Seeking Tweaking Advice-ResScan

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Seeking Tweaking Advice-ResScan

Post by NateS » Wed Feb 08, 2012 11:04 pm

Attention software mavens:

Could you please take a look at this thread:

viewtopic.php?f=1&t=74149&p=681145#p681145

Would any of you be willing to help me out by providing the steps necessary for me to tweak an extra copy of ResScan to see whether it would allow me to see whether my ResMed S9 VPAP Adapt distinguishes between Obstructive vs Central Apneas and, if so to flag and display them separately in ResScan?

This would be my first effort in using foxe242 and of course I would only work on a copy of the program.

Thanks to anyone for their interest and advice.

Best regards, Nate

(On the other hand, if the data is there, why wouldn't SleepyHead be picking it up off the sd card without tweaking?)

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

User avatar
idamtnboy
Posts: 2186
Joined: Mon Nov 01, 2010 2:12 pm
Location: Idaho

Re: Seeking Tweaking Advice-ResScan

Post by idamtnboy » Wed Feb 08, 2012 11:53 pm

NateS wrote:Attention software mavens:

Could you please take a look at this thread:

viewtopic.php?f=1&t=74149&p=681145#p681145

Would any of you be willing to help me out by providing the steps necessary for me to tweak an extra copy of ResScan to see whether it would allow me to see whether my ResMed S9 VPAP Adapt distinguishes between Obstructive vs Central Apneas and, if so to flag and display them separately in ResScan?
Before we do that, do this first, please. Look at the directory on your SD card and list the names of all the files, and sizes, that are on it. Also, is there some place you can upload a copy of the str.edf file so I can download it and look at it? There's no value in mucking with the XML file if the data isn't in the files on the SD card. I really don't think it is or Resmed would have listed it in that data management guide, which is meant for clinicians, not us incompetent and clueless user peons.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: Seeking Tweaking Advice-ResScan

Post by NateS » Thu Feb 09, 2012 7:19 pm

idamtnboy wrote: Before we do that, do this first, please. Look at the directory on your SD card and list the names of all the files, and sizes, that are on it. Also, is there some place you can upload a copy of the str.edf file so I can download it and look at it? There's no value in mucking with the XML file if the data isn't in the files on the SD card. I really don't think it is or Resmed would have listed it in that data management guide, which is meant for clinicians, not us incompetent and clueless user peons.
Excellent point. Here is a what I hope to be a complete directory of the sd disk. I will find someplace to upload the str.edf file for you.

Thanks very much for your interest.

Regards, Nate

Name DATALOG
20120110_032005_EVE.crc 20120110_032005_EVE.edf 20120110_032005_PLD.crc 20120110_032005_PLD.edf 20120110_032005_SAD.crc 20120110_032005_SAD.edf 20120110_104341_EVE.crc 20120110_104341_EVE.edf 20120110_104342_PLD.crc 20120110_104342_PLD.edf 20120110_104342_SAD.crc 20120110_104342_SAD.edf 20120111_013128_EVE.crc 20120111_013128_EVE.edf 20120111_013129_PLD.crc 20120111_013129_PLD.edf 20120111_013129_SAD.crc 20120111_013129_SAD.edf 20120111_052911_EVE.crc 20120111_052911_EVE.edf 20120111_052912_PLD.crc 20120111_052912_PLD.edf 20120111_052912_SAD.crc 20120111_052912_SAD.edf 20120111_070558_EVE.crc 20120111_070558_EVE.edf 20120111_070558_PLD.crc 20120111_070558_PLD.edf 20120111_070558_SAD.crc 20120111_070558_SAD.edf 20120112_021414_EVE.crc 20120112_021414_EVE.edf 20120112_021414_PLD.crc 20120112_021414_PLD.edf 20120112_021414_SAD.crc 20120112_021414_SAD.edf 20120112_021823_EVE.crc 20120112_021823_EVE.edf 20120112_021823_PLD.crc 20120112_021823_PLD.edf 20120112_021823_SAD.crc 20120112_021823_SAD.edf 20120113_011716_EVE.crc 20120113_011716_EVE.edf 20120113_011716_PLD.crc 20120113_011716_PLD.edf 20120113_011716_SAD.crc 20120113_011716_SAD.edf 20120114_021347_EVE.crc 20120114_021347_EVE.edf
Listing of "NO NAME" (Page 1)
Size
39.2 MB 8 bytes 1KB
8 bytes 288 KB
8 bytes 104 KB
8 bytes 548 bytes 8 bytes
3 KB
8 bytes 768 bytes 8 bytes 2KB
8 bytes 155 KB
8 bytes 56 KB
8 bytes 764 bytes 8 bytes 60 KB
8 bytes 22 KB
8 bytes 980 bytes 8 bytes 183 KB
8 bytes 66 KB
8 bytes 548 bytes 8 bytes
3 KB
8 bytes 768 bytes 8 bytes 692 bytes 8 bytes 315 KB
8 bytes 114 KB
8 bytes 728 bytes 8 bytes 322 KB
8 bytes 117 KB
8 bytes 836 bytes

Thursday, February 9, 2012 at 8:02:39 PM
Name 20120114_021348_PLD.crc 20120114_021348_PLD.edf 20120114_021348_SAD.crc 20120114_021348_SAD.edf 20120115_024330_EVE.crc 20120115_024330_EVE.edf 20120115_024330_PLD.crc 20120115_024330_PLD.edf 20120115_024330_SAD.crc 20120115_024330_SAD.edf 20120116_005345_EVE.crc 20120116_005345_EVE.edf 20120116_005345_PLD.crc 20120116_005345_PLD.edf 20120116_005345_SAD.crc 20120116_005345_SAD.edf 20120116_015403_EVE.crc 20120116_015403_EVE.edf 20120116_015403_PLD.crc 20120116_015403_PLD.edf 20120116_015403_SAD.crc 20120116_015403_SAD.edf 20120116_052352_EVE.crc 20120116_052352_EVE.edf 20120116_052352_PLD.crc 20120116_052352_PLD.edf 20120116_052352_SAD.crc 20120116_052352_SAD.edf 20120117_005126_EVE.crc 20120117_005126_EVE.edf 20120117_005126_PLD.crc 20120117_005126_PLD.edf 20120117_005126_SAD.crc 20120117_005126_SAD.edf 20120118_022158_EVE.crc 20120118_022158_EVE.edf 20120118_022159_PLD.crc 20120118_022159_PLD.edf 20120118_022159_SAD.crc 20120118_022159_SAD.edf 20120118_064541_EVE.crc 20120118_064541_EVE.edf 20120118_064542_PLD.crc 20120118_064542_PLD.edf 20120118_064542_SAD.crc 20120118_064542_SAD.edf 20120118_065100_EVE.crc 20120118_065100_EVE.edf 20120118_065101_PLD.crc 20120118_065101_PLD.edf 20120118_065101_SAD.crc
Listing of "NO NAME" (Page 2)
Size
8 bytes 265 KB
8 bytes
96 KB
8 bytes 908 bytes 8 bytes 307 KB
8 bytes 111 KB
8 bytes 548 bytes 8 bytes
38 KB
8 bytes
14 KB
8 bytes 728 bytes 8 bytes 135 KB
8 bytes
49 KB
8 bytes 980 bytes 8 bytes 124 KB
8 bytes
45 KB
8 bytes 800 bytes 8 bytes 339 KB
8 bytes 123 KB
8 bytes 692 bytes 8 bytes 171 KB
8 bytes
62 KB
8 bytes 620 bytes 8 bytes
4 KB
8 bytes 1008 bytes 8 bytes 548 bytes 8 bytes
4 KB
8 bytes

