Almost Drowned in my Sleep!

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
chaswalters
Posts: 17
Joined: Tue Apr 20, 2021 11:48 am

Almost Drowned in my Sleep!

Post by chaswalters » Mon May 03, 2021 7:28 am

Ok, maybe I am being dramatic. After a full week on this machine I decided to give the whole thing a proper cleaning. I is a Resmed Air 10 Auto, with a F-20 full mask. I hung the hose for about 10 hours before using the machine. Everything was fine until about 2:30 am. The piece where the mask and hose attach became very noisy and it was spitting water at me. I could feel it hitting my face. I actually didn't notice. My wife woke me up from the noise of the swivel piece on the mask. My humidity level is set to 7, I have a heated hose. The 3 bif leaks are where I disconnected the hose from the mask to blow out water. After that it would stop doing it.
Here is a pic of mask and Oscar for the night and a pic of my mask this morning.
180069834_145655114188347_6841798157685294787_n.jpg
180069834_145655114188347_6841798157685294787_n.jpg (363.36 KiB) Viewed 1202 times
may2.png
may2.png (87 KiB) Viewed 1202 times
Other than the cleaning, everything was the same as it had been for the first week.
Last edited by chaswalters on Mon May 03, 2021 7:41 am, edited 1 time in total.

User avatar
Pugsy
Posts: 65130
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Almost Drowned in my Sleep!

Post by Pugsy » Mon May 03, 2021 7:33 am

chaswalters wrote:
Mon May 03, 2021 7:28 am
I have a heated hose.
What temperature is it set to?

What temperature is your ambient household temp?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

chaswalters
Posts: 17
Joined: Tue Apr 20, 2021 11:48 am

Re: Almost Drowned in my Sleep!

Post by chaswalters » Mon May 03, 2021 7:40 am

Pugsy wrote:
Mon May 03, 2021 7:33 am
chaswalters wrote:
Mon May 03, 2021 7:28 am
I have a heated hose.
What temperature is it set to?

What temperature is your ambient household temp?
House temp is about 68, hose temp is 81.

User avatar
Julie
Posts: 20052
Joined: Tue Feb 28, 2006 12:58 pm

Re: Almost Drowned in my Sleep!

Post by Julie » Mon May 03, 2021 7:45 am

Is it possible the hose wasn't properly attached... crooked at the screw place if you know what I mean :) , or is your machine located above your head - water can't run back to it that way.

User avatar
Pugsy
Posts: 65130
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Almost Drowned in my Sleep!

Post by Pugsy » Mon May 03, 2021 7:48 am

Increase your hose air temp to prevent condensation/rain out in the mask or hose.

It will go up to 86 degrees. You may or may not need all of that available temp.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

chaswalters
Posts: 17
Joined: Tue Apr 20, 2021 11:48 am

Re: Almost Drowned in my Sleep!

Post by chaswalters » Mon May 03, 2021 8:07 am

Julie wrote:
Mon May 03, 2021 7:45 am
Is it possible the hose wasn't properly attached... crooked at the screw place if you know what I mean :) , or is your machine located above your head - water can't run back to it that way.
The machine is on the hightstand level with my head. The hose drops down in a U shape, down and up to mask.

chaswalters
Posts: 17
Joined: Tue Apr 20, 2021 11:48 am

Re: Almost Drowned in my Sleep!

Post by chaswalters » Mon May 03, 2021 8:09 am

Pugsy wrote:
Mon May 03, 2021 7:48 am
Increase your hose air temp to prevent condensation/rain out in the mask or hose.

It will go up to 86 degrees. You may or may not need all of that available temp.
I will try that. Thank you!

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Almost Drowned in my Sleep!

Post by chunkyfrog » Mon May 03, 2021 8:30 am

What is the relative humidity in your house?
Why is your bedroom temperature at 68?

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

chaswalters
Posts: 17
Joined: Tue Apr 20, 2021 11:48 am

Re: Almost Drowned in my Sleep!

Post by chaswalters » Mon May 03, 2021 8:44 am

chunkyfrog wrote:
Mon May 03, 2021 8:30 am
What is the relative humidity in your house?
Why is your bedroom temperature at 68?
Not sure of the humidity.
Temp is due to menopausal wife. It can get ugly if it gets warm in there!

