Another Medical Adventure: M-Series Auto beeps and blinks!!
Another Medical Adventure: M-Series Auto beeps and blinks!!
Last night my CPAP machine broke down at 4am and I woke to my partner shouting at me "Stop the beeping"
My M-series Auto now waits for a couple of seconds after powering up and then its lights blink and it beeps continuously
I suspect (but don’t know) that it is more than five years old. Luckily I am covered under Original Medicare and a Medicare Supplement. The original supplier , billmyinsurance,(loved you guys) doesn't do it any more.
Is it possible to replace it and pay for it under my Medicare and Medicare Supplement ?
If so what do I need to do to make it happen ?
I am looking for a
a/ Auto machine
b/ with a heated humidifier
c/ with a way of monitoring the data myself on my computer. I currently use a Smart Card reader and Encore Pro
d/ is QUIET. Otherwise my partner throws me out of her bed!!!
Comments and suggestions about procedures on how to get a replacement machine and choice of machine welcomed
Thanks
John M ( yes the John M from years ago)
Feeling grateful I have a Remstar plus for backup
My M-series Auto now waits for a couple of seconds after powering up and then its lights blink and it beeps continuously
I suspect (but don’t know) that it is more than five years old. Luckily I am covered under Original Medicare and a Medicare Supplement. The original supplier , billmyinsurance,(loved you guys) doesn't do it any more.
Is it possible to replace it and pay for it under my Medicare and Medicare Supplement ?
If so what do I need to do to make it happen ?
I am looking for a
a/ Auto machine
b/ with a heated humidifier
c/ with a way of monitoring the data myself on my computer. I currently use a Smart Card reader and Encore Pro
d/ is QUIET. Otherwise my partner throws me out of her bed!!!
Comments and suggestions about procedures on how to get a replacement machine and choice of machine welcomed
Thanks
John M ( yes the John M from years ago)
Feeling grateful I have a Remstar plus for backup
_________________
| Machine | Mask | Humidifier | ||
![]() | ![]() | ![]() | ||
| Additional Comments: Getting tired of feeling exhausted so often during the day and having to nap | ||||
Coming back to learn some more
Re: Another Medical Adventure: M-Series Auto beeps and blinks!!
Mine did that once when I had put the card in the wrong way. Is that possible?
Re: Another Medical Adventure: M-Series Auto beeps and blinks!!
Look at the lcd to see what the message says.JohnMudie wrote:Last night my CPAP machine broke down at 4am and I woke to my partner shouting at me "Stop the beeping"
My M-series Auto now waits for a couple of seconds after powering up and then its lights blink and it beeps continuously
_________________
| Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
| Additional Comments: New users can't remember they can't remember YET! |
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember

If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!
- SleepingUgly
- Posts: 4690
- Joined: Sat Nov 28, 2009 9:32 pm
Re: Another Medical Adventure: M-Series Auto beeps and blinks!!
Is this the model that had the recall on the humidifier? I think there was some light blinking or beeping associated with the humidifier going...
_________________
| Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Rescan 3.10 |
Never put your fate entirely in the hands of someone who cares less about it than you do. --Sleeping Ugly
Re: Another Medical Adventure: M-Series Auto beeps and blinks!!
Beeps and blinks with card outkatcw wrote:Mine did that once when I had put the card in the wrong way. Is that possible?
Thanks for idea
_________________
| Machine | Mask | Humidifier | ||
![]() | ![]() | ![]() | ||
| Additional Comments: Getting tired of feeling exhausted so often during the day and having to nap | ||||
Coming back to learn some more
- SleepingUgly
- Posts: 4690
- Joined: Sat Nov 28, 2009 9:32 pm
Re: Another Medical Adventure: M-Series Auto beeps and blinks!!
Do a google search for the machine you have and add "humidifier recall".
_________________
| Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Rescan 3.10 |
Never put your fate entirely in the hands of someone who cares less about it than you do. --Sleeping Ugly
Re: Another Medical Adventure: M-Series Auto beeps and blinks!!
It'll beep if the card is put in backwards...or if the card is inserted with the wrong side against the electrical contacts (inside....not visible.....just flip the card & re-insert).
- DreamDiver
- Posts: 3082
- Joined: Thu Oct 04, 2007 11:19 am
Re: Another Medical Adventure: M-Series Auto beeps and blinks!!
It shouldn't blink or beep with the card out. Is this an intermittent problem, does it happen now every time you turn it on? Have you tried unplugging and replugging in the machine from the wall socket?JohnMudie wrote:Beeps and blinks with card outkatcw wrote:Mine did that once when I had put the card in the wrong way. Is that possible?
Thanks for idea
You may need to speak with the office of the insurance commissioner within your state to find out what avenues are available to you through medicare.
_________________
| Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions |
| Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF |
Re: Another Medical Adventure: M-Series Auto beeps and blinks!!
Is it possible that water got sloshed around in your humidifier then got into the machine?
Brooke
Re: Another Medical Adventure: M-Series Auto beeps and blinks!!
JohnM,
I'm going to guess that the machine itself beeps twice as well as the blue lights at the same time. On the display it shows a triangle with a ! in the center. If this is the case then there is a major problem with the machine.

