What? Admit that and take up residence in a padded room somewhere? No thanks.blarg wrote:Have you seen gravity?
I might have seen gravity once or twice. Why do you ask??
PeaceSleeper- CONGRATS!!!!!!!!
Interesting, because for a long time I've thought that when doctors prescribe a temporary auto-titrating trial with ANY autopap for a person to use at home, they (the doc, DME, SOMEBODY) ought to do a download within the first few days and EVERY WEEK thereafter. Just to see how things are going -- especially the leak rate -- but also to see if the typical "wide open" setting they prescribe needs to be changed. Tweaks during the trial could make the trial produce much more useful info, imho.PeaceSleeper wrote:There also seems to be some light with a sleep doc in Toronto about monitoring OSA patients much more closely. A doc there is thinking about starting a service---of course in cahoots with ResMed, to pull data from patient's machine weekly for assessment and possible action. Of course, the manufacturers want some of that service bureau revenue so it will be a slow process, but at least someone is moving in the right direction.
PeaceSleeper wrote:Hi All!
Wow, what fun this post started rolling! I have certainly learned even more about respiratory physiology than I planned on.
I am now over a week on my nifty auto BiPAP M series and have given up on the auto titrating function because it definitely wants to push my EPAP up until I start experiencing centrals, no matter what the settings or PSMax.
When I set the unit to IPAP 10.5 cm and EPAP 4.5 cm, my AHI goes consistently below 3 with almost no OAs. In fact, I think the OAs it is reading are just me swallowing or clearing my throat---though I guess those are supposed to be VS events.
Now remember, I was titrated at a reputable sleep lab on bilevel and 10/5 but was not fortunate enough to have the support and help of the kind folks on this forum. I had a super leaky mask and a humidifier that put out way too much H2O to make my asthma do anything but become more reactive. That was 5 years ago and I have been suffering high blood pressure and daily severe headaches since then.
With the help of lots of coaching and knowledge from the forum, I have been able to get my leaks under control, move to a machine that I am quite comfortable using for 6 or 7 hours, and have basically almost no events on a good night.
I just wanted to say---please keep up the discussion, my hope is that people will focus on the issues, and know that you are having a positive impact. Bravo!
Peace
_________________
CPAPopedia Keywords Contained In This Post (Click For Definition): bipap, humidifier, AHI, auto
-SWS. after reading your response, I was curious to learn more and here's some of what I found on ResMed's website:-SWS wrote:Bill, unlike traditional BiLevel, Resmed's EPR algorithmically attempts to "handle events". Specifically, Resmed's EPR tries to suspend lowering pressure during EPAP phase for obstructive apneas.NightHawkeye wrote: My understanding is that EPR simply reduces pressure on exhale while leaving the inhalation pressure or IPAP unchanged.
EPR is simply a BiPAP function, is it not?
Clearly, this is Bi-Level functionality. They go on to say that EPR will be turned OFF during detected apneas, as you indicated, after 15 seconds.ResMed website wrote:How does EPR work?
When the patient exhales, the S8 flow generator detects the beginning of exhalation and reduces motor speed to drop pressure. The patient or clinician chooses one of three comfort levels to determine the degree by which pressure will drop.
Setting 1 = mild comfort (1 cm H20)
Setting 2 = medium comfort (2 cm H20)
Setting 3 = maximum comfort (3 cm H20)
Blarg, I suppose this last one is what confused you about the duration of EPR. Unlike C-flex, EPR is not a transient phenomenon. Instead, EPR is exactly Bi-level operation. It turns OFF on occasion, as -SWS indicated, but that only happens if apneas, hypopneas, etc are detected.ResMed website wrote:EPR Timeout
If a patient's exhalation period exceeds 15 seconds, EPR immediately suspends. The treatment pressure reverts to set CPAP and remains suspended until the next inhalation phase is detected. EPR helps patients experience the highest level of comfort and benefit from their therapy. EPR is available on ResMed S8 EliteTM and the S8 AutoSet VantageTM flow generators.
