ResMed S9 Filter schedule

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
pootsie
Posts: 229
Joined: Mon Apr 29, 2013 6:23 pm

ResMed S9 Filter schedule

Post by pootsie » Thu Aug 01, 2013 10:39 am

Understanding that YMMV, I am curious how often do you S9 users change your filters?

I'm fairly new to this. I went more than 2 months without thinking about it. I changed it last night and the one I pulled out was yucky. Not super-horrible but with a dog and two cats, might have changed sooner.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure: 9 cm H2O. Diagnosis: OSA with AHI 10.6.
You are the Zzz's knees!

User avatar
Pugsy
Posts: 64159
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResMed S9 Filter schedule

Post by Pugsy » Thu Aug 01, 2013 10:54 am

Two to four weeks is pretty much my fine filter changing routine with either the S9 or the PR S1.
Normally I take a quick look at 2 weeks and if it isn't too yucky looking I can wait another couple of weeks but rarely go past 4 weeks.
It kinda depends on what all I have going on here. I live in a tiny and very old farmhouse ...out in the country and if the weather is dry and we have more dust in the air...it gets yucky faster. Plus I do have dogs and a cat.
So not the greatest of environments to start with.
Dusty old house and out in the country where there's even more dust and I have pets...but I suspect that even without the pets the filters would get pretty yucky.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
LSAT
Posts: 13238
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: ResMed S9 Filter schedule

Post by LSAT » Thu Aug 01, 2013 11:08 am

The ResMed S9 manual says replace the filter every 6 months...more often if necessary. Just inspect the filter monthly and if it's dirty replace it.

Kind of like changing the oil in your car. The manufacturer recommends every 7500 miles...most people do it sooner.

User avatar
Denial Dave
Posts: 1335
Joined: Wed Sep 19, 2012 6:45 am
Location: Connecticut

Re: ResMed S9 Filter schedule

Post by Denial Dave » Thu Aug 01, 2013 1:05 pm

How frequently you change your filters will depend on how they look after a few weeks of use.

've gotten into a routine of changing mine the 1 week of every month.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Started at VPAP pressure setting of 20/14.4. I've survived Stage 1 cancer and lost 80+ lbs. Pressure is now 14.5 / 11
if you don't know where you are going... any road will take you there.... George Harrison

User avatar
Sheriff Buford
Posts: 4086
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: ResMed S9 Filter schedule

Post by Sheriff Buford » Thu Aug 01, 2013 1:16 pm

I change mine twice a month (every other week)... but that's me... Sheriff

User avatar
pootsie
Posts: 229
Joined: Mon Apr 29, 2013 6:23 pm

Re: ResMed S9 Filter schedule

Post by pootsie » Thu Aug 01, 2013 1:20 pm

Thanks, guys

I guess really my question is, "how dirty is too dirty?" How do I know its effectiveness is impaired? Ta

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure: 9 cm H2O. Diagnosis: OSA with AHI 10.6.
You are the Zzz's knees!

User avatar
caffeinatedcfo
Posts: 690
Joined: Sun Mar 24, 2013 9:19 am
Location: Upstate NY

Re: ResMed S9 Filter schedule

Post by caffeinatedcfo » Thu Aug 01, 2013 5:50 pm

I used my first one a little over two months - I just take it out and flick any dust off it weekly. It was still white when I replaced it and most likely didn't really need to.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: SleepyHead software; using APAP mode 10-12cm & EPR 3

User avatar
archangle
Posts: 9294
Joined: Sun Mar 27, 2011 11:55 am

Re: ResMed S9 Filter schedule

Post by archangle » Fri Aug 02, 2013 10:17 am

ResMed says 6 months or more often if needed.

Realize that as an air filter gets dirty, it actually filters out more dust, since the fine dust particles act like filters themselves. The problem is that the pressure drop gets higher and higher. The CPAP machine will compensate, but the motor has to work harder and it will eventually start to overheat, or it will not be able to blow all the air you need.

