All S9 users - Is H5i working ??

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
oscar98
Posts: 274
Joined: Fri Apr 16, 2010 1:49 pm
Location: TN - USA

Re: All S9 users - Is H5i working ??

Post by oscar98 » Thu May 06, 2010 4:19 pm

jlgwinn wrote:Well dang! That defeats the purpose of "Climate Control " in a way. On would think that you can turn the huminity up and the hose controls the rain out. That's one of the main reason that I got the machine to take the human intervention out of it.

Thanks
Jeff
Well the good news is that even with the humidity at 6 the rainout I experience was NOTHING like anything I had experienced with my previous humidifier set on 2.5. I would not have even noticed it if it hadn't been spurting tiny water drops out the vent and onto my hand when I was exhaling. If you want lots of humidity then you will probably need the climateline hose. However, I've found the running anywhere between 4 and 5 (depending on the humidity in my room) has worked out great without the fancy hose.

I decided to hold off on getting the climateline because I prefer breathing cool (but moist) air. Once fall/winter rolls around I might change my mind though.

User avatar
Nord
Posts: 565
Joined: Sun Mar 07, 2010 4:30 pm
Location: GTA Canada

Re: All S9 users - Is H5i working ??

Post by Nord » Thu May 06, 2010 5:56 pm

jlgwinn wrote:Well dang! That defeats the purpose of "Climate Control " in a way. On would think that you can turn the huminity up and the hose controls the rain out. That's one of the main reason that I got the machine to take the human intervention out of it.

Thanks
Jeff
Hi Jeff

Humidity is relative and depends on temp of the moist air. With ClimateControl it is automatic and measures by temp from a sensor near the mask. If you set the temp, where you want, it will give you 80% moisture for that temp and thus no rainout.

If you want cooler air and more moisture... you set to manual (human intervention) and possibly some rainout... I have used manual and not got rainout yet but you have to experiment for yourself.

After reading an article provided by KatieW... I have upped the temp to 80 Deg and Auto... only 3 nights so I('m just getting used to it.

Nord

User avatar
oscar98
Posts: 274
Joined: Fri Apr 16, 2010 1:49 pm
Location: TN - USA

Re: All S9 users - Is H5i working ??

Post by oscar98 » Thu May 06, 2010 6:53 pm

Nord wrote:
If you want cooler air and more moisture... you set to manual (human intervention) and possibly some rainout... I have used manual and not got rainout yet but you have to experiment for yourself.

After reading an article provided by KatieW... I have upped the temp to 80 Deg and Auto... only 3 nights so I('m just getting used to it.

Nord
Hey Nord, quick question for ya.

I prefer quite a bit of humidity but with cool air. I like having the humidity at 6 but get rainout without having the climate hose. If I get a climate hose would I be able to control the rainout without heating the air up? I just assumed having the heated hose would mean having warm humidity.

User avatar
KatieW
Posts: 1672
Joined: Sun Sep 27, 2009 11:07 am
Location: southern AZ

H5i Auto versus Manual Setting

Post by KatieW » Thu May 06, 2010 7:28 pm

When I was in Arizona, where the humidity is 25%, I had the H5i on Auto and 80 degrees. I assumed it was doing it's job, because I got no rainout. I had a little mouth dryness, but thought it was just me (meds, not hydrated enough, etc).

However, now that I'm back in Washington, where the humidity is 40%, I was experiencing very dry mouth, with even the roof of my mouth very dry. I tried taping my mouth, to eliminate the possibility of mouth breathing, and it didn't make any difference. The bedroom is usually 70 degrees.

For the past week, I have been experimenting with different Manual settings, and keeping detailed notes.

On Auto and 80 degrees, my H5i was using 6 ounces of water for a 8 hour night. On the Manual setting, it has used from 7 to 9 ounces, depending on the settings. So I am using more water, and getting more humidity on Manual. And, sometimes a little more than I want, with mask rainout.

