Need leak rates for Mirage Quattro FF mask

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
MoneyGal
Posts: 359
Joined: Fri Jan 30, 2009 6:14 pm

Need leak rates for Mirage Quattro FF mask

Post by MoneyGal » Tue Mar 17, 2009 5:21 am

I cannot claim to be addled from lack of sleep any longer, but I don't seem to have the leak rates for my Mirage Quattro mask. I think the rates are available - some of you seem to have them - but I didn't get any documentation with my mask that gives me leak rates.

I assume I am *not* having any leak issues because I am consistently getting AIs of fewer than 2/hr (and HIs of between 2-3). However, I would like to know what the acceptable leak rates are for my mask.

Can anyone help? It feels like this is the last little thing I need to "understand" about my equipment.

luke
Posts: 89
Joined: Wed Oct 08, 2008 12:36 pm
Location: Delaware

Re: Need leak rates for Mirage Quattro FF mask

Post by luke » Tue Mar 17, 2009 5:50 am

You can download the Quattro's user's guide from the Resmed site but here they are:

At pressure of 4 its 22 LPM
at 8 its 32
at 12 its 41
at 16 its 48
at 20 its 54
You can interpolate for in between pressures and also they say the tolerance on those values is +/- 6 LPM
due to manufacturing variations

My pressure is 10 and I found out that by doing the teflon fix I lowered the leak rate by about 4 LPM from 37 to 33



Bill

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Running 11 pressure straight Cpap Flex 1

User avatar
carbonman
Posts: 2523
Joined: Wed Jun 25, 2008 7:57 am

Re: Need leak rates for Mirage Quattro FF mask

Post by carbonman » Tue Mar 17, 2009 5:55 am

MoneyGal wrote:but I don't seem to have the leak rates for my Mirage Quattro mask.
Can anyone help? .
Image

Find the user guide here:

http://www.resmed.com/en-us/products/se ... _multi.pdf
"If your therapy is improving your health but you're not doing anything
to see or feel those changes, you'll never know what you're capable of."
I said that.

User avatar
MoneyGal
Posts: 359
Joined: Fri Jan 30, 2009 6:14 pm

Re: Need leak rates for Mirage Quattro FF mask

Post by MoneyGal » Tue Mar 17, 2009 5:58 am

Thanks to both of you!

Luke - I need to go check my results - I am sure I am well below those rates for my pressure (average of between 9 - 10).

Carbonman - I *do* have that chart...but didn't understand "leak rate" as = to "vent flow rate."

But what does "leak rate" actually mean? What is being measured?

User avatar
LinkC
Posts: 3154
Joined: Wed Oct 29, 2008 1:06 pm
Location: Amelia Island, FL

Re: Need leak rates for Mirage Quattro FF mask

Post by LinkC » Tue Mar 17, 2009 6:16 am

The "leak rate" referred to here is the flow that goes out the vent. It's a normal function of the mask. It's actually NOT a "leak", the term "vent flow" on the chart is more accurate. Any "leak" is airflow above that, i.e. uninterntional flow out of the seals.

_________________
MachineMaskHumidifier
Additional Comments: 11-14 cmH2O
The OSA patient died quietly in his sleep.
Unlike his passengers who died screaming as the car went over the cliff...

luke
Posts: 89
Joined: Wed Oct 08, 2008 12:36 pm
Location: Delaware

Re: Need leak rates for Mirage Quattro FF mask

Post by luke » Tue Mar 17, 2009 6:20 am

Leak rate is the intended "designed" leak for the mask at a certain pressure to assure no breathing in co2.
I think vent flow rate is just different terminology for the same thing.
Actually, I spend 66 % of the night at 9 the other 34 % at 10 and if I get anything under let's say 39 for leak its good.



Bill

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Running 11 pressure straight Cpap Flex 1

User avatar
MoneyGal
Posts: 359
Joined: Fri Jan 30, 2009 6:14 pm

Re: Need leak rates for Mirage Quattro FF mask

Post by MoneyGal » Tue Mar 17, 2009 6:27 am

Ok, wait a minute. I'm still confused.

What am I measuring? If "vent flow rate" = rate to assume no rebreathing of C02, then...what is a "leak rate"? My machine gives me a "leak rate." What is it?

User avatar
LinkC
Posts: 3154
Joined: Wed Oct 29, 2008 1:06 pm
Location: Amelia Island, FL

Re: Need leak rates for Mirage Quattro FF mask

Post by LinkC » Tue Mar 17, 2009 7:06 am

OK, your mask vents CO2 by flushing the mask interior with air during exhale. It all goes out the vent (usually blowing on your sleeping partner's neck... ). The machine sees that as a constant "leak", even tho it's intentional. But it's a "good" leak.

Any unintentional leaks (like the ones that blow constantly in your eyes, or the intermittant "mask farts") are on top of the vent flow. Those are "bad" leaks.

The total "Leak Rate" is both kinds added together.

