CMS50D+ and Sleepyhead 0.9.2-1

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
tide2bama
Posts: 2
Joined: Sun May 15, 2011 7:53 am
Location: Northeast Alabama

CMS50D+ and Sleepyhead 0.9.2-1

Post by tide2bama » Sat Feb 11, 2012 1:23 pm

Can someone tell me if the CMS50D+ memory from previous night can be downloaded to Sleepyhead?

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResScan 3.14, Sleepyhead v0.9.2-1 Use Mirage Swift 11 occasionally

DrPepper00
Posts: 139
Joined: Wed Dec 14, 2011 9:29 am

Re: CMS50D+ and Sleepyhead 0.9.2-1

Post by DrPepper00 » Sat Feb 11, 2012 1:52 pm

Yes. I download the data to SPO2Review and save it. Then I import it to SleepyHead.
Download the cpap data first. Then download the oximetry data. Change the date and start time to match the cpap time and save by clicking the disc icon to the left of the import options. The data may not show up until you exit and return to SleepyHead.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Ruby chinstrap SleepyHead Software

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: CMS50D+ and Sleepyhead 0.9.2-1

Post by archangle » Sat Feb 11, 2012 3:53 pm

I believe you have to set the machine to record and then download it before you turn the pulseox off. Once you turn it off the data is gone.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

yawnny
Posts: 143
Joined: Thu Aug 12, 2010 9:47 pm

Re: CMS50D+ and Sleepyhead 0.9.2-1

Post by yawnny » Sat Feb 11, 2012 4:26 pm

archangle wrote:I believe you have to set the machine to record and then download it before you turn the pulseox off. Once you turn it off the data is gone.
The way my CMS50D+ seems to work is that whatever is in memory stays in memory until the next time that you hit record on. I can turn record off, and exit the settings menu, have the screen go black (off), perform any number of spot checks.

When I want to retrieve the recorded data, all I have to do is enter the settings menu, take no action other than enter the menu. If you don't enter the menu, nothing will download. Plug the data cable into the device, start the SPO2 Review software, and it will happily download the last recorded data. At least that's how it is working for me right now, with data that I recorded a couple of days ago.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: CMS50D+ and CMS50F Pulse Oximeters

DrPepper00
Posts: 139
Joined: Wed Dec 14, 2011 9:29 am

Re: CMS50D+ and Sleepyhead 0.9.2-1

Post by DrPepper00 » Sat Feb 11, 2012 4:42 pm

Yes, I can turn mine off and back on and the data is still there. You do have to turn record off or the program will sit there saying waiting for data.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Ruby chinstrap SleepyHead Software

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: CMS50D+ and Sleepyhead 0.9.2-1

Post by archangle » Sat Feb 11, 2012 10:02 pm

I will have to play with mine when the person I loaned it to returns it. I've managed to lose the data on it several times and haven't been able to get old data back. Maybe I'm just not doing it right.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
Buggstzer
Posts: 22
Joined: Wed Jan 25, 2012 8:22 pm

Re: CMS50D+ and Sleepyhead 0.9.2-1

Post by Buggstzer » Sat Feb 11, 2012 10:45 pm

I have found it difficult to download data since the oximeter shuts off after a few seconds. It seems to download fine and not shut off if you go to one of the menu's before you download. Hope this helps.

_________________
MaskHumidifier
Additional Comments: Res Med S9 VPAP Adapt Contec CMS-50+ Zeo Sleep Manager

User avatar
squid13
Posts: 2301
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: CMS50D+ and Sleepyhead 0.9.2-1

Post by squid13 » Wed Feb 15, 2012 1:14 pm

I downloaded it to SPO2 review but I can't import it to sleepyhead. The file is in the data folder but when you browse for it from sleepy head and I get to the data folder it shows the data folder empty. I right click and check to show hidden files and it 's still empty. I know it's there but just doesn't show from sleepyhead.

