Measured CPAP power consumption
Measured CPAP power consumption
We often get questions about actual power consumption of CPAP machines.
For those of you with measurement capability, what do you get?
Here's one session data.
ResMed S9 AutoSet with heated humidifier. No heated hose.
Auto 18-20, EPR=3, humidifier=4. Fairly humid night. Room temp around 75F.
Kill a Watt power meter measuring AC power into the power brick.
4:32 hours.
0.15 Kilowatt (Correction Kilowatt hours)
Average power 34 watts.
Looking at the instantaneous power readings as I inhale and exhale and the heater runs, power varies all over the place. Between 15 and 90 watts read on the meter.
For those of you with measurement capability, what do you get?
Here's one session data.
ResMed S9 AutoSet with heated humidifier. No heated hose.
Auto 18-20, EPR=3, humidifier=4. Fairly humid night. Room temp around 75F.
Kill a Watt power meter measuring AC power into the power brick.
4:32 hours.
0.15 Kilowatt (Correction Kilowatt hours)
Average power 34 watts.
Looking at the instantaneous power readings as I inhale and exhale and the heater runs, power varies all over the place. Between 15 and 90 watts read on the meter.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Last edited by archangle on Sun Jul 22, 2012 8:56 pm, edited 1 time in total.
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Re: Measured CPAP power consumption
What is a "Kill a Watt power meter" ?
And what's the meaning of this:
4:32 hours.
0.15 Kilowatt
Average power 34 watts.
You don't measure Watts but Amperes and Voltage and calculate the Watts:
Looking at the instantaneous power readings as I inhale and exhale and the heater runs, power varies all over the place. Between 15 and 90 watts read on the meter.
And what's the meaning of this:
4:32 hours.
0.15 Kilowatt
Average power 34 watts.
You don't measure Watts but Amperes and Voltage and calculate the Watts:
Looking at the instantaneous power readings as I inhale and exhale and the heater runs, power varies all over the place. Between 15 and 90 watts read on the meter.
_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6 |
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
-
- Posts: 748
- Joined: Sun Jun 24, 2012 12:39 pm
- Location: Cochise County AZ
Re: Measured CPAP power consumption
"I am a man of peace, but if war comes to my door it will find me home." - Winston Churchill
Re: Measured CPAP power consumption
This is interesting, archangle. I unplug pretty much anything unpluggable when I'm done using them, with the exception of lamps, refrigerator, oven, in an effort to avoid phantom energy drain. I've often wondered to what degree I was canceling that out by using the CPAP all night long.
Epworth Sleepiness Scale: 14
Diagnostic study: overall AHI: 0.2 events/hour; overall RDI: 45 events/hour
Titration study: AHI: 6.1; RDI: 27; CPAP pressures: 5-8cm
Not-tired behind my eyes and with a clear, cool head!
Diagnostic study: overall AHI: 0.2 events/hour; overall RDI: 45 events/hour
Titration study: AHI: 6.1; RDI: 27; CPAP pressures: 5-8cm
Not-tired behind my eyes and with a clear, cool head!
Re: Measured CPAP power consumption
Thanks for showing the Kill a Watt.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6 |
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
-
- Posts: 611
- Joined: Wed Jun 08, 2011 5:20 pm
- Location: Sydney,Australia
Re: Measured CPAP power consumption
I get a deduction from my power bill because I am a cpap user. It is about $19 a qtr. Dont know if it covers it or not.
- chunkyfrog
- Posts: 34545
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Measured CPAP power consumption
We used the same device to determine our power cost for distilling water.
--About 13 cents a gallon in winter; 20 cents in summer.
(seasonal electric rates are different--especially with electric heat)
0
--About 13 cents a gallon in winter; 20 cents in summer.
(seasonal electric rates are different--especially with electric heat)
0
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Measured CPAP power consumption
Here is a Typical Power Consumption of a household. What counts is Kilowatts Hours or Annual kWH:
http://www.oksolar.com/technical/consumption.html
If I take archangel's measurement numbers of:
4:32 hours.
0.15 Kilowatt = 150 Watts
Average power 34 watts
Then it means that the CPAP with the humidifier ON, is using 150 watts. If I multipy it by the 7 hours that I sleep nightly
and by 365 days in a year I get:
150 x 7 x 365 = 383,250 Watts = 383 kiloWatts Hours (KWH) per year.
So it consumes annually as a dishwasher which consume 363 KWH annually or twice as a window fan consuming 170 KWH annually.
If I assume that the average price of residential electricity was 12¢/kWh in the U.S. in December 2011, and ranged from 7.5¢ in Idaho to 36¢ in Hawaii, then the CPAP electrical consumption costs me:
383 KWH x 12 cents/KWH = 4,596 cents = app $46 annually.
http://www.oksolar.com/technical/consumption.html
If I take archangel's measurement numbers of:
4:32 hours.
0.15 Kilowatt = 150 Watts
Average power 34 watts
Then it means that the CPAP with the humidifier ON, is using 150 watts. If I multipy it by the 7 hours that I sleep nightly
and by 365 days in a year I get:
150 x 7 x 365 = 383,250 Watts = 383 kiloWatts Hours (KWH) per year.
So it consumes annually as a dishwasher which consume 363 KWH annually or twice as a window fan consuming 170 KWH annually.
