Travel and dead Resmed power supply. Coincidence?
Travel and dead Resmed power supply. Coincidence?
I just got back from a week long visit to Amsterdam.
When I first plugged my Resmed S9 Autoset into the hotel power in Amsterdam, the unit came on, then immediately went off, then on again, off again a few times before staying on. But then it would turn off semi-randomly during the night. This repeated every night and appeared to be getting worse. I assumed it was some quirk of the local / hotel power supply. (No other equipment had problems, but it was all relatively low power.)
But I just got back and now it wouldn't turn on at all. (The power light barely flashes green, then it turns off, and the cycle repeats.)
Fortunately, my battery backup and DC power supply appear to have no issues, so I'll be using them for a few days.
I am thinking it's likely the issue is with the power supply brick. (A new brick has been ordered from our sponsors.) I've traveled with it to many countries with no issues before this.
It is just under two years old.
So, a few questions:
Who else has had a Resmed power brick fail? About how long is their expected lifetime?
Is there any reason to believe the brick starting to fail the first night of my vacation was anything other than a coincidence? I'm thinking that it could have been starting to fail and that the extra voltage was enough to accelerate that process a little.
When I first plugged my Resmed S9 Autoset into the hotel power in Amsterdam, the unit came on, then immediately went off, then on again, off again a few times before staying on. But then it would turn off semi-randomly during the night. This repeated every night and appeared to be getting worse. I assumed it was some quirk of the local / hotel power supply. (No other equipment had problems, but it was all relatively low power.)
But I just got back and now it wouldn't turn on at all. (The power light barely flashes green, then it turns off, and the cycle repeats.)
Fortunately, my battery backup and DC power supply appear to have no issues, so I'll be using them for a few days.
I am thinking it's likely the issue is with the power supply brick. (A new brick has been ordered from our sponsors.) I've traveled with it to many countries with no issues before this.
It is just under two years old.
So, a few questions:
Who else has had a Resmed power brick fail? About how long is their expected lifetime?
Is there any reason to believe the brick starting to fail the first night of my vacation was anything other than a coincidence? I'm thinking that it could have been starting to fail and that the extra voltage was enough to accelerate that process a little.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Pressure is 11-14. Old CPAP was a Resmed S6 Lightweight. Also have Profile lite mask. ResScan is actually version 3.14. Now I use Sleepyhead. |
- Drowsy Dancer
- Posts: 1271
- Joined: Mon Feb 28, 2011 9:35 am
- Location: here
Re: Travel and dead Resmed power supply. Coincidence?
This power supply stuff is Greek to me. Does this thread have any possible relevance to your problem? It was an airplane power supply, not a hotel, but maybe a similar problem: viewtopic.php?f=1&t=87993&st=0&sk=t&sd=aJointPain wrote:I just got back from a week long visit to Amsterdam.
When I first plugged my Resmed S9 Autoset into the hotel power in Amsterdam, the unit came on, then immediately went off, then on again, off again a few times before staying on. But then it would turn off semi-randomly during the night. This repeated every night and appeared to be getting worse. I assumed it was some quirk of the local / hotel power supply. (No other equipment had problems, but it was all relatively low power.)
But I just got back and now it wouldn't turn on at all. (The power light barely flashes green, then it turns off, and the cycle repeats.)
Fortunately, my battery backup and DC power supply appear to have no issues, so I'll be using them for a few days.
I am thinking it's likely the issue is with the power supply brick. (A new brick has been ordered from our sponsors.) I've traveled with it to many countries with no issues before this.
It is just under two years old.
So, a few questions:
Who else has had a Resmed power brick fail? About how long is their expected lifetime?
Is there any reason to believe the brick starting to fail the first night of my vacation was anything other than a coincidence? I'm thinking that it could have been starting to fail and that the extra voltage was enough to accelerate that process a little.
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: Swift™ FX Bella Nasal Pillow CPAP Mask with Headgears |
Additional Comments: Software: SleepyHead. Pressure: APAP 9.5 min/11 max, A-Flex x2 |
How we squander our hours of pain. -- Rilke
Re: Travel and dead Resmed power supply. Coincidence?
Thanks a lot for the link, but it's a different issue.Drowsy Dancer wrote: This power supply stuff is Greek to me. Does this thread have any possible relevance to your problem? It was an airplane power supply, not a hotel, but maybe a similar problem: viewtopic.php?f=1&t=87993&st=0&sk=t&sd=a
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Pressure is 11-14. Old CPAP was a Resmed S6 Lightweight. Also have Profile lite mask. ResScan is actually version 3.14. Now I use Sleepyhead. |
- Drowsy Dancer
- Posts: 1271
- Joined: Mon Feb 28, 2011 9:35 am
- Location: here
Re: Travel and dead Resmed power supply. Coincidence?
