Couple of ideas.
I wouldn't go changing minimum pressure until I understood that leaks themselves weren't the cause of the elevated AHI.
The leaks may or may not be really bad and I think that needs to be figured out first before increasing the pressure.
Secondly...if leaks are acceptable and more minimum pressure is needed I wouldn't go jumping from 5 cm to 10 cm minimum in one big jump. It's likely to be a difficult adjustment so I suggest going up slowly...like 1 cm for a couple of nights then another 1 cm for a couple nights...watch the AHI data ...might get lucky and not need 10 cm.. Plus I have seen reports where once the minimum is set optimally the 90/95% pressures stabilize much lower than initially seen or thought was needed. Go up slowly if you are going to go up with that minimum. It will be easier to adjust to and you might get lucky and only need 8 cm minimum.
AND most importantly...evaluate the AHI categories...make sure that if you are going to increase the pressure trying to reduce the AHI that the AHI is composed of events that increasing the pressure will fix.
AHI of 6.5....how much of it is central index and how much of it is obstructive apneas and hyponea index?
If it is mainly central...increasing the pressure isn't going to help and might make things worse.
So
1...before worrying about the full face mask thing...make sure that the leaks are really bad and need fixing because of mouth breathing with the over the nose mask. Even if there are a few big leaks it might be better to have 30 minutes in big leak than not being able to sleep for more than 2 hours with a full face mask.
2...before changing the minimum pressure...make sure that we aren't dealing with something that more pressure might make worse. So understand the AHI and it's event categories...don't go on just the AHI alone.
So can you post an image of the detailed nightly report for a typical night?
We talk about it here with some examples.
viewtopic.php?f=1&t=81072&p=737779#p737779
Need to see the AHI bar graphs on the left side (when using SleepyHead) and the graphs on the right (Events, flow rate, pressure and leak) turn off AHI graph as it isn't needed by going to Preferences and graph tab...remove check mark.
If using Encore...just a screen shot of the detailed single night report. Don't do the summary graphs.
Even if you aren't comfortable with changing anything yourself...and until you are...don't go doing it..
it would help if we could see what is going on so we can give you some ideas as to what to discuss with your medical care providers.