S9 UPS solutions w/ home generator

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Cuda
Posts: 336
Joined: Sun Dec 19, 2010 12:42 pm

S9 UPS solutions w/ home generator

Post by Cuda » Wed Dec 29, 2010 3:29 pm

I am new to CPAP stuff only being on the S9 Autoset for a week. We have terrible power problems out here in the country. Lots of flickers and surges as well as times we lose power for days, once even a week. I have a LP fueled generator outside that restores power within 30 seconds. So for just about everything in the house I have a UPS. I am not concerned with how long the UPS can backup power as its never needed for more then a minute. The generator does not power the entire house but I can easily drill a hole to the basement and plug just the S9 into a circuit backed up.

My question is what type of UPS should I have on the S9, again it ONLY needs to power it until the generator kicks on. Would prefer something that does not beep when power is cut.

Thanks in advance.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pad-a-Cheek mask liner. CPAP mode 13cm, EPR: 1 Tube: 75 Humidity: 1.5

Hose_Head
Posts: 804
Joined: Wed Mar 12, 2008 6:43 pm
Location: Ontario Canada

Re: S9 UPS solutions w/ home generator

Post by Hose_Head » Wed Dec 29, 2010 5:50 pm

Cuda wrote:I am new to CPAP stuff only being on the S9 Autoset for a week. We have terrible power problems out here in the country. Lots of flickers and surges as well as times we lose power for days, once even a week. I have a LP fueled generator outside that restores power within 30 seconds. So for just about everything in the house I have a UPS. I am not concerned with how long the UPS can backup power as its never needed for more then a minute. The generator does not power the entire house but I can easily drill a hole to the basement and plug just the S9 into a circuit backed up.

My question is what type of UPS should I have on the S9, again it ONLY needs to power it until the generator kicks on. Would prefer something that does not beep when power is cut.

Thanks in advance.
My advice would be to give it a try for a while with just your cpap and the backup generator. My guess is that you'll find there's very little benefit to adding a UPS into the mix.

The power-out alarm on the UPS will wake you up, anyway. It's possible you'll sleep right through a 30 second interruption until your generator kicks in.
I'm workin' on it.

User avatar
Cuda
Posts: 336
Joined: Sun Dec 19, 2010 12:42 pm

Re: S9 UPS solutions w/ home generator

Post by Cuda » Wed Dec 29, 2010 6:03 pm

I guess I never tried cutting power to the S9, I just assumed that when the power was restored the system would be off waking me up to turn it back on.

Also, some nights the power goes out for a second or two a few hundred times an hour. I have to use UPS systems everywhere to simply level out the power more than provide long term backup.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pad-a-Cheek mask liner. CPAP mode 13cm, EPR: 1 Tube: 75 Humidity: 1.5

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: S9 UPS solutions w/ home generator

Post by avi123 » Wed Dec 29, 2010 8:45 pm

I would give the UPS a try.

Check this:

viewtopic/t57663/ResMed-S9-and-external ... upply.html

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

User avatar
rested gal
Posts: 12881
Joined: Thu Sep 09, 2004 10:14 pm
Location: Tennessee

Re: S9 UPS solutions w/ home generator

Post by rested gal » Thu Dec 30, 2010 2:05 am

Interesting question, Cuda.

I'm adding this topic to...

LINKS to Battery operation, camping, power outage
viewtopic.php?t=9682
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435

User avatar
RandyJ
Posts: 1673
Joined: Thu Apr 21, 2011 7:22 pm
Location: Connecticut, USA

Re: S9 UPS solutions w/ home generator

Post by RandyJ » Sun Nov 13, 2011 9:43 pm

Cuda wrote: Also, some nights the power goes out for a second or two a few hundred times an hour. I have to use UPS systems everywhere to simply level out the power more than provide long term backup.

