Not sure the sleep doctor has this one right!
Not sure the sleep doctor has this one right!
I just took a new home sleep test and it seems to be equally as good as the one I took in a sleep lab in 2008 including sleep EEG for sleep staging, sleep onset and arousal data.Their summary:
The patient had 137 apneas. There were 22 obstructive apneas, no mixes apneas, 12 central apneas, and 103 hypopneas for an AHI of 15.2/hour. Supine 309.2 min (38.8% of sleep 19.5 AHI. Right 327.7 min 60.7% of sleep 12.3 AHI.
Sleep Efficiency. 87%. AHI 15.2/hr. ODI 13.6l/hr. Avg SpO2 93.1%. Highest SpO2 98%. Lowest SpO2 78%. Max HR 95 bpm. Min HR 51bpm. N1&N2 62.8%. N3 5.1%. REM sleep 32.1%. Loud snoring all three nights. Time in bed 616.7 min. 539.6 min test.
Significant Oxygen desaturation observed with a total of 13.8 minutes below 89%. SPO2 longest span <89% was 1.0 minute. Sleep stages of N1, N1, N3, and Rem Sleep observed per EEG lead signals. RERAs, and Clear Air Hypopneas and Clear Air Apneas present.
Dx Severe Obstructive Sleep Apnea and Hypersomnia with OSA.
The Doctor prescribed a CPAP Auto set for 7-14 cmH2O.
Unfortunately I have been trying these and many other settings on Auto CPAP for the past 15 years and when I get rid of the obstructive apneas, they are replaced by central apneas. Can someone help me with settings for a new ResMed AirCurve10 VAuto and give me a chance of getting rid of both obstructive and clear air apneas and let me get rested in 8 hours instead of 10-12 hours?
The patient had 137 apneas. There were 22 obstructive apneas, no mixes apneas, 12 central apneas, and 103 hypopneas for an AHI of 15.2/hour. Supine 309.2 min (38.8% of sleep 19.5 AHI. Right 327.7 min 60.7% of sleep 12.3 AHI.
Sleep Efficiency. 87%. AHI 15.2/hr. ODI 13.6l/hr. Avg SpO2 93.1%. Highest SpO2 98%. Lowest SpO2 78%. Max HR 95 bpm. Min HR 51bpm. N1&N2 62.8%. N3 5.1%. REM sleep 32.1%. Loud snoring all three nights. Time in bed 616.7 min. 539.6 min test.
Significant Oxygen desaturation observed with a total of 13.8 minutes below 89%. SPO2 longest span <89% was 1.0 minute. Sleep stages of N1, N1, N3, and Rem Sleep observed per EEG lead signals. RERAs, and Clear Air Hypopneas and Clear Air Apneas present.
Dx Severe Obstructive Sleep Apnea and Hypersomnia with OSA.
The Doctor prescribed a CPAP Auto set for 7-14 cmH2O.
Unfortunately I have been trying these and many other settings on Auto CPAP for the past 15 years and when I get rid of the obstructive apneas, they are replaced by central apneas. Can someone help me with settings for a new ResMed AirCurve10 VAuto and give me a chance of getting rid of both obstructive and clear air apneas and let me get rested in 8 hours instead of 10-12 hours?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear - Fit Pack (All Sizes Included) |
Additional Comments: Titrated on Auto CPAP at 7/14 cm: Only licensed medical professionals can give medical advice or write prescriptions |
Re: Not sure the sleep doctor has this one right!
We can try to help but need a lot more information including some software reports.
Just how many centrals are we talking about here per hour average?
Do you know about the software available to evaluate things?
Just how many centrals are we talking about here per hour average?
Do you know about the software available to evaluate things?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Not sure the sleep doctor has this one right!
http://www.cpaptalk.com/viewtopic/t1585 ... eview.html
OSCAR https://www.sleepfiles.com/OSCAR/
http://www.apneaboard.com/wiki/index.ph ... stallation
http://www.apneaboard.com/wiki/index.ph ... rpretation
http://www.apneaboard.com/wiki/index.ph ... ganization organize and how to do screen shots.
wiki/index.php/Oscar:organize Wiki
OSCAR https://www.sleepfiles.com/OSCAR/
http://www.apneaboard.com/wiki/index.ph ... stallation
http://www.apneaboard.com/wiki/index.ph ... rpretation
http://www.apneaboard.com/wiki/index.ph ... ganization organize and how to do screen shots.
wiki/index.php/Oscar:organize Wiki
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Not sure the sleep doctor has this one right!
Hi pugsy. I registered on this board a long time ago, so have patience with me.
I now have a new Resmed Aircurve10 and had decided not to chase any more rainbows with the Resmed AirSense10 Auto although it is only about 1 year old. I have OSCAR and it works on both machines. I only have two nights of data, but tell me where to look and I will post the latest night.
I now have a new Resmed Aircurve10 and had decided not to chase any more rainbows with the Resmed AirSense10 Auto although it is only about 1 year old. I have OSCAR and it works on both machines. I only have two nights of data, but tell me where to look and I will post the latest night.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear - Fit Pack (All Sizes Included) |
Additional Comments: Titrated on Auto CPAP at 7/14 cm: Only licensed medical professionals can give medical advice or write prescriptions |
Re: Not sure the sleep doctor has this one right!
hello, ww!ww wrote: ↑Tue Mar 05, 2024 11:30 pmHi pugsy. I registered on this board a long time ago, so have patience with me.
