I'm a little confused over leakage rates. My Australian made Resmed Autoscore which seems to be similar to the US Resmed Elite II, displays mask leakage in L/s (the machine is set for an Activa mask) and the leakage is usually somewhere between 0.2 and 0.48 L/s.
But I've noticed in this forum that leakage rates are usually given in L/m and some report that their leakage is 0.0 L/m.
If you convert my figures to L/m, my leakage is abysmal by comparison and yet I am rarely aware of any leaks from the Activa compared to the Quattro which leaked like a sieve around the eyes.
I'm unsure as to how to solve my quandary but a good starting point would be to ask if US Elite II machines really report in L/m.
Mask leakage
Mask leakage
Machine: Resmed Platinum II Autoscore (probably equivalent to Elite II)
Humidifier: H4i
Mask: Activa LT mask
Worrying about dying doesn't stop you from dying; it merely stops you from living!
Humidifier: H4i
Mask: Activa LT mask
Worrying about dying doesn't stop you from dying; it merely stops you from living!
Re: Mask leakage
My understanding is that the Resmed machines report the excess 95 % ile leak rate in l/sec based on the mask in the menu.
Thus a 0.2 leak rate on the screen corresponds to a 12 l/min leak rate in excess of what would be expected. For example if the mask was a Swift II and the pressure was 12, then the expected leak would be 37 so a reading of .2 would correspond to a 95 %ile leak rate of 49 l/min.
Pressure
(cm H2O)
Flow
(L/min)
4 20
8 29
12 37
16 43
20 49
from page 11 of :
http://www.resmed.com/us/assets/documen ... er_eng.pdf
Thus a 0.2 leak rate on the screen corresponds to a 12 l/min leak rate in excess of what would be expected. For example if the mask was a Swift II and the pressure was 12, then the expected leak would be 37 so a reading of .2 would correspond to a 95 %ile leak rate of 49 l/min.
Pressure
(cm H2O)
Flow
(L/min)
4 20
8 29
12 37
16 43
20 49
from page 11 of :
http://www.resmed.com/us/assets/documen ... er_eng.pdf
Re: Mask leakage
Many thanks for the speedy reply even though it makes my true leakage look like a disaster (and I'd merely thought they were abysmal). I'll go off and look at the linked document now but again, thanks.My understanding is that the Resmed machines report the excess 95 % ile leak rate in l/sec based on the mask in the menu.
Thus a 0.2 leak rate on the screen corresponds to a 12 l/min leak rate in excess of what would be expected. For example if the mask was a Swift II and the pressure was 12, then the expected leak would be 37 so a reading of .2 would correspond to a 95 %ile leak rate of 49 l/min.
Machine: Resmed Platinum II Autoscore (probably equivalent to Elite II)
Humidifier: H4i
Mask: Activa LT mask
Worrying about dying doesn't stop you from dying; it merely stops you from living!
Humidifier: H4i
Mask: Activa LT mask
Worrying about dying doesn't stop you from dying; it merely stops you from living!
- Sleepy Boy
- Posts: 241
- Joined: Thu Nov 06, 2008 5:55 am
- Location: South Branch Michigan
Re: Mask leakage
Hi Capers: The mask I have is a Quattro. There is a chart which shows Leaks Per Minute. It shows what pressure your set at-and a graph with a curve line-anything below that # is not considered a leak by that mask Co. My pressure is set at 13 and on the graph and the leak rate is 43, I checked my leak rate on my machine and it was 31 So as best I know--I'm below the 43--so that should be good. All masks have this, just depends on what kind of mask you have.It's in Leaks per minute.
I think I got this right, Wufman (Den) helped me with this last week. Thanks, good luck, Larry
I think I got this right, Wufman (Den) helped me with this last week. Thanks, good luck, Larry
Sleepy Boy
-
B.Antoine
Re: Mask leakage
Well, that is my first check out to your phorum! We are a group of volunteers and starting a brand new initiative in a regional community in the exact same niche. Your phorum supplied us valuable information to work on. You have done a marvellous task!