Snoredog

Post by dsm » Wed Apr 11, 2007 4:13 am

In another thread you mentioned about the S8 Vantage using a mic to detect snores. This is info I am keen to learn about as it is not what I understood these machines did. In that other thread the question never got dealt with (a load of other unrelated muck did & I am hoping that unfortunate outburst has passed) so I am now asking you as politely as I can, where did you get that info from.

To preempt any misunderstanding, this is not me having a go at you. You posted a comment & I am keen to learn from it. If you don't have any links just say so. If you do, please post.

Thanks

DSM
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

User avatar
StillAnotherGuest
Posts: 1005
Joined: Sun Sep 24, 2006 6:43 pm

His Lips Aren't Moving Either

Post by StillAnotherGuest » Wed Apr 11, 2007 5:58 am

dsm wrote:In another thread you mentioned about the S8 Vantage using a mic to detect snores. This is info I am keen to learn about as it is not what I understood these machines did.
Although the term "transducer" is more commonly used to describe the device that detects snores, in the past, Mike has frequently used "microphone" in his inventions:
In FIG. 3, a CPAP apparatus embodying the invention is illustrated. The CPAP unit comprises a motor 20 which drives a blower 21. The speed of the motor 20 is controlled by an electronic speed control unit 23. As an increase in motor speed also increases the blower speed which in turn increases the output air pressure of the blower 21, the speed control unit can be manipulated to vary the output pressure of the blower 21. The CPAP device also includes a snoring detection means 22 wherein sounds are detected by a microphone 11. In its most general form, the snoring detection means 22 is a pressure detection means and microphone 11 is a differential pressure sensor. The snoring detection means 22 is conveniently in the form of the previously described device 10. Electrical impulses are fed from said microphone 11 to an amplifier/filter/processor unit 26 which generates an electrical signal when snoring sounds occur. The motor speed control means is electrically connected to the snoring detection device 22 and increases the speed of the electric motor 20 by an analogue means in response to the electrical signal generated by the snoring detection device. Accordingly, the output pressure of the CPAP unit increases in response to detection of snoring.

When a snore or sequence of snores is detected by the snoring detection means 22 a signal is generated. The speed control unit 23 increases the speed of the fan motor and the output pressure is increased. As snoring is caused by vibration of the soft palate, it is therefore indicative of an unstable airway and, as previously described, is a warning signal of the imminence of upper airway occlusion in patients that suffer obstructive sleep apnea. Snoring is itself undesirable not only as it is a disturbance to others but it is strongly believed to be connected with hypertension. If the resultant increase in CPAP pressure is sufficient to completely stabilize the airway, snoring will cease. If a further snoring sound is detected, the CPAP pressure is increased again. This process is repeated until the upper airway is stabilized and snoring ceases. Hence, the occurrence of obstructive apnea can be eliminated by application of a minimum appropriate pressure at the time of use.

In order to ensure that the CPAP pressure is maintained at a level as low as practicable to prevent the onset of apnea, the preferred embodiment also includes a means to decrease the pressure if an extended period of snore free breathing occurs. For example, this can be done by automatically reducing the CPAP pressure at a gradual rate as long as snoring is not detected. The rate at which the CPAP pressure is decreased in the absence of snoring is preferably much less than the rate at which it is increased when snoring is detected. This can be achieved, for example, by the amplifier/filter/processor unit 26, in the absence of an electronic signal from the microphone 11, continuously gradually reducing the blower speed over a period of time but increasing the blower speed in incremental steps each time a snore is detected by the microphone 11.
SAG

Image

Aromatherapy may help CPAP compliance. Lavender, Mandarin, Chamomile, and Sweet Marjoram aid in relaxation and sleep. Nature's Gift has these and a blend of all four called SleepEase.

User avatar
GoofyUT
Posts: 1085
Joined: Sun Apr 09, 2006 9:45 am

Citation?

Post by GoofyUT » Wed Apr 11, 2007 6:19 am

SAG-

Could you provide the reference for your quote? I'm simply interested in viewing it more fully.

And, for the rest of us, who's Mike?

