My friend just received a new Respironics RemStar Plus M series with CFLEX to replace his dying Puritan Bennett with 19K hours on it. The P-B was a simple CPAP machine. He doesn't know why they gave him the CFLEX. It wasn't ordered by a doctor.
Anyway, the motor noise of the CFLEX feature ramping up and down is an annoyance for him. He has the CFLEX setting on "1", the lowest, but that doesn't stave off the noise.
So the question... is there any way to turn off the CFLEX feature completely from the clinician menu?
Thanks in advance for any help.
Respironics: can CFLEX be turned off completely?
Although, I'll warn the both of you that the motor sound will still vary quite a bit in pitch even without C-Flex. The machine's job is to provide constant PRESSURE to the airway. To do that, flow needs to change quite a bit from inhale to exhale. If it were me, I'd leave C-Flex on and worry about putting the machine under the bed or in a closet to minimize noise.
I'm a programmer Jim, not a doctor!
- cpapernewbie
- Posts: 685
- Joined: Fri Oct 13, 2006 5:53 pm
TBoson
I can turn off CFlex by going to the clinician menu
My RT hide the AHI, Leak menu and the C-Flex on off menu,
Probably yr friend's RT did the same... they are all trained by the same company anyway...
To go to clinician menu just switch of the CPAP,
push the <- and -> button at the same time while turning it on again
wait until 2 beeps
then by hitting the + and - button you can find all those menus!
I can turn off CFlex by going to the clinician menu
My RT hide the AHI, Leak menu and the C-Flex on off menu,
Probably yr friend's RT did the same... they are all trained by the same company anyway...
To go to clinician menu just switch of the CPAP,
push the <- and -> button at the same time while turning it on again
wait until 2 beeps
then by hitting the + and - button you can find all those menus!
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: CPAP history: dumb tank, auto, PR M, PR System 1, PR BIPAP, PR System 1 model 60, Resmed S9, Resmed S10, Dreamstation |
Actually UNPLUG it, don't just switch it off, hold <- and -> while plugging it back in, wait until two beeps, etc.cpapernewbie wrote:To go to clinician menu just switch of the CPAP,
push the <- and -> button at the same time while turning it on again
wait until 2 beeps
then by hitting the + and - button you can find all those menus!
I'm a programmer Jim, not a doctor!
Hi blarg. It's not just the machine noise he's complaining about, but also the noise conducted up the hose. If the machine is providing constant pressure, there should be no change in motor sound. At least no change by the machine--- but perhaps from backflow or backpressure of exhaled air is what you are talking about?blarg wrote:Although, I'll warn the both of you that the motor sound will still vary quite a bit in pitch even without C-Flex. The machine's job is to provide constant PRESSURE to the airway. To do that, flow needs to change quite a bit from inhale to exhale. If it were me, I'd leave C-Flex on and worry about putting the machine under the bed or in a closet to minimize noise.
The mask makes noise too. If that's part of what's bugging him that can be fixed.TBoson wrote:Hi blarg. It's not just the machine noise he's complaining about, but also the noise conducted up the hose.
That's true if you're talking about a pipe, but we're talking about an airway that happens to be hooked up to lungs.TBoson wrote:If the machine is providing constant pressure, there should be no change in motor sound. At least no change by the machine--- but perhaps from backflow or backpressure of exhaled air is what you are talking about?
On an inhale, the machine has to put out more FLOW to maintain that constant pressure, because the lungs are sucking air out of the pipe. On an exhale, the machine has to put out less FLOW to maintain that constant pressure. That particular machine accomplishes this by varying motor speed. My BiPAP Auto does it with a sliding door shutter so that the motor pitch doesn't vary.
Either way, turn C-Flex off and listen. It'll vary, and quite a bit. It's not an air compressor that puts out constant flow with no regard to what's on the other side, it's an air compressor that specifically monitors what's going on at the output to maintain constant PRESSURE even though the lungs are sucking and pushing by varying FLOW.
Oh, and on the M Series, no, there is no way to turn those bright blue lights off. Sorry.