Standalone Anti-aspyxiation valves
Standalone Anti-aspyxiation valves
Full face masks, hybrid masks, and oral masks have anti-asphyxiation valves. If the power goes off, you can breathe fresh air without removing the mask. On the Oracle oral mask, it's a separate piece that fits on the end of the hose, but it won't fit on a standard nasal mask.
Does anyone make an anti-asphyxiation valve that fits between the mask and the hose? That way, you'd get fresh air with a nasal mask even if the power goes off.
I think this part would work, except the fittings are for the wrong end of the hose and it wouldn't work if you put it on the mask end.
Does anyone make an anti-asphyxiation valve that fits between the mask and the hose? That way, you'd get fresh air with a nasal mask even if the power goes off.
I think this part would work, except the fittings are for the wrong end of the hose and it wouldn't work if you put it on the mask end.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
- chunkyfrog
- Posts: 34545
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Standalone Anti-aspyxiation valves
I understand that hospital-style cpap masks do not have the valve built in; and thus the valve needs to be elsewhere.
Hospital supply places?
Hospital supply places?
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Standalone Anti-aspyxiation valves
Deleted reply.
_________________
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack |
Additional Comments: PR System One Remstar BiPap Auto AS Advanced. |
Last edited by Mary Z on Tue May 31, 2011 8:56 am, edited 2 times in total.
Dog is my copilot
Re: Standalone Anti-aspyxiation valves
No, I'm using the mask in my profile, not a hospital mask.Mary Z wrote:Have you tried your DME? They also might have more information about the mask you're using. I have never heard that hospital masks would differ in that way. Are you using a hospital mask? If so it might be easier and safer to go with a regular mask. Good luck.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
-
- Posts: 122
- Joined: Mon Jun 27, 2011 11:32 am
Re: Standalone Anti-aspyxiation valves
If such a thing would exist it would be great! It is quite easy to invent... I wonder if and when some company will make it.archangle wrote:Does anyone make an anti-asphyxiation valve that fits between the mask and the hose? That way, you'd get fresh air with a nasal mask even if the power goes off.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Quattro FX and Mirage Quattro as backup masks |
Re: Standalone Anti-aspyxiation valves
You could modify the valve or make an adapter.archangle wrote:I think this part would work, except the fittings are for the wrong end of the hose and it wouldn't work if you put it on the mask end.
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Software: SleepyHead 0.9 beta |
Download Sleepyhead here: https://sourceforge.net/projects/sleepyhead/
Re: Standalone Anti-aspyxiation valves
The Oracle's, Forma's and Flexifit's antiasphyxia valve are all the same part. You can take the valve out of the elbow piece and put it in the Opus.
https://www.cpap.com/productpage/fisher ... w-kit.html

You can isert the valve at the end of the hose (right side) instead of the swivel connector.

https://www.cpap.com/productpage/fisher ... w-kit.html

You can isert the valve at the end of the hose (right side) instead of the swivel connector.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Re: Standalone Anti-aspyxiation valves
The none rebreathing valve is made to open at a certified pressure per code. The one- way valve that you mention is probably not made to this standard. Otherwise , you could use a water pipe and put some chashish in it.archangle wrote:Full face masks, hybrid masks, and oral masks have anti-asphyxiation valves. If the power goes off, you can breathe fresh air without removing the mask. On the Oracle oral mask, it's a separate piece that fits on the end of the hose, but it won't fit on a standard nasal mask.
Does anyone make an anti-asphyxiation valve that fits between the mask and the hose? That way, you'd get fresh air with a nasal mask even if the power goes off.
I think this part would work, except the fittings are for the wrong end of the hose and it wouldn't work if you put it on the mask end.

