Data differences in SleepyHead and ResScan

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
nedi
Posts: 49
Joined: Fri Feb 03, 2012 7:06 am
Location: Sint Eustatius, Dutch Caribbean

Data differences in SleepyHead and ResScan

Post by nedi » Wed Feb 08, 2012 10:09 am

So, I tried ResScan 3.12.016 software this morning due to some data oddities in SleepyHead 0.9.1 and also because I was hoping that RS would show more data (a fella can dream, can't he?).
I opened last night's data into both programs and am surprised at the differences. I had the APAP set from 8-10.
SH showed my med and 95% @ 7.10 and 7.60!? While ResScan showed 9.4 and 9.9.
Also, ResScan showed AHI data for all days of about a months worth of data, while SH just showed for 3 with negative values (another ?!).
Anybody else notice these differences? Perhaps its a version problem with SH. Is there a more recent version? How about for ResScan? Where do I get them?
Thanks

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 Escape Auto. Also available is a Mirage Quattro mask.

User avatar
Pugsy
Posts: 65413
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Data differences in SleepyHead and ResScan

Post by Pugsy » Wed Feb 08, 2012 10:24 am

Latest ResScan version is 3.16....available from a link in UncleBob's signature
memberlist.php?mode=viewprofile&u=38643

Latest SleepyHead version is 0.9.2-1 Beta. Links in my signature line. For data oddities Jedimark will need to address those. I have no idea what might be going on. I use Respironics machine and I also have Encore Pro and other than an occasion rounding off difference because Encore only does rounding to tenths..I have not have any remarkable data differences.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
nedi
Posts: 49
Joined: Fri Feb 03, 2012 7:06 am
Location: Sint Eustatius, Dutch Caribbean

Re: Data differences in SleepyHead and ResScan

Post by nedi » Wed Feb 08, 2012 10:29 am

Pugsy wrote:Latest ResScan version is 3.16....available from a link in UncleBob's signature
memberlist.php?mode=viewprofile&u=38643

Latest SleepyHead version is 0.9.2-1 Beta. Links in my signature line. For data oddities Jedimark will need to address those. I have no idea what might be going on. I use Respironics machine and I also have Encore Pro and other than an occasion rounding off difference because Encore only does rounding to tenths..I have not have any remarkable data differences.
Thanks, Pugs! A great help as always!
I am currently downloading both. Will post the results.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 Escape Auto. Also available is a Mirage Quattro mask.

User avatar
Pugsy
Posts: 65413
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Data differences in SleepyHead and ResScan

Post by Pugsy » Wed Feb 08, 2012 10:42 am

I forgot a little hint with the new SleepyHead Beta version.
If you are using Vista or Windows 7....install under the "run as administrator" command or the automatic updates will fail. You can still update by manually opening the program under the "run as administrator" command and then the update will go through.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
nedi
Posts: 49
Joined: Fri Feb 03, 2012 7:06 am
Location: Sint Eustatius, Dutch Caribbean

Re: Data differences in SleepyHead and ResScan

Post by nedi » Wed Feb 08, 2012 11:09 am

Thanks for the tip. I wondered why it wouldn't update.
The updated version shows the same spurious data in SH.
Guess I'll just make the switch to ResScan.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 Escape Auto. Also available is a Mirage Quattro mask.

User avatar
Pugsy
Posts: 65413
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Data differences in SleepyHead and ResScan

Post by Pugsy » Wed Feb 08, 2012 11:12 am

Send Jedimark a note about the discrepancy. He will want to know and see if it is something he needs to address in SleepyHead. SleepyHead is still an infant and not perfect. Only way for him to fix things is when we bring it to his attention.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Pugsy
Posts: 65413
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Data differences in SleepyHead and ResScan

Post by Pugsy » Wed Feb 08, 2012 11:13 am

Forgot to say

Jedimark may see this thread and chime in....otherwise please post problem over in the Beta version thread for Windows version of SH.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
nedi
Posts: 49
Joined: Fri Feb 03, 2012 7:06 am
Location: Sint Eustatius, Dutch Caribbean

Re: Data differences in SleepyHead and ResScan

Post by nedi » Wed Feb 08, 2012 12:19 pm

Pugsy wrote:Forgot to say

Jedimark may see this thread and chime in....otherwise please post problem over in the Beta version thread for Windows version of SH.
Done!

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 Escape Auto. Also available is a Mirage Quattro mask.