I have the same problem with one of mine, this started after it got wet inside.
David
I'm going to guess that the machine itself beeps twice as well as the blue lights at the same time. On the display it shows a triangle with a ! in the center. If this is the case then there is a major problem with the machine.

I have the same problem with one of mine, this started after it got wet inside.
David
_________________
| Mask: ComfortGel Blue Nasal CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Wisp Mask, Pressure is set between 8.5 and 12.5. With a M Series Auto as a backup. Also a CMS-50D+ PulseOX |
Look twice, save a life. You never know when you might see a motorcyclist.
- 2flamingos
- Posts: 439
- Joined: Mon Apr 27, 2009 7:26 pm
- Location: L.A. (Lower Alabama)
Re: Another Medical Adventure: M-Series Auto beeps and blinks!!
Mine did that when the humidifier went out. New humidifier, and everything is as good as new. Have you tried it without the humnidifier attached?
_________________
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Pressure - 12-16 ClimateLine Hose. Zeo Mobile, Chin Strap |
Old enough to know better
Karma is a Wonderful Thing.
Karma is a Wonderful Thing.
- DeVilbiss Marketing
- Posts: 116
- Joined: Sat Mar 28, 2009 9:35 am
Re: Another Medical Adventure: M-Series Auto beeps and blinks!!
This text was taken from the Meidcare Positive Airway Pressure (PAP) Local Coverage Determination (LCD)
"If a PAP device is replaced during the 5 year reasonable useful lifetime (RUL) because of loss, theft, or irreparable damage due to a specific incident, there is no requirement for a new clinical evaluation, sleep test, or trial period.
If a PAP device is replaced following the 5 year RUL, there must be a face-to-face evaluation by their treating physician that documents that the beneficiary continues to use and benefit from the PAP device. There is no requirement for a new sleep test or trial period."
"If a PAP device is replaced during the 5 year reasonable useful lifetime (RUL) because of loss, theft, or irreparable damage due to a specific incident, there is no requirement for a new clinical evaluation, sleep test, or trial period.
If a PAP device is replaced following the 5 year RUL, there must be a face-to-face evaluation by their treating physician that documents that the beneficiary continues to use and benefit from the PAP device. There is no requirement for a new sleep test or trial period."
- DreamDiver
- Posts: 3082
- Joined: Thu Oct 04, 2007 11:19 am
Re: Another Medical Adventure: M-Series Auto beeps and blinks!!
Excellent information - Thanks DeVilbiss Marketing.DeVilbiss Marketing wrote:This text was taken from the Meidcare Positive Airway Pressure (PAP) Local Coverage Determination (LCD)
"If a PAP device is replaced during the 5 year reasonable useful lifetime (RUL) because of loss, theft, or irreparable damage due to a specific incident, there is no requirement for a new clinical evaluation, sleep test, or trial period.
If a PAP device is replaced following the 5 year RUL, there must be a face-to-face evaluation by their treating physician that documents that the beneficiary continues to use and benefit from the PAP device. There is no requirement for a new sleep test or trial period."
_________________
| Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions |
| Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF |
Re: Another Medical Adventure: M-Series Auto beeps and blinks!!
I have been sidetracked by another medical adventure involving nternal bleeding and
hospitalization. I now have a spare of favourite mask packed in my hospital kit so I can jst
hook it up to their machine. I am not out of the hospital and finally talking to my
Respiratory Therapist who is investigating the situation.
So here are my responses to your comments and suggestions. Thank yuo for your support.
John
[quote="GumbyCT"
Look at the lcd to see what the message says.[/quote]
It has an exclamation point in a triabgle and at the top is says "Service Required"
connected or involve
Thanks for idea[/quote]It shouldn't blink or beep with the card out. Is this an intermittent
problem, does it happen now every time you turn it on? Have you tried unplugging and
replugging in the machine from the wall socket?
You may need to speak with the office of the insurance commissioner within your state to
find out what avenues are available to you through medicare.[/quote]
I have plugged and unplugged. Happens every time I turn it on. Going to try my DME before
heading for the insurance commissioner. Nice Thought tho.
quote="bap40"]Is it possible that water got sloshed around in your humidifier then got into
the machine?[/quote]
I am guessing I didnt make sure that all the water was out. I know I emptied it before
moving it. Next time I am going to make real sure ALL the water is out.
quote="drj130"]JohnM,
I'm going to guess that the machine itself beeps twice as well as the blue lights at the
same time. On the display it shows a triangle with a ! in the center. If this is the case
then there is a major problem with the machine.

