015 5:38 pm

As I've understood Hypopnea's they occur when the airway partially collapses so an increase in pressure would be needed.

Is it possible that higher than needed pressure can cause a Hypopnea Apnea? I would thank not but wanted to ask.

Example: If you normally have EPR on at a 1 and then go to 0.
ResMed AirSense S10 Autoset (Ins. Pur. 3/16/2016)
AutoCPAP 10-15 EPR 1.0 (Latest Sleep Study 3-7-16)
Started CPAP March 1995 with a pressure of 11.0

User avatar
BlackSpinner
Posts: 9742
Joined: Sat Apr 25, 2009 5:44 pm
Location: Edmonton Alberta
Contact:

Re: Can higher than needed pressure cause Hypopnea Apena

Post by BlackSpinner » Sun Feb 15, 2015 5:48 pm

No but higher pressures can cause central apnea in some people.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: Quatro mask for colds & flus S8 elite for back up
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Can higher than needed pressure cause Hypopnea Apena

Post by Wulfman... » Sun Feb 15, 2015 5:54 pm

fdw wrote:As I've understood Hypopnea's they occur when the airway partially collapses so an increase in pressure would be needed.

Is it possible that higher than needed pressure can cause a Hypopnea Apnea? I would thank not but wanted to ask.

Example: If you normally have EPR on at a 1 and then go to 0.
I'm not sure if I follow that, but.......
With regard to EPR, it's been often discussed on the forum that IF the point of the end of exhale and beginning of inhale is a source for apneas, then the EPR setting CAN affect the effectiveness of the pressure setting. Or, said another way, if you NEED a pressure of 10 cm. to keep your airway open and your pressure is 12 and your EPR setting is "3", your pressure will fall below the needed pressure at the point of exhale.......and you COULD experience apneas at that point.
So, the general advice has been to either increase the pressure setting or reduce the EPR setting.

With regard to whether higher (than needed) pressures can cause events........"Maybe", but more than likely, "probably not".
Same answer for "pressure induced Centrals". Supposedly, only about 15% of the users are actually susceptible to those.

I believe that with most clinical titrations, they try to find the "sweet spot" where the majority of the events are eliminated with the minimum amount of pressure. Why would anyone want to experience an excessive amount of pressure if you don't actually need it?


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Can higher than needed pressure cause Hypopnea Apena

Post by palerider » Sun Feb 15, 2015 6:10 pm

fdw wrote:Is it possible that higher than needed pressure can cause a Hypopnea Apnea? I would thank not but wanted to ask.

Example: If you normally have EPR on at a 1 and then go to 0.
that would make hypops or obstructives LESS likely.

also, hypopneas and apneas are different things.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
fdw
Posts: 260
Joined: Wed Aug 27, 2014 7:08 am
Location: D/FW TEXAS

Re: Can higher than needed pressure cause Hypopnea Apena

Post by fdw » Sun Feb 15, 2015 6:25 pm

Wulfman... wrote:
fdw wrote:As I've understood Hypopnea's they occur when the airway partially collapses so an increase in pressure would be needed.

Is it possible that higher than needed pressure can cause a Hypopnea Apnea? I would thank not but wanted to ask.

Example: If you normally have EPR on at a 1 and then go to 0.
I'm not sure if I follow that, but.......
With regard to EPR, it's been often discussed on the forum that IF the point of the end of exhale and beginning of inhale is a source for apneas, then the EPR setting CAN affect the effectiveness of the pressure setting. Or, said another way, if you NEED a pressure of 10 cm. to keep your airway open and your pressure is 12 and your EPR setting is "3", your pressure will fall below the needed pressure at the point of exhale.......and you COULD experience apneas at that point.
So, the general advice has been to either increase the pressure setting or reduce the EPR setting.

With regard to whether higher (than needed) pressures can cause events........"Maybe", but more than likely, "probably not".
Same answer for "pressure induced Centrals". Supposedly, only about 15% of the users are actually susceptible to those.

I believe that with most clinical titrations, they try to find the "sweet spot" where the majority of the events are eliminated with the minimum amount of pressure. Why would anyone want to experience an excessive amount of pressure if you don't actually need it?