I believe your statement is awfully heavy-handed there, -SWS. Everything I've read on the matter (excepting Snoredog's unsubstantiated self-serving proclamations) are that Bi-Level (BiPAP) therapy for most folks results in substantially reduced pressure requirements. Here, for instance, is a nearly 20 year old study in the medical journal, Chest (conducted by Respironics), which concludes:-SWS wrote:Quite a few patients can get away with traditional BiLevel EPAP being less than CPAP. But quite a few cannot. That last epidemiological reality is why a BiLevel PSG is so often considered "medically prudent" when switching from CPAP to traditional BiLevel.
Here's the link for anyone interested in reading the original.Chest, Vol 98, 317-324, Copyright © 1990 by American College of Chest Physicians wrote: . . . we have shown that by using a device that permits independent adjustment of EPAP and IPAP, obstructive sleep-disordered breathing can be eliminated at lower levels of expiratory airway pressure compared with conventional nasal CPAP therapy. This may reduce the adverse effects associated with nasal CPAP therapy and improve long-term therapeutic compliance.
Actually, -SWS, I haven't associated C-flex with aerophagia at all. I'm not saying there's not a relationship, just that if there is a connection, it hasn't been of major concern to me. As for the silent GERD, I'm of the opinion that while my own GERD didn't present me with the common "burning" stomach acid sensations, it nevertheless has become more obvious to me recently that there have been more subtle sensations of discomfort which I could have been paying attention to if the problem hadn't been so chronic. (not sure that answers your query at all, -SWS, but that's my experience.)-SWS wrote:Bill, while I have your ear would you mind sharing perceptions? I have always suffered from silent GERD (no perception of my GERD symptoms). I also don't perceive much difference among those various C-Flex settings---aside from increased aerophagia when I have C-Flex turned on. Curious---how does that compare with what you're experiencing?
I don't think one can say "EPR is exactly Bi-level operation" when EPR will be suspended in the presence of more than 15 seconds of no inspiration (inhalation.)NightHawkeye wrote:EPR is exactly Bi-level operation. It turns OFF on occasion, as -SWS indicated, but that only happens if apneas, hypopneas, etc are detected.
Bill, I think you are overlooking an extremely important word in what -SWS said. The word is EPAP.NightHawkeye wrote:I believe your statement is awfully heavy-handed there, -SWS. Everything I've read on the matter ... are that Bi-Level (BiPAP) therapy for most folks results in substantially reduced pressure requirements.-SWS wrote:Quite a few patients can get away with traditional BiLevel EPAP being less than CPAP. But quite a few cannot. That last epidemiological reality is why a BiLevel PSG is so often considered "medically prudent" when switching from CPAP to traditional BiLevel.
If you're saying that based on the study you quoted, I think you might be connecting the dots wrong on that one, Bill -- but maybe I am. I think the researchers were talking only about "lower levels of expiratory airway pressure."NightHawkeye wrote:Here, for instance, is a nearly 20 year old study in the medical journal, Chest (conducted by Respironics), which concludes:Chest, Vol 98, 317-324, Copyright © 1990 by American College of Chest Physicians wrote: . . . we have shown that by using a device that permits independent adjustment of EPAP and IPAP, obstructive sleep-disordered breathing can be eliminated at lower levels of expiratory airway pressure compared with conventional nasal CPAP therapy. This may reduce the adverse effects associated with nasal CPAP therapy and improve long-term therapeutic compliance.NightHawkeye wrote:The preponderance of evidence seemingly indicates that Bi-Level therapy results in a significant overall reduction in pressure requirements for the vast majority of CPAP patients.
I agree, he is misinterpreting the study, I downloaded Bill's study and wasted 10 pages of paper to print it out, still reading and interpreting it, but I agree they are talking about lower levels of "expiratory relief" compared to CPAP which has NONE!rested gal wrote:I don't think one can say "EPR is exactly Bi-level operation" when EPR will be suspended in the presence of more than 15 seconds of no inspiration (inhalation.)NightHawkeye wrote:EPR is exactly Bi-level operation. It turns OFF on occasion, as -SWS indicated, but that only happens if apneas, hypopneas, etc are detected.
