why go to a autopap

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
bank4cat
Posts: 28
Joined: Thu Mar 02, 2006 3:28 pm

why go to a autopap

Post by bank4cat » Tue Sep 12, 2006 4:40 pm

i have a remstar pro2 with cflex, i am curious about the auto machine...does the auto make better numbers and overall theropy or just more comfortable.


_________________
Mask
Additional Comments: BLOWER speed 9 c-flex on, 2 liter 02 added

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Post by Goofproof » Tue Sep 12, 2006 6:21 pm

It's all about options, With the Pro 2, you can get most of the answers to treatment you need, if you invest in the software and reader, lot's of others out there aren't so fortunate.

I used the Pro 2 to find my Sweet Spot. The Doc. set me at 14 Cm. the sleep study had me at 12 Cm., By using the software, I found the spot at 15.5 CM. At 16 CM Centrals entered the picture.

The APAP changed the way finding the best pressure worked. The auto allowed me to spend 3/4 of the night at a pressure 2 or more cm. lower, more comfort, less leaks and discomfort.

It also makes a difference what interface you use, each one needs to be adjusted for it's best operating conditions. I now have my treatment pressure avg in the 12 to 13 cm range 90% of the night, the numbers are good AHI <1.1 .. With the Pro 2, I would have to go back to 15.5 CM for the same results.

If there's two ways to get there, why not take the easiest road! Jim

_________________

CPAPopedia Keywords Contained In This Post (Click For Definition): AHI, auto, APAP

Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

inacpapfog
Posts: 538
Joined: Tue Nov 16, 2004 6:03 pm
Location: SC

Post by inacpapfog » Tue Sep 12, 2006 7:45 pm

I enjoy the ability to sleep with a machine that knows when rise to meet my pressure needs through the night! Like Goofproof points out, a straight cpap keeps prescribed pressure going the complete night. As an apap user, I spend less than 20% of my time at my prescribed pressure. Less pressure through the night means less aerophagia for me!!!


User avatar
GoofyUT
Posts: 1085
Joined: Sun Apr 09, 2006 9:45 am

APAP

Post by GoofyUT » Tue Sep 12, 2006 7:52 pm

More than anything else, an auto-titrating machine will blow the LOWEST pressure that you require to maintain patency of your airway, throughout the night and as such, generally improves compliance (usage) significantly.

APAPs are not thought to be a good idea in certain cases, such as those with a history of heart failure for which the increased pressures that an APAP may get to may aggravate the occurence of central apneas or Cheyne-Stokes respiration (CSR) that oftern accompanies heart failure. For those, ResMed's new VPAP ADAPT ASV or Respironics' HeartPAP may be better choices.

For the rest of us, auto-titrating machines have provided greater comfort and therefore, more complaince in most cases, though there are those that will continue to sleep better with constant pressure from a CPAP. BTW, every model APAP can be set to run in CPAP mode. CPAPs CAN'T be set to run in APAP mode, however.

There are dramatic differences though, in the ways that APAPs from the different manufacturers perform, however. So, you will find that you'll do better on one brand or the other, and you OUGHT to do what you can to at lkeast try the variety of different brands until you find the one that works best for you.

For example, REMStar Autos are universally revered here. But they are NOT for everyone. They tend to be slower in responding to SDB events, and so, these events can cause arousals. ResMeds are very quick and aggressive in their response, but this aggressive attack tends to arouse some folks resulting in poorer sleep. So, make sure that you insist that you be given the opportunity to try them out to find the one that works best for you, if at all possible.

Hope this helps.

Chuck

People are dying every day in Darfur simply for who they are!!! PLEASE HELP THEM!
http://www.savedarfur.org

_______________________________

User avatar
Wulfman
Posts: 12316
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: why go to a autopap

Post by Wulfman » Tue Sep 12, 2006 7:57 pm

bank4cat wrote:i have a remstar pro2 with cflex, i am curious about the auto machine...does the auto make better numbers and overall theropy or just more comfortable.
Are you currently monitoring your therapy with software?
There's nothing wrong with the machine you have.
If you've got the extra cash and want to buy a second machine, then I would say get the Auto (not the M version). If you're thinking about replacing the Pro 2 with an Auto, then I would say "Why bother?"

