ASV Leak Graphs-Impossible to be Horizontal?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

ASV Leak Graphs-Impossible to be Horizontal?

Post by NateS » Wed May 15, 2013 9:52 am

I decided to make this subject into a separate Topic, because it is sort of lost when mixed in with leak discussions on straight CPAP and bi-Pap.

I am wondering whether my thinking on this subject is correct, or incorrect. I am on ASV. I recently posted a leak graph in another discussion (see below) showing that my leaks were consistently under the red ResScan line, thinking out loud that maybe this rules out leaks as being the cause of my dry mouth and that I should look elsewhere for the cause, wherein a discussion ensued about individual differences in what provides satisfaction, quality sleep and waking refreshed or not, etc.

One person on bi-pap posted a comment that, while my very jagged up-and-down line under the red line might have been fine for me, she would have been miserable with all those jumps up and down, even though all were under the red line.

This got me thinking - how is it even possible to have a straight horizontal Leak line on ASV (unless of course you have absolutely ZERO leaks which would admittedly produce a "flat" line)?

Except for ZERO, isn't a leak a fixed percentage of pressure, lost either because of a mask leak or mouth-breathing? Let's say that my leak is a constant 5% loss of pressure. (A purely arbitrary example with a number picked out of thin air.) Now, on ASV, my pressure is constantly jumping up and down, breath by breath. Therefore, since the Leak graph measures QUANTITY of Leak, not PERCENTAGE of Leak, wouldn't my leak line in my Leak graph ALSO have to be jagged? As the percentage of leak stayed constant, wouldn't the quantity of the leak have to vary with the changes in pressure?

Straight CPAP and bi-PAP users have horizontal lines in their Pressure graph because their pressure is constant (except for the difference in inhaling and exhaling for bi-PAP), so if their leak is a constant percentage, their Leak Graph would also be horizontal, not jagged. It would only be jagged if their percentage of leak was constantly changing.

But so long as an ASV machine is set to ASV, wouldn't a constant percentage of leak have to be displayed as a jagged line? How could it possibly be a horizontal line?

Or am I way off base here?

Regards, Nate

Image

Image

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

Tom W
Posts: 223
Joined: Thu Sep 01, 2011 4:59 am
Location: West Central Florida

Re: ASV Leak Graphs-Impossible to be Horizontal?

Post by Tom W » Wed May 15, 2013 10:17 am

NateS wrote: But so long as an ASV machine is set to ASV, wouldn't a constant percentage of leak have to be displayed as a jagged line? How could it possibly be a horizontal line?
Your assumptions seem correct to me since your PS varies and tends to run high (at least much higher than mine) and thus the IPAP is constantly changing (jagged) which effects the leak line.

One thing to consider is that the PS pulses (at least on the PR machine I have) are very short so they perhaps don't skew the average IPAP as much as we think but they do skew the average IPAP and mask flow rate to some degree.

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: ASV Leak Graphs-Impossible to be Horizontal?

Post by JohnBFisher » Wed May 15, 2013 11:27 am

NateS wrote:... How is it even possible to have a straight horizontal Leak line on ASV (unless of course you have absolutely ZERO leaks which would admittedly produce a "flat" line)? ...
Nate, that's one reason I say it's impossible to have a leak free setup with ASV therapy. Instead you try to manage the sound and amount of leaks. All that pressure changing stiffens up the silicone mask cushion. It also tends to blow out the seal. A mask liner seems to help, but it mostly just deadens the sound rather than eliminate the leaks.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: ASV Leak Graphs-Impossible to be Horizontal?

Post by NateS » Wed May 15, 2013 12:26 pm

JohnBFisher wrote:
NateS wrote:... How is it even possible to have a straight horizontal Leak line on ASV (unless of course you have absolutely ZERO leaks which would admittedly produce a "flat" line)? ...
Nate, that's one reason I say it's impossible to have a leak free setup with ASV therapy. Instead you try to manage the sound and amount of leaks. All that pressure changing stiffens up the silicone mask cushion. It also tends to blow out the seal. A mask liner seems to help, but it mostly just deadens the sound rather than eliminate the leaks.
Thanks, John.

It's very reassuring to hear that from you, as I think you are the resident Dean of ASV therapy here!

Regards, Nate

However, just for the fun of it, the challenge is still open! Anyone having a horizontal line for their Leak graph (other than zero) on ASV therapy, please put up your Leak, Events, and Pressure Graph for all to admire!

In fact, even a flat Zero Leak line on ASV would be very interesting to see.

And tell us how you did it!

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: ASV Leak Graphs-Impossible to be Horizontal?

Post by Jay Aitchsee » Wed May 15, 2013 1:20 pm

Nate, that's a very persuasive argument regarding leaks and pressure, but I'm not sure of the correlation in the example you posted. In fact, there seems to be areas where the opposite is occurring, i.e., lower pressures with higher leaks and vice versa. But I do agree that all things being equal, higher pressure should yield a greater leak and, therefore, one could expect some variation in the leak line corresponding to pressure changes.

A change of scale on the leak graph, like with 40 as a max, might be interesting.

Jay

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: ASV Leak Graphs-Impossible to be Horizontal?

Post by NateS » Wed May 15, 2013 1:48 pm

Jay Aitchsee wrote:Nate, that's a very persuasive argument regarding leaks and pressure, but I'm not sure of the correlation in the example you posted. In fact, there seems to be areas where the opposite is occurring, i.e., lower pressures with higher leaks and vice versa. But I do agree that all things being equal, higher pressure should yield a greater leak and, therefore, one could expect some variation in the leak line corresponding to pressure changes.

A change of scale on the leak graph, like with 40 as a max, might be interesting.

Jay
Jay,

Yes, I agree that a fixed percentage of leak rate was presented as a hypothesis. A more precise hypothesis would require sufficient knowledge of the varying efficiency of the mask seal and mouth seal with which to construct an algorithm relating percentage of lost pressure to applied pressure. My hunch is that such an adjustment would still not produce a horizontal leak line, because the range of efficiency between highest and lowest pressures would be in a much narrower band than the range of pressures applied. And yes increased pressure could sometimes cause a drop in lost pressure, but in any such series of events IMHO you are not going to get a horizontal leak line while on ASV, except through a miracle of a most dramatic series of mathematical coincidences over a night's sleep.

As for changing the scale on the Leak graph, that's probably temporarily doable within ResScan, but I'm not sure that would change anything. I'll try it when I have time, and post the resulting graph.

Regards, Nate

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

letsride
Posts: 158
Joined: Sun Mar 24, 2013 2:49 pm

Re: ASV Leak Graphs-Impossible to be Horizontal?

Post by letsride » Wed May 15, 2013 8:21 pm

Here is a look at my stats.
Image

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: ResMed Aircurve 10 ASV

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: ASV Leak Graphs-Impossible to be Horizontal?

Post by NateS » Wed May 15, 2013 8:28 pm

Excellent, letsride!

And also proof that you don't necessarily need a FFM with ASV in all cases.

Regards, Nate

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

letsride
Posts: 158
Joined: Sun Mar 24, 2013 2:49 pm

Re: ASV Leak Graphs-Impossible to be Horizontal?

Post by letsride » Wed May 15, 2013 8:42 pm


_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: ResMed Aircurve 10 ASV
Last edited by letsride on Wed May 15, 2013 8:52 pm, edited 1 time in total.

User avatar
RogerSC
Posts: 1911
Joined: Sat Jul 21, 2012 1:11 pm

Re: ASV Leak Graphs-Impossible to be Horizontal?

Post by RogerSC » Wed May 15, 2013 8:44 pm

I usually shoot for a median leak of 0 for the night, which tells me that there was enough 0 (a ResScan leak of 0 *smile*) leak for me. That's about as good as it gets for me, along with having as much under the red line as possible. I usually have two or three spikes above the red line any given night, but not too bad. ASV is a whole different world with the rapid pressure swings you guys get *smile*. I get pressure swings, but not nearly as rapid or as large in magnitude.

The only time I can see that you would be able to get a flat 0 leak (or just the normal mask vent flow rate for non-ResMed) is if you were a sleeper that didn't move much and where either on straight cpap or apap with a fairly narrow range. My main problem is that I'm an active side sleeper and move around a *lot* at night. I'm not kicking anyone, but I'm changing positions a lot more than most people (including my wife *smile*).
ResMed AirSense 10 Autoset
Philips Respironics Dreamwear nasal mask

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: ASV Leak Graphs-Impossible to be Horizontal?

Post by Jay Aitchsee » Thu May 16, 2013 5:00 am

NateS wrote: And also proof that you don't necessarily need a FFM with ASV in all cases.
And also that letsride probably doesn't suffer mouth leaks

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video