Postby squid13 on Tue Jul 03, 2012 7:31 am

What determines what pressure support number you use. I know it's IPAP minus EPAP. Say that is 4 but could you use a 3 or a 2 and if that is true what effect would it have.

_________________
Mask: Forma Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 VPAP Adapt 36037, ResScan 4.3, SleepyHead V.0.9.3
Last edited by squid13 on Tue Jul 03, 2012 2:14 pm, edited 1 time in total.
US Navy Retired 1973,S9 Adapt 36037, Mode: ASV Auto, Min EPAP: 6.8, Max EPAP: 15.0, Min PS: 3.0, Max PS: 15.0

User avatar
jnk
 
Posts: 5818
Joined: Mon Jun 30, 2008 3:03 pm
Location: Brooklyn Heights
Gender: Male

Re: Setting BiLevel Pressure Support.

Postby jnk on Tue Jul 03, 2012 8:56 am

squid13 wrote:What determines what pressure support number you use. I know it's IPAP minus EPAP. Say that is 4 but could you use a 3 or a 2 and if that is true what effect would it have.

The official answer (although I am nothing official, myself) is that the difference between the prescribed IPAP and the prescribed EPAP is the prescribed PS, as I understand it. The nice thing about your machine is that it can be run as an auto-bilevel while still keeping the PS, the distance between EPAP and IPAP, constant at all times.

Four is the most common PS, since recent AASM suggestions suggest starting at 4 cm for PS during a bilevel titration. A prescription of less than 4 PS these days would be considered nonstandard, I believe.

As far as what effect a different PS would have on a particular person, that would depend on the reasons for the bilevel prescription in the first place for that particular patient.

If you can't help but tinker with it :) , you might want to watch both your AHI AND your O2, using an overnight oximeter. But I would never suggest that anyone tinker with that without having his doc on board with his doing so.

But hey, that's just me.
"All quotes are actually misquotes."-- Never directly attributed to Bartlett.

User avatar
squid13
 
Posts: 1677
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL
Gender: Male

Re: Setting BiLevel Pressure Support in S9 VPAP Auto

Postby squid13 on Tue Jul 03, 2012 9:05 am

I know the ResMed Clinical setup says adjust for patient comfort. So say if you set it for 10 EPAP and 13 IPAP you set it at 3.

_________________
Mask: Forma Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 VPAP Adapt 36037, ResScan 4.3, SleepyHead V.0.9.3
Last edited by squid13 on Tue Jul 03, 2012 2:13 pm, edited 1 time in total.
US Navy Retired 1973,S9 Adapt 36037, Mode: ASV Auto, Min EPAP: 6.8, Max EPAP: 15.0, Min PS: 3.0, Max PS: 15.0

User avatar
Pugsy
 
Posts: 27359
Joined: Thu May 14, 2009 9:31 am
Location: SW Missouri, USA
Gender: Female

Re: Setting BiLevel Pressure Support.

Postby Pugsy on Tue Jul 03, 2012 9:17 am

When I got my bilevel machines I tried to find an explanation as to which pressure support setting might be advisable in what situation and never could find anything. Setting it with a very narrow PS (like 2) sort of makes it work like the Autoset with EPR of 2. While of course not exactly like it because the timing is a bit different it isn't hugely different.
I opted for PS of 4 mainly for comfort and I rest well with it and that seemed to be the most common setting, reports look great so I don't know what I would expect to improve upon by changing PS.
I do suspect that if someone had some bad aerophagia symptoms that perhaps a wider PS might help with that problem but I never really found any documented evidence of it nor any other specific information as to what PS at various settings might do.
SWS or JohnBFisher might have a better grasp on the "whys" because I certainly never could wrap my head around it.

Me....I just picked an average used number and tried it and liked it and felt great. Real scientific heh? :lol: :lol:

_________________
Machine: PR System One REMStar 60 Series BiPAP Auto with Bi-Flex
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: PR System One 60 Series Heated Tube Humidifier with Heated Tube
Additional Comments: S9 Adapt SV being tried now...not available as menu choice

User avatar
squid13
 
Posts: 1677
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL
Gender: Male

Re: Setting BiLevel Pressure Support.