If I assume that the average price of residential electricity was 12¢/kWh in the U.S. in December 2011, and ranged from 7.5¢ in Idaho to 36¢ in Hawaii, then the CPAP electrical consumption costs me:
383 KWH x 12 cents/KWH = 4,596 cents = app $46 annually.
_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6 |
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
- Lizistired
- Posts: 2835
- Joined: Tue Dec 14, 2010 10:47 pm
- Location: Indiana
Re: Measured CPAP power consumption
Breathing all night.... Priceless!
_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Swift FX sometimes, CMS-50F, Cervical collar sometimes, White noise, Zeo... I'm not well, but I'm better. |
ResScan: http://www.resmed.com/int/assets/html/s ... c=patients
ResScan Tutorial- http://montfordhouse.com/cpap/resscan_tutorial/
Machines Video: http://www.cpaplibrary.com/machine-education
ResScan Tutorial- http://montfordhouse.com/cpap/resscan_tutorial/
Machines Video: http://www.cpaplibrary.com/machine-education
Re: Measured CPAP power consumption
Sorry about that. It was 0.15 Kilowatt hours. Blame it on dain bramage.avi123 wrote: 4:32 hours.
0.15 Kilowatt = 150 Watts
Average power 34 watts
For 4:32 hours, that's 34 watts average power usage.
Roughly 0.3 kWh for one night's use. 3-5 cents worth of electricity a night. $10 to 15 a year.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Re: Measured CPAP power consumption
Further information:
My S9 AutoSet power supply draws about 2 watts when the machine is turned off.
My PRS1 Auto power supply appears to draw less than 1 watt when turned off.
My S9 AutoSet power supply draws about 2 watts when the machine is turned off.
My PRS1 Auto power supply appears to draw less than 1 watt when turned off.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Re: Measured CPAP power consumption
Further information
The power supply (Brick) for the S9 comes in 90 watts and 30 watts capacities. I assume that the 90 watts is for use with the humidifier. On the S9 Autoset body it's written 24 volts DC and 3.75 amps (for system). If I multiply the two numbers I get 90 voltampers, which jibes with the 90 watts power supply.
So I am back to my first calculation and I don't need the Kill-a-Watts gadget.
The annual cost of electricity to run the S9 Autoset with the humidifier:
90 watts x 7 hours x 365 days = 229,950 watts hours= 230 KWH
230 KWH x 12 cents/KWH= 2,760 cents = $27.6 annually.
If I don't use the humidifier then its one third as much, about $9.5 annually.
The power supply (Brick) for the S9 comes in 90 watts and 30 watts capacities. I assume that the 90 watts is for use with the humidifier. On the S9 Autoset body it's written 24 volts DC and 3.75 amps (for system). If I multiply the two numbers I get 90 voltampers, which jibes with the 90 watts power supply.
So I am back to my first calculation and I don't need the Kill-a-Watts gadget.
The annual cost of electricity to run the S9 Autoset with the humidifier:
90 watts x 7 hours x 365 days = 229,950 watts hours= 230 KWH
230 KWH x 12 cents/KWH= 2,760 cents = $27.6 annually.
If I don't use the humidifier then its one third as much, about $9.5 annually.
_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6 |
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
Re: Measured CPAP power consumption
The label rating on almost any electrical device will be a maximum number. Actual usage is almost always less.avi123 wrote:Further information
The power supply (Brick) for the S9 comes in 90 watts and 30 watts capacities. I assume that the 90 watts is for use with the humidifier. On the S9 Autoset body it's written 24 volts DC and 3.75 amps (for system). If I multiply the two numbers I get 90 voltampers, which jibes with the 90 watts power supply.
So I am back to my first calculation and I don't need the Kill-a-Watts gadget.
This thread is about actual measured power consumption, not the ratings. In my case, it's about 1/3 of the rated maximum. If I were deciding what size battery I need for emergency back, I'd need a battery 3 times a large if I wanted to use the rated maximum.
If I want to figure out actual yearly cost, I need to use actual consumption, not rated maximum. You can use the rated maximum, but realize you're probably overestimating actual cost.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Re: Measured CPAP power consumption
That means your niebhors are paying $15 extra a quarter, because your electric company allows you a discount that is out of line with your use, Socialism at it's best. Jimozze_dollar wrote:I get a deduction from my power bill because I am a cpap user. It is about $19 a qtr. Dont know if it covers it or not.
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
Re: Measured CPAP power consumption
Once again things seem to be much more expensive Down Under (Aust). Our electricty went up on 1st July due to the Carbon Tax. I received a letter from my supplier today with the new rates.
I have a "use of time" meter and now pay:
Peak Consumption (2pm - 10pm) - 53.59 cents
Shoulder Consumption (8am - 2pm) - 21.82 cents
Off Peak Consumption (10pm - 8am) - 13.42cents
Luckily when I sleep its Off Peak - not that would stop me sleeping now that I actually sleep!
I have a "use of time" meter and now pay:
Peak Consumption (2pm - 10pm) - 53.59 cents
Shoulder Consumption (8am - 2pm) - 21.82 cents
Off Peak Consumption (10pm - 8am) - 13.42cents
Luckily when I sleep its Off Peak - not that would stop me sleeping now that I actually sleep!