respironics SV - alarm E63 Internal watchdog failure

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
kolchak
Posts: 81
Joined: Thu Nov 08, 2007 6:08 pm
Location: Montgomery County MD

respironics SV - alarm E63 Internal watchdog failure

Post by kolchak » Sat May 26, 2012 11:19 am

Hi
Woke up this morning with mask off & (5 yr old) auto SV displaying alarm code E63. I checked manual and discovered code means "Internal watchdog failure" but it doesn't offer any explanation. It sounds serious and I assume it means time to replace the machine or send this somewhere for repair. Either way, I'll be sleeping with a mask on for 1st time in ~ 5 years.
Any ideas or suggestions?
Thank you

User avatar
squid13
Posts: 2301
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: respironics SV - alarm E63 Internal watchdog failure

Post by squid13 » Sat May 26, 2012 11:43 am

It's a E63 Watchdog failure Defective microprocessor Replace Main PCA. I found it here http://www.internetage.ws/cpapdata/manu ... 042508.pdf. Hope this helps, scroll down to you get to error codes.