Unless your machine has you enter the vent rate for the particular mask you use, it's reporting the total and you should only worry about the amount OVER your mask's vent rate. (Sorry, I don't know if Resmeds do the math for you or not.)

_________________
MachineMaskHumidifier
Additional Comments: 11-14 cmH2O
Last edited by LinkC on Tue Mar 17, 2009 12:35 pm, edited 1 time in total.
The OSA patient died quietly in his sleep.
Unlike his passengers who died screaming as the car went over the cliff...

User avatar
MoneyGal
Posts: 359
Joined: Fri Jan 30, 2009 6:14 pm

Re: Need leak rates for Mirage Quattro FF mask

Post by MoneyGal » Tue Mar 17, 2009 7:28 am

OK, just checked my machine. It is giving me leak rates per second. Last night, my rate was 0.24 L/s. For the week, my rate was 0.10 L/s.

How do I convert those to litres per minute? (I know, dumb question, right?) I assume I multiply by 60 (for the 60 seconds in a minute).

So, if I multiply my weeky rate of 0.1 L/s by 60, I get a leak rate of 6. Which is pretty good...even excellent, right?

SharkBait
Posts: 593
Joined: Sun Jan 25, 2009 5:41 pm
Location: Texas -- the ugly part... El Paso? No, not quite THAT ugly...

Re: Need leak rates for Mirage Quattro FF mask

Post by SharkBait » Tue Mar 17, 2009 7:33 am

MoneyGal wrote:OK, just checked my machine. It is giving me leak rates per second. Last night, my rate was 0.24 L/s. For the week, my rate was 0.10 L/s.

How do I convert those to litres per minute? (I know, dumb question, right?) I assume I multiply by 60 (for the 60 seconds in a minute).

So, if I multiply my weeky rate of 0.1 L/s by 60, I get a leak rate of 6. Which is pretty good...even excellent, right?
Yes, you are doing it correctly.

0 is excellent.

6 is certainly good and even last night's 14.4 is manageable. Your machine will compensate the pressure for the leaks up to a point. I'm thinking I read that up to around 20 lpm your machine can maintain your correct pressure. So as long as your leaks aren't bugging the crap out of you or burning your eyes, you're good at 6 lpm,..
Encore Pro 1.8.49; Encore Pro Analyzer 0.8.9 by James Skinner
SnuggleHose - Got the 8 foot and cut it down to 6, used the rest for mask hoses.
Memory Foam Pillow - Cut my own out of my Tempur-pedic pillow. (works great!)
Hose Mgmt - Velcro Tie Strap

User avatar
MoneyGal
Posts: 359
Joined: Fri Jan 30, 2009 6:14 pm

Re: Need leak rates for Mirage Quattro FF mask

Post by MoneyGal » Tue Mar 17, 2009 7:40 am

Hey, this is all awesome. Nothing is bothering me. I feel like I won the jackpot - I put the mask on, turn on the machine, and go to sleep. About 7.5 hours later I wake up, and get on with my day.

User avatar
spitintheocean
Posts: 180
Joined: Thu Feb 26, 2009 10:47 am
Location: Ottawa, Canada

Re: Need leak rates for Mirage Quattro FF mask

Post by spitintheocean » Tue Mar 17, 2009 7:58 am

Since you are using the Resmed Elite II, your machine automatically compensates for the flowrate of that Resmed mask; just ensure that the settings reflect the correct mask in setup and you won't have to fumble with the charts and math.

Paul

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: Settings: 17 - 19 - no EPR; CMS 50F Pulse Oximeter
Life is something to do when you can't get to sleep.
Fran Lebowitz

User avatar
Paul56
Posts: 1016
Joined: Thu Sep 25, 2008 1:38 pm
Location: Texas

Re: Need leak rates for Mirage Quattro FF mask

Post by Paul56 » Tue Mar 17, 2009 8:00 am

MoneyGal wrote:OK, just checked my machine. It is giving me leak rates per second. Last night, my rate was 0.24 L/s. For the week, my rate was 0.10 L/s.

How do I convert those to litres per minute? (I know, dumb question, right?) I assume I multiply by 60 (for the 60 seconds in a minute).

So, if I multiply my weeky rate of 0.1 L/s by 60, I get a leak rate of 6. Which is pretty good...even excellent, right?
0.24 L/s is within acceptable range as Resmed consider anything at or under 0.34 L/s to be within acceptable limits; however, there can be spikes in your leak rate during the night which you can only view with ResScan.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: AHI ~60 / Titrated @ 8 / Operating AutoSet in CPAP mode @ 12

User avatar
MoneyGal
Posts: 359
Joined: Fri Jan 30, 2009 6:14 pm

Re: Need leak rates for Mirage Quattro FF mask

Post by MoneyGal » Tue Mar 17, 2009 8:15 am

Well, last night I mistakenly turned the machine on before the mask was on, so I had 15s or so of "high leak." The weekly rate is probably more indicative of my "normal" leak rate.