S9 AutoSet ?
-
- Posts: 55
- Joined: Thu Feb 25, 2010 2:50 pm
- Location: Ohio
S9 AutoSet ?
Hi all
I'm looking at a new S9 AutoSet, I don't understand the stament below. My min. setting is 11.5 H20, does the stament mean that it will not treat any apneas that I may have? since my min. setting is already set above 10 H20.
Apneas
The APAP algorithm only treats apneas up to 10 cm H2O and does not differentiate between different
types of apneas.
The enhanced AutoSet algorithm detects both obstructive and central apneas. If an apnea occurs, the
device responds appropriately.
Thanks
I'm looking at a new S9 AutoSet, I don't understand the stament below. My min. setting is 11.5 H20, does the stament mean that it will not treat any apneas that I may have? since my min. setting is already set above 10 H20.
Apneas
The APAP algorithm only treats apneas up to 10 cm H2O and does not differentiate between different
types of apneas.
The enhanced AutoSet algorithm detects both obstructive and central apneas. If an apnea occurs, the
device responds appropriately.
Thanks
Re: S9 AutoSet ?
Help us please. Where do you find that statement? Would like to know so we can see the context.The Dreamer wrote:Hi all
I'm looking at a new S9 AutoSet, I don't understand the stament below. My min. setting is 11.5 H20, does the stament mean that it will not treat any apneas that I may have? since my min. setting is already set above 10 H20.
Apneas
The APAP algorithm only treats apneas up to 10 cm H2O and does not differentiate between different
types of apneas.
The enhanced AutoSet algorithm detects both obstructive and central apneas. If an apnea occurs, the
device responds appropriately.
Thanks
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7 |
-
- Posts: 55
- Joined: Thu Feb 25, 2010 2:50 pm
- Location: Ohio
Re: S9 AutoSet ?
It came from a Res Med S9 AutoSet manual on page 7idamtnboy wrote:Help us please. Where do you find that statement? Would like to know so we can see the context.The Dreamer wrote:Hi all
I'm looking at a new S9 AutoSet, I don't understand the stament below. My min. setting is 11.5 H20, does the stament mean that it will not treat any apneas that I may have? since my min. setting is already set above 10 H20.
Apneas
The APAP algorithm only treats apneas up to 10 cm H2O and does not differentiate between different
types of apneas.
The enhanced AutoSet algorithm detects both obstructive and central apneas. If an apnea occurs, the
device responds appropriately.
Thanks
Re: S9 AutoSet ?
That statement is true of the S8 AutoSet and the S7 AutoSet. Both were earlier generations of AutoSet. The enhanced algorithm used in the S9 generation AutoSet models do, indeed, attempt differentiation:
http://www.s9morecomfort.com/s9morecomf ... toset.html
http://www.resmed.com/us/assets/documen ... -paper.pdf
Are you sure that manual is for an S9 generation?
http://www.s9morecomfort.com/s9morecomf ... toset.html
http://www.resmed.com/us/assets/documen ... -paper.pdf
Are you sure that manual is for an S9 generation?
-
- Posts: 55
- Joined: Thu Feb 25, 2010 2:50 pm
- Location: Ohio
Re: S9 AutoSet ?
This statement Came right out of A S9 clinical guide on page 7-SWS wrote:That statement is true of the S8 AutoSet and the S7 AutoSet. Both were earlier generations of AutoSet. The enhanced algorithm used in the S9 generation AutoSet models do, indeed, attempt differentiation:
http://www.s9morecomfort.com/s9morecomf ... toset.html
http://www.resmed.com/us/assets/documen ... -paper.pdf
Are you sure that manual is for an S9 generation?
Page 7.
Apneas
The APAP algorithm only treats apneas up to 10 cm H2O and does not differentiate between different
types of apneas.
The enhanced AutoSet algorithm detects both obstructive and central apneas. If an apnea occurs, the
device responds appropriately.
Contents
Welcome ......................................................................................................1
S9 AutoSet indications for use..............................................................1
S9 AutoSet for Her indications for use..................................................1
S9 Elite indications for use....................................................................1
S9 Escape Auto indications for use.......................................................1
S9 Escape indications for use...............................................................1
S9 contraindications..............................................................................1
S9 adverse effects.................................................................................1
H5i indications for use...........................................................................2
H5i for Her indications for use...............................................................2
H5i contraindications.............................................................................2
S9 at a glance...............................................................................................3
Traveling with the S9.............................................................................3
H5i at a glance..............................................................................................4
Traveling with the H5i............................................................................4
Operating information...................................................................................5
EPR technology.............................................................................................5
EPR checklist.........................................................................................5
AutoSet technology.......................................................................................6
Normal airway.......................................................................................6
Flow limitation.......................................................................................6
Snore.....................................................................................................6
Apneas..................................................................................................7
New to the S9 platform................................................................................8
Climate Control.....................................................................................8
Central sleep apnea...............................................................................9
Sleep quality..........................................................................................9
Ramp.....................................................................................................9
Data management............................................................................... 10
Setup........................................................................................................... 11
Mask and tubing setup........................................................................ 11
Filling the water tub.................................................................................... 12
S9 basics..................................................................................................... 13
Navigating the menus................................................................................. 14
About the menus................................................................................ 14
Home menu ....................................................................................... 14
Changing settings via the Home menu............................................... 15
Treatment screens.............................................................................. 16
Info menu............................................................................................ 16
Setup menu......................................................................................... 17
Reminder menu.................................................................................. 18
Managing Climate Control.......................................................................... 19
Treatment screen and clinical setup menus........................................20
Delivering therapy.......................................................................................23
Adding supplemental oxygen..............................................................23
ii
Data management......................................................................................24
SD card.......................................................................................................24
Removing the card..............................................................................24
Inserting the card................................................................................24
Analyzing the SD card data.................................................................25
Data storage........................................................................................25
Data transmission adapters.................................................................25
Cleaning and maintenance..........................................................................26
Daily....................................................................................................26
Weekly ...............................................................................................26
Monthly...............................................................................................26
Maintenance checklist.........................................................................27
Replacing the air filter.........................................................................27
Antibacterial filters..............................................................................27
Technical specifications...............................................................................28
Warnings and cautions................................................................................32
GENERAL WARNINGS........................................................................32
S9 WARNINGS....................................................................................32
S9 CAUTIONS.....................................................................................33
H5i WARNINGS...................................................................................33
AIR TUBING WARNINGS.....................................................................33
Re: S9 AutoSet ?
That statement is made at page 6 at the top of the page.
_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
Mask: Evora Full Face Mask - Fitpack |
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1 |
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV
Re: S9 AutoSet ?
I think the statement endeavors to to distinguish Resmed's latest algorithm from other APAP algorithms as follows.
The S9 definitely attempts differentiation of central versus obstructive apneas. The PR System One does as well. Most APAP algorithms do not.The Dreamer wrote: This statement Came right out of A S9 clinical guide on page 7
The APAP algorithm only treats apneas up to 10 cm H2O and does not differentiate between different types of apneas.
[ meaning APAP algorithms in general]
The enhanced AutoSet algorithm detects both obstructive and central apneas. If an apnea occurs, the device responds appropriately.
[meaning the S9 algorithm in particular]
Re: S9 AutoSet ?
I believe the cenetral-less algorithm belongs to the S9 Escape.
...other than food...
Re: S9 AutoSet ?
That's true---and it's news to me. A side-by-side comparison of features is available on the specification tab here:deltadave wrote:I believe the cenetral-less algorithm belongs to the S9 Escape.
http://www.resmed.com/us/products/s9_se ... clinicians
Re: S9 AutoSet ?
Cenetral...deltadave wrote:I believe the cenetral-less algorithm belongs to the S9 Escape.
An Italian algorithm...
To fixa youra sleepa apnea...
Ciao!
...other than food...
Re: S9 AutoSet ?
LOL Glad you explained that...deltadave wrote:Cenetral...
An Italian algorithm...
To fixa youra sleepa apnea...
Ciao!
...I incorrectly assumed the word's definition by running with a derivative: "Even the male and female cenetalia need adequate ventilation... Enter the enhanced APAP."
Re: S9 AutoSet ?
Actually, my Canadian accent is much better, eh?
Maybe we can pick up a 2-4, eh?
Beauty, eh?
Maybe we can pick up a 2-4, eh?
Beauty, eh?
...other than food...
Re: S9 AutoSet ?
Very interesting. Looks like a significant difference that we've missed in all our prior comments concerning differences between the Escape and Autoset models. I always thought the Escape Auto did the same thing as the Autoset but without the data recording capability. But this shows if you need a pressure above 10 in APAP mode you MUST have the Autoset.
deltadave wrote:There's that line:
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7 |
Re: S9 AutoSet ?
If you're referring to the S9 AutoSet being mandatory over the S9 Escape Auto in that scenario, then I would have to disagree.idamtnboy wrote: But this shows if you need a pressure above 10 in APAP mode you MUST have the Autoset.
The previous-generation S7 and S8 AutoSet algorithm (named "A10") doesn't respond to apneas above 10cm either. And yet there is plenty of clinical data showing that A10 algorithm treats patients fine who need pressures above 10cm. That A10 algorithm responds to precursor events such as snore and flow limitation to adequately prevent obstructive apneas. That's probably the algorithm used by the S9 Escape Auto.
I would agree that the enhanced S9 AutoSet is preferable in that above-10cm scenario you mentioned. I would prefer the enhanced S9 AutoSet algorithm both below and above 10cm. Obstructive apnea patients who don't present many snore or flow-limitation precursor signals might actually be better off avoiding all APAP algorithms...