Am I having centrals?

Post by fortomorrow » Sun Feb 01, 2009 2:58 pm

First off, hello everyone, long time, no see, etc, etc.

I had gastric bypass surgery three weeks ago. I've lost about 35 lbs.

Now, before surgery my AHI was consistently under 1. I stopped using it for three weeks because I have problems with aerophagia and I was really concerned about inflating my tiny new stomach.

Anyway. I just started using my APAP again three nights ago, and my AHI is up to 4.5. I know that's still really good for most people, but it concerns me. When I look at the reports on Encore Pro it shows that I'm having several apneas at once in a cluster - which indicates centrals, yes?

So I'm wondering - because I've lost a decent amount of weight in a short time, should I try lowering my pressure and see if that doesn't take care of this problem? Or is it merely a matter of getting used to XPAP again?

Thanks guys.
Have you always wanted to try the Liberty? Test drive it here!

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: Am I having centrals?

Post by Wulfman » Sun Feb 01, 2009 3:29 pm

Are any "NR" (Non-Responsive) events showing up on your reports?
Can you determine at what pressure they're occurring?
Are they happening consistently over several nights?
Is it possible that they're "positional" in nature?
Is your pressure range still the same? (if they're happening at lower pressures, maybe the minimum is too low)
Are you using A-Flex or what? What are all of your settings)?

That wasn't quite "20 questions", but it's a start.....


Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

ozij
Posts: 10474
Joined: Fri Mar 18, 2005 11:52 pm

Re: Am I having centrals?

Post by ozij » Sun Feb 01, 2009 3:32 pm

A cluster does not necessarily mean a central apnea.
Is the machine tagging anything as NR (non-responsive ---> possbily central)?
O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

fortomorrow
Posts: 174
Joined: Thu Sep 11, 2008 9:57 am
Location: Sacramento, CA

Re: Am I having centrals?

Post by fortomorrow » Sun Feb 01, 2009 3:37 pm

Wulfman wrote:Are any "NR" (Non-Responsive) events showing up on your reports?
Can you determine at what pressure they're occurring?
Are they happening consistently over several nights?
Is it possible that they're "positional" in nature?
Is your pressure range still the same? (if they're happening at lower pressures, maybe the minimum is too low)
Are you using A-Flex or what? What are all of your settings)?

That wasn't quite "20 questions", but it's a start.....
Thanks Den!

There aren't any NR events.

Pressure range is still 8-20. AFlex is set at 1. Humidity is 1.

The 90% pressure is 9. Average pressure is 8.3.

I'm stumped by the whole thing. My AHI has always been really low!
Have you always wanted to try the Liberty? Test drive it here!

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: Am I having centrals?

Post by Wulfman » Sun Feb 01, 2009 4:35 pm

fortomorrow wrote:
Wulfman wrote:Are any "NR" (Non-Responsive) events showing up on your reports?
Can you determine at what pressure they're occurring?
Are they happening consistently over several nights?
Is it possible that they're "positional" in nature?
Is your pressure range still the same? (if they're happening at lower pressures, maybe the minimum is too low)
Are you using A-Flex or what? What are all of your settings)?

That wasn't quite "20 questions", but it's a start.....
Thanks Den!

There aren't any NR events.

Pressure range is still 8-20. AFlex is set at 1. Humidity is 1.

The 90% pressure is 9. Average pressure is 8.3.

I'm stumped by the whole thing. My AHI has always been really low!
That's good to hear about the NR events.
If the Apneas are occurring at or around your minimum pressure, maybe you need to bump it up (to your 90% number).
With that machine, you've got a lot of options to try, too. Maybe switching from A-Flex to C-Flex.....?

Congratulations on your surgery. Hope everything goes well. My son had that procedure done in December.

Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

track
Posts: 561
Joined: Wed Dec 27, 2006 10:08 am

Re: Am I having centrals?

Post by track » Sun Feb 01, 2009 4:39 pm

It's my understanding having apneas in clusters is not necessarily a sign they are centrals. I can have clusters when I get on my back....move back on my side and they disappear.

_________________
Mask
Additional Comments: Strictly a side sleeper

User avatar
Georgio
Posts: 608
Joined: Thu Aug 28, 2008 5:34 pm
Location: Jacksonville Beach, Fl

Re: Am I having centrals?

Post by Georgio » Sun Feb 01, 2009 5:10 pm

Congratulations, you have a new body now.....so your apap needs may also be changing. Does anyone think that significant weight loss could lead to lower required pressures?

Georgio
DreamStation 2, Oscar
Resmed AirFit P30i Nasal Mask

User avatar
Julie
Posts: 20052
Joined: Tue Feb 28, 2006 12:58 pm

Re: Am I having centrals?

