I cannot get the Sleepy Head program to work

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
iPraiseHim
Posts: 4
Joined: Sun Jun 29, 2014 3:54 am

I cannot get the Sleepy Head program to work

Post by iPraiseHim » Fri Jul 04, 2014 10:18 pm

I cannot get the Sleepyhead program to work.
I have downloaded the file SleepyHead-20140623-Win32-0.9.6-testbuild-Qt5.3.0-BrokenGL-WinXP-MinGW-1.zip several times.
I unzipped it with the 7Zip program into its own folder.

When I run the extracted Sleep head program it say QT5Core.dll is missing from my computer. Yet I can see the file in the folder.

I am stumped. What do I do next?

I operate a windows 8.0 os with a fire fox browser.

I would really like to use this program.
please any suggestions would be helpful.

Thank you

User avatar
Pugsy
Posts: 64168
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: I cannot get the Sleepy Head program to work

Post by Pugsy » Fri Jul 04, 2014 10:28 pm


_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
Sclark08
Posts: 475
Joined: Mon Feb 18, 2013 10:13 am

Re: I cannot get the Sleepy Head program to work

Post by Sclark08 » Sat Jul 05, 2014 5:01 pm

Even if you get it to download properly a notice comes up that says it is test software and some of the software may not work yet

User avatar
jedimark
Posts: 1300
Joined: Fri Feb 18, 2011 12:45 am
Location: Bundaberg, Australia
Contact:

Re: I cannot get the Sleepy Head program to work

Post by jedimark » Sat Jul 05, 2014 9:02 pm

iPraiseHim wrote:I cannot get the Sleepyhead program to work.
I have downloaded the file SleepyHead-20140623-Win32-0.9.6-testbuild-Qt5.3.0-BrokenGL-WinXP-MinGW-1.zip several times.
I unzipped it with the 7Zip program into its own folder.

When I run the extracted Sleep head program it say QT5Core.dll is missing from my computer. Yet I can see the file in the folder.

I am stumped. What do I do next?

I operate a windows 8.0 os with a fire fox browser.

I would really like to use this program.
please any suggestions would be helpful.

Thank you
That should not be happening. Is the Qt5Core.dll file inside the same folder as the SleepyHead.exe executable?

SleepyHead.exe needs to live inside the folder with all it's plugins and DLL's.. If you want a desktop link, you need to right click on SleepyHead.exe, then create a shortcut, then drag the shortcut outside to your desktop/menu/wherever.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS50D+/F Oximeter, S9 VPAP Auto/Adapt, PRS1 Auto, Intellipap Auto, SleepyHead :)
Author of the free, cross platform, open-source sleep tracking software SleepyHead.
Download http://sleepyhead.jedimark.net
Source Code http://gitlab.com/sleepyhead/sleepyhead-code

User avatar
Sclark08
Posts: 475
Joined: Mon Feb 18, 2013 10:13 am

Re: I cannot get the Sleepy Head program to work

Post by Sclark08 » Sun Jul 06, 2014 3:21 am

I have a shortcut on my desktop but that message still comes up across the top of the page. It is highlighted in Red

User avatar
jedimark
Posts: 1300
Joined: Fri Feb 18, 2011 12:45 am
Location: Bundaberg, Australia
Contact:

Re: I cannot get the Sleepy Head program to work

Post by jedimark » Sun Jul 06, 2014 4:29 am

Sclark08 wrote:I have a shortcut on my desktop but that message still comes up across the top of the page. It is highlighted in Red
Um, that message is intentional and part of the program.. it's not an error, it's a reminder you are running pre-release version.

Earlier versions showed it too, just not in red..

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS50D+/F Oximeter, S9 VPAP Auto/Adapt, PRS1 Auto, Intellipap Auto, SleepyHead :)
Author of the free, cross platform, open-source sleep tracking software SleepyHead.
Download http://sleepyhead.jedimark.net
Source Code http://gitlab.com/sleepyhead/sleepyhead-code

iPraiseHim
Posts: 4
Joined: Sun Jun 29, 2014 3:54 am

Re: I cannot get the Sleepy Head program to work

Post by iPraiseHim » Sun Jul 06, 2014 10:57 am

I still get the same message QT5Core.dll missing.

1) I downloaded the ZIP file

2) I extracted the entire folder before I opened it.

3) I opened the folder and double clicked on Sleepyhead.exe

4) I still get the Qt5Core.dll missing message. What am I doing wrong? I have unzipped other zip files with no problem.

5) Where am I going wrong? I also tried the other two suggestions with the same result. I know it works for others so what is my problem?

User avatar
Pugsy
Posts: 64168
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: I cannot get the Sleepy Head program to work

Post by Pugsy » Sun Jul 06, 2014 11:12 am

Can you show me the unzipped folder contents where you are clicking on the SleepyHead application.exe file?
So that I can see exactly which folder you are in when you are clicking on the application.exe file?
Don't know about inserting images? Explained here
Screen shot thread viewtopic.php?f=1&t=81072&p=737779#p737779

You should be inside a folder with around 40 (or more depending on which SH version) files and folders in it and the SleepyHead application.exe file that you should be clicking on to open is the very last file in the long list of files within that folder.

Are you finding and clicking on this file to open/launch SleepyHead? Yours may have a different date as this is an old image the the application.exe file used to open SH is always the last file in the folder and my current version has 47 items in that main folder.

Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

Kinmd
Posts: 1
Joined: Tue Dec 16, 2014 12:05 pm

Re: I cannot get the Sleepy Head program to work

Post by Kinmd » Tue Dec 16, 2014 12:06 pm

I am havingthis exact same problem. Was there any solution found?

User avatar
Pugsy
Posts: 64168
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: I cannot get the Sleepy Head program to work

Post by Pugsy » Tue Dec 16, 2014 12:13 pm

Kinmd wrote:Was there any solution found?
Yes.
Solution in this thread right above your post.
If that doesn't work then you need to tell us exactly what is happening at any messages that you are seeing along with which Sleepyhead version and what the operating system is on your computer.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
OkyDoky
Posts: 2870
Joined: Mon Aug 25, 2014 5:18 pm

Re: I cannot get the Sleepy Head program to work

Post by OkyDoky » Tue Dec 16, 2014 12:28 pm

I was having the same problem because I was trying to load from the 7Zip file without extracting to another folder. Once this was done it worked. Oops sorry, I see you did that. Maybe it will help someone else.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760