Den

(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

Pirate
Posts: 34
Joined: Sat May 27, 2006 9:09 am

Post by Pirate » Wed Sep 13, 2006 5:53 am

I originally had the Remstar Plus machine. My titrated pressure was 14. After several years of use, I began to feel like the machine wasn't blowing as hard, so I bought a manometer. It tested at 11. I reset the machine to 14 using the manometer which was 17 on the machine, but the pressure was too high and I couldn't sleep. So I bought a new Remstar Auto w/ Cflex and set the parameters to 11 - 17. Since May 30, my average pressure is 11.2 and the 90% pressure is 11.6. The pressure graph in Encore Pro indicates an almost rythmic pressure wave between 11 & 12 all night long and it's never gone above 14.

_________________

CPAPopedia Keywords Contained In This Post (Click For Definition): cflex, manometer, Encore Pro, auto


User avatar
Snoredog
Posts: 6397
Joined: Sun Mar 19, 2006 5:09 pm

Post by Snoredog » Wed Sep 13, 2006 10:45 am

The RemstarPro2 is a good machine, it records sleep data just like the Auto.

Depending on what pressure you are at, autopap can result in lower pressure used throughout the night. If you only need 10-minutes worth of high pressure for the night that is all it supplies then it drops back down to the Min. pressure setting.

If your current pressure is around 10cm or less you would not see any benefit over your Pro machine. If your pressure is like 15cm, then you could benefit from the autopap.


arthuranxious
Posts: 120
Joined: Sun Aug 20, 2006 3:19 pm

Recovering from titration

Post by arthuranxious » Wed Sep 13, 2006 2:42 pm

I had my first ever titration study last night.... a mixed experience. My suggestion that because I have a beard I might have less leaks with nasal pillows was accepted and I was tested with a Swift. Not bad, not perfect but didn't leak and didn't hurt although comfortable would be an overstatement. They asked me to sleep on my back because my supine index of apneas was 52 at the diagnostic study. I tried, didn't fall asleep, or so I thought. After an hour or so I turned on my side,my usual sleeping position, and was half asleep. At 2:00 a.m. I began to panic that I would never fall asleep, sat up and asked for help "Couldn't I have sleeping meds, I don't want to be up all night?" Supervisor called, I admitted that I have never taken sleeping pills in my life, and she said, you have been sleeping very nicely until a few minutes ago, just go back to sleep." Eventually I slept fairly well and woke up thinking it was morning but when I looked at my watch it was 10 to 5. A few minutes later, my hose began making horrible noises, and I called the tech. He said its just water in your hose, opened the hose and poured it out. (Why hadn't he heard it if he had really been monitoring me?) After that I dozed on and off, hearing the people coming out of other rooms and so on, but stayed in bed like a good boy until the tech came in at 7:30... I was too tired to move anyway.
Breakfast came I was too exhausted to eat or even sip lukewarm decaf. I went back to bed but couldn't sleep, felt like after transatlantic trip.
Tech asked, did you sleep well? I said no, he marked down Patient seems well rested, patient says slept worse than at home.
ANyway they have me down for 7 cm. and I wonder if I didn't really sleep normally, why would I rely on their estimate, don't I have to get an automatic?


chdurie2
Posts: 642
Joined: Fri Mar 03, 2006 8:25 pm
Location: New Jersey
Contact:

Post by chdurie2 » Wed Sep 13, 2006 8:32 pm

arthur--

please check your pm.

caroline

_________________
Mask
Additional Comments: Large mask collection
caroline

TheWife
Posts: 53
Joined: Wed Aug 30, 2006 5:31 pm
Location: Vancouver, Canada

Post by TheWife » Thu Sep 14, 2006 11:16 pm

Goofproof wrote:It's all about options, With the Pro 2, you can get most of the answers to treatment you need, if you invest in the software and reader, lot's of others out there aren't so fortunate.
How much does it cost to get setup with software? We're in Canada, and my husband is on a trial with a DME supplier, and he's got a RemStar C-Flex machine now, but we don't have the software to read it. The guy at the store said it would cost about $1000 by the time we bought the software, cables, card reader etc. That sounds insane!