I have the same problem with one of mine, this started after it got wet inside.
David[/quote]
Yeom I think this is what is hapeing with mine,
quote="2flamingos"]Mine did that when the humidifier went out. New humidifier, and
everything is as good as new. Have you tried it without the humnidifier attached?[/quote]
Yes, still blinks
hospitalization. I now have a spare of favourite mask packed in my hospital kit so I can jst
hook it up to their machine. I am not out of the hospital and finally talking to my
Respiratory Therapist who is investigating the situation.
So here are my responses to your comments and suggestions. Thank yuo for your support.
John
I pulled the card out and in. Didn't change anythingkatcw wrote:Mine did that once when I had put the card in the wrong way. Is that
possible?
[quote="GumbyCT"
Look at the lcd to see what the message says.[/quote]
It has an exclamation point in a triabgle and at the top is says "Service Required"
I tried to check on this. But I find that it blinks and beeps when the hunidifer is notSleepingUgly wrote:Is this the model that had the recall on the humidifier? I think
there was some light blinking or beeping associated with the humidifier going...
connected or involve
Did that. Didnt seem to be relevantSleepingUgly wrote:Do a google search for the machine you have and add "humidifier
recall".
Beeps and blinks with card outDreamDiver wrote:[quote="JohnMu]Mine did that once when I had put the
card in the wrong way. Is that possible?
Thanks for idea[/quote]It shouldn't blink or beep with the card out. Is this an intermittent
problem, does it happen now every time you turn it on? Have you tried unplugging and
replugging in the machine from the wall socket?
You may need to speak with the office of the insurance commissioner within your state to
find out what avenues are available to you through medicare.[/quote]
I have plugged and unplugged. Happens every time I turn it on. Going to try my DME before
heading for the insurance commissioner. Nice Thought tho.
quote="bap40"]Is it possible that water got sloshed around in your humidifier then got into
the machine?[/quote]
I am guessing I didnt make sure that all the water was out. I know I emptied it before
moving it. Next time I am going to make real sure ALL the water is out.
quote="drj130"]JohnM,
I'm going to guess that the machine itself beeps twice as well as the blue lights at the
same time. On the display it shows a triangle with a ! in the center. If this is the case
then there is a major problem with the machine.

I have the same problem with one of mine, this started after it got wet inside.
David[/quote]
Yeom I think this is what is hapeing with mine,
quote="2flamingos"]Mine did that when the humidifier went out. New humidifier, and
everything is as good as new. Have you tried it without the humnidifier attached?[/quote]
Yes, still blinks
Mine is under 5yrs but getting close.DeVilbiss Marketing wrote:This text was taken from the Meidcare Positive Airway Pressure
(PAP) Local Coverage Determination (LCD)
"If a PAP device is replaced during the 5 year reasonable useful lifetime (RUL) because of
loss, theft, or irreparable damage due to a specific incident, there is no requirement for a
new clinical evaluation, sleep test, or trial period.
If a PAP device is replaced following the 5 year RUL, there must be a face-to-face
evaluation by their treating physician that documents that the beneficiary continues to use
and benefit from the PAP device. There is no requirement for a new sleep test or trial
period."