Does it not Stand Alone?
The other night I sneezed into my F&P #431 mask and blew off the none rebreathing valve leaves. Luckily, I have also the #432 mask from which I borrowed the valve. The next day I fixed those leaves with a pair of twizers.
Ozie is right in this case:
https://www.cpap.com/productpage/fisher ... w-kit.html]
_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6 |
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
Re: Standalone Anti-aspyxiation valves
Yes, that's the right idea, except that it doesn't fit between a normal mask and hose. One end is "normal," but the other end isn't. The right angle also makes it a little difficult to use.ozij wrote:The Oracle's, Forma's and Flexifit's antiasphyxia valve are all the same part. You can take the valve out of the elbow piece and put it in the Opus.
https://www.cpap.com/productpage/fisher ... w-kit.html
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Re: Standalone Anti-aspyxiation valves
If you look closely at the end of this product ( see below ) the anti-asphyxia valve is a separable piece.
The right end fits a normal CPAP hose but the left end is slightly different. It's not so much so that one would have great difficulty in adapting it.
https://www.cpap.com/productpage/Fisher ... Valve.html
I just checked my Oracle 2 mask and easily detached the valve and then replaced it. To make it adapt to something else, I's suggest the infamous "Duck Tape" in your choice of colors.
ozij posted above a different part but it sure looks like the valve is exactly the same and should be removable like on the Oracle 2 mask tubing.
The right end fits a normal CPAP hose but the left end is slightly different. It's not so much so that one would have great difficulty in adapting it.
https://www.cpap.com/productpage/Fisher ... Valve.html
I just checked my Oracle 2 mask and easily detached the valve and then replaced it. To make it adapt to something else, I's suggest the infamous "Duck Tape" in your choice of colors.
ozij posted above a different part but it sure looks like the valve is exactly the same and should be removable like on the Oracle 2 mask tubing.
_________________
Mask: Oracle HC452 Oral CPAP Mask |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV |
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
Re: Standalone Anti-aspyxiation valves
[quote="archangle"]
Does anyone make an anti-asphyxiation valve that fits between the mask and the hose? That way, you'd get fresh air with a nasal mask even if the power goes off.
Isn't this doing what you want?
http://www.easybreathe.com.vhost.zerola ... p2782.html
Take a caliper and check the diameters of the parts.
Does anyone make an anti-asphyxiation valve that fits between the mask and the hose? That way, you'd get fresh air with a nasal mask even if the power goes off.
Isn't this doing what you want?
http://www.easybreathe.com.vhost.zerola ... p2782.html
Take a caliper and check the diameters of the parts.
_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6 |
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
Re: Standalone Anti-aspyxiation valves
Archangle,
I assume this sentence "You can take the valve out of the elbow piece " was not clear enough.
The elbow (right angle piece) and the anti-asphyxia valve are two different parts. As JSS74 said, you can pull them apart.
You are then left with an elbow in one hand, and the anti-asphyxia valve in the other.
You can check the mask manuals and the F&P parts catalog for confirmation.
O.
I assume this sentence "You can take the valve out of the elbow piece " was not clear enough.
The elbow (right angle piece) and the anti-asphyxia valve are two different parts. As JSS74 said, you can pull them apart.
You are then left with an elbow in one hand, and the anti-asphyxia valve in the other.
You can check the mask manuals and the F&P parts catalog for confirmation.
O.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Re: Standalone Anti-aspyxiation valves
Yes, I know, thanks.ozij wrote:Archangle,
I assume this sentence "You can take the valve out of the elbow piece " was not clear enough.
The elbow (right angle piece) and the anti-asphyxia valve are two different parts. As JSS74 said, you can pull them apart.
You are then left with an elbow in one hand, and the anti-asphyxia valve in the other.
You can check the mask manuals and the F&P parts catalog for confirmation.
O.
I took the anti-asphyxia valve out of my oracle mask before I started this thread and even with the right angle piece removed, the connectors don't match up to connect it between the mask and the hose. If I leave the right angle piece in, those connectors don't match up either. I don't have the mask handy right now, but I'll look at it again later and see if I missed anything.
Duct tape isn't a very practical solution because it would be a problem when I clean the mask. It didn't look like an adapter would be very convenient to make and use, but I'll take a look at it again and see if anything looks like it would be practical.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
-
- Posts: 5
- Joined: Sat Jul 02, 2011 9:22 am
Re: Standalone Anti-aspyxiation valves
I'm a PC tech that just got a CPAP recently, so I can't comment on parts, dimensions, and fittings. But I might have a solution, http://www.newegg.com/Product/Product.a ... 6842102117 A battery backup (UPS) will give you power for a brief blackout, and it's warning tone would wake you up before it runs out of power. It's rated to power a full desktop PC (Monitor & Tower) for about 8 minutes, but I imagine it would last significantly longer with a CPAP because of the discrepancy in power drain.
Re: Standalone Anti-aspyxiation valves
Reply:archangle wrote:Full face masks, hybrid masks, and oral masks have anti-asphyxiation valves. If the power goes off, you can breathe fresh air without removing the mask. On the Oracle oral mask, it's a separate piece that fits on the end of the hose, but it won't fit on a standard nasal mask.
Does anyone make an anti-asphyxiation valve that fits between the mask and the hose? That way, you'd get fresh air with a nasal mask even if the power goes off.
I think this part would work, except the fittings are for the wrong end of the hose and it wouldn't work if you put it on the mask end.
Archangel, I think that you make a bit of a fuss about this issue. To start with, it is not called anti-asphyxiation valves but non-rebreathing valves. You would not get asphyxiated during power stoppage b/c CPAP masks have intentional leak holes. But you would breathe air a bit higher with CO2.
About your question: Does anyone make an anti-asphyxiation valve that fits between the mask and the hose? That way, you'd get fresh air with a nasal mask even if the power goes off. Why if you use a nasal mask can't you just open your mouth to breathe in such cases?
_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6 |
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png