The filter is more for protecting the machine than protecting you, anyway. The air going through the CPAP is the same air you breathe all day long.

However, there is the yuck factor.

Respironics PRS1 says wash the grey foam filter every 2 weeks, replace every 6 months. White PRS1 filters should be replaced monthly.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
Sheriff Buford
Posts: 4086
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: ResMed S9 Filter schedule

Post by Sheriff Buford » Mon Aug 05, 2013 4:58 am

pootsie wrote:Thanks, guys

I guess really my question is, "how dirty is too dirty?" How do I know its effectiveness is impaired? Ta
Archangle has a good point. I change mine (bi-weekly) BEFORE the "yuck-factor" kicks in. I can't tell the difference between the old filter and the new one... both still white and no difference in the color. But... that's just me. I shopped around and found cpapfiltersandsupplies.com. They sell a 5 pack of filters for $5.81 and a 50 pack for $51.60... pretty good deal.

Sheriff

MomD
Posts: 1
Joined: Thu Dec 26, 2013 11:15 am

Total newbie question - ResMed S9 Filter change tutorial nee

Post by MomD » Thu Dec 26, 2013 11:26 am

Hello S9 users - Mom uses a Cpap - she and dad had it all figured out. Dad passed away and now we kids are trying to make sense of all this. Mom had only ever changed the hose and face mask - never talked about a filter, so I see them in the box (but the manual we have for the Resmed is very basic and says nothing about the filters. How do I change the filter? (Filter change 101 needed here.) Many thanks -

I'm also reading some on the site about the humidifier - how do I know if she has a humidifier or not? In the ResMed S9 is it part of the one machine? Or is it a separate unit? Mom's having trouble with "knocking" sound and I'm reading that it may relate to humidifier level. Any help greatly appreciated!

User avatar
chunkyfrog
Posts: 34449
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: ResMed S9 Filter schedule

Post by chunkyfrog » Thu Dec 26, 2013 11:34 am

https://www.youtube.com/watch?v=s1RThpOYxt8
The humidifier is a separate box that hooks to the blower unit. It has an oval-shaped window in the top.
The top opens up and the tank inside needs fresh distilled water. Removing to fill is safer.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

craiga
Posts: 19
Joined: Sat Oct 26, 2013 2:17 pm

Re: ResMed S9 Filter schedule

Post by craiga » Thu Dec 26, 2013 9:13 pm

The DME said check it every couple weeks but change it every 2 mo.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Tried Mirage Quattro, Simplus, Quattro Air, Mirage Liberty and I am using sleepyhead software

User avatar
jaybeem
Posts: 189
Joined: Sat May 04, 2013 7:55 pm
Location: CA, USA

Re: ResMed S9 Filter schedule

Post by jaybeem » Thu Dec 26, 2013 10:50 pm

I change the S9 hypo-allergenic filter every two weeks. My asthma is well controlled, thanks to that filter. But if I let the filter go much over two weeks, the asthma is back at night, regardless of how the filter "looks."

_________________
Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure 8. EPR 1. Zero ramp. SleepyHead software.

NeNeen
Posts: 49
Joined: Thu Jul 26, 2012 5:23 pm
Location: Seminole, OK
Contact:

Re: ResMed S9 Filter schedule

Post by NeNeen » Tue Dec 31, 2013 2:28 pm

Medicare authorize 2 white filters every month and 1 black foam filter every 3 months. So I change the white 1 every other week and wash the black one every other week.

_________________
Mask
Additional Comments: EncoreBasic 2.1 Software
Getting my CPAP machine was the best thing ever. I call her my BFF.

User avatar
Jeannh
Posts: 296
Joined: Sun Aug 18, 2013 12:08 pm
Location: CA

Re: ResMed S9 Filter schedule

Post by Jeannh » Tue Dec 31, 2013 3:21 pm

NeNeen wrote:So I change the white 1 every other week and wash the black one every other week.
Can you use both the white and the black filter together?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: EPAP 8, IPAP 14, PS 3.6