That alone makes it not be "exactly" like a true bi-level machine. A true bi-level, as long as it's not an ST machine with a timed backup rate set, will continue delivering the lower EPAP pressure UNTIL the person begins an inhalation. A non-ST bi-level will never "suspend" that lower EPAP pressure and let a higher pressure (the IPAP) in UNTIL the person begins to inhale. Unlike EPR which will let the higher "regular" cpap pressure in again if enough seconds go by without an inhalation.
The only similarity I see in EPR with bi-level is the fact that EPR's pressure drop lasts throughout the entire exhalation. Other than than, EPR is not "exactly Bi-level operation", imho.
Bill, I think you are overlooking an extremely important word in what -SWS said. The word is EPAP.NightHawkeye wrote:I believe your statement is awfully heavy-handed there, -SWS. Everything I've read on the matter ... are that Bi-Level (BiPAP) therapy for most folks results in substantially reduced pressure requirements.-SWS wrote:Quite a few patients can get away with traditional BiLevel EPAP being less than CPAP. But quite a few cannot. That last epidemiological reality is why a BiLevel PSG is so often considered "medically prudent" when switching from CPAP to traditional BiLevel.
"Quite a few patients can get away with traditional BiLevel EPAP being less than CPAP"
The study you quoted is not at odds (imho) with what -SWS carefully explained. In fact, if you look at the words I'll emphasize in red, the study is talking about precisely what -SWS is talking about... about EPAP being set to eliminate obstructive events.
NightHawkeye wrote:Here, for instance, is a nearly 20 year old study in the medical journal, Chest (conducted by Respironics), which concludes:Chest, Vol 98, 317-324, Copyright © 1990 by American College of Chest Physicians wrote: . . . we have shown that by using a device that permits independent adjustment of EPAP and IPAP, obstructive sleep-disordered breathing can be eliminated at lower levels of expiratory airway pressure compared with conventional nasal CPAP therapy. This may reduce the adverse effects associated with nasal CPAP therapy and improve long-term therapeutic compliance.If you're saying that based on the study you quoted, I think you might be connecting the dots wrong on that one, Bill -- but maybe I am. I think the researchers were talking only about "lower levels of expiratory airway pressure."NightHawkeye wrote:The preponderance of evidence seemingly indicates that Bi-Level therapy results in a significant overall reduction in pressure requirements for the vast majority of CPAP patients.
A regular CPAP titration will likely come up with a higher single pressure to be prescribed than a bi-level's EPAP pressure. The straight cpap pressure has to take care of all events. The bi-level's EPAP pressure has to take care of obstructive apneas only.
A bi-level titration of the same person will likely come up with lower EPAP pressure (to take care of apneas.) But...the higher IPAP pressure (to take care of hypopneas, flow limitations and residual snores) arrived at in a bi-level titration could likely be just as high as what would have been reached in a simple cpap titration.
Makes sense for the researchers to say "This may reduce the adverse effects associated with nasal CPAP therapy and improve long-term therapeutic compliance." A very significant "adverse effect" for many CPAP users is difficulty in breathing out against high pressures.
But the "this" that the researchers are talking about is "lower levels of expiratory airway pressure." Same thing -SWS was talking about.
No argument, RG! I believe I noted the differences in detail.rested gal wrote:That alone makes it not be "exactly" like a true bi-level machine.
Not overlooking it all, RG. -SWS chose his words carefully, as did I. (I am not in disagreement with what -SWS said.) My point was simply that the vast majority of folks, at least according to the studies I've seen, obtain the same level of effective therapy with their machines titrated so that EPAP < CPAP, significantly less, rather than with EPAP = CPAP. I simply have not found any studies indicating to the contrary of the ones I posted.rested gal wrote:Bill, I think you are overlooking an extremely important word in what -SWS said.
NightHawkeye wrote:The preponderance of evidence seemingly indicates that Bi-Level therapy results in a significant overall reduction in pressure requirements for the vast majority of CPAP patients.