I was afraid of that. I can practically feel my lips moving when I start to read about power issues...fuses...AC, DC....isn't that a rock band?JointPain wrote:Thanks a lot for the link, but it's a different issue.Drowsy Dancer wrote: This power supply stuff is Greek to me. Does this thread have any possible relevance to your problem? It was an airplane power supply, not a hotel, but maybe a similar problem: viewtopic.php?f=1&t=87993&st=0&sk=t&sd=a
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: Swift™ FX Bella Nasal Pillow CPAP Mask with Headgears |
Additional Comments: Software: SleepyHead. Pressure: APAP 9.5 min/11 max, A-Flex x2 |
How we squander our hours of pain. -- Rilke
- billbolton
- Posts: 2266
- Joined: Wed Jun 07, 2006 7:46 pm
- Location: Sydney, Australia
Re: Travel and dead Resmed power supply. Coincidence?
A couple of questions.....JointPain wrote:When I first plugged my Resmed S9 Autoset into the hotel power in Amsterdam, the unit came on, then immediately went off, then on again, off again a few times before staying on. But then it would turn off semi-randomly during the night. This repeated every night and appeared to be getting worse. I assumed it was some quirk of the local / hotel power supply. (No other equipment had problems, but it was all relatively low power.)
Did you have anything else at all plugged into the same power outlet
What sort of adapter were you using to match a European wall socket to your power plug
Cheers,
Bill
Re: Travel and dead Resmed power supply. Coincidence?
The most likely cause is mechanical problems cause by packing, unpacking, banging, etc. the power supply. Try wiggling all the connectors, flexing the power cord, etc. to see if you can make the problem come and go. Inspect everything by eye, especially the the connectors.
You could also have mechanical damage to the plug on the blower unit as well.
You could have had something like an overvoltage or power surge, but I would expect the power brick to be really tough in that respect.
You could also have mechanical damage to the plug on the blower unit as well.
You could have had something like an overvoltage or power surge, but I would expect the power brick to be really tough in that respect.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Re: Travel and dead Resmed power supply. Coincidence?
Nothing else was plugged into the same power outlet, but there was a desk lamp plugged into another power point immediately above it. I think it's likely that all the outlets in the room were ultimately wired back to one circuit to the room. (All power and lights in the room were controlled from a single insert-your-room-key-here switch next to the door.)billbolton wrote: A couple of questions.....
Did you have anything else at all plugged into the same power outlet
What sort of adapter were you using to match a European wall socket to your power plug
Cheers,
Bill
I used a straight-forward mechanical adapter. In fact, I used two different ones on different nights to rule adapter issues out. I've previously used the same arrangement in Australia without problems.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Pressure is 11-14. Old CPAP was a Resmed S6 Lightweight. Also have Profile lite mask. ResScan is actually version 3.14. Now I use Sleepyhead. |
Re: Travel and dead Resmed power supply. Coincidence?
Thanks, I'll look at that carefully tonight. The power cord just below where it plugs into the blower unit gets a tight bend in it, and a mechanical failure there is a strong possibility.archangle wrote:The most likely cause is mechanical problems cause by packing, unpacking, banging, etc. the power supply. Try wiggling all the connectors, flexing the power cord, etc. to see if you can make the problem come and go. Inspect everything by eye, especially the the connectors.
You could also have mechanical damage to the plug on the blower unit as well.
You could have had something like an overvoltage or power surge, but I would expect the power brick to be really tough in that respect.
I did try unplugging and reconnecting everything a few times to no avail.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Pressure is 11-14. Old CPAP was a Resmed S6 Lightweight. Also have Profile lite mask. ResScan is actually version 3.14. Now I use Sleepyhead. |
- chunkyfrog
- Posts: 34444
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Travel and dead Resmed power supply. Coincidence?
Aha! A de$ign flaw. Just what we can expect from . . (who was that again?)
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
- billbolton
- Posts: 2266
- Joined: Wed Jun 07, 2006 7:46 pm
- Location: Sydney, Australia
Re: Travel and dead Resmed power supply. Coincidence?
It might have just been a faulty connection "behind the wall" to that one socket.JointPain wrote:Nothing else was plugged into the same power outlet....
If you mean you used the same adapters, does that mean you had the ones a variety of different pins on the plug side that can slide out to match the relevant wall socket? If so, do you know what brand they were?JointPain wrote:I've previously used the same arrangement in Australia without problems.
While the switch mode power supplies that are incorporated in most modern CPAP power bricks are pretty tolerant of a lot of things related to the quality of mains suppled power they receive, it is possible that a large number of small power interruptions, due to a faulty connection on the supply side, could damage one
Cheers,
Bill
- StuUnderPressure
- Posts: 1377
- Joined: Mon Jun 18, 2012 10:34 am
- Location: USA
Re: Travel and dead Resmed power supply. Coincidence?
So, does the S9 AutoSet power brick act somewhat like a mini UPS?billbolton wrote:While the switch mode power supplies that are incorporated in most modern CPAP power bricks are pretty tolerant of a lot of things related to the quality of mains suppled power they receive, it is possible that a large number of small power interruptions, due to a faulty connection on the supply side, could damage one
Cheers,
Bill
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Cleanable Water Tub & Respironics Premium Chinstrap |
In Windows 10 Professional 64 bit Version 22H2 - ResScan Version 7.0.1.67 - ResScan Clinician's Manual dtd 2021-02
SD Card Formatter 5.0.2 https://www.sdcard.org/downloads/format ... index.html
SD Card Formatter 5.0.2 https://www.sdcard.org/downloads/format ... index.html
Re: Travel and dead Resmed power supply. Coincidence?
Nope.StuUnderPressure wrote:So, does the S9 AutoSet power brick act somewhat like a mini UPS?billbolton wrote:While the switch mode power supplies that are incorporated in most modern CPAP power bricks are pretty tolerant of a lot of things related to the quality of mains suppled power they receive, it is possible that a large number of small power interruptions, due to a faulty connection on the supply side, could damage one
Cheers,
Bill
The term "Switched Mode Power Supply" references a technology that allows the power supply to work with a wide range of input voltages and frequencies. A UPS might be able to compensate for voltage brown-outs but in the case of switch mode power supplies that's hardly necessary.
An interesting point about switch mode power supplies is that they HATE stepped-approximation sine waves (how many people just pressed the close button?)
That means that if you DO use a UPS (or generator, whatever) and it doesn't have a true sine wave output, it most certainly CAN damage this kind of power supply. Unfortunately, true sine wave UPS' are more expensive.
Sleep loss is a terrible thing. People get grumpy, short-tempered, etc. That happens here even among the generally friendly. Try not to take it personally.
Re: Travel and dead Resmed power supply. Coincidence?
Try checking the pins in the plug connector that attach the s9. Could be that upon inserting the plug, one of the pins was not aligned and got bent, providing intermittent power. If this is the case, you may be able to realign the pin(s) with a small tweezer. I just so happened to be at Sleepquest picking up supplies when a customer came in with a "broken power supply" and they gave him another one but showed him that the problem was with the plug.
_________________
Mask: Mirage Liberty™ Full Face CPAP Mask with Nasal Pillows With Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Quattro FX Full Face Mask |
5thumbs
Re: Travel and dead Resmed power supply. Coincidence?
Does the power brick have a green light when plugged in? I can't look at mine at the moment. The travel S9 brick does have a green indicator if it has power.
Re: Travel and dead Resmed power supply. Coincidence?
They don't appear to be bent.5thumbs wrote:Try checking the pins in the plug connector that attach the s9. Could be that upon inserting the plug, one of the pins was not aligned and got bent, providing intermittent power. If this is the case, you may be able to realign the pin(s) with a small tweezer. I just so happened to be at Sleepquest picking up supplies when a customer came in with a "broken power supply" and they gave him another one but showed him that the problem was with the plug.
In Amsterdam, when the blower was powered up, the green light on the brick was on. When the blower lost power, the green light on the brick was out. The loss of power transition was too fast for me to say which happened first. (I could trigger a power-loss event with a good chance of success by taking a quick breath in.) Back in Houston, the green light comes on, the unit starts to power up, then it loses power and the green light goes out. A moment later the process starts over. If the power brick is not plugged into the blower, the green light stays on.cosmo wrote:Does the power brick have a green light when plugged in? I can't look at mine at the moment. The travel S9 brick does have a green indicator if it has power.
Wiggling the power connectors etc. makes no difference that I can see. The unit was packed carefully inside the S9 travel bag with the hose wrapped around the power supply, and the S9 bag was packed inside my backpack, which I hand carried. I did not use the CPAP while travelling.archangle wrote:The most likely cause is mechanical problems cause by packing, unpacking, banging, etc. the power supply. Try wiggling all the connectors, flexing the power cord, etc. to see if you can make the problem come and go. Inspect everything by eye, especially the the connectors.
You could also have mechanical damage to the plug on the blower unit as well.
You could have had something like an overvoltage or power surge, but I would expect the power brick to be really tough in that respect.
The new power brick has arrived from the site's sponsor (love that service - living in Houston helps). The new brick appears to be working fine.
Thanks a lot everyone for your comments and assistance.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Pressure is 11-14. Old CPAP was a Resmed S6 Lightweight. Also have Profile lite mask. ResScan is actually version 3.14. Now I use Sleepyhead. |