Chuck
People are dying every day in Darfur simply for who they are!!! PLEASE HELP THEM!
http://www.savedarfur.org

_______________________________

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Re: His Lips Aren't Moving Either

Post by dsm » Wed Apr 11, 2007 6:38 am

StillAnotherGuest wrote:
dsm wrote:In another thread you mentioned about the S8 Vantage using a mic to detect snores. This is info I am keen to learn about as it is not what I understood these machines did.
Although the term "transducer" is more commonly used to describe the device that detects snores, in the past, Mike has frequently used "microphone" in his inventions:
In FIG. 3, a CPAP apparatus embodying the invention is illustrated. The CPAP unit comprises a motor 20 which drives a blower 21. The speed of the motor 20 is controlled by an electronic speed control unit 23. As an increase in motor speed also increases the blower speed which in turn increases the output air pressure of the blower 21, the speed control unit can be manipulated to vary the output pressure of the blower 21. The CPAP device also includes a snoring detection means 22 wherein sounds are detected by a microphone 11. In its most general form, the snoring detection means 22 is a pressure detection means and microphone 11 is a differential pressure sensor. The snoring detection means 22 is conveniently in the form of the previously described device 10. Electrical impulses are fed from said microphone 11 to an amplifier/filter/processor unit 26 which generates an electrical signal when snoring sounds occur. The motor speed control means is electrically connected to the snoring detection device 22 and increases the speed of the electric motor 20 by an analogue means in response to the electrical signal generated by the snoring detection device. Accordingly, the output pressure of the CPAP unit increases in response to detection of snoring.

When a snore or sequence of snores is detected by the snoring detection means 22 a signal is generated. The speed control unit 23 increases the speed of the fan motor and the output pressure is increased. As snoring is caused by vibration of the soft palate, it is therefore indicative of an unstable airway and, as previously described, is a warning signal of the imminence of upper airway occlusion in patients that suffer obstructive sleep apnea. Snoring is itself undesirable not only as it is a disturbance to others but it is strongly believed to be connected with hypertension. If the resultant increase in CPAP pressure is sufficient to completely stabilize the airway, snoring will cease. If a further snoring sound is detected, the CPAP pressure is increased again. This process is repeated until the upper airway is stabilized and snoring ceases. Hence, the occurrence of obstructive apnea can be eliminated by application of a minimum appropriate pressure at the time of use.

In order to ensure that the CPAP pressure is maintained at a level as low as practicable to prevent the onset of apnea, the preferred embodiment also includes a means to decrease the pressure if an extended period of snore free breathing occurs. For example, this can be done by automatically reducing the CPAP pressure at a gradual rate as long as snoring is not detected. The rate at which the CPAP pressure is decreased in the absence of snoring is preferably much less than the rate at which it is increased when snoring is detected. This can be achieved, for example, by the amplifier/filter/processor unit 26, in the absence of an electronic signal from the microphone 11, continuously gradually reducing the blower speed over a period of time but increasing the blower speed in incremental steps each time a snore is detected by the microphone 11.
SAG
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

User avatar
tangents
Posts: 750
Joined: Thu Mar 08, 2007 11:03 am
Location: Cleveland, Ohio

Post by tangents » Wed Apr 11, 2007 7:19 am

Thanks for the tutorial, SWS, very interesting. Do any of the other brands try to detect snoring?

User avatar
StillAnotherGuest
Posts: 1005
Joined: Sun Sep 24, 2006 6:43 pm

A Transducer By Any Other Name...

Post by StillAnotherGuest » Wed Apr 11, 2007 8:45 am

Snoring is most often assessed with a pressure transducer (except maybe for a forced oscillation technique-based machine. And you may recall that that one didn't do that well in addressing snoring vs the ResMed algorithm). And even though Respironics claims to have a flow-based algorithm, they say re: addressing snoring:
The Breathing Pattern Recognition also performs a secondary analysis, based on peak flow, to determine the occurrence of breathing problems like apnea and hypopnea, and detects snoring based on a frequency analysis of a pressure signal.
dsm wrote:I know that these were used in earlier Autos but understood that they were dropped infavour of airflow sensing (for vibrations).
When Mike described his snoring algorithm as recently as 2004, he wrote:
CPAP treatment apparatus comprising...
...a pressure transducer to generate input for said processor...
...wherein said processor is programmed to perform the steps of:

determining a snore measure from said pressure transducer indicative of a level of snoring by the patient;
calculating a snore threshold as a function of said treatment pressure;
calculating an adjustment pressure by an amount that is a function of said snore threshold and said snore measure if said snore measure exceeds said snore threshold...
I would be surprised if anybody is using a purely flow-based waveform (one that measures flow directly) to identify snoring. Why would you want to?