User avatar
Pugsy
Posts: 65130
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Almost Drowned in my Sleep!

Post by Pugsy » Mon May 03, 2021 9:05 am

chaswalters wrote:
Mon May 03, 2021 8:44 am
chunkyfrog wrote:
Mon May 03, 2021 8:30 am
What is the relative humidity in your house?
Why is your bedroom temperature at 68?
Not sure of the humidity.
Temp is due to menopausal wife. It can get ugly if it gets warm in there!
Happy wife means happy life. Keep her happy.
Usually all is needed is a bit more warm air in the hose to reduce the condensation. If not there are other options available but I always suggest increasing the hose air temp first because it's the easiest thing to do and works well most of the time. In the winter my bedroom runs around 55 to 60 degrees. Prime recipe for rain out but I manage it with using 86 degrees for hose air temp.
Now in the summer I don't need so much warmth because the house temp is higher.
The condensation in the mask or hose is just Physics in action. Warm air will hold more moisture than cold air.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Almost Drowned in my Sleep!

Post by chunkyfrog » Mon May 03, 2021 10:11 am

Try a hose cozy.
Pad a cheek sells a nice one, or cpap.com, forum sponsors.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

clownbell
Posts: 369
Joined: Fri Apr 02, 2021 8:46 pm

Re: Almost Drowned in my Sleep!

Post by clownbell » Mon May 03, 2021 10:45 am

When moisture in hose condenses, it will run back into humidifier tank IF machine is lower than your head and no U kinks in hose. Consider lowering machine if increasing hose temp does not do the job. Also you can buy a hose cover to insulate the hose from the colder room temp. Amazon or CPAP supplier online. FWIW, I like a cooler room temp too.

https://amazon.com/Resmed-33963-Tubing- ... 343&sr=8-9

I am not affiliated with Amazon in any way. Just a retired guy trying to help.
ResMed AirSense 10 Autoset with built in humidifier
Resmed P10 pillow mask

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Almost Drowned in my Sleep!

Post by Okie bipap » Mon May 03, 2021 11:36 am

We have always kept our bedroom cool at night. My wife grew up in a house using wood heat, so the house would get rather cold at night and she learned to sleep in a cold room. We route our hoses up over the head of the bed and use hose covers from Padacheek, and never have rain out problems. In the winter time, I set my hose temp at 80 degrees and 70 this time of year. The only time I ever had rain out was shortly after beginning treatment and we were on a trip. The hose dipped down quite a bit from the bedside table to the bed. Now, when we travel I carry two Cozy Hoze Boss, one for each of us since my wife now also uses a machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

Janknitz
Posts: 8512
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Almost Drowned in my Sleep!

Post by Janknitz » Mon May 03, 2021 1:57 pm

Wow, we turn our heat down to about 50 at night during the winter and I never use heat in my humidifier. The only time I EVER had moisture in my mask and hose is when I cranked up the heat in the humidifier because I had a nasty head cold. (well, when I used nasal pillows I had some annoying condensation right at the tips because my nose was warmer than the air).

So my advice would be to turn the temperature on the humidifier DOWN to reduce the difference between the room temperature and the hose/mask temperature.

Am I confused?
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
Pugsy
Posts: 65130
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Almost Drowned in my Sleep!

Post by Pugsy » Mon May 03, 2021 5:27 pm

Janknitz wrote:
Mon May 03, 2021 1:57 pm
So my advice would be to turn the temperature on the humidifier DOWN to reduce the difference between the room temperature and the hose/mask temperature.

Am I confused?
Turning down the heater setting to the humidifier is also an option but it also reduces the amount of humidity being delivered to the nasal mucosa and some people find that is a problem. If/when the nasal mucosa gets dried out more than it likes then the person can end up with some rather nasty and annoying symptoms. Reducing the humidity being delivered to the nasal mucosa isn't always a good option for some people.

When we turn down the heat to the humidifier plate it reduces the amount of moisture being added to the air going to the mask.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.