Thursday, February 9, 2012 at 8:02:39 PM
Name 20120118_065101_SAD.edf 20120118_070244_EVE.crc 20120118_070244_EVE.edf 20120118_070244_PLD.crc 20120118_070244_PLD.edf 20120118_070244_SAD.crc 20120118_070244_SAD.edf 20120118_071118_EVE.crc 20120118_071118_EVE.edf 20120118_071119_PLD.crc 20120118_071119_PLD.edf 20120118_071119_SAD.crc 20120118_071119_SAD.edf 20120119_020701_EVE.crc 20120119_020701_EVE.edf 20120119_020701_PLD.crc 20120119_020701_PLD.edf 20120119_020701_SAD.crc 20120119_020701_SAD.edf 20120119_020831_EVE.crc 20120119_020831_EVE.edf 20120119_020831_PLD.crc 20120119_020831_PLD.edf 20120119_020831_SAD.crc 20120119_020831_SAD.edf 20120119_095632_EVE.crc 20120119_095632_EVE.edf 20120119_095632_PLD.crc 20120119_095632_PLD.edf 20120119_095632_SAD.crc 20120119_095632_SAD.edf 20120120_010752_EVE.crc 20120120_010752_EVE.edf 20120120_010752_PLD.crc 20120120_010752_PLD.edf 20120120_010752_SAD.crc 20120120_010752_SAD.edf 20120120_071949_EVE.crc 20120120_071949_EVE.edf 20120120_071949_PLD.crc 20120120_071949_PLD.edf 20120120_071949_SAD.crc 20120120_071949_SAD.edf 20120121_005749_EVE.crc 20120121_005749_EVE.edf 20120121_005749_PLD.crc 20120121_005749_PLD.edf 20120121_005749_SAD.crc 20120121_005749_SAD.edf 20120121_051147_EVE.crc 20120121_051147_EVE.edf
Listing of "NO NAME" (Page 3)
Size
1008 bytes 8 bytes 548 bytes 8 bytes
3 KB
8 bytes 768 bytes 8 bytes 656 bytes 8 bytes 110 KB
8 bytes
40 KB
8 bytes 548 bytes 8 bytes
3 KB
8 bytes 768 bytes 8 bytes 1KB
8 bytes 304 KB
8 bytes 110 KB
8 bytes 548 bytes 8 bytes
10 KB
8 bytes
3 KB
8 bytes 1KB
8 bytes 241 KB
8 bytes
87 KB
8 bytes 764 bytes 8 bytes 121 KB
8 bytes
44 KB
8 bytes 656 bytes 8 bytes 165 KB
8 bytes
60 KB
8 bytes 584 bytes
Modified

Thursday, February 9, 2012 at 8:02:39 PM
Name 20120121_051147_PLD.crc 20120121_051147_PLD.edf 20120121_051147_SAD.crc 20120121_051147_SAD.edf 20120121_051441_EVE.crc 20120121_051441_EVE.edf 20120121_051441_PLD.crc 20120121_051441_PLD.edf 20120121_051441_SAD.crc 20120121_051441_SAD.edf 20120121_060659_EVE.crc 20120121_060659_EVE.edf 20120121_060659_PLD.crc 20120121_060659_PLD.edf 20120121_060659_SAD.crc 20120121_060659_SAD.edf 20120122_024723_EVE.crc 20120122_024723_EVE.edf 20120122_024723_PLD.crc 20120122_024723_PLD.edf 20120122_024723_SAD.crc 20120122_024723_SAD.edf 20120122_120000_EVE.crc 20120122_120000_EVE.edf 20120122_120003_PLD.crc 20120122_120003_PLD.edf 20120122_120003_SAD.crc 20120122_120003_SAD.edf 20120123_020325_EVE.crc 20120123_020325_EVE.edf 20120123_020325_PLD.crc 20120123_020325_PLD.edf 20120123_020325_SAD.crc 20120123_020325_SAD.edf 20120124_022242_EVE.crc 20120124_022242_EVE.edf 20120124_022243_PLD.crc 20120124_022243_PLD.edf 20120124_022243_SAD.crc 20120124_022243_SAD.edf 20120125_022753_EVE.crc 20120125_022753_EVE.edf 20120125_022753_PLD.crc 20120125_022753_PLD.edf 20120125_022753_SAD.crc 20120125_022753_SAD.edf 20120125_023351_EVE.crc 20120125_023351_EVE.edf 20120125_023351_PLD.crc 20120125_023351_PLD.edf 20120125_023351_SAD.crc
Listing of "NO NAME" (Page 4)
Size
8 bytes
4 KB
8 bytes
1 KB
8 bytes 692 bytes 8 bytes 35 KB
8 bytes 12 KB
8 bytes 908 bytes 8 bytes 209 KB
8 bytes 76 KB
8 bytes 1KB
8 bytes 359 KB
8 bytes 130 KB
8 bytes 548 bytes 8 bytes
3 KB
8 bytes 768 bytes 8 bytes 692 bytes 8 bytes 323 KB
8 bytes 117 KB
8 bytes 2KB
8 bytes 349 KB
8 bytes 127 KB
8 bytes 548 bytes 8 bytes
3 KB
8 bytes 768 bytes 8 bytes 656 bytes 8 bytes 156 KB
8 bytes

