help in changing pressure devilbiss intellipap
help in changing pressure devilbiss intellipap
Hi can someone help me i found a site on how to turn up my pressure on devilbiss intellipap dv51au cpap it has been some time that i done it i have forgotten how to do it
It was something like turn on cpap and hold down a few keys but can not remember can someone let me know
Thanks
It was something like turn on cpap and hold down a few keys but can not remember can someone let me know
Thanks
- bearded_two
- Posts: 459
- Joined: Mon Aug 10, 2009 8:01 pm
Re: help in changing pressure devilbiss intellipap
Intellipap setup directions
1. Turn on the machine
2. Turn off the machine
3. Hold down the down arrow down and the ramp button and press the power button.
4. Scroll through the menu using the Select button and use the up and down arrows to change the values.
5. Turn the machine off when done setting.
1. Turn on the machine
2. Turn off the machine
3. Hold down the down arrow down and the ramp button and press the power button.
4. Scroll through the menu using the Select button and use the up and down arrows to change the values.
5. Turn the machine off when done setting.
Re: help in changing pressure devilbiss intellipap
Thanks so much! The A-hole at Devilbiss wouldn't tell me how to do it. She said she would 'get in trouble"!
- chunkyfrog
- Posts: 34420
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: help in changing pressure devilbiss intellipap
I have no doubt that she would get disciplined--or worse.
Manufacturers are in cahoots with the DME's and doctors to protect the power they enjoy (and misuse)
Manufacturers are in cahoots with the DME's and doctors to protect the power they enjoy (and misuse)
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: help in changing pressure devilbiss intellipap
CPAP/APAP/VPAP pressures are prescribed by physicians. These pressures are what are required for effective therapy. If the pressures are changed, the therapy will be ineffective and you might as well not use the PAP equipment. Also, if you are a truck or bus driver and you change pressures you will be fined by the DMV and probably lose your liscense. If a homecare company or others change pressures they will be prosecuted and/or sued.
Re: help in changing pressure devilbiss intellipap
your ignorance is showing.3DCNZ wrote:CPAP/APAP/VPAP pressures are prescribed by physicians. These pressures are what are required for effective therapy. If the pressures are changed, the therapy will be ineffective and you might as well not use the PAP equipment. Also, if you are a truck or bus driver and you change pressures you will be fined by the DMV and probably lose your liscense. If a homecare company or others change pressures they will be prosecuted and/or sued.
or, are you just parroting the bullshit they tell you at the DME you work at?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: help in changing pressure devilbiss intellipap
Can you provide some citations to this? Especially the part where the patient makes the changes.3DCNZ wrote:CPAP/APAP/VPAP pressures are prescribed by physicians. These pressures are what are required for effective therapy. If the pressures are changed, the therapy will be ineffective and you might as well not use the PAP equipment. Also, if you are a truck or bus driver and you change pressures you will be fined by the DMV and probably lose your liscense. If a homecare company or others change pressures they will be prosecuted and/or sued.
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin
Re: help in changing pressure devilbiss intellipap
I would like to see those citations also.Krelvin wrote:Can you provide some citations to this? Especially the part where the patient makes the changes.3DCNZ wrote:CPAP/APAP/VPAP pressures are prescribed by physicians. These pressures are what are required for effective therapy. If the pressures are changed, the therapy will be ineffective and you might as well not use the PAP equipment. Also, if you are a truck or bus driver and you change pressures you will be fined by the DMV and probably lose your liscense. If a homecare company or others change pressures they will be prosecuted and/or sued.
_________________
Mask: SleepWeaver Elan™ Soft Cloth Nasal CPAP Mask - Starter Kit |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Use SleepyHead |
Re: help in changing pressure devilbiss intellipap
one of the downsides of anonymous postings
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: help in changing pressure devilbiss intellipap
I just feel bad for that person's customers/patients who are getting that BS. How many lies do these folks spread? And everyone wonders why cpap compliance is so low.palerider wrote:one of the downsides of anonymous postings
_________________
Mask: SleepWeaver Elan™ Soft Cloth Nasal CPAP Mask - Starter Kit |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Use SleepyHead |
Re: help in changing pressure devilbiss intellipap
that we agree on.49er wrote:I just feel bad for that person's customers/patients who are getting that BS. How many lies do these folks spread? And everyone wonders why cpap compliance is so low.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: help in changing pressure devilbiss intellipap
You're full of crap.3DCNZ wrote:CPAP/APAP/VPAP pressures are prescribed by physicians. These pressures are what are required for effective therapy. If the pressures are changed, the therapy will be ineffective and you might as well not use the PAP equipment. Also, if you are a truck or bus driver and you change pressures you will be fined by the DMV and probably lose your liscense. If a homecare company or others change pressures they will be prosecuted and/or sued.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear |
Additional Comments: Oscar and Rescan |
Mask: bleep and F30i
Machine: Resmed S10 Autoset
Machine: Resmed S10 Autoset
- chunkyfrog
- Posts: 34420
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: help in changing pressure devilbiss intellipap
Those of us who were initially titrated at pressures several cm higher than effective and tolerable--
are looking for our high boots.
are looking for our high boots.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: help in changing pressure devilbiss intellipap
Not true..............you lying piece of equestrian excrement!!!3DCNZ wrote:CPAP/APAP/VPAP pressures are prescribed by physicians. These pressures are what are required for effective therapy. If the pressures are changed, the therapy will be ineffective and you might as well not use the PAP equipment. Also, if you are a truck or bus driver and you change pressures you will be fined by the DMV and probably lose your liscense. If a homecare company or others change pressures they will be prosecuted and/or sued.
Physicians often prescribe ineffective/inaccurate pressures based on one night in a sleep study. I know, it happened to me. These pages also have many, many examples of DME employees incorrectly configuring machines, too. An Auto-CPAP with a wide pressure range (often prescribed by physicians) can also be very ineffective and give horrible therapy. With proper guidance and the use of software with a data-capable PAP machine, the users can easily optimize their own therapy. After all, the patients OWN THEIR machines. It's absolutely NOT improper or illegal for them to change the pressures in THEIR machines. Yes, DMEs and RTs need an "order" from a physician to change the pressures......that's a well-known fact. But, it doesn't apply to the owners/users of the machines.
You have absolutely NO idea what you're talking about.
And, these forum pages are full of idiots like you coming here to demonstrate their stupidity.
Den (adding yet another "pile-on" comment)
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: help in changing pressure devilbiss intellipap
I had a sleep test. Found out the correct pressure to use and set it in a minute.