old respironics bi-flex pro2
old respironics bi-flex pro2
I have an old Respironics Bi-flex pro2 that I would like to keep as a back up, I love this old machine but it just won't start any more, I get an error code? E48 when trying to start it. Does anyone have any idea what that number is?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: ComfortGel Blue Nasal CPAP Mask with Headgear |
| Additional Comments: previous machine Respironics Pro2 |
-
Woodyxxx
Re: old respironics bi-flex pro2
ERR_PV2_sensor_railed
Defective or clogged pressure tube or defective pressure sensor both of these
are internal parts.
Defective or clogged pressure tube or defective pressure sensor both of these
are internal parts.
Re: old respironics bi-flex pro2
I found this : http://www.apneaboard.com/forums/Thread ... AP-machine
_________________
| Machine: DreamStation BiPAP® Auto Machine |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: Settings are IPap 23 EPap 19 |
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: old respironics bi-flex pro2
Contact http://www.acbio.com and see what they have to say.arctic wrote:I have an old Respironics Bi-flex pro2 that I would like to keep as a back up, I love this old machine but it just won't start any more, I get an error code? E48 when trying to start it. Does anyone have any idea what that number is?
They're an authorized Philips/Respironics and ResMed repair facility and a member of this forum.
A number of the forum members have had very good experiences with them.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: old respironics bi-flex pro2
Pretty old machine, that doesn't make it bad, it's a Tank! If you are mechanically inclined, you might open it up,and check for cracked or leaking hoses to the sensors. In the past some hose connectors have broken. J B Weld and Crazy Glue, can work wonders. Jim
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
Re: old respironics bi-flex pro2
Thank you for replies. I love the old tank and would use it instead of Resmed Aircurve 10 Vauto if it would work again