Hawthorne wrote:I see your 90% pressure is 16 cm (meaning you spent 90% of that night at or below 16 cm). You say that was also your titrated pressure.
With your minimum now set at 8, it takes way too long to get to the pressure that will give you proper therapy.
Looks like the minimum pressure is set at 9, but I agree... 8 or 9.. either one is too low and would cause the machine to take wayyyy too long to move up there when 14, 15, or 16 are needed.
Hawthorne wrote:In my opinion (and I am not a medical person of any kind - just someone with sleep apnea) your minimum pressure needs to be way higher. Your maximum has to be higher as well since you are bumping up against the maximum a lot of the night.
I'm not a medical person either, and I agree on both counts.
Especially about raising the minimum
wayyyyy higher.
Hawthorne wrote:I don't know if you can stand to be at a higher minimum but, if it were me, I would set the minimum at 10 cm or even 12 cm would be better and put the maximum at 20 cm for a start. I think you can stand a higher minimum as I look at your pressure (top section), your leak, and the apneas.
In my opinion, your minimum is going to have to go higher than 10 cm or even 12 cm but you may have to work up, gradually, to a better minimum for you. Your minimum should eventually be maybe around 14 cm to 16 cm. I don't think you should jump up there right away. I am suggesting you start with a minmum of 10 cm or 12 cm and a maximum of 20 cm (as high as this machine will go) and, after a few nights see what your data looks like. You could use the ramp feature to get you to sleep before the pressure gets high.
The only thing I'd differ on is, I wouldn't work up gradually. Given the extreme daytime sleepiness Jeff is fighting, I'd go for therapeutic pressure now. I'd do one of these things now:
1. Keep the machine in auto mode, but set it to work like a straight CPAP by setting the min and max pressures the same. At a pressure of 16 that the sleep study came up with.
or
2. Keep the machine in auto mode, set for a range of 14 (or 15, or 16) to 20. Set the minimum pressure near or at the prescribed single pressure from the sleep study (pressure of 16) so that the machine can do a better, quicker job preventing events instead of the machine trying to play slow catch-up and losing the race.
Hawthorne wrote:It could also be that you should be on a bipap with the high pressures but you don't know, at this point how high your pressure will go.
Actually, I'd want to be on bipap even if the pressure doesn't have to go a lot higher. But, if I went to bipap, I'd still want the "lower" pressure setting (EPAP..the exhale pressure, in the case of a bipap machine) to be set up very close to, or at, 16. Perhaps bipap set for IPAP 18 / EPAP 14 for starters -- gather some data and see how that looks.
Hawthorne wrote:It also could be that you should set your machine in cpap mode at your titrated pressure of 16 cm.
Good idea. Instead of "cpap mode", I'd use auto mode with both the min and max set at 16 cm. It would "treat" as if it were straight cpap then, but would still gather "flow limitations" data, which is the only data that would be missing if set for just "cpap mode."
Hawthorne wrote:Do you know if you had centrals in your sleep study? I don't think this data shows any.
Right. The machine won't show centrals. If a few showed up, no big deal. If there had been a significant number, let's hope the sleep doctor was aware. It's always good, though, to get the full NPSG (nocturnal polysomnogram) report...not just the doc's summary page about the study.
Hawthorne wrote:You do have a lot of leaks going on as well and that may be responsible for some of your apneas. You do have to get those leaks under control. The leak line is not too bad for some of the night but there are those large spikes. Good thing about the leaks is that you are not having any large leaks, at least you weren't this night.
It might be easier to get the mask sealing better if you are working on it (puttying it, adjusting it, etc.) at the pressure it's going to have to withstand some or most of the night -- at 16. Getting it to fit and seal at a starting pressure of that minimum of 9 is one thing. Getting it to seal at 15 or 16 can be a whole other ballgame.
I'd work on "fitting" the mask while on your back, with mouth hanging open -- which is probably what happens when you sleep. When the jaw drops wayyyy down with an open mouth, the whole fit of a FF mask can change drastically. The area on both sides of the open mouth sinks in. And gravity is changing the geometry of the face when you are laying down, especially on your back. Working on getting it "leakproof" while sitting up on the side of the bed at a pressure of 9 and with mouth closed or just barely open won't be
anything like working on the leak situation during worst case scenario -- on your back, mouth hanging completely open, and a pressure of 16 blowing in.
I'd also think about trying the ResMed Ultra Mirage FF mask in several cushion sizes, including the "shallow" shape cushion in various sizes. The UM FF mask works well for many people. I'm not sure the newer Quattro is really an improvement on that older tried and true FF mask. But everyone is different, so one FF mask can always work better than another -- Quattro is the ticket for some, and Ultra Mirage FF for others.
Couple of things you might try with your current FF mask:
1. Use a chin strap. If it's the jaw dropping down too much, changing shape of face around the mouth, keeping the jaw up some might help. Doesn't have to be so tight that you can't mouth breathe easily within your FF mask. Just keep from having so much of a jaw drop during the complete relaxation of sleep.
2. Put an extra strap (like an Ace bandage or...the leg cut off a pair of tights...anything soft and stretchy) around the front of the mask, tied behind your head with a flat "double start knot":
DoubleStart Knot How to tie a flat knot behind head when using a homemade strap.
http://www.fieggen.com/shoelace/doublestartknot.htm
To keep the extra strap from sliding up or down on the front of a FF mask, put some adhesive backed velcro "dots" or strips on the plastic front to make the fabric catch and not slip. It's amazing what a homemade extra strap across the front can do for many masks to steady them in place. The extra strap doesn't have to be very tight. Just "being there" can help stop leaks, much like when you press your fingers lightly against the front of a mask to stop a leak.
If not doing this already, you might want to think about using some kind of hose hanging rig to prevent the main air hose from tugging on the mask when you turn over or move your head.
LINKS to Hose hangers and methods of managing the air hose
viewtopic.php?t=10640
Hawthorne wrote:Remember your pressure setting has nothing to do with the severity of your sleep apnea. It is just the pressure that is needed to keepYOUR airway open.
Right. Absolutely right.
ozij finished her post while I was still typing. Reading ozij's comments now, I agree so much with this:
ozij wrote:at 16, your doctor should to start thinking of a Bi-Leverl machine - not about provigil.
With a titration of 16, putting you at a minimum of 8 on the auto is asking for trouble.
and with this:
ozij wrote:All of the above assumes you have regular OSA -severe and in need of high pressure. I also assume those 19 central apneas in the first titration were the result of arousals.
However, since those central apneas may be an inidcation having complex sleep apnea -- where centrals appear under cpap - having the results of your new study is important - because that may indicate you need a new titration - on either a bi-level, or even on a third type of machine, an ASV one.