However, my original point is that a microphone is nothing more than a pressure transducer.
SAG

Image

Aromatherapy may help CPAP compliance. Lavender, Mandarin, Chamomile, and Sweet Marjoram aid in relaxation and sleep. Nature's Gift has these and a blend of all four called SleepEase.

User avatar
Snoredog
Posts: 6399
Joined: Sun Mar 19, 2006 5:09 pm

Post by Snoredog » Wed Apr 11, 2007 10:03 am

what are you there at Resmed? Just the janitor?

That would explain how you get those machines to recycle

Like SAG said, its part of the pressure transducer circuit and even included in the Clinical manual:
The flow sensor, located in the AUTOSET SPIRIT unit, enables detection of inspiratory flow limitation and apneas. The pressure sensor, also located in the unit, enables measurement of pressure and snore.
More on AutoSet (from Clinical Manual):
Figure 3: Inspiratory flow/time curve affected by snore
The AutoSet algorithm assigns an arbitrary value between 0.0 and 2.0 to the
average amplitude of the snoring
detected for the past 5 breaths. A value of 1.0 is equivalent to approximately 75dBA measured 10cm from the nares. Treatment pressure increases by up to 0.2 cm H2O per second (proportional to the severity of the snore) for snore above 0.2 snore units. When snore is less than 0.2 snore units, therapy is reduced towards the minimum pressure with a 20-minute time constant.

An apnea is defined as a greater than 75% decrease in ventilation. The AutoSet algorithm scores an apnea if the 2-second moving average entilation drops below 25% of the recent time average (time constant 100 seconds) for at least 10 consecutive seconds. Treatment pressure increases based on the duration of the apnea. The pressure will not rise above 10 cm H2O when an apnea is detected, to prevent an inappropriate response to central apneas. Initial pressure increases are rapid, but the rate of increase diminishes as the pressure approaches 10 cm H2O.
When no further apneas are detected, therapy is reduced towards the minimum pressure with a 20-minute time constant.

A hypopnea is defined as a 50 to 75% drop in ventilation. A hypopnea is scored if the 8-second moving average ventilation drops below 50%, but not below 25%, of the recent average for 15 consecutive seconds. In order to avoid falsely responding to central hypopneas, the AutoSet algorithm does not respond to hypopneas but rather to the associated snore or flow limitation.
and SAG was probably quoting from the ResCare patent dated September 21, 1993:
http://tinyurl.com/2j77op

Last edited by Snoredog on Wed Apr 11, 2007 11:43 am, edited 2 times in total.
someday science will catch up to what I'm saying...

-SWS
Posts: 5301
Joined: Tue Jan 11, 2005 7:06 pm

Post by -SWS » Wed Apr 11, 2007 10:16 am

tangents wrote:Thanks for the tutorial, SWS, very interesting. Do any of the other brands try to detect snoring?
That wasn't me earlier in the thread. But on behalf of SAG, you're very welcome!

This post deviates just a bit off topic:

While that revered pneumotacographic sensor used in APAP machines is a "flow sensor" according to conventional use and medical definition, it really measures voltage induced by differential pressure across sensor heads. Again, that pneumotacographic sensor is considered the "flow sensor" inside many APAP models today, because it is used to derive flow and volume measurements.

While this document describes the Respironics Novametric Series 3 flow sensor in particular, it also highlights various details about pneumotacographic transduction in general:
http://oem.respironics.com/Downloads/Fl ... rement.pdf


User avatar
tangents
Posts: 750
Joined: Thu Mar 08, 2007 11:03 am
Location: Cleveland, Ohio

Post by tangents » Wed Apr 11, 2007 11:09 am

Thanks, SWS, both for the additional information, and for pointing out my SAG-ging memory! Thanks to you, SAG, for your tutorial.

I read the article on the Respironics Flow Measurements (you just gotta love pv=nRt), and started wondering about Purital Bennett. I found a White paper describing the GoodKnight 420E system, maybe you have already read it. It claims ~95% sensitivity in snoring detection. They also use a pneumotachograph.

I'm still trying to decide which machine to buy, so all this information is good stuff!


User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Post by Wulfman » Wed Apr 11, 2007 11:41 am

tangents wrote:Thanks, SWS, both for the additional information, and for pointing out my SAG-ging memory! Thanks to you, SAG, for your tutorial.