What other option is there to read the data? Go back to the store every time for them to download it? Are there machines where you can read the data directly from the machine without purchasing software?

Theresa


cflame1
Posts: 3311
Joined: Sat Mar 11, 2006 6:55 am
Location: expat Canadian in Kentucky

Post by cflame1 » Fri Sep 15, 2006 5:26 am

you can get the software and the reader from cpap.com for less than
$200


User avatar
oldgearhead
Posts: 1243
Joined: Thu Mar 30, 2006 9:53 am
Location: Indy

Post by oldgearhead » Fri Sep 15, 2006 5:35 am

How much does it cost to get setup with software? We're in Canada, and my husband is on a trial with a DME supplier, and he's got a RemStar C-Flex machine now
Theresa - The software/reader package is 170 USD from CPAP dot com.
I believe there is little restriction on sending Respironics products to Canada. However, I think ResMed products cannot be sent to Canada.

You MUST have one of these units and a 'Smart Card' or the software
is of very little value:
RemStar Pro 2 (plain Pro won't work)
RemStar Auto w/C-Flex
RemStar 'M' Pro
RemStar 'M' Auto

+ Aussie heated hose.
....................................................................

People have more fun than anybody..

User avatar
oldgearhead
Posts: 1243
Joined: Thu Mar 30, 2006 9:53 am
Location: Indy

Post by oldgearhead » Fri Sep 15, 2006 5:41 am

...does the auto make better numbers and overall theropy or just more comfortable.
Re-read GoofProof's post on this subject. His battle at high pressures is
more commendable than the rest of us at 7 - 12 cm/H2O.

+ Aussie heated hose.
....................................................................

People have more fun than anybody..

TheWife
Posts: 53
Joined: Wed Aug 30, 2006 5:31 pm
Location: Vancouver, Canada

Post by TheWife » Fri Sep 15, 2006 10:16 am

oldgearhead wrote:You MUST have one of these units and a 'Smart Card' or the software
is of very little value:
RemStar Pro 2 (plain Pro won't work)
RemStar Auto w/C-Flex
RemStar 'M' Pro
RemStar 'M' Auto
Right now, he's trying out a RemStar C-Flex. Is the Smart Card the little card that I see inserted into the side of the machine? Does the software install on a PC? Without the software, I guess that extra feature is useless?

$200 to $1000 is a huge markup, even with the exchange rate to Canadian $$. And I don't think my insurance company would pay for it.

_________________

CPAPopedia Keywords Contained In This Post (Click For Definition): C-FLEX, auto, Smart Card

Last edited by TheWife on Fri Sep 15, 2006 11:54 pm, edited 1 time in total.

snoregirl
Posts: 1318
Joined: Fri Apr 07, 2006 3:36 pm

Post by snoregirl » Fri Sep 15, 2006 10:20 am

I don't think that there is such a thing as just a Remstar Cflex

It would be Remstar Pro or Remstar Pro2 or Remstar Plus or Remstar AUTO

There is a red (I think that is the color) word after Remstar on the machine. That will tell you if it can use software. Just Pro or Plus can't. Pro2 or AUTO can.

The software goes on the computer. The slot is for the smart card. Think of the smart card like the card that you put in a digital camera. You take it out and read it on your computer with the card reader and software. All the current Remstar macines have the smart card, but only the Pro2 and Auto (from the older series non-M) have the ability to collect data more than just straight compliance -- hours used).

$170 US is the price at CPAP.com. not $1000
The $170 is total for the card and reader

First you have to determine which model of Remstar you have to see if you can use the software.

You are very right, it is doubtful if insurance will pay for software



_________________

CPAPopedia Keywords Contained In This Post (Click For Definition): cflex, auto, Smart Card

_________________

CPAPopedia Keywords Contained In This Post (Click For Definition): cpap.com, cflex, auto, Smart Card

_________________

CPAPopedia Keywords Contained In This Post (Click For Definition): cpap.com, cflex, auto, Smart Card

_________________

CPAPopedia Keywords Contained In This Post (Click For Definition): cpap.com, cflex, auto, Smart Card