Is a basic machine best, or merely adequate?
-
SleeplessInIL
- Posts: 13
- Joined: Tue Jan 29, 2008 8:00 am
Is a basic machine best, or merely adequate?
I've been on cpap for one week now. All in all, it's going ok (as long as I use a little sleep aid when I turn in). Before I tell the DME that this machine will work for me (ie go ahead and start the billing), i would like opinions from those of you who are experienced.
The DME provided me with the Remstar M series basic model under the rent to own program that my insurance uses. My prescription pressure is 7cm.
Is there any benefit at my pressure to insist on c-flex capability?
And, there doesn't seem to be any mention in the users guide about retrieving detailed data to see how I'm doing. I would like to know more about how it's working for me, but this machine seems limited.
Thanks
The DME provided me with the Remstar M series basic model under the rent to own program that my insurance uses. My prescription pressure is 7cm.
Is there any benefit at my pressure to insist on c-flex capability?
And, there doesn't seem to be any mention in the users guide about retrieving detailed data to see how I'm doing. I would like to know more about how it's working for me, but this machine seems limited.
Thanks
_________________
| Mask: Opus 360 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: 7cm prescribed, ramp from 5.5 my choice |
The "MINIMUM" machine you want to get is a REMstar M Series Pro. That is a straight pressure CPAP machine that collects data.....that you WILL be able to use to monitor your therapy.
The next steps up are the Auto/APAP machines. All of these CPAP/APAP machines have the same insurance code and your insurance provider will pay exactly the same amount regardless of which one the DME gives you. Consequently, your DME will (probably) fight you tooth and nail to give you the cheapest equipment in order to maximize their profits.
If you don't at least get the "Pro" NOW.......you'll wish at some point (probably sooner than later) that you'd have gotten it.
Den
The next steps up are the Auto/APAP machines. All of these CPAP/APAP machines have the same insurance code and your insurance provider will pay exactly the same amount regardless of which one the DME gives you. Consequently, your DME will (probably) fight you tooth and nail to give you the cheapest equipment in order to maximize their profits.
If you don't at least get the "Pro" NOW.......you'll wish at some point (probably sooner than later) that you'd have gotten it.
Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Post subject: Is a basic machine best, or merely adequate?
Actually, to answer your initial question......."NEITHER".
It sure as Hell isn't "Best" and would only be "adequate" as a backup machine once you've determined that your therapy is working for you.
Den
Actually, to answer your initial question......."NEITHER".
It sure as Hell isn't "Best" and would only be "adequate" as a backup machine once you've determined that your therapy is working for you.
Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
-
SleeplessInIL
- Posts: 13
- Joined: Tue Jan 29, 2008 8:00 am
Thanks Den.
Yep, the DME supplier said they will not provide anything but the basic. If I want to upgrade, I have to go somewhere else (which I'm prepared to do) or I suppose I could pay the difference.
Yep, the DME supplier said they will not provide anything but the basic. If I want to upgrade, I have to go somewhere else (which I'm prepared to do) or I suppose I could pay the difference.
_________________
| Mask: Opus 360 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: 7cm prescribed, ramp from 5.5 my choice |
I SO agree! My sleep study indicated that my pressure needed to be set at 6.0. A year later, after finding this forum, getting software and a card reader, I found that I was not really getting any effective CPAP therapy. So, I purchased an M Series Auto w/APAP, found my ACTUAL pressure setting is 12.0. Hoping you'll be able to at least get a data capable machine, or even better, and M Series auto. It's a "sweet dream" machine!!

"You are getting sleepy . . ."

"You are getting sleepy . . ."
With their attitude.....I would "walk" (actually run) from them and make it very well known WHY.SleeplessInIL wrote:Thanks Den.
Yep, the DME supplier said they will not provide anything but the basic. If I want to upgrade, I have to go somewhere else (which I'm prepared to do) or I suppose I could pay the difference.
Have you checked with your insurance provider about them reimbursing YOU if you purchase the stuff out-of-pocket?
There are many of us who have used that method and it gets brought up frequently.
In the finally summary.....you need to ask yourself how much your health is worth and be willing to shell out a few more bucks if necessary.
Good luck,
Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
If your insurance is contracted w/other local DME suppliers, give this supplier the option of giving you the Respironics M Series Pro w/C-Flex and Smart Card or you will take your business elsewhere. BUT DO IT NOW, ASAP, BEFORE the first 30 days have passed!!!!!! Or you are pretty much STUCK w/what you have now.
Call your insurance and ask them what other local DME CPAP suppliers they are contracted with so you know what and who your options are. DO IT NOW!
Call your insurance and ask them what other local DME CPAP suppliers they are contracted with so you know what and who your options are. DO IT NOW!
Den is RIGHT ON! The bold emphasis is mine.Wulfman wrote: ... All of these CPAP/APAP machines have the same insurance code and your insurance provider will pay exactly the same amount regardless of which one the DME gives you. Consequently, your DME will (probably) fight you tooth and nail to give you the cheapest equipment in order to maximize their profits.
If you don't at least get the "Pro" NOW.......you'll wish at some point (probably sooner than later) that you'd have gotten it.
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator |
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.
My computer says I need to upgrade my brain to be compatible with its new software.
To look at this a different way ask yourself this.
Do you want to have to get PSG everytime you lose or gain 20 pounds? Or if you change something in your life like how you drink or smoke or have some other medical issue come up and have no way to see how they effect your OSA?
Do you want to be limited in your treatment by not having the ability to run the unit from time to time in Auto mode to check for and to be able to adjust your treatment to fit whatever changes come your way in life?
With the Apap and Data recording ability you can adapt to whatever life throws at you as well as run as a straight Cpap. A non-data recording Cpap can be adjusted pressure wise but only in a hit or miss fashion like a shot in the dark. Do you really wish to guess at it?
Get an Auto with Data recording it is the most sensible and cost effective way to insure the best treatment of your life threatening condition. Even if you have to get it from some place else or have to make them order it special for you or pay extra for it, fight for it. In the end at least it will help you take the medical leach off your pocketbook.
Do you want to have to get PSG everytime you lose or gain 20 pounds? Or if you change something in your life like how you drink or smoke or have some other medical issue come up and have no way to see how they effect your OSA?
Do you want to be limited in your treatment by not having the ability to run the unit from time to time in Auto mode to check for and to be able to adjust your treatment to fit whatever changes come your way in life?
With the Apap and Data recording ability you can adapt to whatever life throws at you as well as run as a straight Cpap. A non-data recording Cpap can be adjusted pressure wise but only in a hit or miss fashion like a shot in the dark. Do you really wish to guess at it?
Get an Auto with Data recording it is the most sensible and cost effective way to insure the best treatment of your life threatening condition. Even if you have to get it from some place else or have to make them order it special for you or pay extra for it, fight for it. In the end at least it will help you take the medical leach off your pocketbook.
_________________
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: I swap out the Swift FX as needed with the Mirage Quatro Full Face with Headgear. |
Personally I'm always ready to learn, although I do not always like being taught.
Sir Winston Churchill
I’m not asleep… but that doesn’t mean I’m awake.
- Albert Camus
Sir Winston Churchill
I’m not asleep… but that doesn’t mean I’m awake.
- Albert Camus
Wow after reading threw this I am more nervious than befor......
I have not sat doun withe the DME I am still waiting for insurence aproval
but I was wondering Is there a CPAP with recording and presure ajustsments? or am I wanting an APAP?? the Dr. Perscrption just calls for CPAP Unit with Heated Humidifier and a setting of 9CWP (not sure what CWP means other than presure) that is all that is on the prscription.
Am I able to go in to the DME and tell them what I want or are they going to say this is it?
_________________
CPAPopedia Keywords Contained In This Post (Click For Definition): humidifier, CPAP, DME, APAP
I have not sat doun withe the DME I am still waiting for insurence aproval
but I was wondering Is there a CPAP with recording and presure ajustsments? or am I wanting an APAP?? the Dr. Perscrption just calls for CPAP Unit with Heated Humidifier and a setting of 9CWP (not sure what CWP means other than presure) that is all that is on the prscription.
Am I able to go in to the DME and tell them what I want or are they going to say this is it?
_________________
CPAPopedia Keywords Contained In This Post (Click For Definition): humidifier, CPAP, DME, APAP
- DreamStalker
- Posts: 7503
- Joined: Mon Aug 07, 2006 9:58 am
- Location: Nowhere & Everywhere At Once
CWP = centimeters water pressure.TiredTex wrote:Wow after reading threw this I am more nervious than befor......