I read the article on the Respironics Flow Measurements (you just gotta love pv=nRt), and started wondering about Purital Bennett. I found a White paper describing the GoodKnight 420E system, maybe you have already read it. It claims ~95% sensitivity in snoring detection. They also use a pneumotachograph.

I'm still trying to decide which machine to buy, so all this information is good stuff!
Do you snore?

Den

(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

-SWS
Posts: 5301
Joined: Tue Jan 11, 2005 7:06 pm

Post by -SWS » Wed Apr 11, 2007 1:28 pm

Wulfman wrote: Do you snore?
Thanks, Den!! Don't mind if I do...

<zzzzzzzz... gurgle.... gasp...... choke... once again wide awake>

How about yourself? Care for an after-lunch snore?

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Post by Wulfman » Wed Apr 11, 2007 1:39 pm

-SWS wrote:
Wulfman wrote: Do you snore?
Thanks, Den!! Don't mind if I do...

<zzzzzzzz... gurgle.... gasp...... choke... once again wide awake>

How about yourself? Care for an after-lunch snore?
Thanks, Steve.

Don't mind if I do, too......ZZZZZZZZZzzzzzzzzZZZZZZZZzzzzzzzzz

Den (trained to nose-breath when napping in a chair)
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Post by dsm » Wed Apr 11, 2007 4:11 pm

The data discussed thus far points to pressure sensors as being the device used to measure snoring. I don't think these can be called microphones in the way that the early Autos did use genuine audio transducers. Those machines lacked the airflow sensors and an audio transducer was all they had.

Snoredog, Thanks for the reply - that was all that was needed. The other remarks weren't - please (asking kindly) don't make claims about my employment that are ficticious. I work here http://www.metcash.com/ as their IT Technical Architect & have been there since returning from HK in 2003 where I was a principal architect in HK Telecom (who used to be H.K. Cable & Wireless when I started there). I don't really want to be posting my personal info here but it is the only way I can see to end the disinformation. Can we now end those remarks - I *don't* work for Resmed & never have.

getting back to the issue of audio detection of snores. Is the feedback here saying ...

1) That Autos only use audio snore detection
2) That Autos only use airflow detectors (pressure transducers) to detect pressure vibrations
3) That they use pressure transducers and audio detectors
4) That they use only use pressure transducers


(I vote for 4)

DSM

xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

User avatar
Snoredog
Posts: 6399
Joined: Sun Mar 19, 2006 5:09 pm

Post by Snoredog » Wed Apr 11, 2007 4:42 pm

[quote="dsm"]The data discussed thus far points to pressure sensors as being the device used to measure snoring. I don't think these can be called microphones in the way that the early Autos did use genuine audio transducers. Those machines lacked the airflow sensors and an audio transducer was all they had.

Snoredog, Thanks for the reply - that was all that was needed. The other remarks weren't - please (asking kindly) don't make claims about my employment that are ficticious. I work here http://www.metcash.com/ as their IT Technical Architect & have been there since returning from HK in 2003 where I was a principal architect in HK Telecom (who used to be H.K. Cable & Wireless when I started there). I don't really want to be posting my personal info here but it is the only way I can see to end the disinformation. Can we now end those remarks - I *don't* work for Resmed & never have.

getting back to the issue of audio detection of snores. Is the feedback here saying ...

1) That Autos only use audio snore detection
2) That Autos only use airflow detectors (pressure transducers) to detect pressure vibrations
3) That they use pressure transducers and audio detectors
4) That they use only use pressure transducers


(I vote for 4)

DSM

someday science will catch up to what I'm saying...

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Post by dsm » Wed Apr 11, 2007 5:19 pm

Snores,

Take a look at these photos, these are not microphones. They are pressure sensors in a tandem configuration capable of detecting the difference and intensity of air prssure from two predetermined points in the air tube. The electronics then use that data to determine the volume of air passing through a known diameter tube (the internal air circuit) and at a known outlet pressure (see the third pressure transducer used to detect exit pressure).

Remstar dissasembled (in particular see photos 6 & 7)
http://www.internetage.com/cpapinfo/remstar-auto-dis-1/

Resmed S7 dissasembled (photo 11 & 32)
http://www.internetage.com/cpapinfo/resmed-s7-dis-1/

Look at the three sensors in each unit.

DSM

_________________

CPAPopedia Keywords Contained In This Post (Click For Definition): resmed

xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)