Postby squid13 on Tue Jul 03, 2012 9:41 am

Mine is set at EPAP 10 and IPAP 14 with PS of 4. I know you can tweak your pressures up or down to what ever is the best for you but just trying to just understand the PS when doing so. I didn't sleep well when I had my BiPap titration and they came up with 10 and 14. I also read somewhere where 10 and 14 is an average setting.

_________________
Mask: Forma Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 VPAP Adapt 36037, ResScan 4.3, SleepyHead V.0.9.3
US Navy Retired 1973,S9 Adapt 36037, Mode: ASV Auto, Min EPAP: 6.8, Max EPAP: 15.0, Min PS: 3.0, Max PS: 15.0

User avatar
jnk
 
Posts: 5818
Joined: Mon Jun 30, 2008 3:03 pm
Location: Brooklyn Heights
Gender: Male

Re: Setting BiLevel Pressure Support.

Postby jnk on Tue Jul 03, 2012 9:52 am

4.3.1.6 The recommended minimum IPAP-EPAP differential is 4 cm H2O and the recommended maximum IPAP-EPAP differential is 10 cm H2O (Consensus).
This recommendation is based on consensus agreement by the PAP Titration Task Force and Guideline-Level evidence . . . for the minimum IPAP-EPAP differential in adult patients.

http://www.aasmnet.org/Resources/clinic ... 040210.pdf
"All quotes are actually misquotes."-- Never directly attributed to Bartlett.

User avatar
squid13
 
Posts: 1677
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL
Gender: Male

Re: Setting BiLevel Pressure Support.

Postby squid13 on Tue Jul 03, 2012 9:58 am

jnk thanks for that article, I'll sit down and read it after lunch.

_________________
Mask: Forma Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 VPAP Adapt 36037, ResScan 4.3, SleepyHead V.0.9.3
US Navy Retired 1973,S9 Adapt 36037, Mode: ASV Auto, Min EPAP: 6.8, Max EPAP: 15.0, Min PS: 3.0, Max PS: 15.0

User avatar
robysue
 
Posts: 5717
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Gender: Female

Re: Setting BiLevel Pressure Support.

Postby robysue on Tue Jul 03, 2012 10:05 am

On the Resmed Auto VPAPs, that PS is going to be the constant difference IPAP and EPAP. As the machine adjusts the pressure up and down, the IPAP and EPAP will both be adjusted at the same time and by the same amount.

It may help to consider a few examples: Let's say that we have Min EPE = 8 and MAX IPAP = 16. Now lets consider what happens with different PS levels:

With PS = 2, we have the following ranges 8 <= EPAP <= 14 and 10 <= IPAP <= 16
With PS = 3, we have the following ranges 8 <= EPAP <= 13 and 11 <= IPAP <= 16
With PS = 4, we have the following ranges 8 <= EPAP <= 12 and 12 <= IPAP <= 16
With PS = 5, we have the following ranges 8 <= EPAP <= 11 and 13 <= IPAP <= 16
With PS = 6, we have the following ranges 8 <= EPAP <= 10 and 14 <= IPAP <= 16

So in "picking" a PS, what's important?

Well comfort is one consideration. The basic premise of bi-level is that providing a sufficiently large enough drop in pressure on exhale should make the machine more comfortable to use. But if the drop is too large, that can cause comfort (and leak) problems for some individuals. Really large PS settings may leave you feeling as machine is blasting air down you gut with every inhalation, which may not be very sleep inducing.

But therapy needs are also important: That EPAP needs to be able to increase far enough to deal with nasty clusters of OAs and significant snoring during the times when you are running at max pressure. In other words, the PS needs to be set so that Max EPAP = Max IPAP - PS should be at least 1-2cm above your 95% EPAP. And the initial IPAP = Min EPAP + PS needs to be high enough for you to not feel like you are air-starved. And that initial IPAP should be high enough to deal with most of you flow limitations and hypopneas. In other words, Min EPAP + PS should probably be within 2-4cm of your 95% IPAP.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: PR System One Heated Humidifier
Additional Comments: PR System One REMstar BiPAP Auto (Series 50). Max IPAP = 8, Min EPAP=4, and Rise time setting = 3

User avatar
RandyJ
 
Posts: 1641
Joined: Thu Apr 21, 2011 7:22 pm
Location: Connecticut, USA
Gender: Male

Re: Setting BiLevel Pressure Support.

