I have a Resmed S9 and just started using Sleepyhead beta 0.95. Rescan 4.3 crashes my desktop during navagation. I uninstalled their drivers, but Win 7 crashes so bad I have to unplug my computer to recover.
When viewing Sleepyhead daily summary I get: AHI 1, Hypopnea(duration 1.5hrs), pressure and mask pressure do not change in response to the Hypop. The S9 is set to APAP mode. Why isn't the pressure changing? How accurate are the Sleepyhead graphs compared to Rescan?
AHI and Pressure
Re: AHI and Pressure
Please post an image of your detailed report so we can see what you are seeing. If you don't know how see this thread as we discussed it and there's some examples. It helps us help you if we can see what you are seeing.
Screen shot thread viewtopic.php?f=1&t=81072&p=737779#p737779
The SleepyHead graphs are usually spot on in terms of how they compare to ResScan graphs.
Screen shot thread viewtopic.php?f=1&t=81072&p=737779#p737779
The SleepyHead graphs are usually spot on in terms of how they compare to ResScan graphs.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: AHI and Pressure
Not sure what you mean by "Hypopnea(duration 1.5hrs)". A single hypopnea is not going to last 1.5 hours.2bitwise wrote: When viewing Sleepyhead daily summary I get: AHI 1, Hypopnea(duration 1.5hrs), pressure and mask pressure do not change in response to the Hypop.
I'm going to hazard a guess on this one: I suspect that you are looking at an "isolated hypopnea" or an "isolated obstructive apnea". The thing is: The auto algorithm is programed to NOT respond to isolated Hs and OAs if there is no snoring or flow limitation being detected. It takes two or more obstructive events (Hs or OAs) that happen very close to each other (like within 5 minutes of each other) to trigger the machine to increase the pressure for the events.The S9 is set to APAP mode. Why isn't the pressure changing? How accurate are the Sleepyhead graphs compared to Rescan?
This mimics the AASM guidelines for manual titration of CPAP/BiPAP on attended titration sleep studies. The reason the guideline is there is at least partially to keep people from being over titrated.
_________________
Machine: DreamStation BiPAP® Auto Machine |
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5 |
Re: AHI and Pressure
I used F12 in SH, and I've got a PNG image. I copied it, but there is no paste function when I try to attach it here. (Win 7)
Re: AHI and Pressure
You need to upload the png image to a photo sharing site like Photbucket and then make the link from there. The url for the image needs to be between an [img] and an [/img] command.
_________________
Machine: DreamStation BiPAP® Auto Machine |
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5 |
Re: AHI and Pressure
So the only attachments can be a url.
That ties it.
That ties it.