NYCzzz wrote:I'm confused -- on sleepyhead how can i have an "obstructive apnea" without a flow limitation... as detailed below... shouldn't an apnea without any flow limitation be classified as a central?
Your equipment is listed as a Resmed S9 AutoSet. If you've listed the equipment correctly, the way your machine decides whether to score an apnea as an OA or a CA depends only on the results of the
Forced Oscillation Technique (FOT) algorithm and not at all on what's going on in the Flow Limitation graph. If the results from the FOT algorithm indicate the airway is likely to be blocked (obstructed), the apnea is scored as an OA. If the results of the FOT algorithm indicate the airway is likely to be clear (not obstructed), the apnea is scored as a CA.
It is possible for the airway to suddenly collapse without much warning in the sense of significant flow limitations or snoring being present. Sometimes this kind of an obstructive apnea is referred to as a "frank OA" meaning that it came without warning.
An in the data you posted there is evidence of one or two flow limited breaths right before the apnea starts:
The inhalations that I've circled have some evidence of flow limited shape---particularly the one right before the apnea itself. As to why there's no bump in the Flow Limitations graph: It may be that the distortions in the inhalations were not quite bad enough to meet the S9's FL scoring criteria to cause a bump on the FL graph. Or it could be that there was not enough of them in a row for the S9 to decide to increase the value in the FL graph. Like everything else with our machines, the algorithm used to score FLs is not perfect and is not going to pick up every single minor flow limitation.