New info on Airsense/Aircurve 11 problems
- vandownbytheriver
- Posts: 403
- Joined: Sat Feb 03, 2024 11:42 pm
New info on Airsense/Aircurve 11 problems
News from Uncle Nick on the 11 problem... check your tube plug carefully!
Here's the first video describing complaints:
https://www.youtube.com/watch?v=WiMPOaZBqVs
Here's the second video where the problems have been discovered:
https://www.youtube.com/watch?v=ZAs01WRPtGk
There are two points to the new information:
a) The Climateline for the AS11 can slip out of being fully plugged in... a slight pull to the side and the clip catch jumps out... the tube is now leaking. This can actually cause the machine to turn itself off if the leak is large enough and you have set Auto Start/Stop.
b) There is a firmware update coming... turns out that high leaking can cause the humidity system (when in Auto) to fry itself. If it's already fried then you're SOL, needs a hardware repair. People that allow the machine to blow before putting on the mask (I assume minutes not seconds) can cause this problem. Large leaks with humidity in Auto mode can cause this problem. The solution is to use Manual humidity mode until the firmware update can go out.
I don't have an AS11, those that have one can check their hose connection with a wiggle.
Here's the first video describing complaints:
https://www.youtube.com/watch?v=WiMPOaZBqVs
Here's the second video where the problems have been discovered:
https://www.youtube.com/watch?v=ZAs01WRPtGk
There are two points to the new information:
a) The Climateline for the AS11 can slip out of being fully plugged in... a slight pull to the side and the clip catch jumps out... the tube is now leaking. This can actually cause the machine to turn itself off if the leak is large enough and you have set Auto Start/Stop.
b) There is a firmware update coming... turns out that high leaking can cause the humidity system (when in Auto) to fry itself. If it's already fried then you're SOL, needs a hardware repair. People that allow the machine to blow before putting on the mask (I assume minutes not seconds) can cause this problem. Large leaks with humidity in Auto mode can cause this problem. The solution is to use Manual humidity mode until the firmware update can go out.
I don't have an AS11, those that have one can check their hose connection with a wiggle.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: I use O2Ring, Oscar, SleepHQ, and Cover Roll Stretch mouth tape. |
Re: New info on Airsense/Aircurve 11 problems
——-vandownbytheriver wrote: ↑Wed Apr 03, 2024 11:09 pmNews from Uncle Nick on the 11 problem... check your tube plug carefully!
Here's the first video describing complaints:
https://www.youtube.com/watch?v=WiMPOaZBqVs
Here's the second video where the problems have been discovered:
https://www.youtube.com/watch?v=ZAs01WRPtGk
There are two points to the new information:
a) The Climateline for the AS11 can slip out of being fully plugged in... a slight pull to the side and the clip catch jumps out... the tube is now leaking. This can actually cause the machine to turn itself off if the leak is large enough and you have set Auto Start/Stop.
b) There is a firmware update coming... turns out that high leaking can cause the humidity system (when in Auto) to fry itself. If it's already fried then you're SOL, needs a hardware repair. People that allow the machine to blow before putting on the mask (I assume minutes not seconds) can cause this problem. Large leaks with humidity in Auto mode can cause this problem. The solution is to use Manual humidity mode until the firmware update can go out.
I don't have an AS11, those that have one can check their hose connection with a wiggle.
Unless I waaaayyyy over stretch while in bed, my AS 11 is stationary on my nightstand and there’s no need to worry about my hose slipping out. And on the matter of another Nicko issue, because it’s stationary, there’s little chance it’ll roll over because of design issues. I have had to question what I’d need to be doing for the hose to pull loose or the machine to roll over anyway.
![Question :?:](./images/smilies/icon_question.gif)
Last edited by Ebonyeyez on Mon Apr 29, 2024 12:00 pm, edited 1 time in total.
- vandownbytheriver
- Posts: 403
- Joined: Sat Feb 03, 2024 11:42 pm
Re: New info on Airsense/Aircurve 11 problems
Does your hose show the wiggle problem? Do you run your heater in Manual? Did you run an Ax10 before?
Glad you're problem-free, there are many others with and without issues. In particular, the water tub metal design appears improved over the 10... I'm working on that issue right now, going to repair a leaky 10 tub and see how that works.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: I use O2Ring, Oscar, SleepHQ, and Cover Roll Stretch mouth tape. |
Re: New info on Airsense/Aircurve 11 problems
I live in Florida and because of that I don't have as much of a need for humidification because I don't wake up with dry mouth but I do wake up with a dry jaw every once in awhile if I'm sleeping on my side. I don't know if I need a heated tube for the same reason but I should have one in case I go up north and the same is true for unification although it might help the joy issue. But I wonder if Medicare pays for the heated tube which it probably does and I know it pays for humidification but if you want the heated tube it probably has to be a doctors request And then Medicaid will pay for it.
However, what if you live in Florida and the doctor says you should have battery backup, with Medicare pay for that? Or you might need it for traveling.
I'm still waiting for the ResMed air curve 11.
Alan
However, what if you live in Florida and the doctor says you should have battery backup, with Medicare pay for that? Or you might need it for traveling.