Den

.
So in my case. I'm on straight CPAP 13.6 with EPR 1, last official titration 13.0 (8 years ago). If I turn off EPR should I just leave my pressure alone? The reason I'm asking is because my AHI numbers look great. However sometime I'll have Hypopnena lasting up to 30 plus seconds long......like AHI of 0.60 but time in Apena 2 plus minutes.
ResMed AirSense S10 Autoset (Ins. Pur. 3/16/2016)
AutoCPAP 10-15 EPR 1.0 (Latest Sleep Study 3-7-16)
Started CPAP March 1995 with a pressure of 11.0

User avatar
LSAT
Posts: 13353
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Can higher than needed pressure cause Hypopnea Apena

Post by LSAT » Sun Feb 15, 2015 6:29 pm

[/quote]So in my case. I'm on straight CPAP 13.6 with EPR 1, last official titration 13.0 (8 years ago). If I turn off EPR should I just leave my pressure alone? The reason I'm asking is because my AHI numbers look great. However sometime I'll have Hypopnena lasting up to 30 plus seconds long......like AHI of 0.60 but time in Apena 2 plus minutes.[/quote]

Personally I would not change a thing....a 30 second hypopnea is no big deal

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Can higher than needed pressure cause Hypopnea Apena

Post by Wulfman... » Sun Feb 15, 2015 6:32 pm

fdw wrote:So in my case. I'm on straight CPAP 13.6 with EPR 1, last official titration 13.0 (8 years ago). If I turn off EPR should I just leave my pressure alone? The reason I'm asking is because my AHI numbers look great. However sometime I'll have Hypopnena lasting up to 30 plus seconds long......like AHI of 0.60 but time in Apena 2 plus minutes.
Theoretically, yes.
However, that may or may not account for or prevent "positional" events. (sleeping supine, chin tucked down to the chest, and any number of other possibilities)
If turning off EPR doesn't fix it, bumping up one (or more) centimeter(s) might be in order.

Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
fdw
Posts: 260
Joined: Wed Aug 27, 2014 7:08 am
Location: D/FW TEXAS

Re: Can higher than needed pressure cause Hypopnea Apena

Post by fdw » Sun Feb 15, 2015 6:43 pm

Wulfman... wrote:
fdw wrote:So in my case. I'm on straight CPAP 13.6 with EPR 1, last official titration 13.0 (8 years ago). If I turn off EPR should I just leave my pressure alone? The reason I'm asking is because my AHI numbers look great. However sometime I'll have Hypopnena lasting up to 30 plus seconds long......like AHI of 0.60 but time in Apena 2 plus minutes.
Theoretically, yes.
However, that may or may not account for or prevent "positional" events. (sleeping supine, chin tucked down to the chest, and any number of other possibilities)
If turning off EPR doesn't fix it, bumping up one (or more) centimeter(s) might be in order.

Den

.
Great....I think I'll try it for a few night just to see how i feel.
ResMed AirSense S10 Autoset (Ins. Pur. 3/16/2016)
AutoCPAP 10-15 EPR 1.0 (Latest Sleep Study 3-7-16)
Started CPAP March 1995 with a pressure of 11.0

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Can higher than needed pressure cause Hypopnea Apena

Post by Wulfman... » Sun Feb 15, 2015 7:00 pm

fdw wrote:
Wulfman... wrote:
fdw wrote:So in my case. I'm on straight CPAP 13.6 with EPR 1, last official titration 13.0 (8 years ago). If I turn off EPR should I just leave my pressure alone? The reason I'm asking is because my AHI numbers look great. However sometime I'll have Hypopnena lasting up to 30 plus seconds long......like AHI of 0.60 but time in Apena 2 plus minutes.
Theoretically, yes.
However, that may or may not account for or prevent "positional" events. (sleeping supine, chin tucked down to the chest, and any number of other possibilities)
If turning off EPR doesn't fix it, bumping up one (or more) centimeter(s) might be in order.

Den

.
Great....I think I'll try it for a few night just to see how i feel.
I've been using straight 12 and 13 for the last couple of years with no exhale relief. I used a C-Flex setting of "2" for years and then turned it off after a few nap "trials" and found out I didn't really need it.......and wondered why I waited so long to turn it off.

Good luck with your experiment.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
fdw
Posts: 260
Joined: Wed Aug 27, 2014 7:08 am
Location: D/FW TEXAS

Re: Can higher than needed pressure cause Hypopnea Apena

Post by fdw » Mon Feb 16, 2015 6:45 am

Theoretically, yes.
If turning off EPR doesn't fix it, bumping up one (or more) centimeter(s) might be in order.

Den

Great....I think I'll try it for a few night just to see how i feel.
I've been using straight 12 and 13 for the last couple of years with no exhale relief. I used a C-Flex setting of "2" for years and then turned it off after a few nap "trials" and found out I didn't really need it.......and wondered why I waited so long to turn it off.

Good luck with your experiment.

Den


Den, So you change your pressure from 12 to 13 (back and forth), how do you know when it's time to change?

Last night. Not what I was hoping to see on my report this morning. Looks like the Hypopnea's nearly doubled with the longest at 44 seconds and shortest 18 seconds. Flow Limitations also doubled from previous nights. And finally Time in Apnea was nearly 4 minutes. I feel fine, but will see as the day goes on.

Den, You think I Should wait a few more days before increasing the pressure?

Image
Last edited by fdw on Mon Feb 16, 2015 8:37 am, edited 3 times in total.
ResMed AirSense S10 Autoset (Ins. Pur. 3/16/2016)
AutoCPAP 10-15 EPR 1.0 (Latest Sleep Study 3-7-16)
Started CPAP March 1995 with a pressure of 11.0

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Can higher than needed pressure cause Hypopnea Apena

Post by Julie » Mon Feb 16, 2015 7:15 am

You feel good, your AHI's under 1, stop looking at every number and wait for a weekly wrap-up. Don't get obsessed with every little change overnight on every graph.

User avatar
fdw
Posts: 260
Joined: Wed Aug 27, 2014 7:08 am
Location: D/FW TEXAS

Re: Can higher than needed pressure cause Hypopnea Apena

Post by fdw » Mon Feb 16, 2015 8:01 am

Julie wrote:You feel good, your AHI's under 1, stop looking at every number and wait for a weekly wrap-up. Don't get obsessed with every little change overnight on every graph.
Im sorry Julie....I don't want to come off as rude, a know it all, or un-apprecitave. But my question was directed to Den. I certinally value others (yours) opinions and recommendations.

Please forgive me Julie if Im out of line!
ResMed AirSense S10 Autoset (Ins. Pur. 3/16/2016)
AutoCPAP 10-15 EPR 1.0 (Latest Sleep Study 3-7-16)
Started CPAP March 1995 with a pressure of 11.0

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Can higher than needed pressure cause Hypopnea Apena

Post by palerider » Mon Feb 16, 2015 9:19 am

Julie wrote:You feel good, your AHI's under 1, stop looking at every number and wait for a weekly wrap-up. Don't get obsessed with every little change overnight on every graph.
ah, the obsessive quest for perfect and it's attendant spiral into madness.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Can higher than needed pressure cause Hypopnea Apena

Post by Julie » Mon Feb 16, 2015 9:32 am

Not out of line - you have a right to talk to anyone - good luck with all.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Can higher than needed pressure cause Hypopnea Apena

Post by Wulfman... » Mon Feb 16, 2015 2:09 pm

fdw wrote:
Theoretically, yes.
If turning off EPR doesn't fix it, bumping up one (or more) centimeter(s) might be in order.

Den

Great....I think I'll try it for a few night just to see how i feel.
I've been using straight 12 and 13 for the last couple of years with no exhale relief. I used a C-Flex setting of "2" for years and then turned it off after a few nap "trials" and found out I didn't really need it.......and wondered why I waited so long to turn it off.

Good luck with your experiment.

Den


Den, So you change your pressure from 12 to 13 (back and forth), how do you know when it's time to change?

Last night. Not what I was hoping to see on my report this morning. Looks like the Hypopnea's nearly doubled with the longest at 44 seconds and shortest 18 seconds. Flow Limitations also doubled from previous nights. And finally Time in Apnea was nearly 4 minutes. I feel fine, but will see as the day goes on.

Den, You think I Should wait a few more days before increasing the pressure?
Looks really good to me. I'd let it ride for awhile. You're very close to "perfection". The numbers and types of events can vary from night to night and you can choose to watch those for any trends........and then you can decide whether to tweak them or not. I know, for me, at this time of the year my AHI numbers can be just a bit higher for some reason.

This is a little complicated, but........ With regard to my pressures, I typically use a setting of 12 cm. and have since late 2012. However, in early 2011, I acquired two "cheap" ( $15.00 each ) machines from a source who used to sell on the CPAPAuction.com site. Later in 2011, I acquired a gauge manometer from CPAP.COM to check all my machines. I discovered that one of those "cheapies" was putting out 13 cm. at a setting of 12 cm., and the original Auto I bought (second-hand) in 2006 was putting out nearly 15 cm. at a setting of 12 cm. Well, that "cheapie" I use for naps (it lost its data-recording capabilities before I got it and that's probably why it was so cheap), but, it does blow air and that's why I use it for naps and don't have to download the nap data. With the Auto, I decided to use it this year (except for a few years, I've rotated my machines in some manner.....yearly, monthly, etc.) and in order to get close to 12 cm., using my manometer, I set it to 10 cm., which essentially gives me straight 13 cm. So, this will be the first YEAR I've used straight 13 cm., so I'll see how that works. In the past years when I used that Auto, I had it set to straight 12 and it was actually delivering about 15, so I was actually using a higher pressure than I thought I was using. I actually contemplated sending that Auto machine to one of the places that does repairs and have them recalibrate it, but decided it would be cheaper if I just set the pressure according to the manometer.......which was the way they set the machines years ago. Anyway, so far, so good.

As far as changing MY settings (or machines or whatever), over the years it's become longer and longer trial periods. What used to be "a week or more", became "a month or more" and is now "a year or more". I find it interesting to watch the trends throughout a year or even years. And, that's another reason I use straight pressure. It's a "constant" and other things/factors become the "variables".......instead of the pressure.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05