Thursday, February 9, 2012 at 8:02:39 PM
Name 20120125_023351_SAD.edf 20120125_063203_EVE.crc 20120125_063203_EVE.edf 20120125_063203_PLD.crc 20120125_063203_PLD.edf 20120125_063203_SAD.crc 20120125_063203_SAD.edf 20120125_063844_EVE.crc 20120125_063844_EVE.edf 20120125_063845_PLD.crc 20120125_063845_PLD.edf 20120125_063845_SAD.crc 20120125_063845_SAD.edf 20120125_064059_EVE.crc 20120125_064059_EVE.edf 20120125_064100_PLD.crc 20120125_064100_PLD.edf 20120125_064100_SAD.crc 20120125_064100_SAD.edf 20120126_014119_EVE.crc 20120126_014119_EVE.edf 20120126_014119_PLD.crc 20120126_014119_PLD.edf 20120126_014119_SAD.crc 20120126_014119_SAD.edf 20120128_023718_EVE.crc 20120128_023718_EVE.edf 20120128_023718_PLD.crc 20120128_023718_PLD.edf 20120128_023718_SAD.crc 20120128_023718_SAD.edf 20120128_023906_EVE.crc 20120128_023906_EVE.edf 20120128_023906_PLD.crc 20120128_023906_PLD.edf 20120128_023906_SAD.crc 20120128_023906_SAD.edf 20120128_065033_EVE.crc 20120128_065033_EVE.edf 20120128_065033_PLD.crc 20120128_065033_PLD.edf 20120128_065033_SAD.crc 20120128_065033_SAD.edf 20120130_022410_EVE.crc 20120130_022410_EVE.edf 20120130_022411_PLD.crc 20120130_022411_PLD.edf 20120130_022411_SAD.crc 20120130_022411_SAD.edf 20120131_004659_EVE.crc 20120131_004659_EVE.edf
Listing of "NO NAME" (Page 5)
Size
56 KB
8 bytes 584 bytes 8 bytes
6 KB
8 bytes
2 KB
8 bytes 548 bytes 8 bytes
4 KB
8 bytes 1008 bytes 8 bytes 2KB
8 bytes 151 KB
8 bytes
54 KB
8 bytes 2KB
8 bytes 340 KB
8 bytes 123 KB
8 bytes 548 bytes 8 bytes
3 KB
8 bytes 768 bytes 8 bytes 620 bytes 8 bytes 162 KB
8 bytes
58 KB
8 bytes 764 bytes 8 bytes 158 KB
8 bytes
57 KB
8 bytes 2KB
8 bytes 348 KB
8 bytes 126 KB
8 bytes 836 bytes