I devised a fleece hoodie for my Headrest Nasal pillows, and alternate with the SoftGel rather than the Activa Lt, since it's more prone to rainout, until I can figure out how to make a cover for it.

My best results so far are with a setting of 5 and 78 degrees. At that setting my mouth and throat are only a little dry, and minimal rainout (just a few small drops in the mask, but not in my nose).

The experiment will continue.

_________________
Mask: Pico Nasal CPAP Mask with Headgear
Additional Comments: Resmed AirCurve 10 ASV and Humidifier, Oscar for Mac
KatieW

User avatar
Nord
Posts: 565
Joined: Sun Mar 07, 2010 4:30 pm
Location: GTA Canada

Re: All S9 users - Is H5i working ??

Post by Nord » Thu May 06, 2010 8:22 pm

oscar98 wrote: Hey Nord, quick question for ya.

I prefer quite a bit of humidity but with cool air. I like having the humidity at 6 but get rainout without having the climate hose. If I get a climate hose would I be able to control the rainout without heating the air up? I just assumed having the heated hose would mean having warm humidity.
It depends on your room setting. The "secret" is that warmer air holds more moisture, so... a heated hose will allow more moisture through the hose until it hits something colder. If you keep your room cold or even cooler... then you might get some rainout. Generally, I have found that the S9 does not put as much water into the hose... probably due to smaller diameter hose and obstructions of ClimateLine which slows down Flow and thus moisture.

You would have to experiment and compare as Katie is doing, to see if ClimateLine with some heat or SlimLine with a cozy, is better for getting the moisture to you nose w/o droplets. One other fly in the ointment: according to the latest article that I have read... the nose collects the moisture on the exhale so, the moist hot air on exhale, meets the moist cool air from the mask and adds more moisture inside the mask... probably rain

This is some theory and some fact... experimentation is the only sure way to tell. And that will change every night...

Sorry about the bad news.

Nord
Last edited by Nord on Fri May 07, 2010 6:50 am, edited 1 time in total.

User avatar
oscar98
Posts: 274
Joined: Fri Apr 16, 2010 1:49 pm
Location: TN - USA

Re: All S9 users - Is H5i working ??

Post by oscar98 » Thu May 06, 2010 8:53 pm

Nord wrote: One other fly in the ointment: according to the latest article that I have read... the nose collects the moisture on the exhale so, the moist hot air on exhale, meets the moist cool air from the mask and adds more moisture inside the mask... probably rain

Nord
Thanks for pointing that out. I bet that's what my problem is because the most prominent place for my moisture problem is the little puddle I'm getting in the bottom of my mask.

User avatar
elena88
Posts: 1650
Joined: Sun Feb 28, 2010 2:25 pm
Location: california

Re: All S9 users - Is H5i working ??

Post by elena88 » Thu May 06, 2010 10:09 pm

So far so good with mine, just been a few days though!

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: sleep study: slept 66 min in stage 2 AHI 43.3 had 86 spontaneous arousals I changed pressure from 11 to 4cm now no apap tummy sleeping solved apnea

User avatar
ww
Posts: 523
Joined: Sun Mar 16, 2008 11:58 am

Re: All S9 users - Is H5i working ??

Post by ww » Thu May 06, 2010 10:19 pm

I am using the S9 autoset with H5i humidifier and ClimateControl Hose. I really like the way it works. Set on Auto and 80 degrees and it uses about 1/2 tank of distilled water. I am sure that everyone realizes that if your room temperature is 75 degrees and you set your humidifier to deliver 75 degree air, that very very little water will be added to the air. The heated hose works very well and was one of the many reasons the S9 is proving so popular with many of us.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear - Fit Pack (All Sizes Included)
Additional Comments: Titrated on Auto CPAP at 7/14 cm: Only licensed medical professionals can give medical advice or write prescriptions

User avatar
elena88
Posts: 1650
Joined: Sun Feb 28, 2010 2:25 pm
Location: california

Re: All S9 users - Is H5i working ??

Post by elena88 » Fri May 07, 2010 3:50 am

SPOKE TO SOON!

ITS NOT WORKING AT ALL.. HEATING PLATE IS FREEZING COLD..


Would not have known, if hubby was not in the dog house tonight, and he let me torture him with my apap for
an hour..

I have had the humidity up to five for three hours, checked everything... its now a nice fish tank for artic krill.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: sleep study: slept 66 min in stage 2 AHI 43.3 had 86 spontaneous arousals I changed pressure from 11 to 4cm now no apap tummy sleeping solved apnea

User avatar
oscar98
Posts: 274
Joined: Fri Apr 16, 2010 1:49 pm
Location: TN - USA

Re: All S9 users - Is H5i working ??

Post by oscar98 » Fri May 07, 2010 5:19 am

elena88 wrote:SPOKE TO SOON!

ITS NOT WORKING AT ALL.. HEATING PLATE IS FREEZING COLD..


Would not have known, if hubby was not in the dog house tonight, and he let me torture him with my apap for
an hour..

I have had the humidity up to five for three hours, checked everything... its now a nice fish tank for artic krill.
Try running the prewarm cycle to confirm that it does heat up.

User avatar
Nord
Posts: 565
Joined: Sun Mar 07, 2010 4:30 pm
Location: GTA Canada

Re: All S9 users - Is H5i working ??

Post by Nord » Fri May 07, 2010 8:07 am

If you are in doubt about your H5i working properly... depending on your DME... they will want to know whats wrong...

Its worth re-posting some tests that started this thread and got my machine replaced... twice.

Nord

Nord wrote:Are you sure your H5i is working the way it should ???
Mine hasn't and there has been a sprinkling of others with concerns...

Past several nights awakened with dry throat and cold air/ so cold it hurts a little deep in your throat. House temp set for 67 deg at night and 71 deg during the day... no windows open.
Set for Auto Climate on S9 set at 82 deg... then at 76... then 68... then 86... then I'm up. Feel the plate below the tank and its cold, very cold.
Never get rainout/ moisture in tube or mask.
Uses very little water and with Humidifier always ON... I have never run out and water would last, perhaps a week, if I left it.

I replaced the H5i about 3 weeks ago when the original humidifier did the same thing.
The following week replaced the S9 to make sure that wasn't the problem.
Contacted the DME to have the ResMed rep get some answers... still waiting since last week.

This AM I did some tests:
Removed the ClimateLine to "wipe" the contacts and make sure that part was working... it registers on the S9 as being engaged.
Removed and replaced the H5i to again "wipe" the contacts... it registers on the S9 as being engaged.
(before you ask) Power outlet running full 120 volts to single outlet in good working order.

TEST #1
Set to Climate Control to ON... ClimateLine Engaged and set temp to 80 deg and turn on Flow for S9...
1. run 10 minutes at pressure with dry resistance at end of the tube
2. feel heated ClimateLine... no warmth to touch.
3. Stop S9, lift tank out and feel heater plate below... COLD, not cool, COLD

TEST #2
Set to Climate Control to ON... ClimateLine Engaged and set temp to 86 deg (Max) and turn on Flow for S9...
1. run 10 minutes at pressure with dry resistance at end of the tube
2. feel heated ClimateLine... no warmth to touch.
3. Stop S9, lift tank out and feel heater plate below... COLD, not cool, COLD

TEST #3
Set to Climate Control to ON... ClimateLine Engaged and set temp to 72 deg and turn on Flow for S9...
1. run 10 minutes at pressure with dry resistance at end of the tube
2. feel heated ClimateLine... no warmth to touch.
3. Stop S9, lift tank out and feel heater plate below... COLD, not cool, COLD

TEST #4
Set to Climate Control to MANUAL... ClimateLine Engaged and set temp to 86 deg; Set Humidity to OFF; and turn on Flow for S9...
1. run 10 minutes at pressure with dry resistance at end of the tube
2. feel heated ClimateLine... no warmth to touch.
3. Stop S9, lift tank out and feel heater plate below... COLD, not cool, COLD

