24 L/M = 0.4 L/s
Awwwrrrrgghh!!! Math is NOT my forte!
60 (seconds in a minute) x .80 per second = 48.00 or 48 L/m
60 x .78 = 46.8 L/M
And then we have to add the 10 L/M so those two 95the percentile leak rates would be 58 L/M and 56.8 L/M.
Is that right, BillBolton???
understanding data on my elite 2
Re: understanding data on my elite 2
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator |
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.
My computer says I need to upgrade my brain to be compatible with its new software.
Re: understanding data on my elite 2
ok thanks
i am new to this
question
my leak rate is going to be off because i am not using a resmed mask
i know that now .i cant change that i have a good seal .78 l/s
will my other reading be somewhat acurate
ahi 22 ah 13 hi 9
my pressure reads no data for the last 3 nights ...not sure why ....leak rate too high ???
also if i can get a detailed reading with my smart card on a computer resdata 3,7 will i be able to see a more detailed readout
times of my ah and ai perhaps leak rate
i thankyou dan
i am new to this
question
my leak rate is going to be off because i am not using a resmed mask
i know that now .i cant change that i have a good seal .78 l/s
will my other reading be somewhat acurate
ahi 22 ah 13 hi 9
my pressure reads no data for the last 3 nights ...not sure why ....leak rate too high ???
also if i can get a detailed reading with my smart card on a computer resdata 3,7 will i be able to see a more detailed readout
times of my ah and ai perhaps leak rate
i thankyou dan
_________________
Mask: FlexiFit HC407 Nasal CPAP Mask with Headgear |
Additional Comments: pressure ..set at 9 ramp start 4 ..10 min to 9 |
Re: understanding data on my elite 2
You do NOT have a good seal yet w/your mask. Your reported Leak rate is .78 L/s.
If you have your
1] Mask Selection correctly set at Standard for your Fisher & Paykel HC 407 nasal cushion mask
2] and your pressure is set at 9 cms
3] the allowed vent rate for the HC 407 at 9 cms is 34 L/M
4] the Resmed subtracts 24 L/M from the leak rate BEFORE reporting the 95th percentile on the LCD screen when Mask Selection is set to Standard
5] so a reported Leak rate of more than 10 L/M is excessive and will somewhat distort your AHI data
6] your Resmed is reporting the 95th percentile Leak rate in L/s rather than in L/M
7] so you need to "translate" that 10 L/M into L/s to know what L/s is excessive Leak for you
If you have your
1] Mask Selection correctly set at Standard for your Fisher & Paykel HC 407 nasal cushion mask
2] and your pressure is set at 9 cms
3] the allowed vent rate for the HC 407 at 9 cms is 34 L/M
4] the Resmed subtracts 24 L/M from the leak rate BEFORE reporting the 95th percentile on the LCD screen when Mask Selection is set to Standard
5] so a reported Leak rate of more than 10 L/M is excessive and will somewhat distort your AHI data
6] your Resmed is reporting the 95th percentile Leak rate in L/s rather than in L/M
7] so you need to "translate" that 10 L/M into L/s to know what L/s is excessive Leak for you
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator |
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.
My computer says I need to upgrade my brain to be compatible with its new software.
Re: understanding data on my elite 2
Also, the Elite II CPAP gives no pressure data for most of us. That bit only exists in the menu because they use the same software in one of their APAP machines, which does give pressure data. Just ignore the no pressure data.
Re: understanding data on my elite 2
thanks for your quick responce
i'm getting a PHD in sleep apnea ..here
cheers dan
i'm getting a PHD in sleep apnea ..here
cheers dan
_________________
Mask: FlexiFit HC407 Nasal CPAP Mask with Headgear |
Additional Comments: pressure ..set at 9 ramp start 4 ..10 min to 9 |