I'm still waiting for the ResMed air curve 11.
Alan
Re: New info on Airsense/Aircurve 11 problems
You could have had an AC 10 by now.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Happiness is being on Dog Slobber pr & zonkers foe list
CrankyGranny is Whale Road + many other ids
They are here to help.
CrankyGranny is Whale Road + many other ids
They are here to help.
zonkers + palerider aka GrumpyHere wrote: ↑What exactly do you think you're adding to this thread?
Re: New info on Airsense/Aircurve 11 problems
Yes, I know, I could have gotten one 5 years ago when the 10 was already a 5-year-old machine because I got my S9 few weeks before it was discontinued.
ResMed was putting out new machines every 5 years, and I should have thought of that before I took the S9. If you assume that relatively shortly you'll get a new generation of machine, at a time when the S10 was already 5 years old, why not wait a few months which became a few years and the longer you wait the more ludicrous it is after waiting 10 years to go back to the 10.
A 10 was delivered to me and I specifically said at least a 20 times that I want the AC 11 and the salesperson said yes no problem at least
10 times and then when the delivery guy came and I made him wait until I opened it up sure enough it was the 10 so I didn't accept it. If I could get the 10 until the 11s come out I would do it but I don't want to be stuck with a 10 for five more years. Already we know that the 11 is much better energy wise and that could make a difference after a hurricane in Florida. We already know that the 11 can do on air firmware updates instead of having to send the machine back to ResMed and waiting weeks.
I don't understand after all these years how they couldn't get the bugs out of these machines and start producing them in high numbers. Look at how much money they are losing as they lost the whole 5-year cycle of machines but they were making ventilators so, they didn't care.
But what about the questions I asked?
Will Medicare pay for the heated hose and a battery backup if the doctor puts it in the prescription?
Re: New info on Airsense/Aircurve 11 problems
Medicare (and other insurance) WILL pay for the heated hose. It is considered a covered expense and there is an appropriate HCPCS billing code for it.
HCPCS billing code A4604 and allowed 1 per 3 months.
That would be the top one on the list "tube with integrated heating element".
Medicare (and other insurance) normally do NOT consider a battery backup as a covered expense and those won't pay for it though someone might get it covered under some special coding or exception thing.
Below is the replacement allowance schedule for Medicare and most other insurance for cpap stuff.
If you notice there is NO billing code for batteries of any sort.
Also please note that travel machines (like the tiny machines) aren't covered either.
![Image](http://i.imgur.com/jMWaqNBl.jpg)
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: New info on Airsense/Aircurve 11 problems
Call your bipap dealer to ask them. Really
Same call your bipap dealer to ask them. Really it does work
Me Im waiting for the 12 Really
You Can't Fix Stupid Really
- vandownbytheriver
- Posts: 403
- Joined: Sat Feb 03, 2024 11:42 pm
Re: New info on Airsense/Aircurve 11 problems
The S9 was a great machine. I still have mine. Many thousands of hours on it.
The 10 is the best machine RM ever made. I own five of them. The firmware is unlocked, you can make these things do whatever you want. Replacement motors are 130$ on Amazon. I have a machine I rebuilt that has 22,500 hours on it... ready to go another 22,000.ResMed was putting out new machines every 5 years, and I should have thought of that before I took the S9. If you assume that relatively shortly you'll get a new generation of machine, at a time when the S10 was already 5 years old, why not wait a few months which became a few years and the longer you wait the more ludicrous it is after waiting 10 years to go back to the 10.
Do some research, the 11 is a dog, hose connection and heating circuit problems, tiny motor, too many problems to list. The 10 is (was!) being blown out at ludicrous prices, I bought a C2C Autoset during Black Friday for 399USD, I bought an Elite (*with* 4G modem) for 299USD! Those days are gone, starting to see the prices for the 10 and 11 equalize... I recommend everyone get a new 10 while they still can. If you don't already have a generator you're not prepared for a hurricane... and if you have a generator (I have two, we freeze over often here in TX) you don't care about 30 extra watts consumption.
A 10 was delivered to me and I specifically said at least a 20 times that I want the AC 11 and the salesperson said yes no problem at least
10 times and then when the delivery guy came and I made him wait until I opened it up sure enough it was the 10 so I didn't accept it. If I could get the 10 until the 11s come out I would do it but I don't want to be stuck with a 10 for five more years. Already we know that the 11 is much better energy wise and that could make a difference after a hurricane in Florida. We already know that the 11 can do on air firmware updates instead of having to send the machine back to ResMed and waiting weeks.
The 10 is the most popular machine on the planet IMO. It is the open-source platform of the future for CPAP. You can tell I'm giddy over them, right?
I don't understand after all these years how they couldn't get the bugs out of these machines and start producing them in high numbers. Look at how much money they are losing as they lost the whole 5-year cycle of machines but they were making ventilators so, they didn't care.
Heated hose yes, battery no. One is clinical, one situational.
But what about the questions I asked?
Will Medicare pay for the heated hose and a battery backup if the doctor puts it in the prescription?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: I use O2Ring, Oscar, SleepHQ, and Cover Roll Stretch mouth tape. |