Post by Julie » Sun Feb 01, 2009 8:54 pm

Tell you a secret - but from what I've read, seen and heard, having bypass usually eliminates the need for Cpap altogether, cures GERD and possibly even diabetes (and I'm not talking about snake-oil stories, but real serious research). You may be just in the early stages of things, so wait a while and see where things stand.

fortomorrow
Posts: 174
Joined: Thu Sep 11, 2008 9:57 am
Location: Sacramento, CA

Re: Am I having centrals?

Post by fortomorrow » Mon Feb 02, 2009 7:59 am

Julie wrote:Tell you a secret - but from what I've read, seen and heard, having bypass usually eliminates the need for Cpap altogether, cures GERD and possibly even diabetes (and I'm not talking about snake-oil stories, but real serious research). You may be just in the early stages of things, so wait a while and see where things stand.
Yes, I know. It cures GERD and acid reflux because there's no longer any acid in the stomach. They haven't quite figured out how it cures type 2 diabetes (not type 1 insulin dependent though) yet - although the results are instantaneous. Apparently type 2 diabetes goes away the day you have surgery - I wouldn't know, I didn't have it. It also seems to cure hypertension as well.

As for eliminating the need for CPAP - that's only the case if your OSA is caused by obesity, and as we all know, not ALL sleep apnea is the direct result of being fat. There are quite a few people on my WLS support message board that have stopped using their CPAP, some as directed by a doctor after a sleep study, and some not. I worry about the ones who assume that they've been cured of their sleep apnea, because it's not necessarily the case.

Especially as it seems that I've lost weight and I need to INCREASE my pressure. BOOOOOOOO.
Have you always wanted to try the Liberty? Test drive it here!

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: Am I having centrals?

Post by Wulfman » Mon Feb 02, 2009 11:25 am

I had read a "blip" sometime back about the procedure "curing" Type 2 diabetes, so your posts about it made me curious. I just did a Google search on the following phrase and got lots of links. I haven't read them all yet, but glanced through a few.

"how gastric bypass cures Type 2 diabetes"

Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

fortomorrow
Posts: 174
Joined: Thu Sep 11, 2008 9:57 am
Location: Sacramento, CA

Re: Am I having centrals?

Post by fortomorrow » Tue Feb 03, 2009 7:27 am

Okay, I bumped my pressure up to 9 last night. AHI was 3.6 which is still pretty high considering, and my 90% pressure went up to 10. Does that mean I should move the pressure up to 10 as well?

Or should I just leave it and accept 3.6? I feel okay this morning. I didn't wake up at all last night for the first time in weeks (thank god). But looking at the charts just makes me feel unhappy. Over 4 minutes in apnea and all those events...

Image

I think it's lame that I'm losing weight and my pressure is going up. Everyone else I know got to go down in pressure, or of CPAP completely.

Or maybe it's a longer term side effect of general anaesthetic. Who knows?

I'll stop whining now...
Have you always wanted to try the Liberty? Test drive it here!

User avatar
Engransan
Posts: 70
Joined: Sat May 24, 2008 10:36 am
Location: Southeast USA

Re: Am I having centrals?

Post by Engransan » Tue Feb 03, 2009 8:33 am

Following my coronary artery bypass surgery I had about two weeks of erratic data.
Ken
"OLD AGE AIN'T NO PLACE FOR SISSIES"

User avatar
Hawthorne
Posts: 3972
Joined: Tue Oct 19, 2004 4:46 am
Location: London Ontario -Canada

Re: Am I having centrals?

Post by Hawthorne » Tue Feb 03, 2009 9:08 am

I think your body is still recovering from surgery and all that is involved in the procedure.

Engransan is right. You need 6 to 8 weeks or maybe even more for your body to recover. Your data is bound to be erratic during that recovery time.

Bumping your minimum to 9 seems to be helpful for now. You may be able to lower than again as your body heals and adjusts. You could raise it to 10 but, in my opinion, things are going to be very erratic for awhile yet. Your AHI, at 3.6, isn't to bad so maybe you just need to leave the settings for a week or so , as we all should do when making any pressure changes.

You've done a major thing to help your health! Good for you! Give it time.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments:  Backups- FX Nano masks. Backup machine- Airmini auto travel cpap

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: Am I having centrals?

Post by Wulfman » Tue Feb 03, 2009 9:41 am

Your call.
You might leave it where it's at for a few more nights to see if it's consistent. (and then, if it is, bump it up)

I've read a few posts by forum members that DID have to bump up their pressure when they lost weight.

Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow