Hi Guys,
Five days into the therapy, I am seeing 3-4 OSA events every night pressure range (6-14, PS=0).
The events seem to be happening at different pressures in the same nights and between nights (7, 8, 9, 10).
The overall AHI has been less that 1 all nights.
The 90% pressure has also been a variable - overall average of 5 days is 9.8.
So; should I be increasing the pressure (CPAPmin) all the way upto 10+ to see if it eliminates all OSAs or be content with a nice low AHI?
By the way, I have started noticing 3-4 CA (centrals) every night and I am suspecting thats a side effect of CPAP or just a machine interpretation of something else.
thanks!
Five days into the therapy, I am seeing 3-4 OSA events every night pressure range (6-14, PS=0).
The events seem to be happening at different pressures in the same nights and between nights (7, 8, 9, 10).
The overall AHI has been less that 1 all nights.
The 90% pressure has also been a variable - overall average of 5 days is 9.8.
So; should I be increasing the pressure (CPAPmin) all the way upto 10+ to see if it eliminates all OSAs or be content with a nice low AHI?
By the way, I have started noticing 3-4 CA (centrals) every night and I am suspecting thats a side effect of CPAP or just a machine interpretation of something else.
thanks!
Last edited by bones20 on Mon Aug 24, 2015 2:22 pm, edited 1 time in total.
Re: What should be your CPAPmin?
Hi, first of all if your 'centrals' occur just as you're falling asleep and waking up, those would be normal unless there were very many more and more likely throughout the night. Secondly, if you otherwise feel good and AHI's remain low, unless you're aware of major leaks, I would think you could leave things just as they are. I don't think lightening would strike if you raised your min. pressure to 7 for a few days to see if things change, but pursuing an AHI of zero is probably a bit unrealistic.
Re: What should be your CPAPmin?
Thanks Julie! Just to clarify, I am not talking about AHI but just AI
I will stop chasing the goal of 0 AI. And it makes sense, since less than 5 is considered normal.
Also I did raise the CPAPmin to 7 last night with pretty much the same result.
I will stop chasing the goal of 0 AI. And it makes sense, since less than 5 is considered normal.
Also I did raise the CPAPmin to 7 last night with pretty much the same result.
Re: What should be your CPAPmin?
Don't expect to eliminate all OSA events...you will drive yourself crazy...3-4 a night is great. An AHI of <1 is great...especially for a new user. My opinion would be to leave your pressure alone for several weeks....load Sleepyhead and study your data.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is a new AS10. |
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: What should be your CPAPmin?
The general consensus around here from past discussions over many years......is to set the minimum to the point where it eliminates almost all of the Obstructive events.bones20 wrote:Hi Guys,
Five days into the therapy, I am seeing 3-4 OSA events every night pressure range (6-14, PS=0).
The events seem to be happening at different pressures in the same nights and between nights (7, 8, 9, 10).
The overall AHI has been less that 1 all nights.
The 90% has also been a variable - overall average of 5 days is 9.8.
So; should I be increasing the pressure (CPAPmin) all the way upto 10+ to see if it eliminates all OSAs or be content with a nice low AHI?
By the way, I have started noticing 3-4 CA (centrals) every night and I am suspecting thats a side effect of CPAP or just a machine interpretation of something else.
thanks!
Since you're having some Centrals, you need to study your reports and note the pressure(s) at which they're occurring.
Some people have been known to have "pressure-induced Centrals", so if they're occurring at higher pressures, it might be a good idea to limit the maximum pressure below that threshold.
On the other hand, some people have "sleep onset Centrals", which have been described as "no big deal".
Since you're very new to this therapy, it can take awhile to see some consistent numbers. We don't sleep the same night after night. At some point, you may even want to try straight pressure and see how that works for you. For some people the changing pressures can be disturbing to their sleep.
Edit:
Rereading your post and profile, I see you're using a bi-level machine. So, in that regard, (again, and the same as before) the recommended minimum setting for EPAP is to where it eliminates almost all of the Obstructive events.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: What should be your CPAPmin?
Thanks Wuffman.
Sorry, but I am confused now. The response from the other gentleman indicates that 2-3 OE are fine? Also, from what I hear, an AHI < 5 does even merit SA treatment, right?
Also, the centrals are actually all in the low pressure range.
Sorry, but I am confused now. The response from the other gentleman indicates that 2-3 OE are fine? Also, from what I hear, an AHI < 5 does even merit SA treatment, right?
Also, the centrals are actually all in the low pressure range.
Wulfman... wrote:The general consensus around here from past discussions over many years......is to set the minimum to the point where it eliminates almost all of the Obstructive events.bones20 wrote:Hi Guys,
Five days into the therapy, I am seeing 3-4 OSA events every night pressure range (6-14, PS=0).
The events seem to be happening at different pressures in the same nights and between nights (7, 8, 9, 10).
The overall AHI has been less that 1 all nights.
The 90% has also been a variable - overall average of 5 days is 9.8.
So; should I be increasing the pressure (CPAPmin) all the way upto 10+ to see if it eliminates all OSAs or be content with a nice low AHI?
