Any ideas on what's going on

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
RiverDave
Posts: 470
Joined: Mon Jun 25, 2007 6:46 pm
Location: Central Texas, USA

Any ideas on what's going on

Post by RiverDave » Thu Apr 12, 2018 8:20 pm

Not strictly an apea question, but I'm acutely interested in opinions. Interested in the 01:50 - 02:40 time frame.
Capture.PNG
Capture.PNG (76.59 KiB) Viewed 2910 times

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Any ideas on what's going on

Post by palerider » Thu Apr 12, 2018 9:33 pm

RiverDave wrote:
Thu Apr 12, 2018 8:20 pm
Not strictly an apea question, but I'm acutely interested in opinions. Interested in the 01:50 - 02:40 time frame.
That part looks normal. The 23:10-1:50 part looks wacky.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
RiverDave
Posts: 470
Joined: Mon Jun 25, 2007 6:46 pm
Location: Central Texas, USA

Re: Any ideas on what's going on

Post by RiverDave » Thu Apr 12, 2018 9:44 pm

Hmm. That's what most of my nights look like.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Any ideas on what's going on

Post by palerider » Thu Apr 12, 2018 9:54 pm

RiverDave wrote:
Thu Apr 12, 2018 9:44 pm
Hmm. That's what most of my nights look like.
Unfortunately, you cut off the parts that might be helpful... is that a FL chart I see peeking out above?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
RiverDave
Posts: 470
Joined: Mon Jun 25, 2007 6:46 pm
Location: Central Texas, USA

Re: Any ideas on what's going on

Post by RiverDave » Thu Apr 12, 2018 10:04 pm

here is the flow limitation graph.
Attachments
Capture2.PNG
Capture2.PNG (105.96 KiB) Viewed 2887 times

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Any ideas on what's going on

Post by palerider » Thu Apr 12, 2018 10:45 pm

RiverDave wrote:
Thu Apr 12, 2018 10:04 pm
here is the flow limitation graph.
Can we see the regularly requested graphs?

Flags,
Flow,
Pressure
Leaks
Snore
Flow Limitations.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
RiverDave
Posts: 470
Joined: Mon Jun 25, 2007 6:46 pm
Location: Central Texas, USA

Re: Any ideas on what's going on

Post by RiverDave » Fri Apr 13, 2018 12:11 am

here you go
Attachments
Capture3.PNG
Capture3.PNG (114.8 KiB) Viewed 2864 times

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Any ideas on what's going on

Post by palerider » Fri Apr 13, 2018 12:25 am

RiverDave wrote:
Fri Apr 13, 2018 12:11 am
here you go
Sorry, I shoulda said "and don't cut off the left panel"..

You appear to have your machine at a constant pressure... any reason for that?

It looks like you're still having significant flow limitations, which are making you work harder for your air... and probably degrading your sleep quality.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
RiverDave
Posts: 470
Joined: Mon Jun 25, 2007 6:46 pm
Location: Central Texas, USA

Re: Any ideas on what's going on

Post by RiverDave » Fri Apr 13, 2018 12:31 am

Sorry for my ingnorance. 3rd time's a charm (hopefully). I've used CPAP for 9 of the 11 years I've been on the hose. Had a bit of an issue with changing pressures over the night.
Attachments
Capture5.PNG
Capture5.PNG (183.89 KiB) Viewed 2860 times

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Any ideas on what's going on

Post by palerider » Fri Apr 13, 2018 12:35 am

RiverDave wrote:
Fri Apr 13, 2018 12:31 am
Sorry for my ingnorance. 3rd time's a charm (hopefully)
There's a lot of good info over there too... both in the charts and in the left panel. it all adds up.

Have you tried auto, or a higher pressure? The AHI looks good, obviously, but it doesn't look like easy sleep, the roughness of the flow line (I bet it's interesting zoomed in...)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
RiverDave
Posts: 470
Joined: Mon Jun 25, 2007 6:46 pm
Location: Central Texas, USA

Re: Any ideas on what's going on

Post by RiverDave » Fri Apr 13, 2018 12:42 am

I got a new machine (S9 to A10) about 10 days ago, I just started using Sleepyhead when I got the new machine). I dropped the pressure from 16 to 14 just to see the difference. I will admit my sleep quality has declined for the last 2 months. Tell me the time frame you want zoomed in on and I will post it

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Any ideas on what's going on

Post by palerider » Fri Apr 13, 2018 1:06 am

RiverDave wrote:
Fri Apr 13, 2018 12:42 am
I got a new machine (S9 to A10) about 10 days ago, I just started using Sleepyhead when I got the new machine). I dropped the pressure from 16 to 14 just to see the difference. I will admit my sleep quality has declined for the last 2 months. Tell me the time frame you want zoomed in on and I will post it
How about just zipping up the sd cards from both machines (I'm hoping you still have the s9 sd card) and uploading them to:

https://www.dropbox.com/request/0LalWvEsijRIgX8sdM7H

Full SD cards for both, please?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Ron AKA
Posts: 133
Joined: Mon Apr 02, 2018 10:36 pm
Location: Canada

Re: Any ideas on what's going on

Post by Ron AKA » Fri Apr 13, 2018 8:22 am

RiverDave wrote:
Fri Apr 13, 2018 12:42 am
I got a new machine (S9 to A10) about 10 days ago, I just started using Sleepyhead when I got the new machine). I dropped the pressure from 16 to 14 just to see the difference. I will admit my sleep quality has declined for the last 2 months. Tell me the time frame you want zoomed in on and I will post it
Is there a reason you are running the machine in CPAP mode with a fixed pressure?

User avatar
Pugsy
Posts: 65127
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Any ideas on what's going on

Post by Pugsy » Fri Apr 13, 2018 8:38 am

Ron AKA wrote:
Fri Apr 13, 2018 8:22 am
Is there a reason you are running the machine in CPAP mode with a fixed pressure?
RiverDave wrote:
Fri Apr 13, 2018 12:31 am
I've used CPAP for 9 of the 11 years I've been on the hose. Had a bit of an issue with changing pressures over the night.
Apparently one of those people who doesn't like the variations of pressure when in apap mode.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
RiverDave
Posts: 470
Joined: Mon Jun 25, 2007 6:46 pm
Location: Central Texas, USA

Re: Any ideas on what's going on

Post by RiverDave » Fri Apr 13, 2018 10:24 am

Full SD cards for both, please?
Unfortunately, I was comfortable enough with my therapy, I never put a card in the S9 (didn't know about sleepyhead, or I would have - I'm a data geek). I'll get the A10 data to you once I figure out how to zip it.

Thanks.