TEST #5
Set to Climate Control to MANUAL... ClimateLine Engaged and set temp to 86 deg; Set Humidity to 0.5; and turn on Flow for S9...
1. run 10 minutes at pressure with dry resistance at end of the tube
2. feel heated ClimateLine... no warmth to touch.
3. Stop S9, lift tank out and feel heater plate below... COLD, not cool, COLD

TEST #6
Set to Climate Control to MANUAL... ClimateLine Engaged and set temp to 86 deg; Set Humidity to 6.0; and turn on Flow for S9...
1. run 10 minutes at pressure with dry resistance at end of the tube
2. feel heated ClimateLine... no warmth to touch.
3. Stop S9, lift tank out and feel heater plate below... COLD, not cool, COLD

TEST #7
Set Warm-up to ON for 3 minutes and Temp to 86 deg
1. run 5 minutes
2. feel heated ClimateLine... warm to touch.
3. Stop S9, lift tank out and feel heater plate below... HOT, not warm, HOT

Any thoughts... have you checked yours ???

Nord

User avatar
elena88
Posts: 1650
Joined: Sun Feb 28, 2010 2:25 pm
Location: california

Re: All S9 users - Is H5i working ??

Post by elena88 » Fri May 07, 2010 10:40 am

thanks for all of that, working on it now..
I have a sore throat now from all that cold air..
no wonder I couldnt sing!

EDIT.. We got it to work now, it kept shutting off everytime I checked a different setting, so have to fix that?
thank you soooooooooo much!

another edit: tried for a long time to get it to work last night, no luck!
still trouble shooting..

I only have a regular hose, dme said slim line was not included with the machine. We dont use our central heating, but its been
chilly at night, so the air in the tube is REALLY cold! This is so unpleasant!
I disconnected the humidifier last night..

We also dont have the heated hose, we turned off the climate control so we could set the humidifier independantly, and it still
wont work. The plate gets warm, but not warm enough to heat the air or humidify it..

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: sleep study: slept 66 min in stage 2 AHI 43.3 had 86 spontaneous arousals I changed pressure from 11 to 4cm now no apap tummy sleeping solved apnea
Last edited by elena88 on Sat May 08, 2010 12:51 pm, edited 2 times in total.

jlgwinn
Posts: 27
Joined: Thu Mar 12, 2009 2:05 pm

Re: All S9 users - Is H5i working ??

Post by jlgwinn » Fri May 07, 2010 8:28 pm

All test failed for me..even #7, my hose never got warm to the touch even after running for 15 minutes. The base however did get warm to the touch on warm up. Guess I'll be making a call Monday to my DME.

Jeff

Nord wrote:If you are in doubt about your H5i working properly... depending on your DME... they will want to know whats wrong...

Its worth re-posting some tests that started this thread and got my machine replaced... twice.

Nord

Nord wrote:Are you sure your H5i is working the way it should ???
Mine hasn't and there has been a sprinkling of others with concerns...

Past several nights awakened with dry throat and cold air/ so cold it hurts a little deep in your throat. House temp set for 67 deg at night and 71 deg during the day... no windows open.
Set for Auto Climate on S9 set at 82 deg... then at 76... then 68... then 86... then I'm up. Feel the plate below the tank and its cold, very cold.
Never get rainout/ moisture in tube or mask.
Uses very little water and with Humidifier always ON... I have never run out and water would last, perhaps a week, if I left it.

I replaced the H5i about 3 weeks ago when the original humidifier did the same thing.
The following week replaced the S9 to make sure that wasn't the problem.
Contacted the DME to have the ResMed rep get some answers... still waiting since last week.

This AM I did some tests:
Removed the ClimateLine to "wipe" the contacts and make sure that part was working... it registers on the S9 as being engaged.
Removed and replaced the H5i to again "wipe" the contacts... it registers on the S9 as being engaged.
(before you ask) Power outlet running full 120 volts to single outlet in good working order.

TEST #1
Set to Climate Control to ON... ClimateLine Engaged and set temp to 80 deg and turn on Flow for S9...
1. run 10 minutes at pressure with dry resistance at end of the tube
2. feel heated ClimateLine... no warmth to touch.
3. Stop S9, lift tank out and feel heater plate below... COLD, not cool, COLD

TEST #2
Set to Climate Control to ON... ClimateLine Engaged and set temp to 86 deg (Max) and turn on Flow for S9...
1. run 10 minutes at pressure with dry resistance at end of the tube
2. feel heated ClimateLine... no warmth to touch.
3. Stop S9, lift tank out and feel heater plate below... COLD, not cool, COLD

TEST #3
Set to Climate Control to ON... ClimateLine Engaged and set temp to 72 deg and turn on Flow for S9...
1. run 10 minutes at pressure with dry resistance at end of the tube
2. feel heated ClimateLine... no warmth to touch.
3. Stop S9, lift tank out and feel heater plate below... COLD, not cool, COLD

TEST #4
Set to Climate Control to MANUAL... ClimateLine Engaged and set temp to 86 deg; Set Humidity to OFF; and turn on Flow for S9...
1. run 10 minutes at pressure with dry resistance at end of the tube
2. feel heated ClimateLine... no warmth to touch.
3. Stop S9, lift tank out and feel heater plate below... COLD, not cool, COLD

TEST #5
Set to Climate Control to MANUAL... ClimateLine Engaged and set temp to 86 deg; Set Humidity to 0.5; and turn on Flow for S9...
1. run 10 minutes at pressure with dry resistance at end of the tube
2. feel heated ClimateLine... no warmth to touch.
3. Stop S9, lift tank out and feel heater plate below... COLD, not cool, COLD

TEST #6
Set to Climate Control to MANUAL... ClimateLine Engaged and set temp to 86 deg; Set Humidity to 6.0; and turn on Flow for S9...
1. run 10 minutes at pressure with dry resistance at end of the tube
2. feel heated ClimateLine... no warmth to touch.
3. Stop S9, lift tank out and feel heater plate below... COLD, not cool, COLD

TEST #7
Set Warm-up to ON for 3 minutes and Temp to 86 deg
1. run 5 minutes
2. feel heated ClimateLine... warm to touch.
3. Stop S9, lift tank out and feel heater plate below... HOT, not warm, HOT

Any thoughts... have you checked yours ???

Nord

robmax
Posts: 49
Joined: Fri Jan 01, 2010 11:25 am

Re: All S9 users - Is H5i working ??

Post by robmax » Sat May 08, 2010 10:03 am

Sounds like the vaunted $1000+ S9 needs some more tweaking. Back to the lab, Resmed Boys and Girls. Right now I am so happy I got the S8 (at the reduced price) and didn't wait for the new one.

manas
Posts: 48
Joined: Sat Mar 27, 2010 3:54 pm

Re: All S9 users - Is H5i working ??

Post by manas » Sat May 08, 2010 10:48 am

Nord: Did you hear from the rep *why* your newer H5i is supposed to be working better?

My ClimateLine was on backorder when I got my S9 Apr 6, so I used 10 year old hose with my diamond-plated S9 for the first 2 weeks, with a manual humidity level of 3.0. When I got my ClimateLine hose 2 weeks ago, I thought the first night was better, but since then have never found it to be warm, and saw drops of water inside my FX today. My humidifier/water is cold every morning.

I have done the following tests:

a) disconnect ClimateLine, set humidity to 6.0, and hit WarmUp. The water warmed up.
b) recommect ClimateLine, set at Auto, remove water container. Warm air through the hose.

So I assumed things were working, but I have no way of proving that the ClimateLine really works. The Bay Area has humidity in the 35%-45% all spring/summer, so maybe the H5I decides that no heating is needed.

Hence my query. How do you know why it is working now?

Manas

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Activa LT backup mask