By the way, I have started noticing 3-4 CA (centrals) every night and I am suspecting thats a side effect of CPAP or just a machine interpretation of something else.
thanks!
Since you're having some Centrals, you need to study your reports and note the pressure(s) at which they're occurring.
Some people have been known to have "pressure-induced Centrals", so if they're occurring at higher pressures, it might be a good idea to limit the maximum pressure below that threshold.
On the other hand, some people have "sleep onset Centrals", which have been described as "no big deal".
Since you're very new to this therapy, it can take awhile to see some consistent numbers. We don't sleep the same night after night. At some point, you may even want to try straight pressure and see how that works for you. For some people the changing pressures can be disturbing to their sleep.
Edit:
Rereading your post and profile, I see you're using a bi-level machine. So, in that regard, (again, and the same as before) the recommended minimum setting for EPAP is to where it eliminates almost all of the Obstructive events.
Den
.
Re: What should be your CPAPmin?
Ignore the centrals unless there are very large numbers in clusters....You have no problem.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is a new AS10. |
-
- Posts: 1116
- Joined: Wed Feb 05, 2014 11:10 am
- Location: Rochester, MN
Re: What should be your CPAPmin?
The goal of cpap therapy is for AHI <5... 1,2,3, and 4 are fine; it is very rare to bring it down to O.bones20 wrote:Thanks Wuffman.
Sorry, but I am confused now. The response from the other gentleman indicates that 2-3 OE are fine? Also, from what I hear, an AHI < 5 does even merit SA treatment, right?
Also, the centrals are actually all in the low pressure range.
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: AirFit™ F10 Full Face Mask with Headgear |
Additional Comments: Sleepyhead |
Now using AirFit F10 mask; Quattro Air is backup mask. RemZzzz mask liners with both.
Re: What should be your CPAPmin?
SHADDUP!bones20 wrote:The overall AHI has been less that 1 all nights.
I have started noticing 3-4 CA (centrals) every night
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: What should be your CPAPmin?
To reiterate, you're new to this therapy and it usually takes some time to get the therapy settled in and the user adjusted to this new way of sleeping. You're going to have "events" no matter what your settings are and no matter how long you've been using this therapy......period.bones20 wrote:Thanks Wuffman.
Sorry, but I am confused now. The response from the other gentleman indicates that 2-3 OE are fine? Also, from what I hear, an AHI < 5 does even merit SA treatment, right?
Also, the centrals are actually all in the low pressure range.
I'm not clear about what you mean when you ask about the AHI less than 5 not meriting SA treatment.
Did you mean with or without therapy?
Personally, I'd feel like crap if my AHI was that high. My AHI when I had my sleep study was in the mid-40s....."severe".
Many people don't get XPAP prescribed if their AHI is less than 5.0, but depending on the circumstances, the person being diagnosed may also be just starting to experience OSA and their numbers might get worse over time.
If your (treated) AHI is typically less than 1.0 every night, that's GREAT!
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: What should be your CPAPmin?
He's asking simple questions. You are making it way too damn complicated.Wulfman... wrote:To reiterate, you're new to this therapy and it usually takes some time to get the therapy settled in and the user adjusted to this new way of sleeping. You're going to have "events" no matter what your settings are and no matter how long you've been using this therapy......period.
I'm not clear about what you mean when you ask about the AHI less than 5 not meriting SA treatment.
Did you mean with or without therapy?
Personally, I'd feel like crap if my AHI was that high. My AHI when I had my sleep study was in the mid-40s....."severe".
Many people don't get XPAP prescribed if their AHI is less than 5.0, but depending on the circumstances, the person being diagnosed may also be just starting to experience OSA and their numbers might get worse over time.
Re: What should be your CPAPmin?
Thanks Guys! I will continue with this as suggested and get back if I see something new.
My diagnosed AHI was 62 (still find it hard to believe). So, the machine is obviously working well it seems bringing it down to less than 1.
My diagnosed AHI was 62 (still find it hard to believe). So, the machine is obviously working well it seems bringing it down to less than 1.
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: What should be your CPAPmin?
I apologize if I was making things sound "complicated", but at least I won't tell you to "SHADDUP!".bones20 wrote:Thanks Guys! I will continue with this as suggested and get back if I see something new.
My diagnosed AHI was 62 (still find it hard to believe). So, the machine is obviously working well it seems bringing it down to less than 1.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: What should be your CPAPmin?
Not at all! Details and depth is exactly I am looking for.Wulfman... wrote:I apologize if I was making things sound "complicated", but at least I won't tell you to "SHADDUP!".bones20 wrote:Thanks Guys! I will continue with this as suggested and get back if I see something new.
My diagnosed AHI was 62 (still find it hard to believe). So, the machine is obviously working well it seems bringing it down to less than 1.
Den
.
I am just a newbie. Somethings I might say may sound immature or uninformed.
Re: What should be your CPAPmin?
Just like some of the ones who have been here for years and made thousands of posts.bones20 wrote:I am just a newbie. Somethings I might say may sound immature or uninformed.