Search found 265 matches
- Fri Jul 16, 2010 6:48 am
- Forum: CPAP and Sleep Apnea Message Board
- Topic: Mask for mouth-breathing, side-sleeping, guy with goatee.
- Replies: 5
- Views: 1968
Re: Mask for mouth-breathing, side-sleeping, guy with goatee.
I wear a full beard and moustache, although through my time on CPAP I have at times only had a Van Dke, and I do quite well with the Ultra Mirage FF. It took some time to get used to sleeping on my side with the mask, but with a little adjustment on my pillow I have make it work quite well.
- Thu Jul 15, 2010 12:37 pm
- Forum: CPAP and Sleep Apnea Message Board
- Topic: Error code 58 Reso bipap auto
- Replies: 3
- Views: 1377
Re: Error code 58 Reso bipap auto
If its the "Tank" model, E58 indicates the pressure sensor is bad. It could just be a loose or blocked sensing line, or the sensor is bad, in which case the Main PCA would have to be replaced. Probably not financially feasible to repair if the sensor is actually bad.
- Thu Jul 15, 2010 8:44 am
- Forum: CPAP and Sleep Apnea Message Board
- Topic: Error code 58 Reso bipap auto
- Replies: 3
- Views: 1377
Re: Error code 58 Reso bipap auto
Which Respironics machine is it.....the tank, the m-series or the PR series. Which one will make a difference in what the code means.
- Mon Mar 22, 2010 11:40 am
- Forum: CPAP and Sleep Apnea Message Board
- Topic: Will I have to shave my beard?
- Replies: 31
- Views: 21702
Re: Will I have to shave my beard?
I wear a very full and sometimes quite shaggy beard and have used the UMFF since day one without a bit of problems. My leakage is a little bit above what many seem to think is normal, but the machine is designed to make up for quite a bit of leakage. As long a the little bit of air escaping from the...
- Wed Mar 10, 2010 3:50 pm
- Forum: CPAP and Sleep Apnea Message Board
- Topic: Translate BiPap settings to APap settings
- Replies: 11
- Views: 1959
Re: Translate BiPap settings to APap settings
These settings are for an Auto BiPap. They cannot really be translated to an APAP. If this is the way the script was sent to the DME, you should end up with and Auto BiPap.
- Thu Jan 14, 2010 8:00 am
- Forum: CPAP and Sleep Apnea Message Board
- Topic: Beards and masks
- Replies: 26
- Views: 11479
Re: Beards and masks
I have a very full and most of the time quite bushy beard and I have successfully used the UMFF since day one.
- Fri Nov 20, 2009 3:04 pm
- Forum: CPAP and Sleep Apnea Message Board
- Topic: Doc doesn't care to look at data - need advice
- Replies: 17
- Views: 2259
Re: Doc doesn't care to look at data - need advice
Using the equipment that the OP claims to have in this thread viewtopic/t46839/Taking-control-of-my-f ... st-II.html, will give them much more information than is available on any CPAP.
- Fri Nov 20, 2009 9:50 am
- Forum: CPAP and Sleep Apnea Message Board
- Topic: Taking control of my families sleep Apnea - Stardust II
- Replies: 4
- Views: 1362
Re: Taking control of my families sleep Apnea - Stardust II
Just curious........where were you able to acquire one of these systems, and where do you get the disposables for it????? Dit it come with the necessary software to download the data from the recorder?????
- Tue Nov 17, 2009 10:33 am
- Forum: CPAP and Sleep Apnea Message Board
- Topic: mycpap.org downloads CardUtilityInstaller
- Replies: 11
- Views: 2639
Re: mycpap.org downloads CardUtilityInstaller
In order for this software to be able to restore the data to the card, it first had to be backed up to your computer before you took it to the DME and they erased it. For you to be able to do this you will have to get the card reader. From your situation, you cannot restore the data....it is gone fo...
- Wed Nov 04, 2009 8:25 am
- Forum: CPAP and Sleep Apnea Message Board
- Topic: Hospital stays and CPAP
- Replies: 28
- Views: 6608
Re: Hospital stays and CPAP
I guess I can weigh in as professional in the Biomedical/Clinical engineering field. Yes, we are required to check all electrically operated equipment that may come into contact with a patient. This requirement is mandated by NFPA 99. Any electrical equipment that is used in the "patient care vicini...
- Wed Oct 14, 2009 10:42 am
- Forum: CPAP and Sleep Apnea Message Board
- Topic: Best Full Face Mask for Facial Hair
- Replies: 8
- Views: 2708
Re: Best Full Face Mask for Facial Hair
Full beard and moustache.....I use the UMFF....always have.
- Fri Oct 02, 2009 7:50 am
- Forum: CPAP and Sleep Apnea Message Board
- Topic: Sleep study question
- Replies: 10
- Views: 1058
Re: Sleep study question
Very curious to see if any one knows anything about this relationship. . Cant say that I know anything about the relationship, however, I have been successfully treating my OSA for over 3.5 years, and was just recently diagnosed with low testosterone. Dr started me on transdermal testosterone and t...
- Thu Sep 24, 2009 9:07 am
- Forum: CPAP and Sleep Apnea Message Board
- Topic: Sleep apnea without the apnea.
- Replies: 62
- Views: 7672
Re: Sleep apnea without the apnea.
It's a RemStar Pro 2. I first took it apart, to look for a suitable point with a voltage that indicated some useful information. I first thought I might have to connect to the wires which drive the motor, but then I noticed a lot of "test points" on the board. ("test points" are points in the circu...
- Mon Sep 14, 2009 8:02 am
- Forum: CPAP and Sleep Apnea Message Board
- Topic: Any military veterans use VA health care ?
- Replies: 34
- Views: 3703
Re: Any military veterans use VA health care ?
For you guys/gals who got machines through the VA, is your sleep apnea considered service connected? I'm a Desert Storm vet with sleep apnea, but since I wasn't diagnosed while in service mine was declared "not service connected." I'm wondering if I could still get a machine through the VA, even th...
- Fri Sep 11, 2009 10:25 am
- Forum: CPAP and Sleep Apnea Message Board
- Topic: Any military veterans use VA health care ?
- Replies: 34
- Views: 3703
Re: Any military veterans use VA health care ?
wlenz wrote:Sounds like I am missing some great benefits. I believe I am eligible for VA benefits, but never applied. Can I do it online?
Cant say on that one. I signed up about 20 years ago...werent much for "on line" back then. Check out http://www.va.gov.