Thursday, February 9, 2012 at 8:02:39 PM
Name 20120131_004700_PLD.crc 20120131_004700_PLD.edf 20120131_004700_SAD.crc 20120131_004700_SAD.edf 20120131_025602_EVE.crc 20120131_025602_EVE.edf 20120131_025602_PLD.crc 20120131_025602_PLD.edf 20120131_025602_SAD.crc 20120131_025602_SAD.edf 20120202_001551_BRP.crc 20120202_001551_BRP.edf 20120202_001551_EVE.crc 20120202_001551_EVE.edf 20120202_001551_PLD.crc 20120202_001551_PLD.edf 20120202_001551_SAD.crc 20120202_001551_SAD.edf 20120202_002020_BRP.crc 20120202_002020_BRP.edf 20120202_002020_EVE.crc 20120202_002020_EVE.edf 20120202_002020_PLD.crc 20120202_002020_PLD.edf 20120202_002020_SAD.crc 20120202_002020_SAD.edf 20120203_013415_BRP.crc 20120203_013415_BRP.edf 20120203_013415_EVE.crc 20120203_013415_EVE.edf 20120203_013416_PLD.crc 20120203_013416_PLD.edf 20120203_013416_SAD.crc 20120203_013416_SAD.edf 20120204_020419_BRP.crc 20120204_020419_BRP.edf 20120204_020419_EVE.crc 20120204_020419_EVE.edf 20120204_020419_PLD.crc 20120204_020419_PLD.edf 20120204_020419_SAD.crc 20120204_020419_SAD.edf 20120205_012804_BRP.crc 20120205_012804_BRP.edf 20120205_012804_EVE.crc 20120205_012804_EVE.edf 20120205_012804_PLD.crc 20120205_012804_PLD.edf 20120205_012804_SAD.crc 20120205_012804_SAD.edf 20120206_014658_BRP.crc
Listing of "NO NAME" (Page 6)
Size
8 bytes 84 KB
8 bytes 30 KB
8 bytes
1 KB
8 bytes 189 KB
8 bytes 68 KB
8 bytes 768 bytes 8 bytes 548 bytes 8 bytes
3 KB
8 bytes 768 bytes 8 bytes 3.2 MB
8 bytes
2 KB
8 bytes 365 KB
8 bytes 132 KB
8 bytes 2.7 MB
8 bytes
2 KB
8 bytes 303 KB
8 bytes 110 KB
8 bytes 3.2 MB
8 bytes
3 KB
8 bytes 363 KB
8 bytes 132 KB
8 bytes 2.5 MB
8 bytes 800 bytes 8 bytes 285 KB
8 bytes 103 KB
8 bytes

Thursday, February 9, 2012 at 8:02:39 PM
Name 20120206_014658_BRP.edf 20120206_014658_EVE.crc 20120206_014658_EVE.edf 20120206_014659_PLD.crc 20120206_014659_PLD.edf 20120206_014659_SAD.crc 20120206_014659_SAD.edf 20120207_025529_BRP.crc 20120207_025529_BRP.edf 20120207_025529_EVE.crc 20120207_025529_EVE.edf 20120207_025530_PLD.crc 20120207_025530_PLD.edf 20120207_025530_SAD.crc 20120207_025530_SAD.edf 20120208_023634_BRP.crc 20120208_023634_BRP.edf 20120208_023634_EVE.crc 20120208_023634_EVE.edf 20120208_023634_PLD.crc 20120208_023634_PLD.edf 20120208_023634_SAD.crc 20120208_023634_SAD.edf 20120208_063359_BRP.crc 20120208_063359_BRP.edf 20120208_063359_EVE.crc 20120208_063359_EVE.edf 20120208_063400_PLD.crc 20120208_063400_PLD.edf 20120208_063400_SAD.crc 20120208_063400_SAD.edf 20120209_015403_BRP.crc 20120209_015403_BRP.edf 20120209_015403_EVE.crc 20120209_015403_EVE.edf 20120209_015403_PLD.crc 20120209_015403_PLD.edf 20120209_015403_SAD.crc 20120209_015403_SAD.edf
Identification.crc Identification.tgt Journal.dat SETTINGS
ABR.crc ABR.log AGL.crc AGL.tgt BGL.crc BGL.tgt CGL.crc CGL.tgt
Listing of "NO NAME" (Page 7)
Size
3.1 MB
8 bytes
2 KB
8 bytes 348 KB
8 bytes 126 KB
8 bytes 2.6 MB
8 bytes
1 KB
8 bytes 301 KB
8 bytes 109 KB
8 bytes 1.3 MB
8 bytes 944 bytes 8 bytes 153 KB
8 bytes
55 KB
8 bytes 1.7 MB
8 bytes
1 KB
8 bytes 191 KB
8 bytes
69 KB
8 bytes 2.3 MB
8 bytes 1016 bytes 8 bytes 261 KB
8 bytes
94 KB
4 bytes 205 bytes 32 KB
512 KB
8 bytes 172 bytes 4 bytes
36 bytes
4 bytes 226 bytes 4 bytes
25 bytes

