There's nothing magical about 15cm pressure.pcstud wrote: ↑Sun May 12, 2019 3:07 pmI always thought so. My observation is that the problem appears at a pressure higher than 15cm. That's why I asked here if anyone has similar observations.
ResMed apnea type CA/OA recognition problem
Re: ResMed apnea type CA/OA recognition problem
Get OSCAR 
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
						Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: ResMed apnea type CA/OA recognition problem
You've never been this old before, ya know? things change.pcstud wrote: ↑Sun May 12, 2019 3:06 pmI have been following my results for a long time. There is a limit value above which I do not snore, I do not have hypopnea or apnea. This time the pressure was much higher, so I do not suspect obstruction. Well, unless too high pressure can cause obstructive apnea?
Get OSCAR 
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
						Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: ResMed apnea type CA/OA recognition problem
Classic obstructives followed by recovery breaths.
Get OSCAR 
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
						Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- katestyles
 - Posts: 610
 - Joined: Sun Jan 06, 2019 9:08 am
 
Re: ResMed apnea type CA/OA recognition problem
I looked at your screenshots, and I see obstructive apnea and central apnea both flagged above 15cms. Are you thinking that they are all central? How can you be sure?  A machine fault exclusively above 15cms is highly improbable. I think that it is more probable that something else has changed and the machine is reporting correctly.
			_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier | 
| Mask: Bleep DreamPort CPAP Mask Solution | 
| Additional Comments: Back up mask - anything in the drawer | 
Re: ResMed apnea type CA/OA recognition problem
How do you distinguish between OA and CA in this chart?
Re: ResMed apnea type CA/OA recognition problem
I'm not sure, that's why I asked if anyone has similar observations. It is strange for me to have OA at such high pressures and it is quite easy for me to explain CA. The logical explanation is the incorrect qualification of the type of apnea, which I did not observe at lower values. It's just a machine. Nobody realizes that he will never make a mistake.katestyles wrote: ↑Sun May 12, 2019 4:30 pmI looked at your screenshots, and I see obstructive apnea and central apnea both flagged above 15cms. Are you thinking that they are all central? How can you be sure? A machine fault exclusively above 15cms is highly improbable. I think that it is more probable that something else has changed and the machine is reporting correctly.
Re: ResMed apnea type CA/OA recognition problem
- Attachments
 - 
			
		
				
- Zrzut ekranu 2019-05-13 o 00.40.50.png (65.89 KiB) Viewed 1229 times
 
 - 
			
		
				
- Zrzut ekranu 2019-05-13 o 00.41.17.png (67.26 KiB) Viewed 1229 times
 
 
Re: ResMed apnea type CA/OA recognition problem
There's more flow showing up due to FOT in the central flagged event, indicating that you've got an open airway.
You should be able to see a difference in the corresponding Mask Pressure chart too. they're not huge differences, they're subtle, but they're there.
Get OSCAR 
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
						Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: ResMed apnea type CA/OA recognition problem
Hey Kate...tell him about OAs happening at your pressures.   
  
 
If he thinks 15 is high...tell him about yours.
IMHO...the zoomed in screen shot...the only thing that wasn't real was maybe the central/CA flagged event. That looks like arousal breathing to me. I think the cluster of OAs caused an arousal (may or may not remember waking up) and that central flagged was a not asleep central flagged because of a pause in the breathing because of being awake and the cluster of real OAs caused the arousal.
Just need more pressure at that time for some reason or other...and when that happens the 2 usual culprits....supine sleeping and REM or a combination of both causing someone to just need more pressure than they need at other times during the night.
OAs usually are followed by a big recovery breath...a big gulp of air as the body tries to recuperate from the apnea event. Reflex action more than anything else.
Central apneas don't normally have the big gulp of air so much.
			If he thinks 15 is high...tell him about yours.
IMHO...the zoomed in screen shot...the only thing that wasn't real was maybe the central/CA flagged event. That looks like arousal breathing to me. I think the cluster of OAs caused an arousal (may or may not remember waking up) and that central flagged was a not asleep central flagged because of a pause in the breathing because of being awake and the cluster of real OAs caused the arousal.
Just need more pressure at that time for some reason or other...and when that happens the 2 usual culprits....supine sleeping and REM or a combination of both causing someone to just need more pressure than they need at other times during the night.
OAs usually are followed by a big recovery breath...a big gulp of air as the body tries to recuperate from the apnea event. Reflex action more than anything else.
Central apneas don't normally have the big gulp of air so much.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier | 
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ | 
I may have to RISE but I refuse to SHINE.
						- Jay Aitchsee
 - Posts: 2936
 - Joined: Sun May 22, 2011 12:47 pm
 - Location: Southwest Florida
 
Re: ResMed apnea type CA/OA recognition problem
Given the time of night (morning), I think it's possible that you've entered a period of disturbed sleep punctuated by arousals, maybe starting with an OA, which leads to arousals which leads to breath holding which the machine usually flags as CA. Events during light disturbed sleep with arousals are sometimes referred to as Sleep Wake Junk (SWJ). During these types of periods I have seen breath holding classified as OA instead of CA.
In a lab, events following an arousal would not be classified as either OA or CA.
Since you believe these events are caused by P greater than 15, why not limit Pmax to 15 and see what happens. You could decrease PS, to raise EPAP if you thought it necessary.
			In a lab, events following an arousal would not be classified as either OA or CA.
Since you believe these events are caused by P greater than 15, why not limit Pmax to 15 and see what happens. You could decrease PS, to raise EPAP if you thought it necessary.
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear | 
| Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video | 
Re: ResMed apnea type CA/OA recognition problem
I know the theory ... but I really do not see the difference on the graph.
- Attachments
 - 
			
		
				
- Zrzut ekranu 2019-05-13 o 00.55.46.png (59.42 KiB) Viewed 1220 times
 
 
Re: ResMed apnea type CA/OA recognition problem
This situation happened to me for the first time. I will make it exactly as you say. Thank you for your explanation. That sounds reasonable.Jay Aitchsee wrote: ↑Sun May 12, 2019 4:53 pmGiven the time of night (morning), I think it's possible that you've entered a period of disturbed sleep punctuated by arousals, maybe starting with an OA, which leads to arousals which leads to breath holding which the machine usually flags as CA. Events during light disturbed sleep with arousals are sometimes referred to as Sleep Wake Junk (SWJ). During these types of periods I have seen breath holding classified as OA instead of CA.
In a lab, events following an arousal would not be classified as either OA or CA.
Since you believe these events are caused by P greater than 15, why not limit Pmax to 15 and see what happens. You could decrease PS, to raise EPAP if you thought it necessary.
Where can I read more about these arousals?
Re: ResMed apnea type CA/OA recognition problem
Thank you for your explanation.Pugsy wrote: ↑Sun May 12, 2019 4:51 pm
IMHO...the zoomed in screen shot...the only thing that wasn't real was maybe the central/CA flagged event. That looks like arousal breathing to me. I think the cluster of OAs caused an arousal (may or may not remember waking up) and that central flagged was a not asleep central flagged because of a pause in the breathing because of being awake and the cluster of real OAs caused the arousal.
Just need more pressure at that time for some reason or other...and when that happens the 2 usual culprits....supine sleeping and REM or a combination of both causing someone to just need more pressure than they need at other times during the night.
OAs usually are followed by a big recovery breath...a big gulp of air as the body tries to recuperate from the apnea event. Reflex action more than anything else.
Central apneas don't normally have the big gulp of air so much.
					Last edited by pcstud on Sun May 12, 2019 5:10 pm, edited 1 time in total.
									
			
									
						- Jay Aitchsee
 - Posts: 2936
 - Joined: Sun May 22, 2011 12:47 pm
 - Location: Southwest Florida
 
Re: ResMed apnea type CA/OA recognition problem
If the scale if decreased to enlarge the FOT wave and then The FOT from the Flow rate graph is compared to the corresponding FOT from the Mask Pressure graph, there is usually a phase shift apparent with one type of apnea and not the other. I forget which now.
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear | 
| Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video | 
					Last edited by Jay Aitchsee on Sun May 12, 2019 5:12 pm, edited 1 time in total.
									
			
									
						- Jay Aitchsee
 - Posts: 2936
 - Joined: Sun May 22, 2011 12:47 pm
 - Location: Southwest Florida
 
Re: ResMed apnea type CA/OA recognition problem
viewtopic.php?f=1&t=113409&st=0&sk=t&sd=a#p1094245
viewtopic/t114197/Flow-Rate-Waveform-Ex ... l#p1104979
https://www.bing.com/search?q=sleep+wak ... 92b9a4eea2
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear | 
| Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video | 