ResScan Graph and Leak Question

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
jmw5430
Posts: 4
Joined: Tue Aug 13, 2013 7:48 am

ResScan Graph and Leak Question

Post by jmw5430 » Tue Aug 13, 2013 8:06 am

Hello,

First off, thanks to all of you for providing your knowledge to create this invaluable resource. It's been really helpful during my first months on CPAP.

I have a question that I was hoping some community member could help me out with. My wife has recently brought to my attention that occasionally at night, I'll let small puffs of air out of my mouth. My leak rates and all other relevant data appear to be well within the acceptable limits, and was wondering if this might be an issue. Also, I noticed last night that I woke up at one point last night and my mouth was open. No air was leaking out since my tongue was blocking my airway, but I can't say with certainty that that's always the case.

Also, If there's anything strange you see in the graphs, I'd love to hear about that as well.

I use a Resmed S9 Autoset & Swift FX Nasal Pillows. My graph from last night is below. I'd really appreciate any insight you could provide.

Thanks for the help!

Image
Image

_________________
MaskHumidifier

User avatar
Pugsy
Posts: 65246
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResScan Graph and Leak Question

Post by Pugsy » Tue Aug 13, 2013 9:02 am

I can't see your images.

Missing a step somewhere maybe??? Check this thread to see what you may be missing.
viewtopic.php?f=1&t=81072&p=737779#p737779

Use the preview button to confirm that your image shows up.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

jmw5430
Posts: 4
Joined: Tue Aug 13, 2013 7:48 am

Re: ResScan Graph and Leak Question

Post by jmw5430 » Tue Aug 13, 2013 9:24 am

Must have missed a step in there somewhere. Let's try this again:

Happy to provide zoomed images, different timeframes, etc. if it's helpful. Also wanted to note that my pressure is set wide open from 4-20.

Image

Image

_________________
MaskHumidifier

User avatar
Pugsy
Posts: 65246
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResScan Graph and Leak Question

Post by Pugsy » Tue Aug 13, 2013 9:40 am

Well if you are opening your mouth to do a little mouth breathing it sure isn't bad enough to show up on the leak graph.
I know what you mean by opening your mouth but no air escapes...I do that too sometimes. I also do a little mouth breathing for sure and sometimes see a little increases in the leak graph.

The red line is 24 L/min and that's what you want to avoid and you do that very nicely. Unless leaks or open mouth issues are waking you up and disturbing your sleep...they aren't large enough to negatively impact therapy and I wouldn't worry about them.
A little leak the machine can compensate for and you have a lot of room before leaks become a problem for you.
Those little small puffs of air that your wife notices and you don't...not hurting anything except maybe her sleep...

Your leak line...I would be doing a happy dance for sure if I had that line....mine is almost always more exciting.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: ResScan Graph and Leak Question

Post by avi123 » Tue Aug 13, 2013 9:43 am

Removed my Stats and graphs.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
Last edited by avi123 on Tue Aug 13, 2013 11:12 am, edited 1 time in total.

User avatar
Pugsy
Posts: 65246
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResScan Graph and Leak Question

Post by Pugsy » Tue Aug 13, 2013 10:16 am

If your leak line is your only question and it never really gets close to 24 L/min red line in ResScan...there's no need to produce volumes of reports for review. If your data last night is pretty much what you see the other nights...then there isn't much to discuss unless you have other questions.
Looks like what few events you do have are likely mostly centrals and not enough of them to warrant worry.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

jmw5430
Posts: 4
Joined: Tue Aug 13, 2013 7:48 am

Re: ResScan Graph and Leak Question

Post by jmw5430 » Tue Aug 13, 2013 10:19 am

This is great information. Thanks to both of you.

Good to know that small leaks don't have a serious impact on the therapy. I was concerned that since the behavior was physically observed by my wife, but not really recorded by the CPAP, that the analytics available on the S9 weren't entirely accurate.

Thanks again for the support and for contributing to such a valuable resource.

_________________
MaskHumidifier

sleepy_learner
Posts: 5
Joined: Tue Aug 06, 2013 10:28 am

Re: ResScan Graph and Leak Question

Post by sleepy_learner » Tue Aug 13, 2013 10:22 am

pillows often leak with high pressure settings. your pressures are not really very high. do you wash them every day. usually if the seal is poor, the starts leaking. if it bothers consider a nasal mask.

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: ResScan Graph and Leak Question

Post by avi123 » Tue Aug 13, 2013 11:11 am

jmw5430 wrote:This is great information. Thanks to both of you.

Good to know that small leaks don't have a serious impact on the therapy. I was concerned that since the behavior was physically observed by my wife, but not really recorded by the CPAP, that the analytics available on the S9 weren't entirely accurate.

Thanks again for the support and for contributing to such a valuable resource.

IMO, the data from the S9 on ResScan, is very accurate, but home CPAP's are not meant to replace sleep clinic's instrumentations.

I am going to remove my own graphs, now.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

User avatar
DoriC
Posts: 5214
Joined: Sat Sep 13, 2008 9:28 pm
Location: NJ

Re: ResScan Graph and Leak Question

Post by DoriC » Tue Aug 13, 2013 6:27 pm

Your data looks very good to me so the only questions I have are is your sleep undisturbed and how are you feeling during the day?

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: 14/8.4,PS=4, UMFF, 02@2L,
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08

jmw5430
Posts: 4
Joined: Tue Aug 13, 2013 7:48 am

Re: ResScan Graph and Leak Question

Post by jmw5430 » Wed Aug 14, 2013 11:31 am

Thanks, Dori.

I do tend to wake up 1-2 times throughout the course of the night. Not sure what to attribute that to, necessarily. I've still only been using the machine for a couple of months, so perhaps I'm still in the phase where I'm getting accustomed to it.

Overall, I feel less tired during the day, but have noticed some interesting pattern changes. Previously, I would be pretty lethargic in the morning, wake up a bit through the course of the day, and crash relatively early in the evening. What I've been noticing now is that I feel great in the morning, hit a lull shortly after lunch, and then catch a second wind around 2-3 in the afternoon. Maybe this is more of a normal cycle.

I'll be curious to see what happens in the coming months. From the research I've done on this board, it sounds like paying off a "sleep debt" can take months!

Thanks again!

_________________
MaskHumidifier