User avatar
b360155
Posts: 163
Joined: Sat Oct 22, 2011 7:46 pm
Location: North Richland Hills, TX

Re: Data differences in SleepyHead and ResScan

Post by b360155 » Wed Feb 08, 2012 1:20 pm

viewtopic.php?f=1&t=73749&st=0&sk=t&sd=a

See my earlier thread. I've been having issues as far as maximum leak rates and usage hours being out of sync between the two software programs that Jedimark is looking into. On the leak rates, my daily detail graphs look the same between the two but ResScan keeps coming up with a maximum leak of 22.8 or 24.0, substantially higher than the actual maximum showing up in the daily detail graphs which Sleepyhead appears to be capturing correctly. On the usage hours, ResScan comes up with the right amount of total hours and Sleepyhead is coming up with some crazy total usage hours which I've posted in the Bug Tracker section of the SourceForge.net site here:

http://sourceforge.net/tracker/?func=de ... id=2089181

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Using ResScan Version 3.16 Software and Sleepyhead 0.9.2

User avatar
idamtnboy
Posts: 2186
Joined: Mon Nov 01, 2010 2:12 pm
Location: Idaho

Re: Data differences in SleepyHead and ResScan

Post by idamtnboy » Wed Feb 08, 2012 3:05 pm

nedi wrote:I opened last night's data into both programs and am surprised at the differences. I had the APAP set from 8-10.
SH showed my med and 95% @ 7.10 and 7.60!? While ResScan showed 9.4 and 9.9.
I've seen this same sort of discrepancy when doing the average, median, maximum, and 95% calculations in Lotus from the raw data from the S9. These same calculations are performed inside the S9 and the results are part of the summary data file, which is separate from the raw data files. So far I haven't figured out why, unless it has to do with the S9 using a few more seconds of data than what get recorded in the data file.

Take a read through the topic b360155 links to in his comment. I share my observations there in a fair amount of detail.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7

User avatar
Papit
Posts: 844
Joined: Sun Dec 18, 2011 4:47 pm
Location: USA CA

Re: Data differences in SleepyHead and ResScan

Post by Papit » Wed Feb 08, 2012 9:42 pm

Hi, I've been lurking and following this thread conversation. I'm wondering what might be the attraction for downloading SH or any other available CPAP software to a user who already has ResScan. Are there any additional features or functionality in SH that go beyond what ResScan has? Thanks.

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: Machine: AirCurve 10 ASV (37043), Software:ResScan 5.7.0.9477, SleepyHead V1.00BETA2, Oximeter:CMS-50i

DrBucky
Posts: 31
Joined: Mon Feb 06, 2012 7:41 pm

Re: Data differences in SleepyHead and ResScan

Post by DrBucky » Wed Feb 08, 2012 9:50 pm

Papit wrote:Hi, I've been lurking and following this thread conversation. I'm wondering what might be the attraction for downloading SH or any other available CPAP software to a user who already has ResScan. Are there any additional features or functionality in SH that go beyond what ResScan has? Thanks.
Mainly if you have a mac/linux and don't want to purchase(or find if you are a pirate) a copy of windows/go through the bother of booting into a different OS just to look at your XPAP data.

Also it integrates with the CM50D+ (and higher) series of pulse ox monitors. Although to get the data off of a CM50D+ you have to use windows, but sleepy head does let you match up the pulse ox data to the CPAP data (assuming you have the time set correctly on both devices). Such is probably the only time you'd use sleepy head on a windows machine.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: climate line hose rocks

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Data differences in SleepyHead and ResScan

Post by archangle » Thu Feb 09, 2012 2:35 am

Papit wrote:Hi, I've been lurking and following this thread conversation. I'm wondering what might be the attraction for downloading SH or any other available CPAP software to a user who already has ResScan. Are there any additional features or functionality in SH that go beyond what ResScan has? Thanks.
Try it and you'll see the differences.

Unfortunately, no one can be told what The Matrix is. You have to see it for yourself.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

DrPepper00
Posts: 139
Joined: Wed Dec 14, 2011 9:29 am

Re: Data differences in SleepyHead and ResScan

Post by DrPepper00 » Thu Feb 09, 2012 11:38 am

Papit wrote:Hi, I've been lurking and following this thread conversation. I'm wondering what might be the attraction for downloading SH or any other available CPAP software to a user who already has ResScan. Are there any additional features or functionality in SH that go beyond what ResScan has? Thanks.
It is the difference between a black and white movie and a Technicolor 3-D movie.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Ruby chinstrap SleepyHead Software