Indeed, RG, that is exactly what they were saying, and that is exactly the point I have been making. It is also the issue which -SWS keyed in on when he revived this thread. (It occurs to me as I look at this just prior to hitting the send button, that perhaps you misinterpreted what I said above. Note that I never said lower IPAP than CPAP. -my apologies for any clumsy wording.)rested gal wrote:If you're saying that based on the study you quoted, I think you might be connecting the dots wrong on that one, Bill -- but maybe I am. I think the researchers were talking only about "lower levels of expiratory airway pressure."
I understand that to be your interpretation, RG.rested gal wrote:A regular CPAP titration will likely come up with a higher single pressure to be prescribed than a bi-level's EPAP pressure. The straight cpap pressure has to take care of all events. The bi-level's EPAP pressure has to take care of obstructive apneas only.
Only issue I have here is that you seem to imply a BiPAP titration will result in a higher IPAP level than CPAP level. This is not what research shows, RG. The first study I referenced in this thread stated, as I recall, that for 89% of CPAP patients, the IPAP titration value was essentially the same as the CPAP titration value, while the EPAP titration value was only 60% of that. Those are very significant findings, RG. It is not a small difference, but a very large difference. Rather than setting EPAP equal to CPAP as some here claim, researchers found it should be set to only 60% of that value in 89% of patients. In other words, if one is titrated on CPAP at 10 cm, then the researchers found the equivalent BiPAP titration to be 10 cm IPAP, 6 cm EPAP - for most patients.rested gal wrote:A bi-level titration of the same person will likely come up with lower EPAP pressure (to take care of apneas.) But...the higher IPAP pressure (to take care of hypopneas, flow limitations and residual snores) arrived at in a bi-level titration could likely be just as high as what would have been reached in a simple cpap titration.
You bet! That 40% reduction in EPAP, without any increase in IPAP, apparently relieves a lot of grief!rested gal wrote:Makes sense for the researchers to say "This may reduce the adverse effects associated with nasal CPAP therapy and improve long-term therapeutic compliance." A very significant "adverse effect" for many CPAP users is difficulty in breathing out against high pressures.
Yes! On that we are in full agreement, RG.rested gal wrote:But the "this" that the researchers are talking about is "lower levels of expiratory airway pressure." Same thing -SWS was talking about.
Ok, I thought when you said "exactly" you meant "exactly." Glad to see that you didn't exactly mean it when you chose that exact word.NightHawkeye wrote:Not overlooking it all, RG. -SWS chose his words carefully, as did I.rested gal wrote:Bill, I think you are overlooking an extremely important word in what -SWS said.
Good. Kind'a sounded like you were when you said, "I believe your statement is awfully heavy-handed there, -SWS." Even with the "wink" included.NightHawkeye wrote:(I am not in disagreement with what -SWS said.)
Agreed, now that you specify "EPAP" instead of talking in general about "a significant overall reduction in pressure requirements for the vast majority of CPAP patients." Thanks for clarifying what you had in mind.NightHawkeye wrote: My point was simply that the vast majority of folks, at least according to the studies I've seen, obtain the same level of effective therapy with their machines titrated so that EPAP < CPAP, significantly less, rather than with EPAP = CPAP.
Not sure what you are taking issue with there or why you think I implied that.NightHawkeye wrote:Only issue I have here is that you seem to imply a BiPAP titration will result in a higher IPAP level than CPAP level.rested gal wrote:A bi-level titration of the same person will likely come up with lower EPAP pressure (to take care of apneas.) But...the higher IPAP pressure (to take care of hypopneas, flow limitations and residual snores) arrived at in a bi-level titration could likely be just as high as what would have been reached in a simple cpap titration.
NightHawkeye wrote:The first study I referenced in this thread stated, as I recall, that for 89% of CPAP patients, the IPAP titration value was essentially the same as the CPAP titration value
I have no problem with that. If I were gonna set the machine, I'd probably put the EPAP at 7 in that example...just to be sure.NightHawkeye wrote:Rather than setting EPAP equal to CPAP as some here claim, researchers found it should be set to only 60% of that value in 89% of patients. In other words, if one is titrated on CPAP at 10 cm, then the researchers found the equivalent BiPAP titration to be 10 cm IPAP, 6 cm EPAP - for most patients.
RG,rested gal wrote:I don't think one can say "EPR is exactly Bi-level operation" when EPR will be suspended in the presence of more than 15 seconds of no inspiration (inhalation.)NightHawkeye wrote:EPR is exactly Bi-level operation. It turns OFF on occasion, as -SWS indicated, but that only happens if apneas, hypopneas, etc are detected.
That alone makes it not be "exactly" like a true bi-level machine. A true bi-level, as long as it's not an ST machine with a timed backup rate set, will continue delivering the lower EPAP pressure UNTIL the person begins an inhalation. A non-ST bi-level will never "suspend" that lower EPAP pressure and let a higher pressure (the IPAP) in UNTIL the person begins to inhale. Unlike EPR which will let the higher "regular" cpap pressure in again if enough seconds go by without an inhalation.
The only similarity I see in EPR with bi-level is the fact that EPR's pressure drop lasts throughout the entire exhalation. Other than than, EPR is not "exactly Bi-level operation", imho.
<snip>NightHawkeye wrote:I believe your statement is awfully heavy-handed there, -SWS. Everything I've read on the matter ... are that Bi-Level (BiPAP) therapy for most folks results in substantially reduced pressure requirements.-SWS wrote:Quite a few patients can get away with traditional BiLevel EPAP being less than CPAP. But quite a few cannot. That last epidemiological reality is why a BiLevel PSG is so often considered "medically prudent" when switching from CPAP to traditional BiLevel.
That's exactly right, Bill. I revived this thread because I felt you were making some key points that were ignored. More in a later post on those two medical studies and even an EPAP=CPAP statement issued by no other than renowned Dr. Rappaport. In short I felt your contributions to this particular topic were too important to get "short-circuited".NightHawkeye wrote:Indeed, RG, that is exactly what they were saying, and that is exactly the point I have been making. It is also the issue which -SWS keyed in on when he revived this thread.
At least the design objective of "event handling" makes EPR BiLevel sound inherently safer than traditional BiLevel.dsm wrote: I do agree that the suspension during detected events makes it a 'different' Bilevel but it is Bilevel
Truer words were never written!PeaceSleeper wrote:Since I'm on my second swing at the bat with CPAP/BiPAP, I cannot help but comment that if a patient is not compliant, it doesn't matter what method of airway splinting they are prescribed---because they aren't using it.
I'd guess that if you are having a central and your breathing starts again before the "timer" (so to speak) of the three little pressure nudges is finished, the machine could think it had been working on an obstructive and would mark it as an obstructive.PeaceSleeper wrote:When in auto and my results show higher EPAP pressures during the night, my apneas get longer---so I believe the machine is believing it is treating an OA event when in reality it is a CA event.
Good question. Dunno.PeaceSleeper wrote:And what are the demographics in the population for this behavior, versus what I guess is the more "normal" behavior of almost 100% obstructive events for other people.
That's how many of us feel...thank goodness for help from the message boards!PeaceSleeper wrote:Without this forum I would be in a bad way physically and mentally, but now my outlook is pretty darned positive and I know I'm a hose head for life. Wahooooo!
SWS,-SWS wrote:That's exactly right, Bill. I revived this thread because I felt you were making some key points that were ignored. More in a later post on those two medical studies and even an EPAP=CPAP statement issued by no other than renowned Dr. Rappaport. In short I felt your contributions to this particular topic were too important to get "short-circuited".NightHawkeye wrote:Indeed, RG, that is exactly what they were saying, and that is exactly the point I have been making. It is also the issue which -SWS keyed in on when he revived this thread.
At least the design objective of "event handling" makes EPR BiLevel sound inherently safer than traditional BiLevel.dsm wrote: I do agree that the suspension during detected events makes it a 'different' Bilevel but it is Bilevel