I think this is interesting. Made me wonder if the power supply to the S9 would take care of fluctuations in voltage and protect the S9 itself? One day a few weeks ago a motorist struck a telephone pole down the street from where I live and my power was interrupted twice in the space of a few minutes, cutting and coming back on within 30 seconds each time. I thought of my S9 and went to unplug it in case the power cut again, but it made me wonder how sensitive it could be to such occurrences, if the S9 were to have to reboot several times in a few minutes.

I wasn't concerned for my a/v equipment as it is all plugged into UPSes but my S9 (at the moment) is plugged directly into a wall outlet.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Backup & Travel Machines: PR System One Bipap Auto, S9 VPAP Auto, S9 Autoset, Oximeter CMS-50E
Diagnosed March 2011, using APAP 14 - 16.5 cm, AFlex+ 2
Alt masks Swift FX pillows, Mirage FX nasal mask, Mirage Quattro full face mask

moresleep
Posts: 371
Joined: Sun Jul 24, 2011 12:14 am

Re: S9 UPS solutions w/ home generator

Post by moresleep » Sun Nov 13, 2011 10:15 pm

All equipment with electronic parts is susceptible to damage from surges and spikes--and those can be caused by lots of things... I doubt that the S9's power supply has much in the way of surge/spike suppression circuitry; but, that would be a good questions to email Resmed. Otherwise, if you are not going to run it off a UPS system of some kind, an investment in a $30 surge/spike suppressor power strip (make sure it has a light to indicate the circuitry is still working) would be wise.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: S9 UPS solutions w/ home generator

Post by archangle » Sun Nov 13, 2011 11:29 pm

Some gadgets get confused by short power outages and don't "reset" properly. If it really takes around 30 seconds for the generator to turn on, most gadgets will restart gracefully, but I haven't experimented with the S9 in particular.

Interestingly enough, on the PRS1 Auto, if the power goes out for a short time, it will start blowing immediately when power comes back if the power has only been off for a short time. (If the machine was on when power went off.) If power is off for a longer period of time, it will NOT start back up automatically. I forget how long the power has to be off before it doesn't restart, but I think it's something around 5 minutes. I think the PRS1 will restart if you have autostart turned on and breathe into it.

You might want to experiment with the S9 if it matters to you.

Note that S8 humidifiers may have problems with UPSs or even some generators. More and more generators have inverters to generate 120VAC and may have MSW waveforms. S9 humidifiers are MSW safe. I haven't found an official statement from Respironics on their machines.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
billbolton
Posts: 2264
Joined: Wed Jun 07, 2006 7:46 pm
Location: Sydney, Australia

Re: S9 UPS solutions w/ home generator

Post by billbolton » Sun Nov 13, 2011 11:41 pm

moresleep wrote:I doubt that the S9's power supply has much in the way of surge/spike suppression circuitry
I doubt that it doesn't have it..... Resmed has been building medical equipment for global use for a long time now.

Cheers,

Bill

_________________
MachineMask
Additional Comments: Airmini, Medistrom Pilot 24, CMS 60C Pulse Oximeter, ResScan 6

moresleep
Posts: 371
Joined: Sun Jul 24, 2011 12:14 am

Re: S9 UPS solutions w/ home generator

Post by moresleep » Mon Nov 14, 2011 11:17 pm

billbolton wrote:
moresleep wrote:I doubt that the S9's power supply has much in the way of surge/spike suppression circuitry
I doubt that it doesn't have it..... Resmed has been building medical equipment for global use for a long time now.

Cheers,

Bill
I'd be surprised if the power supply of any Cpap, of whatever price or manufacturer, has more than basic spike suppression circuitry. They are unlikely to spend the money or weight on it. And, unless you know for sure, why risk it? A power strip with with reasonable spike/surge suppression is not that expensive. Get one with indicators to tell you that the suppression circuitry is still working, as involved components tend to blow out... Pay at least mid-price from a name you recognize. Note that this is still a far cry from a real line conditioner.

User avatar
jamiswolf
Posts: 851
Joined: Wed Jul 20, 2011 4:08 pm

Re: S9 UPS solutions w/ home generator

Post by jamiswolf » Tue Nov 15, 2011 7:54 am

Hi Cuda,
With such unreliable power, I'd be seriously looking to operate the machine on battery power all the time. Then the frequent power outages would only be disrupting the charging circuitry which is less critical.

