Dreamwear. An elbow not a 'ball and socket'?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Controlshift
Posts: 68
Joined: Sat Mar 05, 2016 11:09 am

Dreamwear. An elbow not a 'ball and socket'?

Post by Controlshift » Wed Sep 14, 2016 4:01 pm

Does anyone else wish the Dreamwear mask has a ball and socket type attachment instead of the elbow.
When I move in my sleep of the hose doesn't want to be at right angles to my head it put pressure on the joint
Popping it out once

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Amara View Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Fixed pressure 22/16, heated hose, insulated hose. Devastation AVAPS 30

User avatar
LSAT
Posts: 13352
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Dreamwear. An elbow not a 'ball and socket'?

Post by LSAT » Wed Sep 14, 2016 4:15 pm

Wishing won't make it happen...most masks have the elbow.

User avatar
Grace~~~
Posts: 662
Joined: Tue May 10, 2016 3:27 pm
Location: Florida

Re: Dreamwear. An elbow not a 'ball and socket'?

Post by Grace~~~ » Wed Sep 14, 2016 4:43 pm

LSAT wrote:Wishing won't make it happen...most masks have the elbow.

LOL

I am not sure if *I* even know for sure what a 'ball and socket' is?
(can I now blame sleep apnea and not blondness?)

I guess ControlShift needs to ask {*someone*} how much a mask with a ball and socket would cost?
... and a minimum order?


It's kinda like wishing for the tooth fairy.
More contractual wishing?

(I love being a contractual wishing fiduciary!)


***I'd like to know that too <if you find out ControlShift?>
How much would you think it would be WORTH above the elbow 'that most masks have?'

I have NO IDEA as yet how to value anything XPAP.
Began XPAP May 2016. Autoset Pressure min. 8 / max 15. Ramp off. ERP set at 2. No humidity. Sleepyhead software installed and being looked at daily, though only beginning to understand the data.

jimsp1
Posts: 37
Joined: Tue Nov 03, 2015 4:14 pm

Re: Dreamwear. An elbow not a 'ball and socket'?

Post by jimsp1 » Wed Sep 14, 2016 4:59 pm

I use a hose hanger so that the hose comes from above and elbow works just fine. I imagine if it came from the side it might be more of a problem.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Dreamwear. An elbow not a 'ball and socket'?

Post by palerider » Wed Sep 14, 2016 5:05 pm

Controlshift wrote:Does anyone else wish the Dreamwear mask has a ball and socket type attachment instead of the elbow.
When I move in my sleep of the hose doesn't want to be at right angles to my head it put pressure on the joint
Popping it out once
just how would that work, and still maintain unrestricted airflow?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
RogerSC
Posts: 1912
Joined: Sat Jul 21, 2012 1:11 pm

Re: Dreamwear. An elbow not a 'ball and socket'?

Post by RogerSC » Wed Sep 14, 2016 5:29 pm

I'm currently using a Tap Pap that does have the ball and socket joint (actually it's more like a swivel, but allows more movement than a simple swivel). It just doesn't work very well. When I'm first lying down for the night, the ball and socket moves with jerks instead of smoothly, and it makes noise and it's disruptive when the tubing from the mask jerks and it hits me in the head as a result of the sudden movement *smile*. I really don't see that it improves my experience with the mask...those plastic ball and socket joints aren't really all they're cracked up to be. Or maybe it's just the poorly made one on the Tap Pap. Even using a hose hanger doesn't help a lot, it still bothers me.

After a little research, I might try some KY Jelly, Blixtex, or olive oil to lubricate that stupid joint *smile*. I haven't done that yet, since I was looking for a low-impact lubricant since the joint is in my airflow path.

User avatar
Gasper62
Posts: 982
Joined: Tue Mar 10, 2015 1:22 pm
Location: Central Maryland

Re: Dreamwear. An elbow not a 'ball and socket'?

Post by Gasper62 » Wed Sep 14, 2016 6:30 pm

I've often wondered why a simple, shop-type air hose can have an "any way" swivel, but PAP hoses/masks can not. Don't make good sense ! After all.... they're really not all that complex in design. >>> Link to example : http://public.snapon.com/R_RRD/Objects_ ... YA502M.jpg

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Dreamwear. An elbow not a 'ball and socket'?

Post by Goofproof » Wed Sep 14, 2016 9:33 pm

I've been using elbow equipt cpap masks fot 11 years, only had one elbow come out of a mask, that was from trying to force the hose off the elbow. My fault for not knowing it acts like chinese finger handcuffs, we had as kids.

I run the hose over my headboard, over my head. I move all over like rover in my sleep, never have a problem with hose control, but I stopped doing gymastics in bed long ago. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Dreamwear. An elbow not a 'ball and socket'?

Post by chunkyfrog » Wed Sep 14, 2016 9:41 pm

I believe the Easyfit Silkgel masks from DeVillBiss have a ball joint at the elbow.
This gives the connection more flexibility than a straight 360 degree swivel.
Of course, this kind of feature is proprietary to a particular maker;
and is not available on other brands of masks.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Controlshift
Posts: 68
Joined: Sat Mar 05, 2016 11:09 am

Re: Dreamwear. An elbow not a 'ball and socket'?

Post by Controlshift » Thu Sep 15, 2016 2:52 pm

I'll give a hose hanger a go until Philips take note of my post and fit a ball and socket joint

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Amara View Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Fixed pressure 22/16, heated hose, insulated hose. Devastation AVAPS 30

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Dreamwear. An elbow not a 'ball and socket'?

Post by palerider » Thu Sep 15, 2016 2:53 pm

Controlshift wrote:I'll give a hose hanger a go until Philips take note of my post and fit a ball and socket joint
I wouldn't advise holding your breath.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

library lady
Posts: 1116
Joined: Wed Feb 05, 2014 11:10 am
Location: Rochester, MN

Re: Dreamwear. An elbow not a 'ball and socket'?

Post by library lady » Thu Sep 15, 2016 7:14 pm

My F10 has an elbow that swivels... is that a combo elbow/ball? Seriously though, I also hang it on my headboard and that keeps the hose where I want it.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments:  Sleepyhead
Now using AirFit F10 mask; Quattro Air is backup mask. RemZzzz mask liners with both.

tlohse
Posts: 208
Joined: Fri Jun 27, 2014 5:32 pm
Location: West Bend, Wisconsin
Contact:

Re: Dreamwear. An elbow not a 'ball and socket'?

Post by tlohse » Thu Sep 15, 2016 10:18 pm

I think a ball and socket would be something like your hip where your leg meets your body. I don't know of any CPAP'S that have anything like that.
Thomas Lohse
Machine: ResMed AirSense 10 Auto Set With Heated hose and Humidifier
Mask: ResMed AirFit P10 Size M Nasal Pillow with headgear
Mask: ResMed AirFit N30i/P30i
Original Mask: ResMed Mirage FX Nasal
CPAP USER SINCE JUNE 2013

User avatar
Controlshift
Posts: 68
Joined: Sat Mar 05, 2016 11:09 am

Re: Dreamwear. An elbow not a 'ball and socket'?

Post by Controlshift » Fri Sep 16, 2016 6:25 am

palerider wrote: I wouldn't advise holding your breath.
That's what got me here

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Amara View Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Fixed pressure 22/16, heated hose, insulated hose. Devastation AVAPS 30

marnne108
Posts: 34
Joined: Fri Mar 04, 2016 9:21 am

Re: Dreamwear. An elbow not a 'ball and socket'?

Post by marnne108 » Fri Sep 16, 2016 6:36 am

palerider wrote:
Controlshift wrote:Does anyone else wish the Dreamwear mask has a ball and socket type attachment instead of the elbow.
When I move in my sleep of the hose doesn't want to be at right angles to my head it put pressure on the joint
Popping it out once
just how would that work, and still maintain unrestricted airflow?
Palerider hit the nail on the head. The elbow is part of the two vent system on this mask. A ball and socket wouldn't work without some venting system in place. Not sure how that could be done.

_________________
MachineMaskHumidifier