Thursday, February 9, 2012 at 8:02:39 PM
Listing of "NO NAME" (Page 8 )
Name Size DGL.crc 4 bytes DGL.tgt 79 bytes EGL.crc 4 bytes EGL.tgt 49 bytes ERR.crc 24 bytes ERR.log 172 bytes IGL.crc 4 bytes IGL.tgt 59 bytes MGL.crc 4 bytes MGL.tgt 275 bytes NGL.crc 4 bytes NGL.tgt 15 bytes PGL.crc 4 bytes PGL.tgt 126 bytes SGL.crc 4 bytes SGL.tgt 51 bytes TXE.crc 8 bytes TXE.log 172 bytes VGL.crc 4 bytes VGL.tgt 47 bytes XGL.crc 4 bytes XGL.tgt 47 bytes ZYL.crc 8 bytes ZYL.log 172 bytes
STR.crc 8 bytes STR.edf 14 KB

Thursday, February 9, 2012 at 8:02:39 PM

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: Seeking Tweaking Advice-ResScan

Post by NateS » Thu Feb 09, 2012 7:36 pm

And I have now sent you a pm to a link which contains the STR.edf file.

Nate

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

User avatar
idamtnboy
Posts: 2186
Joined: Mon Nov 01, 2010 2:12 pm
Location: Idaho

Re: Seeking Tweaking Advice-ResScan

Post by idamtnboy » Fri Feb 10, 2012 3:04 am

NateS wrote:And I have now sent you a pm to a link which contains the STR.edf file.

Nate
Here are what are called the signals in the summary file. I won't explain them all as I'm sure you can decipher most, if not all, of them. The last four are the apnea signals. As you can see there is no CA and OA in the list. That means your ASV is not deciphering and recording CAs and OAs separately but just lumping them together in the AI number. The UAI number would be for unknown cause apneas.

So unfortunately, editing the XML file will do not good as there's no way to retrieve numbers that don't exist. Just for a double check upload one of the EVE.edf files also. They're small. That's the individual mask session file that has the flags for all the events. Even if there are no events to be flagged for a particular session the file will have a header and signal portion that defines what data values are recorded. I don't expect to find an OA and CA entry in it but let's make sure.

Signal Label
1 Mask On
2 Mask Off
3 Mask Events
4 Mask Dur
5 Mode
6 Set Pressure
7 EPAP
8 Max PS
9 Min PS
10 Mask Pres Med
11 Mask Pres 95
12 Mask Pres Max
13 Insp Pres Med
14 Insp Pres 95
15 Insp Pres Max
16 Exp Pres Med
17 Exp Pres 95
18 Exp Pres Max
19 Leak Med
20 Leak 95
21 Leak Max
22 Min Vent Med
23 Min Vent 95
24 Min Vent Max
25 RR Med
26 RR 95
27 RR Max
28 Tid Vol Med
29 Tid Vol 95
30 Tid Vol Max
31 Targ Vent Med
32 Targ Vent 95
33 Targ Vent Max
34 AHI
35 HI
36 AI
37 UAI

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7

User avatar
idamtnboy
Posts: 2186
Joined: Mon Nov 01, 2010 2:12 pm
Location: Idaho

Re: Seeking Tweaking Advice-ResScan

Post by idamtnboy » Fri Feb 10, 2012 3:20 am

I have no idea what all those files are that are in the "page 8" part of your list, except the STR ones. None of them exist in the files generated by the S9 Autoset or Elite APAPs. The PLD, EVE, BRP, and SAD files are probably the same as for the Autoset and Elite.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7

