Ramp in CPAP and APAP
-
ArtCominio
- Posts: 3
- Joined: Fri Jun 28, 2019 10:06 am
Ramp in CPAP and APAP
I joined this board specifically because I have seen much BAD information about using the ramp feature for a ResMed AirSense 10.
Ramp is straightforward to understand in CPAP mode, but less so in APAP mode. Many have stated they turn ramp off. This may make perfect sense, but may be very useful to lower AHI scores in either CPAP or APAP modes.
RAMP WORKS DIFFERENTLY IN CPAP AND APAP MODES.
I have my machine set for APAP with a starting pressure of 6 and a max pressure of 15. Most would claim it makes no sense to turn ramp on in APAP mode.
With ramp off, I often see many events (using SleppyHead) during the period while I am still awake (because of irregular breathing pattern).
Because events are not even monitored during ramp time, using ramp allows the machine to ignore events while you are falling asleep.
Even those who use APAP will benefit from using ramp. it will cause the pressure to remain low for the entire ramp time. Then auto sense will kick in and start to adjust the pressure. BUT all events prior to ramp time will be ignored.
Those of us who use APAP and see events while falling asleep can significantly lower our AHI score by using ramp even in APAP mode.
SO RAMP CAN BE VERY USEFUL EVEN FOR APAP USERS!
Even CPAP users can see lower scores by turning ramp on to ignore events during the time they are falling asleep.
I know I am a new poster and expect much hate mail, to which I shall probably not respond.
Art Cominio
Ramp is straightforward to understand in CPAP mode, but less so in APAP mode. Many have stated they turn ramp off. This may make perfect sense, but may be very useful to lower AHI scores in either CPAP or APAP modes.
RAMP WORKS DIFFERENTLY IN CPAP AND APAP MODES.
I have my machine set for APAP with a starting pressure of 6 and a max pressure of 15. Most would claim it makes no sense to turn ramp on in APAP mode.
With ramp off, I often see many events (using SleppyHead) during the period while I am still awake (because of irregular breathing pattern).
Because events are not even monitored during ramp time, using ramp allows the machine to ignore events while you are falling asleep.
Even those who use APAP will benefit from using ramp. it will cause the pressure to remain low for the entire ramp time. Then auto sense will kick in and start to adjust the pressure. BUT all events prior to ramp time will be ignored.
Those of us who use APAP and see events while falling asleep can significantly lower our AHI score by using ramp even in APAP mode.
SO RAMP CAN BE VERY USEFUL EVEN FOR APAP USERS!
Even CPAP users can see lower scores by turning ramp on to ignore events during the time they are falling asleep.
I know I am a new poster and expect much hate mail, to which I shall probably not respond.
Art Cominio
Re: Ramp in CPAP and APAP
It's not a contest, 'lower scores' aren't going to win you anything.ArtCominio wrote: ↑Fri Jun 28, 2019 10:28 amThose of us who use APAP and see events while falling asleep can significantly lower our AHI score by using ramp even in APAP mode.
SO RAMP CAN BE VERY USEFUL EVEN FOR APAP USERS!
Even CPAP users can see lower scores by turning ramp on to ignore events during the time they are falling asleep.
I know I am a new poster and expect much hate mail, to which I shall probably not respond.
Ramp is only actually "useful" if it helps you get to sleep easier, not if it 'improves your score"... heck, if you want, I'll give you a gold star right now and you can move on to focusing on more important things.
Your choice.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Ramp in CPAP and APAP
ResMed machines don't flag anything in regular ramp...nothing.
So lower numbers doesn't mean they didn't happen...just they didn't get flagged.
We don't know if the machine is ignoring asleep real events or awake sleep/wake/junk events....so it is possible that the lower AHI means sub optimal therapy if the person goes right to sleep and a truckload of real events happen while the machine is ramping up.
Now the AirSense machine do have AutoRamp available and that's a whole different can of worms...and that's assuming it works as planned.
Now Respironics machines...depends on the mode and the settings and the model machine. There is some potential for flagging during ramp depending on parameters available.
So lower numbers doesn't mean they didn't happen...just they didn't get flagged.
We don't know if the machine is ignoring asleep real events or awake sleep/wake/junk events....so it is possible that the lower AHI means sub optimal therapy if the person goes right to sleep and a truckload of real events happen while the machine is ramping up.
Now the AirSense machine do have AutoRamp available and that's a whole different can of worms...and that's assuming it works as planned.
Now Respironics machines...depends on the mode and the settings and the model machine. There is some potential for flagging during ramp depending on parameters available.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- Dog Slobber
- Posts: 4260
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Ramp in CPAP and APAP
This is not true.ArtCominio wrote: ↑Fri Jun 28, 2019 10:28 am
Because events are not even monitored during ramp time, using ramp allows the machine to ignore events while you are falling asleep.
Even those who use APAP will benefit from using ramp. it will cause the pressure to remain low for the entire ramp time. Then auto sense will kick in and start to adjust the pressure. BUT all events prior to ramp time will be ignored.
When in AutoSet mode, while in AutoRamp mode the device does monitor events. It monitors them so it can end Ramp prematurely (if needed) and enter traditional AutoSet mode.
Specifically, AutoRamp has 4 triggers where it can exit Ramp mode:
- When it believes you're asleep (30 breaths of stable breathing)
- 30 Minutes
- 5 consecutive snore breaths
- 3 OAs or hypopnea in 2 minutes
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Last edited by Dog Slobber on Fri Jun 28, 2019 11:27 am, edited 1 time in total.
Battery Backup: EcoFlow Delta 2
Re: Ramp in CPAP and APAP
I hate you.
Just wanted to fulfill your expectations.
We aim to please.
JPB
Just wanted to fulfill your expectations.
We aim to please.
JPB
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Sleep on a Buckwheat Hull Pillow. |
Re: Ramp in CPAP and APAP
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Ramp in CPAP and APAP
So kind, considerate and accommodating!
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Ramp in CPAP and APAP
Well when Julie ripped you for being so mean, I took it to heart.
If just in spirit(or sport?).
JPB
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Sleep on a Buckwheat Hull Pillow. |
Re: Ramp in CPAP and APAP
Hurt my feelings, that did!
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Ramp in CPAP and APAP
while we wait patiently for links and data to prove this-ArtCominio wrote: ↑Fri Jun 28, 2019 10:28 amI joined this board specifically because I have seen much BAD information about using the ramp feature for a ResMed AirSense 10.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- billbolton
- Posts: 2264
- Joined: Wed Jun 07, 2006 7:46 pm
- Location: Sydney, Australia
Re: Ramp in CPAP and APAP
That's only likely to be an issue if a very long ramp time is set.
For a 5 minute ramp it is quite inconsequential.
Re: Ramp in CPAP and APAP
True especially if the OSA is on the mild side but we often don't know all the particulars.billbolton wrote: ↑Fri Jun 28, 2019 5:18 pmThat's only likely to be an issue if a very long ramp time is set.
For a 5 minute ramp it is quite inconsequential.![]()
I have had 17 events in 17 minutes....I wouldn't want to be letting those slide unless I had no choice and I for sure would want to know about them.
For someone with an untreated AHI of 100....there is the potential for quite a few events to happen even in 5 minutes. That's almost 2 a minute and even in 5 minutes the potential for oxygen desats.
Not the end of the world but not something I would want to happen.
I just offer the information...up to the user how they want to apply it to their situation.
One of the few features about the Respironics that I find superior...in ramp it still flags everything.
I wish ResMed did because I don't like guessing.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Ramp in CPAP and APAP
Most people turn off ramp because they feel like they cannot breathe at lower pressures.
If someone has a higher than expected score, a look at the data will help determine what may be happening. I've never seen what you describe on data charts I've seen.
If someone has a higher than expected score, a look at the data will help determine what may be happening. I've never seen what you describe on data charts I've seen.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: DreamWear Nasal CPAP Mask with Headgear |
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm



