A NewBie with Hypopnea Concerns

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Guest

A NewBie with Hypopnea Concerns

Post by Guest » Sun Jan 18, 2015 2:56 pm

I have just joined this forum and I hope for some solid information from people with experience dealing with Sleep Apnea.

During last summer I would have spells where I would feel like I was going to fall asleep while driving. My PA ordered a sleep study. I was diagnosed with Mild Sleep Apnea (AHI 6.3). A Sleep Doctor prescribed CPAP at 7% Pressure. This has reduced my symptoms by more than one half. But after 89 days with CPAP I was still having mild symptoms and I returned to the Sleep Doctor. The doctor did not understand my concern because my AHI was below 5 and in his opinion was acceptable. After reviewing my download, the only suggestions he had was to repeat the Sleep Study or change the CPAP to APAP with 20/5 pressure. He wrote a prescription and the machine was changed beginning on the night of 1/7/2015. After eleven days with APAP I feel very little change.

My Symptoms are:
31% were Irritable (along with some depression) days
52% were Headache days
37% were Tired days
54% were Foggy Head (memory issues) days
17% were Zoned Out (a feeling of being asleep with eyes open) days

I use the Sleepyhead software and I see it does provide a ton of information. I am very surprised a sleep doctor could not make sense for all of this information. Here is the Summary data from my downloaded data.

SleepyHead Usage Statistics: (98 days of CPAP Data, between Fri Oct 10 2014 and Sat Jan 17 2015)
Average Hours per Night……..……….…...9:15
AHI…………………………………………..…........ 1.98
Obstructive Index…………………………..….. 0.21
Hypopnea Index……………………………….... 1.73
Clear Airway Index……………………….….... 0.05
Flow Limitation Index…………………….…... 0.02
RERA Index……………………………………...... 1.03
% of time in Cheyne Stokes Respiration…….. 0.24%
Average Leak Rate……………………………... 0.50
90% Leak Rate…………………………...……..1.00
% of time above Leak Rate threshold…..…..0.00%
Average Pressure………………………...…..6.72
Min Pressure……………………………..……..4.00
Max Pressure…………………………..……….9.20
90% Pressure……………………………….….7.00

After reviewing the Sleephead data I see, on the average, 87% of the AHI is caused from Hypopnea and 11% from Obstructive. I am concerned about the Hypopnea because, to my limited knowledge, a pressure change will not prevent Hypopnea. I am hoping someone in this group can shed some light on this for me. I have shown the eleven days of APAP data below.

Thank you.
++++++++++++++++++++++++++++++++++++++++++++++

Prescription Change (1/7/2015): APAP (Variable) Min 5 Max 20 (cmH2O)
…………………………………....…………............Cheyne ……………………………………………………..................…... Vibratory…………………………
……………………….....………... Periodic …....Stokes…………………………………………..................……Vibratory….Snore… Pressure…..90%....
Date.………Hours…..AHI…. Breathing …Respiration........…Obstructive...Hypopnea...RERA…Snore…....(VS2)…….Pulse……Pressure
1/7/2015…….9:23…..2.66…..1.00%........0.48%...............0.11….…...….2.45…....…0.96...…0.00……...0.21………0.64………6.5%
1/8/2015…….7:59…..1.13…..0.00%........0.00%...............0.13…....….…1.00…..…..1.00…...0.00………0.00……...0.38………6.0%
1/9/2015…….9:35…..1.15…..0.00%........0.00%...............0.10….…...….1.04…..…..0.83…...0.10………0.52……...0.31………6.5%
1/10/2015…...9:23…..1.38…..0.00%......0.00%...............0.00…....….…1.28…..…..0.75...…0.00……...0.11………0.32………6.5%
1/11/2015……9:08…..1.75…..0.00%.......0.00%...............0.11….....……1.64…..…..1.53...…0.22……...0.44………0.55………8.0%
1/12/2015……8:43…..3.09…..0.00%.......0.52%...............0.23…....…....2.75….…..0.69...…0.00……...0.11………0.57………6.0%
1/13/2015……9:00…..3.11…..0.00%.......0.34%...............0.00…....…....2.78….…..1.67...…0.00……...0.11………0.56………7.5%
1/14/2015……8:21…..0.84…..0.00%.......0.00%...............0.00…....…....0.84….…..0.48...…0.00……...0.12………0.24………6.0%
1/15/2015……8:21…..1.20…..0.00%.......0.00%...............0.24….......….0.96……...0.96...…0.00……...0.12………0.24………6.0%
1/16/2015……9:00…..1.55…..0.00%.......0.00%...............0.00….....……1.44…..…..1.22...…0.00……...0.00………0.55………6.0%
1/17/2015……8:59…..1.11…..0.00%......0.00%................0.33….....……0.78……....0.67...…0.00……...0.56………0.44………5.5%

User avatar
GearBear70
Posts: 19
Joined: Fri Jan 16, 2015 10:07 pm

Re: A NewBie with Hypopnea Concerns

Post by GearBear70 » Sun Jan 18, 2015 3:14 pm

I screwed up somewhere. In the above post, it was signed by "Guest" but it should have been by "GearBear70".

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Eson™ Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirSense 11 Autoset. Using Sleepyhead software CPAP Elite. APAP settings 20/8 cm H2O. Using SleepHead software and have downloaded Oscar

Sleeprider
Posts: 1562
Joined: Tue May 06, 2008 5:57 pm
Location: Murrysville, PA

Re: A NewBie with Hypopnea Concerns

Post by Sleeprider » Sun Jan 18, 2015 3:29 pm

Pretty easy in this case. Hypopneas are like incomplete obstructive apneas. They are very disruptive to sleep and oxygen levels, so you want to limit them. Your machine pressure is set to a minimum of 4. Go zoom in on the hypopnea events and focus on the pressures at which they occur. Move your minimum pressure above that.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead software. Just changed from PRS1 BiPAP Auto DS760TS

User avatar
Pesser
Posts: 547
Joined: Sat Jan 11, 2014 4:41 pm
Location: Nova Scotia Canada

Re: A NewBie with Hypopnea Concerns

Post by Pesser » Sun Jan 18, 2015 3:44 pm

My numbers were very similar to yours and so were my circumstances. I moved my minimum pressure to 7 and left my maximum at 20. I did this to eliminate any undetected UARS. It worked.

_________________
MachineMaskHumidifier
Additional Comments: Fear of devil with halo

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: A NewBie with Hypopnea Concerns

Post by palerider » Sun Jan 18, 2015 4:08 pm

GearBear70 wrote:I screwed up somewhere. In the above post, it was signed by "Guest" but it should have been by "GearBear70".
that's what happens when you forget to log in

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
GearBear70
Posts: 19
Joined: Fri Jan 16, 2015 10:07 pm

Re: A NewBie with Hypopnea Concerns

Post by GearBear70 » Sun Jan 18, 2015 7:22 pm

Sleeprider wrote:Pretty easy in this case. Hypopneas are like incomplete obstructive apneas. They are very disruptive to sleep and oxygen levels, so you want to limit them. Your machine pressure is set to a minimum of 4. Go zoom in on the hypopnea events and focus on the pressures at which they occur. Move your minimum pressure above that.
Thanks for this idea. I will go through the last few days and record the pressure at the Hypopneas.

After a quick look, I see when I was under CPAP Pressure 7 cmH2O with a Ramp of 4 cmH2O for 20 minutes, I had a lot of Hypopnea. Example on 12/5/2014: AHI = 3.65, Cheyne = 1.45%, Clear Airway = 0.20, Obstructive = 0.10, Hypopnea = 3.36, RERA = 1.09, Vibratory Snore = 0.00, Vibratory Snore (VS2) = 0.30 and Pressure Pulse = 0.79.

Another example 1/14/2015 with APAP 20 cmH2O/5 cmH2O with no ramp: AHI = 0.84, Cheyne = 0.00%, Clear Airway = 0.00, Obstructive = 0.00, Hypopnea = 0.84, RERA = 0.48, Vibratory Snore = 0.00, Vibratory Snore (VS2) = 0.12 and Pressure Pulse = 0.24. The pressure for everyone of these Hypopnea was 5 cmH2O. Raising the lower pressure maybe worth a try.

Thank you.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Eson™ Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirSense 11 Autoset. Using Sleepyhead software CPAP Elite. APAP settings 20/8 cm H2O. Using SleepHead software and have downloaded Oscar
Last edited by GearBear70 on Sun Jan 18, 2015 10:23 pm, edited 1 time in total.

User avatar
LSAT
Posts: 13377
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: A NewBie with Hypopnea Concerns

Post by LSAT » Sun Jan 18, 2015 8:44 pm

With an AHI of 1.98...I would not give it another thought.

User avatar
GearBear70
Posts: 19
Joined: Fri Jan 16, 2015 10:07 pm

Re: A NewBie with Hypopnea Concerns

Post by GearBear70 » Sun Jan 18, 2015 9:50 pm

LSAT wrote:With an AHI of 1.98...I would not give it another thought.
I would agree if my symptoms were not such a bother. Each day I have recorded how I have felt for that day and the AHI. To help you better understand my concerns I show my AHI and daily remarks for the month of January. My history shows when my AHI is 2.0 or higher I can have a couple of bad days trying to get over the symptoms.

Thank you for your comment.

++++++++++++++++++++++++++++++++++++++++

Date………..……AHI
1/1/2015……….0.6…….Remarks: I woke feeling rested and my head was clear. I finished the day feeling good.
1/2/2015……….1.3…….Remarks: I have had a headache since 3 am & have taken pills. My head is foggy but I otherwise I feel kind of good. My headache went away after the pills & my head cleared up before noon. In all it was a good day.
1/3/2015……….1.2……Remarks: I woke up feeling somewhat rested & with a clear head. It was a pretty good day.
1/4/2015……….2.2……Remarks: I woke feeling rested but with a slight headache. My head is a little foggy. By late evening I was tired but in all a pretty good day.
1/5/2015……….1.7……Remarks: I woke feeling some rested and with a clear head. We had 6 inches of snow & the snowblower would not start. I cleaned a lot of the driveway with a shovel & the neighbor cleaned the rest. But I was not very tired at day's end.
1/6/2015……….1.5……Remarks: I woke feeling some rested and with a clear head. I ran the snow blower again today and was not tired in the evening.
1/7/2015……….2.7……Remarks: I woke feeling tired, with a headache and somewhat foggy feeling. I had a couple zoned out periods in the afternoon but felt better that evening. Last night was the first night with the new prescription Auto-CPAP 20/5 per Dr. James Cafaro.
1/8/2015……….1.1……Remarks: I woke feeling rested but my head is a little foggy. When the day ended I felt pretty good.
1/9/2015……….1.1……Remarks: I woke feeling rested but with a slight headache and a little foggy. The headache & foggy was gone by lunch time. At the end of the day I felt good.
1/10/2015……..1.4…….Remarks: I woke feeling somewhat rested and with a clear head. The day ended good.
1/11/2015……..1.8…….Remarks: I woke feeling pretty good. By evening I was very tired.
1/12/2015……..3.1…….Remarks: I woke tired, irritable and with a foggy head. The foggy stayed with me through the afternoon. By evening I was very tired and took a nap around 6:00 pm. I woke up choking.
1/13/2015……..3.1…….Remarks: I woke tired, irritable and with a foggy head. I ended the day irritable and tired.
1/14/2015……..0.8…….Remarks: I woke somewhat rested but a little foggy. It was a pretty good day.
1/15/2015……..1.2…….Remarks: I woke with a headache, feeling foggy and a little irritable. I took pills twice to shake the headache. I had a few zoned out spells during the day. I spent the day easily irritated and depressed.
1/16/2015…….1.6……..Remarks: I woke feeling somewhat rested but with a slight headache and a little foggy.
1/17/2015…….1.1……..Remarks: I woke feeling rested and with a clear head. At the end of the day I was tired but still with a clear head.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Eson™ Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirSense 11 Autoset. Using Sleepyhead software CPAP Elite. APAP settings 20/8 cm H2O. Using SleepHead software and have downloaded Oscar

Sleeprider
Posts: 1562
Joined: Tue May 06, 2008 5:57 pm
Location: Murrysville, PA

Re: A NewBie with Hypopnea Concerns

Post by Sleeprider » Sun Jan 18, 2015 10:04 pm

GearBear70 wrote:
Sleeprider wrote:Pretty easy in this case. Hypopneas are like incomplete obstructive apneas. They are very disruptive to sleep and oxygen levels, so you want to limit them. Your machine pressure is set to a minimum of 4. Go zoom in on the hypopnea events and focus on the pressures at which they occur. Move your minimum pressure above that.
Thanks for this idea. I will go through the last few days and record the pressure at the Hypopneas.

After a quick look, I see when I was under CPAP Pressure 7.0 cmH2O with a Ramp of 4 cmH2O for 20 minutes, I had a lot of Hypopnea. Example on 12/5/2014: AHI = 3.65, Cheyne = 1.45%, Clear Airway = 0.20, Obstructive = 0.10, Hypopnea = 3.36, RERA = 1.09, Vibratory Snore = 0.00, Vibratory Snore (VS2) = 0.30 and Pressure Pulse = 0.79.

Another example 1/14/2015 with APAP 20 cmH2O/5 cmH2O with no ramp: AHI = 0.84, Cheyne = 0.00%, Clear Airway = 0.00, Obstructive = 0.00, Hypopnea = 0.84, RERA = 0.48, Vibratory Snore = 0.00, Vibratory Snore (VS2) = 0.12 and Pressure Pulse = 0.24. The pressure for everyone of these Hypopnea was 5 cmH2O. Raising the lower pressure maybe worth a try.

Thank you.
I have edited your post to show correct terminology. Pressure is in cmH2O water not percent. The results of your data review show a clear path to improvement. All hypopnea (and I bet most RERA) is at 5 cmH2O pressure or nearly that. Raise the minimum pressure to 6.0. Keep the ramp function OFF. If that doesn't erase the hypopnea and RERA, continue to increase minimum pressure in 0.5 cmH2O increments. Good luck!

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead software. Just changed from PRS1 BiPAP Auto DS760TS

User avatar
GearBear70
Posts: 19
Joined: Fri Jan 16, 2015 10:07 pm

Re: A NewBie with Hypopnea Concerns

Post by GearBear70 » Sun Jan 18, 2015 10:20 pm

Thank you 'Sleeprider'. I can use all kinds of help.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Eson™ Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirSense 11 Autoset. Using Sleepyhead software CPAP Elite. APAP settings 20/8 cm H2O. Using SleepHead software and have downloaded Oscar

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: A NewBie with Hypopnea Concerns

Post by Wulfman... » Sun Jan 18, 2015 11:50 pm

Without being able to look at your reports, I'm also going to suggest that your "symptoms" are the result of inadequate minimum pressure. But, along with that, I also suspect that they're a result of your pressures changing all the time during the night. When you enter a deep sleep stage or REM, your breathing can change.......that may trigger a pressure increase and bump you out of that needed sleep stage and into a lighter one......and leave you feeling unrested. If that's happening all night long, it's no wonder you're having those symptoms. You ain't getting any sound sleep!


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: A NewBie with Hypopnea Concerns

Post by Julie » Mon Jan 19, 2015 5:48 am

Agree, but would suggest you raise the low pressure even more - 6 is about as 'bare minimum' a pressure as anyone can really breathe at, and I'd go for at least 8 and see how you feel after a few days (one is not enough). You can always lower it again if it's higher than you're comfortable with, but for the record, most of us seem to end up closer to 10 than otherwise.

User avatar
GearBear70
Posts: 19
Joined: Fri Jan 16, 2015 10:07 pm

Re: A NewBie with Hypopnea Concerns

Post by GearBear70 » Mon Jan 19, 2015 4:36 pm

Sleeprider wrote:
GearBear70 wrote:
Sleeprider wrote:Pretty easy in this case. Hypopneas are like incomplete obstructive apneas. They are very disruptive to sleep and oxygen levels, so you want to limit them. Your machine pressure is set to a minimum of 4. Go zoom in on the hypopnea events and focus on the pressures at which they occur. Move your minimum pressure above that.
Thanks for this idea. I will go through the last few days and record the pressure at the Hypopneas.

After a quick look, I see when I was under CPAP Pressure 7.0 cmH2O with a Ramp of 4 cmH2O for 20 minutes, I had a lot of Hypopnea. Example on 12/5/2014: AHI = 3.65, Cheyne = 1.45%, Clear Airway = 0.20, Obstructive = 0.10, Hypopnea = 3.36, RERA = 1.09, Vibratory Snore = 0.00, Vibratory Snore (VS2) = 0.30 and Pressure Pulse = 0.79.

Another example 1/14/2015 with APAP 20 cmH2O/5 cmH2O with no ramp: AHI = 0.84, Cheyne = 0.00%, Clear Airway = 0.00, Obstructive = 0.00, Hypopnea = 0.84, RERA = 0.48, Vibratory Snore = 0.00, Vibratory Snore (VS2) = 0.12 and Pressure Pulse = 0.24. The pressure for everyone of these Hypopnea was 5 cmH2O. Raising the lower pressure maybe worth a try.

Thank you.
I have edited your post to show correct terminology. Pressure is in cmH2O water not percent. The results of your data review show a clear path to improvement. All hypopnea (and I bet most RERA) is at 5 cmH2O pressure or nearly that. Raise the minimum pressure to 6.0. Keep the ramp function OFF. If that doesn't erase the hypopnea and RERA, continue to increase minimum pressure in 0.5 cmH2O increments. Good luck!
I have completed my analysis of the Pressure during a Hypopnea for the time I have been on the APAP. I recorded the Date, the Current Hypopnea, a calculated New Hypopnea based on only counting those Pressures greater than 7 cmH2O, the Current AHI and a New AHI (assuming all of the other elements that make up the AHI stay the same).

For me this is an encouraging exercise. After leaving the sleep doctor's office I was very very discouraged. But after reading everyone's comments here I have hope for a better night's sleep. I will find a way to get either the sleep doctor or my Physician's Assistant to write a new prescription for APAP 20 cmH2O / 7 cmH2O with no Ramp. I will try the 20 cmH2O / 7 cmH2O for a period of time and if no improvement I will ask for the lower 7 cmH2O to be increased.

Thank you Everyone for your help.

+++++++++++++++++++++++++++++++

………………..Current….…New…………Current….New
Date…………Hypopnea…Hypopnea…….AHI…..…AHI
1/7/2015……….2.45…..….0.11………….2.66….....0.32
1/8/2015……….1.00……...0.00………….1.13…..…0.13
1/9/2015……….1.04……...0.10………….1.15…..…0.21
1/10/2015……...1.28……..0.11………….1.38…..…0.21
1/11/2015………1.64…….0.66…………..1.75…..…0.77
1/12/2015………2.75…….0.00…………..3.09…..…0.34
1/13/2015………2.78…….0.33…………..3.11…..…0.66
1/14/2015………0.84…….0.00…………..0.84…..…0.00
1/15/2015..……0.96….….0.00…………..1.20…..…0.24
1/16/2015…..…1.44….….0.00…………..1.55…..…0.11
1/17/2015…..…0.78….….0.00…………..1.11…..…0.33
1/18/2015……..1.58….….0.00…………..1.81……..0.23

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Eson™ Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirSense 11 Autoset. Using Sleepyhead software CPAP Elite. APAP settings 20/8 cm H2O. Using SleepHead software and have downloaded Oscar

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: A NewBie with Hypopnea Concerns

Post by palerider » Mon Jan 19, 2015 4:41 pm

GearBear70 wrote: I will find a way to get either the sleep doctor or my Physician's Assistant to write a new prescription for APAP 20 cmH2O / 7 cmH2O with no Ramp. I will try the 20 cmH2O / 7 cmH2O for a period of time and if no improvement I will ask for the lower 7 cmH2O to be increased.
I won't pretend that I know what your results mean... but I will point out that you can, should you choose to, change the pressures of your own machine in just a few minutes, should you choose to do so. instructions are readily available, just check the bottom of one of sleepriders posts

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
GearBear70
Posts: 19
Joined: Fri Jan 16, 2015 10:07 pm

Re: A NewBie with Hypopnea Concerns

Post by GearBear70 » Mon Jan 19, 2015 4:52 pm

Wulfman... wrote:Without being able to look at your reports, I'm also going to suggest that your "symptoms" are the result of inadequate minimum pressure. But, along with that, I also suspect that they're a result of your pressures changing all the time during the night. When you enter a deep sleep stage or REM, your breathing can change.......that may trigger a pressure increase and bump you out of that needed sleep stage and into a lighter one......and leave you feeling unrested. If that's happening all night long, it's no wonder you're having those symptoms. You ain't getting any sound sleep!


Den

.
Please help me to understand what you mean by "reports" and how would I get that to you. I may be able to learn how to do this.

Thank you for your comment.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Eson™ Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirSense 11 Autoset. Using Sleepyhead software CPAP Elite. APAP settings 20/8 cm H2O. Using SleepHead software and have downloaded Oscar