User avatar
esangston
Posts: 51
Joined: Tue Jan 31, 2012 1:13 pm

Re: Seeking Tweaking Advice-ResScan

Post by esangston » Fri Feb 10, 2012 4:03 am

New to all this cpap data stuff, but both resscan and sleepyhead record and show each separate for me.... is the auto really that much more difficult being that it's supposed to be the better choice from what I understand

_________________
Mask: Mirage™ FX For Her Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Started cpap on 1/20/2012 and so far so good!
Yes, I'm married to man with a bad case of narcolepsy. And yup, I have 3 kids under age 4... Adding OSA just seemed like the right thing to do! :-)

User avatar
idamtnboy
Posts: 2186
Joined: Mon Nov 01, 2010 2:12 pm
Location: Idaho

Re: Seeking Tweaking Advice-ResScan

Post by idamtnboy » Fri Feb 10, 2012 11:33 am

esangston wrote:New to all this cpap data stuff, but both resscan and sleepyhead record and show each separate for me.... is the auto really that much more difficult being that it's supposed to be the better choice from what I understand
This topic is dealing with the S9 VPAP Adapt, a whole different machine from the Elite and Autoset, both in operation and price, something like $7000 to $8000 for the Adapt vs. ~$800 for the others. The Autoset is pretty much the same as the Elite, except for its auto adjusting feature.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: Seeking Tweaking Advice-ResScan

Post by NateS » Fri Feb 10, 2012 1:02 pm

idamtnboy wrote: Here are what are called the signals in the summary file. I won't explain them all as I'm sure you can decipher most, if not all, of them. The last four are the apnea signals. As you can see there is no CA and OA in the list. That means your ASV is not deciphering and recording CAs and OAs separately but just lumping them together in the AI number. The UAI number would be for unknown cause apneas.

So unfortunately, editing the XML file will do not good as there's no way to retrieve numbers that don't exist. Just for a double check upload one of the EVE.edf files also. They're small. That's the individual mask session file that has the flags for all the events. Even if there are no events to be flagged for a particular session the file will have a header and signal portion that defines what data values are recorded. I don't expect to find an OA and CA entry in it but let's make sure.
idamtnboy wrote:I have no idea what all those files are that are in the "page 8" part of your list, except the STR ones. None of them exist in the files generated by the S9 Autoset or Elite APAPs. The PLD, EVE, BRP, and SAD files are probably the same as for the Autoset and Elite.
Many thanks for all your interest and efforts!

It seems so strange that a machine which I understand to be specifically designed to treat central apneas does not record and report individual events as being either centrals or obstructives. Seems like an odd design decision.

The files on page 8 are inside a folder called: "SETTINGS"

I am sending you a new pm with a link to a recent EVE.edf as you suggested.

Best regards and with much appreciation,

Nate

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

User avatar
idamtnboy
Posts: 2186
Joined: Mon Nov 01, 2010 2:12 pm
Location: Idaho

Re: Seeking Tweaking Advice-ResScan

Post by idamtnboy » Fri Feb 10, 2012 2:15 pm

NateS wrote:I am sending you a new pm with a link to a recent EVE.edf as you suggested.
Just as I thought. Your machine lumps OA and CA events together. Here's the data list from your file.

Onset Duration Annotation

0 0 Recording starts
4045 16 Apnea
7858 14 Hypopnea
10299 19 Apnea
10328 15 Apnea
10396 11 Apnea
10431 11 Apnea
15116 19 Apnea
15326 15 Apnea
15385 25 Hypopnea
17741 12 Hypopnea
21603 10 Hypopnea
21838 10 Hypopnea
23822 23 Hypopnea

Compare that to the data from one of my Autoset EVE files.

Onset Duration Annotation
0 0 Recording starts
6012 11 Obstructive apnea
11979 14 Central apnea
12542 10 Obstructive apnea
18743 15 Central apnea
21260 11 Central apnea
24461 12 Central apnea
25421 12 Central apnea
25501 14 Central apnea
25530 10 Central apnea
25736 10 Central apnea

See the difference? The Adapt lists "Apnea" where the Autoset lists "Obstructive Apnea" and "Central Apnea".