Postby RandyJ on Tue Jul 03, 2012 10:06 am

I found this post by restedgal helpful when I was setting my machine. She advocates using the highest PS setting:

viewtopic.php?f=1&t=45059&p=450776&hilit=+pressure+support#p450776

I think this is so IPAP has more freedom to range in responding to events without necessarily causing EPAP to rise as well.

_________________
Machine: PR System One REMStar 60 Series Pro CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: PR System One 60 Series Heated Tube Humidifier with Heated Tube
Additional Comments: Backup & Travel Machines: PR System One Bipap Auto, S9 VPAP Auto, S9 Autoset & S8 Elite, Oximeter CMS-50E
Diagnosed March 2011, using CPAP 10cm, CFlex+ 3
Alt masks Swift FX pillows, Mirage FX nasal mask, Mirage Quattro full face mask

User avatar
Pugsy
 
Posts: 27359
Joined: Thu May 14, 2009 9:31 am
Location: SW Missouri, USA
Gender: Female

Re: Setting BiLevel Pressure Support.

Postby Pugsy on Tue Jul 03, 2012 10:15 am

RandyJ wrote:I found this post by restedgal helpful when I was setting my machine. She advocates using the highest PS setting:


I believe she was using a Respironics BiPap at the time she wrote that.
Respironics BiPap machines allows for a roaming pressure support...so it can vary up to the 8 cm pressure support. This is totally different way of working than the ResMed bilevel machines. So she was really leaving the PS wide open to range from 2 cm to 8 cm as the machine deemed necessary.

ResMed machines have a fixed pressure support and there is no varying or roaming about. EPAP and IPAP are tied to each other by the pressure support which won't/can't vary.

This roaming vs fixed pressure support raises another question which I never found an answer to..
When is one method preferable over another?

_________________
Machine: PR System One REMStar 60 Series BiPAP Auto with Bi-Flex
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: PR System One 60 Series Heated Tube Humidifier with Heated Tube
Additional Comments: S9 Adapt SV being tried now...not available as menu choice

User avatar
robysue
 
Posts: 5717
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Gender: Female

Re: Setting BiLevel Pressure Support.

Postby robysue on Tue Jul 03, 2012 10:33 am

Pugsy wrote:Respironics BiPap machines allows for a roaming pressure support...so it can vary up to the 8 cm pressure support. This is totally different way of working than the ResMed bilevel machines. So she was really leaving the PS wide open to range from 2 cm to 8 cm as the machine deemed necessary.

ResMed machines have a fixed pressure support and there is no varying or roaming about. EPAP and IPAP are tied to each other by the pressure support which won't/can't vary.

This roaming vs fixed pressure support raises another question which I never found an answer to..
When is one method preferable over another?
I think the anwser to that is "It depends."

For people like you who appreciate a 4cm drop in pressure while they're awake and who are not troubled by aerophagia when the EPAP is high, the fixed PS is probably more comfortable.

For people like me who don't mind starting out with IPAP = EPAP + 2 and who have problems with aerophagia when the EPAP is near its max, the roaming PS will maximize the time spent with EPAP at or near the Min EPAP setting and hence be more comfortable.

For most people? My guess is that it probably doesn't much matter.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: PR System One Heated Humidifier
Additional Comments: PR System One REMstar BiPAP Auto (Series 50). Max IPAP = 8, Min EPAP=4, and Rise time setting = 3

User avatar
jnk
 
Posts: 5818
Joined: Mon Jun 30, 2008 3:03 pm
Location: Brooklyn Heights
Gender: Male

Re: Setting BiLevel Pressure Support.

Postby jnk on Tue Jul 03, 2012 11:17 am

The variable PS would appeal to me IF there was a way to set minimum PS.

For patients with OHS issues, or other hypoventilation issues, 2 cm ain't gonna cut it, IMO as a patient.

The issue, also merely IMO, is not so much the brand of machine used but that RTs don't know the difference between brands well enough and too often set up machines wrong when attempting to set up autobilevels to run in auto mode. I have no evidence of that beyond the anecdotal evidence of what I've experienced, read, and heard from the mouths of RTs, though.

Discussions of setting the PS in a ResMed autobilevel are moot, though, unless we know for sure a poster's machine is in VAUTO mode with some room to vary. I believe that autobilevel machines are most often handed out set up to run in plain old bilevel (S) mode, or are programmed with min EPAP set at the exact prescribed EPAP and the max IPAP set at the exact prescribed IPAP and therefore do not function as autos at all, even though the doc and RT may not be aware that is the case.

Those are merely my opinions about my personal assumptions. :lol:
"All quotes are actually misquotes."-- Never directly attributed to Bartlett.

User avatar
squid13
 
Posts: 1677
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL
Gender: Male

Re: Setting BiLevel Pressure Support.

Postby squid13 on Tue Jul 03, 2012 11:29 am

robysue wrote:It may help to consider a few examples: Let's say that we have Min EPE = 8 and MAX IPAP = 16. Now lets consider what happens with different PS levels:

With PS = 2, we have the following ranges 8 <= EPAP <= 14 and 10 <= IPAP <= 16
With PS = 3, we have the following ranges 8 <= EPAP <= 13 and 11 <= IPAP <= 16
With PS = 4, we have the following ranges 8 <= EPAP <= 12 and 12 <= IPAP <= 16
With PS = 5, we have the following ranges 8 <= EPAP <= 11 and 13 <= IPAP <= 16
With PS = 6, we have the following ranges 8 <= EPAP <= 10 and 14 <= IPAP <= 16

robysue I must be having a senor moment cause this chart confuses me. Is that 8 below equals EPAP,below equals 14 and 10, below equals IPAP below equals 16

_________________
Mask: Forma Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 VPAP Adapt 36037, ResScan 4.3, SleepyHead V.0.9.3
US Navy Retired 1973,S9 Adapt 36037, Mode: ASV Auto, Min EPAP: 6.8, Max EPAP: 15.0, Min PS: 3.0, Max PS: 15.0

User avatar
GumbyCT
 
Posts: 5789
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Gender: None specified

Re: Setting BiLevel Pressure Support.

Postby GumbyCT on Tue Jul 03, 2012 12:05 pm

squid13 wrote:
robysue wrote:It may help to consider a few examples: Let's say that we have Min EPE = 8 and MAX IPAP = 16. Now lets consider what happens with different PS levels:

With PS = 2, we have the following ranges 8 <= EPAP <= 14 and 10 <= IPAP <= 16
With PS = 3, we have the following ranges 8 <= EPAP <= 13 and 11 <= IPAP <= 16
With PS = 4, we have the following ranges 8 <= EPAP <= 12 and 12 <= IPAP <= 16
With PS = 5, we have the following ranges 8 <= EPAP <= 11 and 13 <= IPAP <= 16
With PS = 6, we have the following ranges 8 <= EPAP <= 10 and 14 <= IPAP <= 16

robysue I must be having a senor moment cause this chart confuses me. Is that 8 below equals EPAP,below equals 14 and 10, below equals IPAP below equals 16

Try this
8 <= EPAP <= 14
and
10 <= IPAP <= 16
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!

User avatar
RandyJ
 
Posts: 1641
Joined: Thu Apr 21, 2011 7:22 pm
Location: Connecticut, USA
Gender: Male

Re: Setting BiLevel Pressure Support.

Postby RandyJ on Tue Jul 03, 2012 12:12 pm

Pugsy wrote:I believe she was using a Respironics BiPap at the time she wrote that.
Respironics BiPap machines allows for a roaming pressure support...so it can vary up to the 8 cm pressure support. This is totally different way of working than the ResMed bilevel machines. So she was really leaving the PS wide open to range from 2 cm to 8 cm as the machine deemed necessary.

ResMed machines have a fixed pressure support and there is no varying or roaming about. EPAP and IPAP are tied to each other by the pressure support which won't/can't vary.



Oops... missed that part.... maybe the OP could add "Resmed S9 VPAP" to the thread title....

_________________
Machine: PR System One REMStar 60 Series Pro CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: PR System One 60 Series Heated Tube Humidifier with Heated Tube
Additional Comments: Backup & Travel Machines: PR System One Bipap Auto, S9 VPAP Auto, S9 Autoset & S8 Elite, Oximeter CMS-50E
Diagnosed March 2011, using CPAP 10cm, CFlex+ 3
Alt masks Swift FX pillows, Mirage FX nasal mask, Mirage Quattro full face mask

Next