I have not sat doun withe the DME I am still waiting for insurence aproval
but I was wondering Is there a CPAP with recording and presure ajustsments? or am I wanting an APAP?? the Dr. Perscrption just calls for CPAP Unit with Heated Humidifier and a setting of 9CWP (not sure what CWP means other than presure) that is all that is on the prscription.
Am I able to go in to the DME and tell them what I want or are they going to say this is it?
It's your money! You need to decide if you are buying your health needs or if are going to let them take your money regardless of what your health needs are.
It's either your health or their bottom line ... you decide.
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.
I like the way you thinkDreamStalker wrote:
CWP = centimeters water pressure.
It's your money! You need to decide if you are buying your health needs or if are going to let them take your money regardless of what your health needs are.
It's either your health or their bottom line ... you decide.
But I dont know what I need
and I have been reading this site for a week and it has been infomative but I am still unclear on so many difrent things
- DreamStalker
- Posts: 7503
- Joined: Mon Aug 07, 2006 9:58 am
- Location: Nowhere & Everywhere At Once
First get a copy of your PSG (sleep lab study) and the original copy of your prescription ... before signing any papers regarding machines or masks.
Then dermine your insurance copay if insurance is involved.
Then you have the minimum info to start shopping and demanding your health needs. You are then ready to demand your DME what you want/need and if they cannot provide you go to another ... you may discover that paying out of pocket is actually a better deal when you consider all of the headaches that insurance/DME will give you.
You will need to shop for a machine ... it's better to get an "auto titrating/adjusting" machine called an APAP. An APAP can be set to work just like a plain CPAP but CPAP cannot work as APAP ... and all APAP machines provide monitoring data which is essential to your treatment. Many here like the Respironics "M" series APAP with AFLEX ... also, you will be better of with a stand alone heated humidifer rather than the integrated one offered with the machine. The F&P HC 150 is a good choice for the HH.
The mask will be the most difficult to find since everyone has a different size/shape face and head ... you will just need to try enough of them to find what works best for you.
Then dermine your insurance copay if insurance is involved.
Then you have the minimum info to start shopping and demanding your health needs. You are then ready to demand your DME what you want/need and if they cannot provide you go to another ... you may discover that paying out of pocket is actually a better deal when you consider all of the headaches that insurance/DME will give you.
You will need to shop for a machine ... it's better to get an "auto titrating/adjusting" machine called an APAP. An APAP can be set to work just like a plain CPAP but CPAP cannot work as APAP ... and all APAP machines provide monitoring data which is essential to your treatment. Many here like the Respironics "M" series APAP with AFLEX ... also, you will be better of with a stand alone heated humidifer rather than the integrated one offered with the machine. The F&P HC 150 is a good choice for the HH.
The mask will be the most difficult to find since everyone has a different size/shape face and head ... you will just need to try enough of them to find what works best for you.
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.
-
Guest
I SO agree! I love my machine (Respironics M Series Pro w/CFLEX), but I so wish I'd known more "back then" and requested an auto. I don't think I'd like to use it in auto mode all the time, but just use that to test myself at certain points in my life, like mentioned above, and then switch it to CPAP for normal use.hades161 wrote:To look at this a different way ask yourself this.
Do you want to have to get PSG everytime you lose or gain 20 pounds? Or if you change something in your life like how you drink or smoke or have some other medical issue come up and have no way to see how they effect your OSA?
Do you want to be limited in your treatment by not having the ability to run the unit from time to time in Auto mode to check for and to be able to adjust your treatment to fit whatever changes come your way in life?
With the Apap and Data recording ability you can adapt to whatever life throws at you as well as run as a straight Cpap. A non-data recording Cpap can be adjusted pressure wise but only in a hit or miss fashion like a shot in the dark. Do you really wish to guess at it?
I recently lost around 20 pounds (and intend on losing more) and I've already made adjustments to mine, but they are guesses. And with my high deductible insurance, I can't afford to have a sleep study done every time something, like weight loss, happens in my life
I would think you could at least get the one I have, which is a great machine with data capability. I didn't even have to ask for it (even though I was prepared with all my research to show why I wanted that one!). It was waiting for me on the DME's table when I came in for the appointment. (I was so consumed with getting treated, I totally passed by researching autos and just stuck to cpap. If only I knew then what I know now. )
Pam