Hey, I'm posting this on behalf of another forum member. It occurs to me if she had the clinician's menu for the machine she has (Resmed 6 lightweight), life might be slightly better for her.
viewtopic/t37339/Doctor-frustrations.html
Can anyone help?
Cheers,
Barbara
Need clinician's menu for Resmed 6 lightweight
Need clinician's menu for Resmed 6 lightweight
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Additional Comments: Started XPAP 04/20/07. APAP currently wide open 10-20. Consistent AHI 2.1. No flex. HH 3. Deluxe Chinstrap. |
I currently have a stash of Nasal Aire II cannulas in Small or Extra Small. Please PM me if you would like them. I'm interested in bartering for something strange and wonderful that I don't currently own. Or a Large size NAII cannula.
Re: Need clinician's menu for Resmed 6 lightweight
See if this works.....
Den
ResMed cpap S6
1. machine must be shut off
2. press START and TIMER 20 MINUTES buttons at the same time
3. switch on machine using master switch at the backside of the machine
4. using TIMER 5 MINUTES and TIMER 10 MINUTES buttons increase or decrease pressure (pressure setting is shown on display at reverse side of machine)
5. complete setting by pushing STOP button once
6. now the machine is set to the new pressure and will remember that when being shut off on switched on again
And, the manual may also be on Doug's website to download, too.
http://www.internetage.ws/cpapdata/manuals/
Den
ResMed cpap S6
1. machine must be shut off
2. press START and TIMER 20 MINUTES buttons at the same time
3. switch on machine using master switch at the backside of the machine
4. using TIMER 5 MINUTES and TIMER 10 MINUTES buttons increase or decrease pressure (pressure setting is shown on display at reverse side of machine)
5. complete setting by pushing STOP button once
6. now the machine is set to the new pressure and will remember that when being shut off on switched on again
And, the manual may also be on Doug's website to download, too.
http://www.internetage.ws/cpapdata/manuals/
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: Need clinician's menu for Resmed 6 lightweight
Dude!!! You continually surprise me. I had no idea you collected Resmed info too. I'm just gonna go straight to you with all questions from now on.
Thanks buddy!!!!
B.
Thanks buddy!!!!
B.
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Additional Comments: Started XPAP 04/20/07. APAP currently wide open 10-20. Consistent AHI 2.1. No flex. HH 3. Deluxe Chinstrap. |
I currently have a stash of Nasal Aire II cannulas in Small or Extra Small. Please PM me if you would like them. I'm interested in bartering for something strange and wonderful that I don't currently own. Or a Large size NAII cannula.