I now have a new Resmed Aircurve10 and had decided not to chase any more rainbows with the Resmed AirSense10 Auto although it is only about 1 year old. I have OSCAR and it works on both machines. I only have two nights of data, but tell me where to look and I will post the latest night.
do you see the links that pugsy posted? if you click on them, they'll give you the instructions you need to get the proper Oscar graphs and how to post them here.
if after reading those links, you still have questions then let us know how we can help.
good luck!
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Not sure the sleep doctor has this one right!
lThanks.zonker wrote: ↑Tue Mar 05, 2024 11:55 pmhello, ww!ww wrote: ↑Tue Mar 05, 2024 11:30 pmHi pugsy. I registered on this board a long time ago, so have patience with me.
I now have a new Resmed Aircurve10 and had decided not to chase any more rainbows with the Resmed AirSense10 Auto although it is only about 1 year old. I have OSCAR and it works on both machines. I only have two nights of data, but tell me where to look and I will post the latest night.
do you see the links that pugsy posted? if you click on them, they'll give you the instructions you need to get the proper Oscar graphs and how to post them here.
if after reading those links, you still have questions then let us know how we can help.
good luck!
I think I got the screen capture file, but it has more than the amount of graphs shown. Now how do you upload the screen capture file to my account ?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear - Fit Pack (All Sizes Included) |
Additional Comments: Titrated on Auto CPAP at 7/14 cm: Only licensed medical professionals can give medical advice or write prescriptions |
Re: Not sure the sleep doctor has this one right!
One thing he (gen non-specific) got wrong was
If you remove the central apneas your obstructive AHI is 13.9, or mild obstructive apnea.
Just sayin'.
By definition, severe obstructive sleep apnea AHI > 30. Moderate obstructive sleep apnea AHI between 15 and 30. Mild obstructive sleep apnea AHI is between 5 and 15.
If you remove the central apneas your obstructive AHI is 13.9, or mild obstructive apnea.
Just sayin'.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
- chunkyfrog
- Posts: 34461
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: In the abyss that is Nebraska--wish me luck!
Re: Not sure the sleep doctor has this one right!
Maybe OP's doctor has a different definition of "severe".
Then again, simple numbers may be weighted by comorbidities.
--as they should be.
Then again, simple numbers may be weighted by comorbidities.
--as they should be.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Not sure the sleep doctor has this one right!
Different than The Standards of Practice?chunkyfrog wrote: ↑Wed Mar 06, 2024 9:49 amMaybe OP's doctor has a different definition of "severe".
Don't walk away.
Run.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
Re: Not sure the sleep doctor has this one right!
I finally got the attachment uploaded. Question 1: Why do I sleep so long? Question 2: Are there any better settings to reduce snore and obstructive apneas and not cause central apneas?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear - Fit Pack (All Sizes Included) |
Additional Comments: Titrated on Auto CPAP at 7/14 cm: Only licensed medical professionals can give medical advice or write prescriptions |
Last edited by ww on Wed Mar 06, 2024 11:47 am, edited 1 time in total.
Re: Not sure the sleep doctor has this one right!
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear - Fit Pack (All Sizes Included) |
Additional Comments: Titrated on Auto CPAP at 7/14 cm: Only licensed medical professionals can give medical advice or write prescriptions |
- chunkyfrog
- Posts: 34461
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: In the abyss that is Nebraska--wish me luck!
Re: Not sure the sleep doctor has this one right!
So, no doctor has ever placed his OPINION above professional standards?Rubicon wrote: ↑Wed Mar 06, 2024 9:58 amDifferent than The Standards of Practice?chunkyfrog wrote: ↑Wed Mar 06, 2024 9:49 amMaybe OP's doctor has a different definition of "severe".
Don't walk away.
Run.
It happens all the time.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Not sure the sleep doctor has this one right!
There's not enough central apneas on that report to worry about. Do you have other reports with a much higher number of CAs/Central Apneas?
It's normal to have an occasional central apnea anyway.
Do you take any medications of any kind? If so, what?
Do you have any other physical or mental health issues? If so, what?
Are you waking often during the night for some reason or other? If so, why?
It's normal to have an occasional central apnea anyway.
Do you take any medications of any kind? If so, what?
Do you have any other physical or mental health issues? If so, what?
Are you waking often during the night for some reason or other? If so, why?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Not sure the sleep doctor has this one right!
you asked "why do I sleep so long?". now i'm not trying to be a smart aleck here, but it may be because your body needs it.
or are you saying you're sleeping longer now than at any other time in your life?
or are you saying you're sleeping longer now than at any other time in your life?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Not sure the sleep doctor has this one right!
Great, then show some specific sleep-related examples.chunkyfrog wrote: ↑Wed Mar 06, 2024 11:55 amSo, no doctor has ever placed his OPINION above professional standards?Rubicon wrote: ↑Wed Mar 06, 2024 9:58 amDifferent than The Standards of Practice?chunkyfrog wrote: ↑Wed Mar 06, 2024 9:49 amMaybe OP's doctor has a different definition of "severe".
Don't walk away.
Run.
It happens all the time.
Easy to shoot from the hip with unfounded generalizations.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.