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV

DrPepper00
Posts: 139
Joined: Wed Dec 14, 2011 9:29 am

Re: CMS50D+ and Sleepyhead 0.9.2-1

Post by DrPepper00 » Wed Feb 15, 2012 1:26 pm

squid13 wrote:I downloaded it to SPO2 review but I can't import it to sleepyhead. The file is in the data folder but when you browse for it from sleepy head and I get to the data folder it shows the data folder empty. I right click and check to show hidden files and it 's still empty. I know it's there but just doesn't show from sleepyhead.
From SleepyHead, are you setting the download folder to the C:\ProgramFiles(x86)\ SPO2\Data folder? I had to change it from the D: that SleepyHead downloads CPAP from.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Ruby chinstrap SleepyHead Software

wolftracker
Posts: 315
Joined: Sun May 14, 2006 8:20 am

Re: CMS50D+ and Sleepyhead 0.9.2-1

Post by wolftracker » Wed Feb 15, 2012 1:42 pm

I am unable to get sleepy head to download the entire memory
it starts and stops after a few seconds and if you start it again
it down loads the same thing over again

so can not get a full nite download
but the info is there til you download or re record

my .02

User avatar
squid13
Posts: 2301
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: CMS50D+ and Sleepyhead 0.9.2-1

Post by squid13 » Wed Feb 15, 2012 2:16 pm

DrPepper00 wrote:From SleepyHead, are you setting the download folder to the C:\ProgramFiles(x86)\ SPO2\Data folder? I had to change it from the D: that SleepyHead downloads CPAP from.
That was what I was using but it says the data folder is empty and it is empty. Where the data is at is C:\MyName\AppData\Local\VirtualStore\ProgramFiles(x86)\SPO2\Data When you check that location it's there but when you browse to that location from SleepyHead it's empty.

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV

DrPepper00
Posts: 139
Joined: Wed Dec 14, 2011 9:29 am

Re: CMS50D+ and Sleepyhead 0.9.2-1

Post by DrPepper00 » Wed Feb 15, 2012 2:25 pm

squid13 wrote:
DrPepper00 wrote:From SleepyHead, are you setting the download folder to the C:\ProgramFiles(x86)\ SPO2\Data folder? I had to change it from the D: that SleepyHead downloads CPAP from.
That was what I was using but it says the data folder is empty and it is empty. Where the data is at is C:\MyName\AppData\Local\VirtualStore\ProgramFiles(x86)\SPO2\Data When you check that location it's there but when you browse to that location from SleepyHead it's empty.
That sounds like a temporary location. Did you close SPO2 Review before you opened SleepyHead? It might not write it to the final location until you close.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Ruby chinstrap SleepyHead Software

User avatar
Uncle_Bob
Posts: 2777
Joined: Tue Feb 24, 2009 12:10 pm
Location: Arizona

Re: CMS50D+ and Sleepyhead 0.9.2-1

Post by Uncle_Bob » Wed Feb 15, 2012 2:40 pm

archangle wrote:I believe you have to set the machine to record and then download it before you turn the pulseox off. Once you turn it off the data is gone.
Well thats how the CMS50E behaved if i remember correctly.

User avatar
squid13
Posts: 2301
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: CMS50D+ and Sleepyhead 0.9.2-1

Post by squid13 » Wed Feb 15, 2012 2:53 pm

I have a CMS50D+ and that's where SPO2 Review downloaded it to. Like I said I can browse to that location from SleepyHead and it shows the data folder empty. When I go to that location in the computer the data folder has the information. It just doesn't show in SleepyHead. No problem downloading it to SPO2 Review closed review before trying to SleepyHead.

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV

User avatar
Lizistired
Posts: 2835
Joined: Tue Dec 14, 2010 10:47 pm
Location: Indiana

Re: CMS50D+ and Sleepyhead 0.9.2-1

Post by Lizistired » Wed Feb 15, 2012 3:53 pm

Squid, I had that problem and I think it was the "permissiosn" that were assigned to the folder. First the contents were hidden and then I had to change the permissions or rights on the files/folders to access them. Hope that helps but I don't remember the specifics now.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Swift FX sometimes, CMS-50F, Cervical collar sometimes, White noise, Zeo... I'm not well, but I'm better.