And depending on the humidity levels there...if you could get used to a passover level of humidification...or even add a second humidifier so that you didn't have to run a "heated" humidifier, you would reduce the current draw from the cpap considerably.

Your electrical reliability (or lack thereof) reminds me of when I lived in rural SE Iowa. Outages all the time.
Jamis

McSleepy
Posts: 485
Joined: Fri Nov 26, 2010 10:32 pm
Location: USA

Re: S9 UPS solutions w/ home generator

Post by McSleepy » Tue Nov 15, 2011 8:42 am

Given that you have already invested in a costly LP stand-by generator and a multitude of UPSs around the house, it should not be a stretch to spend another $100-$150 for a solution that will give you complete peace of mind. If you obtain a pure sine wave UPS, place it in the basement so the alarm doesn't wake you up (many UPSs are auto-resetting), and hook up both your CPAP and the humidifier to it, it will serve as a full-time power-line conditioner and will guarantee clean power to your sensitive equipment at all times - from short brown-outs to long outages. You might be able to get by without it, or with a lesser solution, but why take a chance and waste time experimenting?
McSleepy

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Previous machine: ResMed S9 VPAP Auto 25 BiLevel. Mask: Breeze with dilator pillows. Software: ResScan ver. 5.1
ResMed AirCurve 10 VAuto; Puritan-Bennett Breeze nasal pillow mask; healthy, active, middle-aged man; tall, athletic build; stomach sleeper; on CPAP since 2003; lives @ 5000 ft; surgically-corrected deviated septum and turbinates; regular nasal washes

moresleep
Posts: 371
Joined: Sun Jul 24, 2011 12:14 am

Re: S9 UPS solutions w/ home generator

Post by moresleep » Tue Nov 15, 2011 2:18 pm

Cuda wrote:I am new to CPAP stuff only being on the S9 Autoset for a week. We have terrible power problems out here in the country. Lots of flickers and surges as well as times we lose power for days, once even a week. I have a LP fueled generator outside that restores power within 30 seconds. So for just about everything in the house I have a UPS. I am not concerned with how long the UPS can backup power as its never needed for more then a minute. The generator does not power the entire house but I can easily drill a hole to the basement and plug just the S9 into a circuit backed up.

My question is what type of UPS should I have on the S9, again it ONLY needs to power it until the generator kicks on. Would prefer something that does not beep when power is cut.

Thanks in advance.
Does the Resmed 12-volt adapter power both the S9 Cpap and the humidifer? If so, I think I would get the 12 volt adapter, and power the machine and humidifier through that using an 80 AH or larger 12 volt AGP deep cycle battery, charged by a Battery Tender Plus. That way, your generator would keep the battery charged; but, even without the generator, you'd have several days of run time. Anyway, an alternative to consider...

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: S9 UPS solutions w/ home generator

Post by archangle » Thu Nov 17, 2011 9:39 pm

moresleep wrote:Does the Resmed 12-volt adapter power both the S9 Cpap and the humidifer?
Yes for the S9 series, I believe. I think the answer is "No" for S8. And it's a different adapter as well.

[edit 2/25/2013 - Change "No" for S9 to "No" for S8]

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Last edited by archangle on Mon Feb 25, 2013 7:23 pm, edited 1 time in total.
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: S9 UPS solutions w/ home generator

Post by archangle » Thu Nov 17, 2011 9:43 pm

billbolton wrote:
moresleep wrote:I doubt that the S9's power supply has much in the way of surge/spike suppression circuitry
I doubt that it doesn't have it..... Resmed has been building medical equipment for global use for a long time now.

Cheers,

Bill
Even if it's not heavily "surge proof," a lot of these external power supplies are sort of "sacrificial" protection. A surge may damage the power supply brick, but it usually only damages the brick. It's a lot cheaper and easier to replace just the brick than to disassemble and service the machine.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.