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7

User avatar
Grand-PAP
Posts: 338
Joined: Sat Dec 03, 2011 5:28 am

Re: Seeking Tweaking Advice-ResScan

Post by Grand-PAP » Fri Feb 10, 2012 3:21 pm

idamtnboy wrote:Just as I thought. Your machine lumps OA and CA events together.
Hi idamtnboy,

That was very impressive and informative!!

Next Wednesday I have an appointment with my sleep Doc to discuss the possibility of getting a ResMed S9 VPAP Adapt or a Respironics System One BiPAP autoSV.

Prior to reading this thread, I think I would have been more interested in the VPAP. However, obviously, my concern is CAs and if VPAP doesn't identify them, maybe I should take a closer look at the autoSV.

QUESTION:

Do you, or anyone else out there, know if the System One BiPAP SV records CAs?

Thanks!

User avatar
idamtnboy
Posts: 2186
Joined: Mon Nov 01, 2010 2:12 pm
Location: Idaho

Re: Seeking Tweaking Advice-ResScan

Post by idamtnboy » Fri Feb 10, 2012 3:31 pm

Grand-PAP wrote:Do you, or anyone else out there, know if the System One BiPAP SV records CAs?
To quote Sargent Schultz, "I know nothing, I know nothing!"

I don't know a bloomin' thing about the other machines' data systems.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: Seeking Tweaking Advice-ResScan

Post by NateS » Fri Feb 10, 2012 3:44 pm

idamtnboy wrote:
NateS wrote:I am sending you a new pm with a link to a recent EVE.edf as you suggested.
Just as I thought. Your machine lumps OA and CA events together. Here's the data list from your file.

Onset Duration Annotation

0 0 Recording starts
4045 16 Apnea
7858 14 Hypopnea
10299 19 Apnea
10328 15 Apnea
10396 11 Apnea
10431 11 Apnea
15116 19 Apnea
15326 15 Apnea
15385 25 Hypopnea
17741 12 Hypopnea
21603 10 Hypopnea
21838 10 Hypopnea
23822 23 Hypopnea

Compare that to the data from one of my Autoset EVE files.

Onset Duration Annotation
0 0 Recording starts
6012 11 Obstructive apnea
11979 14 Central apnea
12542 10 Obstructive apnea
18743 15 Central apnea
21260 11 Central apnea
24461 12 Central apnea
25421 12 Central apnea
25501 14 Central apnea
25530 10 Central apnea
25736 10 Central apnea

See the difference? The Adapt lists "Apnea" where the Autoset lists "Obstructive Apnea" and "Central Apnea".
So I see!

Amazing!

Any thoughts you might have as to why an expensive machine made to treat Central Apnea doesn't bother to distinguish between Centrals and Obstructives???

Nate

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

User avatar
Pugsy
Posts: 65115
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Seeking Tweaking Advice-ResScan

Post by Pugsy » Fri Feb 10, 2012 9:34 pm

Grand-PAP wrote:Do you, or anyone else out there, know if the System One BiPAP SV records CAs?
http://www.healthcare.philips.com/asset ... cSheet.pdf

According to the machine information at above link the PR S1 BiPap AutoSV Advanced...does gather CA.

All I could find for the AVAPS and S/T models is this brochure but it mentions access to clear airway data.
http://bipapst.respironics.com/PDF/BiPAP_Brochure.pdf

I had a link that showed the data available on all the machines but it is dead now and of course I wasn't smart enough to make a copy. Darn it. I get a message that it has been moved..but I have no idea where.

I am thinking the the new PR S1 high end line of ASV machines all gather Clear Airway data. I will continue to look for that spec sheet that had everything listed in a nice little package.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
DreamDiver
Posts: 3082
Joined: Thu Oct 04, 2007 11:19 am

Re: Seeking Tweaking Advice-ResScan

Post by DreamDiver » Fri Feb 10, 2012 10:46 pm

NateS wrote:It seems so strange that a machine which I understand to be specifically designed to treat central apneas does not record and report individual events as being either centrals or obstructives. Seems like an odd design decision.
Do you think it's actually hiding all clear-airway apneas, or do you suppose all apneas have been lumped together in one category, emulating what it might look like on the S8?

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF